Release Notes 5.8
Jump to: navigation, search
Introduction
We are proud to deliver you Version 5.8 of our RadiMation® EMC Test software. This 5.8 version is the result of several improvements we have done.
This letter contains all the information that is needed before your system is upgraded. If you have any additional questions or remarks please let us know through your reseller, or contact radimation-support@raditeq.com.
Installation
Backup
  • Correction files
  • TSF files
  • Sequence files
  • Limit Line files
  • Device driver configuration files
Multi version installation
Be advised that:
  • It is not possible to run two (or more) different versions at the same time without risking a software crash. Close other instances of RadiMation® before starting another version!
  • Raditeq cannot be held accountable if information is lost because two different versions of RadiMation® are used at the same time.
  • The user needs to have the rights as described in User rights and installation.
Installation
Installing software updates
Updating device drivers
Verification and Validation
1) Automatic knob pushing program
2) Validation and usage in a test lab
3) Validation and usage in a test house
4) Validation by end-users
  • Correct control of the used measurement equipment
  • Validation of the software according to internal quality standards
  • Possible verification and comparison of measurements on a 'known' EUT
Solved items
See below the list of all the issues that have been solved in RadiMation® version 5.8 since version 5.6.
RadiMation version 5.8.38
  • #4409: Switchmatrix device driver is not notified when polarization is finished
  • #4338: Communication with a Narda NBM550 is very slow
  • #4235: Pulse modulation with R&S SMT
  • #4221: Actions don't work in Test Equipments menu
  • #4201: Driver for the 120S1G4
  • #4200: Driver: EMControl Turntable
  • #4196: Total testtime of immunity test with all RadiCentre equipment is too long
  • #4194: Turntable INNCO CO1000 is not working anymore in 5.8.37
  • #4187: Control of a 3 axis antenna system from a R&S receiver
  • #4173: RadiCentre RSW1004A’s – Comport is already open error.
  • #4156: Driver: AVL Turntable / EUT Controller
RadiMation version 5.8.37
  • #4462 Error You are not licensed to use the Report Generator
  • #4461 Driver updates from the Installation and Training visit
  • #4452 Driver: default RS232 speed RadiCentre RPR1006A changed
  • #4429 Is LLF information saved independently with test data
  • #4420 Questions validation and calibration
  • #4412 General questions about Report Generator
  • #4401 Compatibility controlling RadiPower units with old RM
  • #4400 Statements for sorting columns with a space
  • #4391: Driver: ETS-Lindgren 7604
  • #4307: R&S ESU40: set measurement bandwidth to 6dB
  • #4305: Cause of the difference in noise in region below 1 kHz
  • #4304: Driver: no communication PMM EP601
  • #4302: Request for the latest version of RadiMation on CD
  • #4300: Possibility to set amount of Reflected Power during Power Delivery Test
  • #4283: Driver: Milmega 80RF1000-175 Amplifier
  • #4268: Driver: wrong frequency range Agilent N5181A-503
  • #4265: Implementing ESPC driver with option B3
  • #4243: Wrong step size (rounding) in MB
  • #4193: Issue auto perform CS114
  • #4189: Show limit curve on graph during Immunity tests
  • #4188: Performing RS101 susceptibility test described in MILSTD F
  • #4183: Driver for R&S ESR
  • #4176: Driver: AR PM2003
  • #4168: RadiSwitch problem during initialisation
  • #4154: Report generator crashes when a single band emission graph is included
  • #4149: Report generator error 91
  • #4148: Driver: AR 5000A225
  • #4135: Wrong forward power measured with NRVD
  • #4099: INNCO 2000 that only controls the polarization
  • #4086: TDEMI driver with option for multisampling
  • #4058 Driver: Luthi CDN L-801 AF4
  • #544: Include RI and CI manual mode directly in main menu
RadiMation version 5.8.36
  • #4141: Driver: TTi TGA1230
  • #4138: Driver: Frankonia FC 05
  • #4136: RadiSense device driver raises a 'no response received' error when controlled in a USB controlled RadiCentre 1
  • #4134: Advantest TR4172 not working correctly in Multiband
  • #4133: DLL for R&S SMF100 cannot be found (RSSMF_32.DLL)
  • #4131: Agilent N9038A only activates the low frequency band preamplifier
  • #4130: Configurable Signal generator device driver should turn carrier off during the deinitialization
  • #4129: When turntable overshoots it's destination angle, RadiMation keeps waiting
  • #4127: Selecting input2 of the Agilent N9038A raises a frequency error
  • #4126: Driver: PMM 9060 and PMM 9180
  • #4125: Errormessage "Com port is already open" for SOPM01 and SOPM03
  • #4124: Unable to use the 10 MHz RBW on the FSP30
  • #4123: Error messages: Unable to find device driver for ...
  • #4121: Version number of the device driver DLL's are confusing
  • #4120: FSEK 30 only has a single input
  • #4118: Driver: Powermeter for IFI SMC 100
  • #4117: Driver: EMCO1085 (as a turntable controller)
  • #4116: R&S ESMI spectrum analyser shows strange emission shapes
  • #4115: Driver: Maturo MCU as clamp positioner
  • #4113: Problem with identifier of Agilent powermeter
  • #4110: Report for a small frequency span from a large scan
  • #4107: PMM timeout when cancelling multiband test
  • #4106: Error 113 on the display of SMA 100A
  • #4105: Agilent N9038A with multiple detectors
  • #4104: Agilent 9038 in receiver mode in GTEM-test
  • #4097: Keyword for test level in EFT tests
  • #4085: Driver problem with Agilent N9038A
  • #4082: Driver: Rohde & Schwarz ESL6
  • #4075: Turntable from Frankonia uses unit cm
  • #4047: First point from final measurement is not measured correctly
  • #4041: Antenna mast Frankonia FC 02 not working
  • #4030: TDEMI controlled LISNs
  • #3954: PMM uses wrong RBW and Measurement time in conducted emission tests
  • #3953: Error when input from PMM is switched from low to high (9 kHz RBW in high input)
  • #3935: Driver: PMM Controlled LISN
  • #3775: wrong spelling in advanced settings for Maturo MC32 antenna tower
  • #3723: Configurable CDN
  • #3613: driver for R&S ESL 3
RadiMation version 5.8.35
  • #4102: LISN Emission graph shows a large amount of zero's for the frequency on the x-axis
  • #4093: Command error (4198) during start of report generation
  • #4083: Driver: SMC 100A via usb/ethernet
  • #4077: the PULSE modulation is not working with R&S SMP02 generator
  • #4073: Keithley2000 driver as multimeter
  • #4070: System calibration for attenuation/gain (Cable calibration) is complaining of amplifier protection
  • #4054: Zerospan command is not allowed with Agilent N9000A
  • #4024: The canoe driver is giving zero back instead of signal value
  • #3993: Support for more then 8 channels in the configurable ad convertor
RadiMation version 5.8.34
  • #4091: 40 GHz end freqency is too low for measurement.
  • #4076: RadiSense III not working
RadiMation version 5.8.33
  • #4065: Limit line is not shown between 5 and 30 MHz in a CE LISN test
  • #4055: Older embedded software of the R&S SMX needs the space removed for the frequency
  • #4048: Driver: Prana DT25DC RF
  • #4045: ESU 8 Input 2 not working as sensor powermeter
  • #4044: lower start frequency for the vector/signal generator Agilent E8267D which supports downrange
  • #4043: PMM 9010/30P is giving error back for setting the RBW while by hand this is possible
  • #4040: Driver for signal generator Agilent 33220A
  • #4039: Driver for the AR 350AH1
  • #2728: Driver: Switch Matrix VSE Electronic ISF – 1HE 2/4-2
  • #1565: Possibility to specify attenuation of coupler as absolute value and not as a correction of the factory default value
RadiMation version 5.8.32
  • #4211: Error report generator
  • #4037: Power substitution emission measurements in dBm cannot use turntable optimization
  • #4035: Save EUT file as on a different location does not work on Windows 7
  • #4034: Excel Exporter crashes when some specific EUT data is exported
  • #4029: N9000A driver is crashing at about 70% of the sweep when using over lan
  • #4025: Using N9020A with configurable number of points equal to 10001 and higher result in complete crash of Radimation.
  • #2505: Bonn Elektronik BLWA 2010-50 and BSA 0122-100 driver is not working
RadiMation version 5.8.31
  • #3098: Agilent 4417 device driver is not working correctly anymore
  • #3099: Driver for the AR 25S1G4A doesn't work correctly anymore
  • #3102: Using the HP 33250 as an external modulation states that wrong GPIB commands are used
  • #3407: Driver: SC1000
  • #3733: Driver: Milmega AS0104-200/200
  • #4000: Driver for spectrum analyser Agilent N9320A
  • #4010: Limit lines are not displayed correctly in the graph
  • #4016: ESU error: More points received from the receiver/analyser than calculated
  • #4021: RE multi band 360 degrees area does not work all the time
RadiMation version 5.8.30
  • #3863: report generator code horizontal and vertical max peak are not working correctly as the complete frequency range is not displayed.
  • #3884: The configuration window of the configurable switch is not closing when pressing ok
  • #3886: CANoe Analyser CLSID cannot be found
  • #4007: Final measurement IF overload is not detected correctly
  • #4013: Configurable amplifier is crashing when settings the frequency range, driverversion 2011-6-7
RadiMation version 5.8.29
  • #2616: First point of a calibration is too low
  • #3887: Agilent E4446A analyser should use the correct RBW setting for MIL-STD
  • #4002: Power meter device driver Advantest 3361B is giving event Time out
  • #4011: Reviewing Harmonic test is crashing RadiMation
RadiMation version 5.8.28
  • #3867: Net power and calculated current on the CAN-bus
  • #3999: Driver Spectrum Analyser HP E7405A does not work any more after return from agilent service point
  • #4006: No prescan is done with FSV40
  • #4009: No final measurement is done with FSP 3 and LISN in multiband
RadiMation version 5.8.27
  • #2601: Test initialization window should also show the name of the test-site that is being initialized
  • #3976: Agilent N9020A: Configurable number of points
  • #3998: Driver for AR500A250A
  • #4004: Report generator code for emission tests in multiband
  • #4008: Measurement with HF-probe
RadiMation version 5.8.26
  • #4003: Closing the Visa window is crashing RadiMation with new gcc compiler
  • #3990: Driver: FSEA20
  • #3986: Configurable power meter driver can not find expression value while the value is send over the GPIB bus
  • #1942: No check for connected device is included in device driver of Marconi 6960
RadiMation version 5.8.25
  • #3543: Unclear message when the RadiPower is not connected to the RadiCentre
  • #3783: Advantest Analyser is not measuring CE peaks correctly with version 5.7.4
  • #3903: Advantest R3361B analyser raises timeout errors in a Multiband Emission test
  • #3937: Scope driver Multi band is decreasing the scale when the returned value is zero and the scale is still maximum
  • #3988: Report Generator crashes when a ) is included in the template filename
RadiMation version 5.8.24
  • #3920: Driver: Marconi 2390
  • #3965: Is not possible to choose between EXT-AC and EXT-DC using the siggen HP8648A.
  • #3979: Compression test fails when the amplifier has correction files attached
  • #3980: device driver for amplifier AR 60S1G4
  • #3982: Engineer system in RadiMation allows impossible selections and removals.
  • #3983: Modulation is always marked red after stopping and continuing RI multiband test
  • #3984: Allow more than 8 AD-channels
RadiMation version 5.8.23
  • #3969: Selecting the lowest location type in a multiband Lisn test configuration results in a crash of the software
  • #3975: Configured sweeptime of a band is not used, due to sweeptime/MHz calculation
  • #3977: Changing reference level is not always allowed
  • #3978: Agilent 4446A: Sweeptime is changing when detector is selected
RadiMation version 5.8.22
  • #3964: Pauses in a sequence are crashing the software
  • #3966: Peak not detected in conducted emission current probe test
RadiMation version 5.8.21
  • #2162: Create base class for the RadiCentre drivers
  • #3795: Adding the device driver version to the about screen of RadiMation
  • #3921: Extend driver of RadiTower to control a seperate RadiPol
  • #3967: Pause test can not be used; problem during creating it
  • #3968: Multiband emission test files cannot be opened anymore.
RadiMation version 5.8.20
  • #3960: HD 100 Antenna tower driver is always in max speed.
  • #3948: Agilent E4446A is not keeping 50 sec sweeptime in multiband test
RadiMation version 5.8.19
  • #3910: External AM sine and square is not working with the Agilent PSG E8257D
  • #3932: Driver: AR FP4080 controlled as an AR FP3000
  • #3949: Printout is not generating the same plot twice, and the printout during the test is a combination of both
  • #3652: Conducted emission - current probe TSF is saved in conducted immunity folder
  • #3956: Agilent E4440A as a powermeter
  • #3957: Registration of STDOLE2.TLB Error message during installation on Windows 7
  • #3958: Some device drivers are missing
  • #3959: Report generator code AVERAGE DETECTOR uses information of Peak Detector
RadiMation version 5.8.18
  • #3925: R&S FSV13 does supports scans from 10Hz
  • #3926: RBW is not correctly set on the R&S FSV13
  • #3927: Raditurn is giving already in progress after motor error
  • #3934: Several Multiband Conducted Emission Current probe problems
  • #3938: Incorrect limit line when using Merge function
  • #3947: AD/dc selection in power meter driver advantest is giving problems
  • #3950: Zoom-out an emission graph by drawing a reverse zoom-in box
RadiMation version 5.8.17
  • #3922: Copyright is 2010 instead of 2011
  • #3941: Agilent E4417A power meter is limited to 18 GHz
  • #3942: Driver: Agilent N9000A
  • #3943: Driver: Agilent N9030A
  • #3944: FAQ is not explaining why the measured value of the field probe can be higher with the am modulation on.
  • #3945: How does the cable calibration works in the wiki page
  • #3946: explain what is the difference between a printout during the test and a print out in the eut window
RadiMation version 5.8.16
  • #3905: Test sites are lost in from time to time without any leads.
RadiMation version 5.8.15
  • #3924: Sweeptime is not set in the Agilent N9020A
RadiMation version 5.8.14
  • #3876: Gaps in emisson graph after final measurement
  • #3894: Function "Continuous Measure" can block RadiMation with an error warning
  • #3895: Test-site from a band can not be chosen after pre-scan
  • #3904: Setting of the sweep time of the HP 4395A is not possible
  • #3909: Pre amplifier is not yet turned on while sweeping
  • #3917: Nonexistent peak and wrong final measurement with PMM during conducted emission multiband tests
  • #3918: FSH 3 is setting wrong detector
  • #3928: Driver: HP 4396A
RadiMation version 5.8.13
  • #1856: No possibility to specify the modulation depth and frequency for external AM modulation
  • #3196: Driver: Agilent N9020A
  • #3875: Question: Radiated Emission Multiband: Order of measured peaks in the final measurement
  • #3896: Older test results show a low resolution Multiband Emission graph
  • #3902: The command ||LISN LINE|| is not recognized anymore by the report generator
  • #3906: Scaling of the graphs can make the AD convertor graphs unusable
  • #3907: Angle and height of detected peaks are always the same (in singleband emission)
  • #3908: Agilent 4446A cannot be used below 9 kHz
  • #3911: Driver: Agilent 83640A
  • #3914: Report of MB emission tests contains a large list of detected peaks, if no peaks are detected
  • #3915: Use a Microsoft SQL Server database to retrieve customer information
  • #3916: External PM is not allowing the selection of modulation out side the dwell time
  • #3919: Possibility to draw graphs via report generator with the full dataset
RadiMation version 5.8.12
  • #3857: PMM 9030 driver is giving time out during measurement of peak
  • #3877: Slow final measurement with PMM 9010/30P via RS232
  • #3878: PMM 9010/30P hangs up in final measurement
  • #3885: Peak and Average measurement at the same time with PMM not possible
  • #3891: TSF does not save measure- and observation time from "optimize" - settings
  • #3892: Wrong detector is used in final measurement
  • #3899: Y-axis unit is not shown anymore in singleband emission manual mode
  • #3900: Observation time of a singleband emission test is always 1 second
  • #3901: Opening Multiband Emission tests is slower
RadiMation version 5.8.11
  • #3308: Measure with ESVS10/ESHS10 with PK and AVG detectors together: average trace is not visible until the sweep is finished and the method is saved, closed and open again
  • #3868: Creating new 16 point verification on a chinese pc is displaying a message box and not crashing
  • #3888: Graphic of correction file not displayed
  • #3889: RadiMation manual contains a lot of unrelated text
  • #3890: High peaks in the emission graphs of the R&S ESHS
RadiMation version 5.8.10
  • #3869: Saving a cor file is giving problems with ) at the end with a chinese pc
  • #3882: CE Current probe test is crashing
  • #3883: Multiband Emission graph is showing Bars
  • #3913: list of detected peaks does not have all the Average values specified
RadiMation version 5.8.9
  • #3791: Peak detection wrong in CE-MB (LISN) test
  • #3843: Multiband emsion data file cannot be loaded anymore. It seems to be corrupt
  • #3872: Reselecting the report is giving the .doc instead of .txt in the result file
RadiMation version 5.8.8
  • #2915: Report generator template should be selectable in the file selection window
  • #3757: Report Generator to ASCII
RadiMation version 5.8.7
  • #457: Possibility to perform MIL-STD emission measurements in dBuA
  • #469: Not possible to perform broadband measurements (dBuV/MHz)
  • #475: Vertical scale of emission graph is not possible in dBuA
  • #552: Configuration of the emission graph lines
  • #784: Radiated emission graph dBuV should be dBuV/m
  • #1300: Merge function - Scaling is wrong in Report Generator
  • #1688: Sorting in unit configuration window is not working as expected
  • #1704: IMaster unit transformer
  • #1834: Possibility of measuring a magnetic field with a loop antenna (Unit dBµA/m)
  • #2063: Changing the unit in a CE Current clamp test, doesn't change the graph
  • #2205: Is fast scan mode supported in the multiband emission
  • #2248: RE magnetic field MIL STD test reference value is always zero.
  • #2274: Multiband Conducted Emission current probe
  • #2327: RE multiband test can not be started from restart last test button
  • #2413: Radiated emission substitution method test
  • #2656: A multiband graphline is included multiple times on the graph
  • #2966: Possibility to see the internal attenuation of a coupler device driver
  • #3184: The unit of the reference level and the Emission is the same, so the correct reference level can not be reported
  • #3381: Traces are not converted to the correct amplitude unit
  • #3395: Limit lines hidden during pre-scan
  • #3415: The sequence of the Multiband is according to the way it's been entered
  • #3420: AdvantestTR4172 device driver is missing for input 1 50Hz - 1800MHz
  • #3596: Limit Lines not shown during CE measurement in Multiband
  • #3607: CE Multi band can only be performed with a LISN
  • #3699: Possibility to set a default unit for radiated Emission and conducted Emission
  • #3718: Limit line with unit
  • #3719: Emission Test: Results are not converted to chosen unit - just the display is upgraded
  • #3731: sx32w.dll is causing problems with the dongle in version 5.7.2
  • #3784: RI multiband test is not turning on the amplifiers in version 5.7.4 while this was working in version 5.6.13
  • #3830: Physical properties do not match with the units in the menus when using German language
  • #3831: Support for the mil std dbuV/Mhz/M according to the 462C standard
  • #3832: changing dbuV to dbuV/m does not change the reading in the display
  • #3833: Narrow band / broad band detection according to the Mil std 462C standard
  • #3835: Specification of the limit line in the configuration of the limit line
  • #3844: dbuV/m/MHz axis is not displayed is not displayed in the mil std test
  • #3845: BCI multiband is not displaying the lines correctly when the levels are identical
  • #3852: Saving a file doesn't show a progress bar
  • #3855: ESHS10 is scanning the frequency range 3 times
  • #3856: Opening older (< 5.8) multi band lisn test is causing a crash
  • #3859: AC coupling on input 1 decreases the start frequency of the ESU26.
  • #3861: Communication Problem with configurable AD converter via TCP/IP
RadiMation version 5.8.6
  • #3418: Driver: Advantest R3261A as powermeter
  • #3604: RadiSwitch is generating sometimes is generating an error during the test.
  • #3653: Driver: IFI 404
  • #3656: Driver: OPHIR 5019E
  • #3681: RadiMation tries to put FSP 3 in Receiver Mode
  • #3803: CARAD - Timeout
  • #3804: CARAD - debug window
  • #3805: CARAD - clear buffer after timeout
  • #3806: CARAD - error message when using wrong IP-address
  • #3809: Not possible to configure stepsize in RE-MB RC test
  • #3818: several questions regarding CAPL
  • #3824: Configurable alarm detection is always reporting alarm, even when the correct response is returned.
  • #3829: Driver for Amplifier Bonn BSA1050-250/100
  • #3840: Copyright is 2009 instead of 2010
  • #3841: Closed loop band before substitution band is checking the calibration information of the substitution band
  • #3842: end-of-line character for AD Converter via TCP/IP
  • #3846: Cannot select multiple cables for a specific part in a testsite
  • #3849: Am modulation is not turned off when selecting no modulation in RI multiband test.
RadiMation version 5.8.5
  • #2600: Undocumented error 0x300000C9 when Prana amplifier was already switched off
  • #2775: Measurement of current by using the current sensor in combination with a multimeter or oscilloscope is wrong
  • #3190: Driver: Anritsu MS2621B
  • #3447: Conducted Immunity Tests are slow in 5.5.7 in comparison to 5.3.18
  • #3636: Driver: Logimetrics AN8200G
  • #3637: Driver: Logimetrics AN 8100Q
  • #3651: Driver: Kalmus 122FC
  • #3652: Driver: Kalmus LA200U
  • #3732: Driver: Milmega Rf1000
  • #3734: Driver: schwarzbeck antenna's FESP 5132 + FESP 5133-7/41
  • #3735: Driver: SCHWARZBECK TEMZ 5231
  • #3819: Network Analyser EUT Calibration, using a R&S ZVK, crashes with an overflow
  • #3828: Unit is single band graph do not change after changing the unit in the unit configuration of the result file
  • #3838: Why is it required to specify a magnetic flux density file for a MIL STD Calibration?
  • #3839: Radiated Immunity MIL STD Calibration is always using a 70 dBuV Reference level
RadiMation version 5.8.4
  • #1290: Radiated Immunity not completely usable for H-Field measurements
  • #2604: Full support for Magnetic field tests in Radiated Immunity module
  • #3095: RadiTurn controller does not allow a destination angle below 0 or above 360 degrees
  • #3147: Memory leak in opening/closing RE MB TSF
  • #3283: Modulation command is not retransmitted by configurable device driver if modulation parameter is changed
  • #3686: AD Converter is read out before modulation is switched on
  • #3720: During emission test: crosses instead of line
  • #3779: Calculated field is not send to switch matrix in multiband test
  • #3781: Driver: Farnell PSG 1000B
  • #3789: Memory leak in reviewing Multiband Emission test
  • #3802: Memory leakage in Multiband Emission test, when test is opened and closed
  • #3823: TDS 600 Series scope driver reports unknown device driver when pressing advanced in the device driver settings window
  • #507: Overload indication of an analyser/receiver is not detected
  • #2764: Optimize IDataIdentifier comparing
  • #3183: Crash of RadiMation Version 5.3.18 when CARAD driver is included
  • #3354: CARAD interface: two commands get combined to one
  • #3372: ESU8 returned an error message during Radiated Emission MB measurement
  • #3535: *OPC for ZVK with antenna measurement
  • #3544: Device driver for Narda NBM550 uses the wrong baudrate
  • #3615: Multiband-Conducted Immunity: data is not displayed
  • #3780: Driver: Pre amplifier Schwarzbeck BBV 9718
  • #3782: Driver: Antenna Schwarzbeck STLP 9128D and Elektromectrics EM 6954
  • #3794: Driver for amplifier Bonn BLWA-0820-160/75/50D
  • #3814: Driver: AR 15S1G4
  • #3816: *OPC problems for ZVK (as signal generator) when setting the power level
  • #2152: Peak search for turntable angle should be configurable to include another angle range than the configured turntable
  • #2153: Peak search for antenna tower height should be configurable to include another height range than the configured antennatower
  • #2843: Incrementing the turntable angle, should wrap to the corresponding value if the maximum angle is reached
  • #2849: When the turntable angle is changed, the direction with the shortest angle difference should be used to turn the turntable
  • #3545: Automatic Angle and Height optimization should use the physical limitations of the turntable and antenna tower
  • #3592: Turn Table Action in Manual Mode not possible with EMCO 1050
  • #3593: Antenna Tower Action in Manual Mode not possible with EMCO 1050
  • #3755: RadiMation is setting up the ESPI with incorrect RBW during CI-testing
  • #3201: Dificulties with OATS measurements
  • #3563: Limit lines should be shown in polar plot and height graph
  • #3701: Multiband Radiated Emission: Default value in dropdown list for Turntable Angle specifies 235°
  • #3744: When 'Detect Peaks' is pressed, already detected peaks should be placed on the angle/height with the maximum emission
  • #3745: Turntable doesn't turn a complete turn when it starts on a negative angle
  • #3746: Peak zoom-in doesn't seem to work in multiband emission
  • #3747: Improve peak detection by splitting the span in chunks
  • #3748: Improve EMCO1060 turntable controller
  • #3749: Click on the graph to add peak, should allow multiple peaks on the same frequency but with a different polarization
  • #3751: Frequency marker is not moved on the graph when it is dragged
  • #3752: Clicking on the turntable CW button to release it again doesn't respond
  • #3753: HP4395A is always using 10 Hz RBW when it is used for a voltage measurement
  • #3754: Climate chamber is not controlled correctly by RadiMation.
RadiMation version 5.8.0
  • #1713: Continuous scanning is not possible in Multiband Emission Manual mode
  • #2512: Zooming into emission graph while measurement still running
  • #2597: In multiband immunity the cancel button only reacts on the end of the dwelltime
  • #3203: Manual mast control does not work correctly, Up is not always Up
  • #3386: the Radisense ctr1001B driver does not allow you to set the filter when using in gpib mode.
  • #3490: Multi band emission manual mode is generating error when selecting the test site
  • #3456: Manual mode multiband emission test does not have an initial graph which results in a crash
  • #3547: First graph in a multiband emission measurements shows a bottom y-axis value of -1000 dBuV/m
  • #3552: Spurious peaks in polar plot as a result of the initial measurement
  • #3566: Change the frequency/angle/height of a detected peak by moving the marker in the graph
  • #3611: Turning range of the antenna tower when changing polarization
  • #3709: zero-span mode in radiated emission multiband
  • #3714: Emission Multiband: Scan analysis while scan is running
  • #3724: Radisense 4 over RS232 does not set the filter setting
  • #3729: CE multi band measurement is not showing the information below 1 MHz
  • #3736: I can not select apply modulation outside dwell time when selecting pulse
RadiMation version 5.7.8
RadiMation version 5.7.7
RadiMation version 5.7.6
  • #3853: CARAD monitor: settings for AD Input
  • #3854: When cancelling CARAD timeout the "Save EUT As" window pops up
RadiMation version 5.7.5
  • #3614: Selecting no detector in the initial trace will result in a error 91 in multi band
  • #3684: RadiMation and CAPL (CANalyzer 5.2.7)
  • #3820: graphs are smoother in version 5.7.5 than in version 5.5.7
RadiMation version 5.7.4
  • #3546: Manual mode multiband emission test does not have an initial graph which results in a crash
RadiMation version 5.7.3
  • #3674: UP300 - generator output
  • #3704: Driver for RadiSense IV (RadiCentre) not working
  • #3725: ESU 26 power meter driver is reporting unknown function during amplifier harmonics measurement
  • #3726: Signal generator goes to +24 dbm while the limit is set to +10 dBm with a RI subst test
RadiMation version 5.7.2
  • #2204: Add a peak by clicking on the emission graph
  • #2920: disable zoom in does not work directly
  • #3073: Trace gaps on emission plots
  • #3497: Keithley 2000 scanner does not detect connected device correctly
  • #3502: RadiMation stays in a never ending loop to zoom the peak
  • #3565: Negative turntable angles should be allowed in combination with the "measure during turn" option
  • #3584: Wrong value readout of LeCroy 9361 scope
  • #3628: Overloaded driver of the ESU is crashing then the device *IDN? description is not in capitals
  • #3629: Changing the test order before reporting is crashing the RadiMation
  • #3630: Driver: Maturo NCD
  • #3634: Scaling in CE multiband is going to -1000 during the test.
  • #3639: Driver: AH system SAS-570
  • #3640: Driver: AH system SAS-571
  • #3642: Driver: AH system SAS-550-1B
  • #3643: Driver: AH system SAS-560
  • #3649: Driver: HP 8673E
  • #3650: Driver: HP 8662A
  • #3654: Driver: IFI SMCC10
  • #3655: Driver: OPHIR 5163
  • #3657: Driver: HP 8341B
  • #3658: Driver: Solar 6338-5-PJ-50N
  • #3659: Driver: Solar 9233-50TS-50N
  • #3661: Driver: Solar 7334-1
  • #3662: Driver: Electro metrics EM-6892
  • #3663: Driver: Electro metrics EM-6960
  • #3664: Driver: Electro metrics EM-6961
  • #3665: Driver: Electro metrics EM-6876
  • #3668: Driver: Solar 9251-1
  • #3673: Peak selection is not working as expected after deselecting all the peaks
  • #3675: Configuration for optimize angle and height has problems with the measure- and observation time of different detectors
  • #3676: During Optimize Height actual defined settings are not used but settings defined a step before
  • #3677: When you press <<Detect Peaks>> after a radiated emission test in multiband you can not select more than one peak afterwards.
  • #3678: Peak detected in first band but not in second band
  • #3680: Configurable device signal generator device driver is not taking the correct codes in version 5.6.9
  • #3682: Emission multiband is slow when 200 peak sweeps are used
  • #3692: Prana amplifier does not turn off when the amplifier is changed during manual mode
RadiMation version 5.7.1
  • #1948: No gridlines are shown from 1 GHz to 2 Ghz in logarithmic mode
  • #2599: Order of tested bands should have a preference for the frequency order
  • #2681: Editing a testlevel when multiple bands are selected, only changed the variable test-level for the first selected band
  • #2754: Dongle is not detected when using Remote Desktop.
  • #2756: Frequency Label of the x-axis of the graph in a correction file is displayed too low
  • #2866: How can scaling of frequency axis be configured when logarithmic scaling is choosen?
  • #2911: dBmA/m unit in multi band is giving wrong measurement information
  • #2999: Order of the bands in Multiband immunity is random
  • #3141: Test equipment wrong in Sequence mode
  • #3416: AM modulation indication on the HP 83630 is not activated
  • #3454: Term MHz in Graph while the values are also in MHz.
  • #3617: Driver: Anapico APSIN300HC
  • #3621: No virtual multi meter available in version 5.7.0
  • #3622: January and February dates are not detected on a non-english PC
  • #3623: RI Magnetic calibration are not displayed after RadiMation is restarted
  • #3625: New bands in a multiband test always have the color black
  • #3626: Driver: R & S FSV 3
  • #3627: Driver: R & S FSU 3
  • #3632: Carrier level jumps directly from -100 dBm to +1 dBm when it should go up 1 dBm
RadiMation version 5.7.0
  • #1937: Esi startup problem(s)
  • #1986: Switching to the next band takes more then 2 minutes
  • #2479: Driver: 1000W1000C
  • #2557: FSP30 as power meter driver
  • #3164: Cable correction between the signal generator and the amplifier is performed in the wrong direction
  • #3494: ESU Time Domain Scan Fails when using lan
  • #3551: R&S ESxx receivers: Perform short calibration before the final measurements
  • #3560: R&S ESxx receivers: Enable demodulator
  • #3568: SMR 20 does not support the external PM modulation
  • #3569: strange behaviour after driver package update ESVS10
  • #3575: Driver: Emco 3104C antenna
  • #3576: Problems with deinitialization of ESVS10
  • #3585: For a configurable LISN no manual mode is included
  • #3586: CE multiband LISN test does not perform the band measurement
  • #3587: driver for R&S SMC-B103
  • #3588: The note on the multiband test window cannot be changed
  • #3589: Note textbox on multiband tests, is not showing scrollbars
  • #3590: Final peak measurement is performed too often
  • #3591: Report generator is using more than 500 MB of memory
  • #3618: Driver: 30S1G4
  • #3619: Driver: ETS EMCO 3142D
  • #3620: Driver: Narda 6630
RadiMation version 5.6.12
RadiMation version 5.6.11
RadiMation version 5.6.10
  • #3631: Dongle is only detected correctly the first time on a windows 2000 SP4 and version 5.6.9 of RadiMation
RadiMation version 5.6.9
  • #3270: Graphs in MultiBand
RadiMation version 5.6.8
RadiMation version 5.6.7
  • #1845: Change the size of the Select a TSF file window
  • #3548: List of selected equipment in testsite selection window is grayed
  • #3851: RE mil std test is not displaying the markers correctly
  • #3582: Difference to peak limit is incorrect in the re magnetic test
  • #3583: The reference level of the RE magnetic field test is 0 or auto
RadiMation version 5.6.6
  • #3479: driver: PMM EP600
  • #3532: SMR20 should generate an error when going below the minimum value of the signal generator
  • #3536: Crash when a graph is shown
  • #3542: Attachments in the External test results test are not shown in the table
  • #3550: Configuration directory path uses a limited path length
  • #3558: R&S NRP-Z91 device driver crashes when R&S DLL's are not installed
  • #3571: HP 8562A frequency range is incorrect, the top frequency is 22 GHz.
  • #3574: Bonn amplifier BLMA 1080-60/35/20D is not working correctly when using the USB interface.
  • #3577: ESU analyser driver does not set the pre-amplifier and the pre selector.
  • #3578: EN 61000-4-3 2008 version should be added to uniform field area calculation
  • #3579: RS232 communication with RadiSense plugin card is problematic
  • #3580: PMM OR 03 device connection check is not working always
RadiMation version 5.6.5
  • #3471: regulation problem when leveling on net-power while limiting forward power
  • #3538: Driver: Agilent N5183A
RadiMation version 5.6.4
  • #2945: Inserting Table is crashing the Software
  • #3202: Table to clipboard button does not work, right click on the table is ok.
  • #3205: If 'peaks' is selected extra graphs appear. There does not appear to be any way of removing them after they have been created.
  • #3221: There are problems with importing results from multiband Tsfs.
  • #3474: Improvements to Polar plot and height plot to show information of the selected peak
  • #3539: Readout of Amplifier Research FI 7000 does not work
  • #3540: Crash when the sorting of the columns in the report generator window is changed
  • #3541: Not possible to include an attached photo from an external test results test into the report
RadiMation version 5.6.3
  • #2626: Capl interface is lacking the turn table and the antenna polarisation
  • #2796: R&S ESI cannot be used as an powermeter from 20 Hz
  • #2925: Driver for Yokogawa Oscilloscope DL1740
  • #3383: AC/ DC selection is not possible with the ESU
  • #3393: Radimation Multiband Anechoic Chamber dBuV - dBm projection is incorrect
  • #3443: Reseller logo emitec
  • #3468: peak measured at wrong frequency with ESU 8 in Time Domain
  • #3484: possibility to configure scan resolution for ESU in analyser mode
  • #3485: HP 8542E does not allow the filter selection for the second input
  • #3491: TEST ENGINEER keyword in the report generator does not work for a multiband
  • #3492: ESIB 26 is generating waiting for OPC errors and is not reacting to the GPIB bus any more, restore only with mains cycle
  • #3493: Graph line update with real hardware is causing > 2 GB of memory usage, while virtual is much lower
  • #3496: hi 4455 configuration window does not work correctly
  • #3498: Merge function does not support the G-tem test.
  • #3500: file does not want to be saved when the disk is full of write protected
  • #3501: Not every version of RadiMation is displayed in the control panel
  • #3506: 5.1 dB limitation does not work correctly with ci fixed current test.
  • #3507: EUT controller not working as expected
  • #3510: Memory usage with the ESU in scanning receiver mode is increasing
  • #3511: ESU scanning receiver is using port 1 while 2 is selected
  • #3512: Signalgenerator Agilent/HP E8257C not working anymore
  • #3515: Frequency band user control is looks funny when 2 band are after one another.
  • #3524: Amount of samples does not work in the Attenuation system calibration when the value is not changed
  • #3529: Hp 34401A configuration window does not work correctly, the settings are not saved
  • #3531: Agilent N9310A has got problems doing the init after updating the Visa lib
RadiMation version 5.6.2
  • #2134: AD Convertor device driver that monitors received measurement values on the serial port
  • #2447: Emco 2090 driver update after install
  • #2645: Call was rejected by the callie
  • #3028: option to remove the overrange warning from the 4455 driver
  • #3053: Driver: Pmm 8051 + probe BA05
  • #3423: the "Print Report"-Box in the Report Generator is activated by default.
  • #3481: Import rsus substitution file is incorrect
  • #3482: Release note is not generated in the automated build scripts
  • #3486: FSEM 30 Input 1 and input 2 are incorrect because the device only has 1 input.
  • #3488: Multi band manual mode generates an error of start stop frequency during the init of the test
  • #3478: An error message is shown when the configuration window of the Agilent 4440 is closed
  • #3477: Using a sub template as an template for reporting after the tests results in an empty graph
  • #3472: Sorting peaks according to numbers does not work correctly
  • #3470: peaks wrongly detected & merged graphs wrongly displayed
  • #3469: Limitation of net-power instead of forward power
  • #3466: Second band of a emission multiband test is using the same step frequency as the first band
  • #3464: Crash (error 91) when the configuration window is closed
  • #3463: Kalmus 757LC amplifier is not working. Test seems to hang
  • #3462: SMA100A is going to 0 dBm all the time
  • #3452: Driver Antenna AR AT5080 and Schwarzbeck HFRA9150
  • #3451: Driver: Teseq CBA 1G-1000 and milmega AS0104-400/200
  • #3439: Starting level of a multiband test is -100 while the start level setting of the signal generator is -50 in the configuration
  • #3375: Error message from ESU8 in CE/RE-MB test "more points received than expected"
  • #3094: Using a measure and observation time of both 5 seconds is raising problems with the R&S ESU
  • #3017: ESU-40 driver issues in FFT mode
  • #2589: R&S ESU (GPIB controlled) and R&S ESU (LAN controlled) should be combined into a single driver
  • #2745: Problems with ESU 8 when doing Time Domain Scan
  • #2871: Determination of interference threshold does not work at BCI closed loop
  • #2958: R&S SME 03 AM 0% instead of off to prevent the relay from dying
RadiMation version 5.6.0
  • #420: Option to limit the power in a conducted immunity substitution test, based on the calibrated forward power
  • #421: Configurable device driver for reading out an AD-convertor
  • #954: Step frequency in RE test
  • #1302: EMI test with log. Frequency steps
  • #1741: PSA limit not the only one describing power limitation on forward power calibration
  • #1803: Device driver for AR 500W1000 is conflicting for different end-users
  • #1811: Switch matrix device driver for the HP 3488A
  • #2145: Amplifier is switch in POWER OFF after BCI test
  • #2450: Changing description 'PSA-Limit' to 'Power Limit' in BCI
  • #2538: Not possible to level on the current in CI substitution manual mode
  • #2699: During a 'Check' for a device, all the other controls should be disabled
  • #2774: Question to CE measurement (problem with accuracy of detected peak)
  • #2881: Driver: Advantest R3131A as powermeter
  • #2929: A small misspell in the CE test.
  • #2957: RadiSwitch 6 port does not have an option to not switch the relay
  • #3111: How to level on the current during BCI substitution tests?
  • #3178: 500W1000 is not switched to OPERATE in version 5.3.18
  • #3182: RadiMation sets Anritsu 2623B to wrong settings
  • #3185: Driver: SUNOL CONT94
  • #3231: 150A400 "needs" a reset command when leaving manual mode
  • #3248: Selecting square wave AM modulation does not change the waveform on the Agilent 83640B, and it doesn't raise an error
  • #3264: Speaker of the HP 8591E analyser is sometimes activated
  • #3318: files are getting bigger, customer can't open them anymore
  • #3320: wrong remeasured frequencies in Radiated Immunity (singleband)- Manual Mode
  • #3334: Peak not detected when zooming in Multiband Emisson graph
  • #3337: How is the Ramp in Mult. Im. configured?
  • #3338: Problems starting Cond. Im. Multiband
  • #3379: Final measurement with QP detector using a PMM 9030 receiver results in a software crash
  • #3391: Clean install crashes when opening device driver configuration window
  • #3392: clean install sets the configuration directory incorrectly to Radimationconfdvdr
  • #3399: HP8657 signal generator driver does not support 400 Hz AM.
  • #3405: Rs-232 support for the configurable switch matrix
  • #3406: signal generator sml03 sometimes is not responding to opc?
  • #3408: Driver FC03, turn table and antenna tower controller
  • #3413: Carad driver is reporting mcItestData is nothing
  • #3417: Using the ESU tracking generator as a signal generator for the attenuation measurement
  • #3422: PSA preselector need option to set the preselector off so that you can measure above the 1 GHz.
  • #3426: Driver: Agilent 87405C pre amplifier
  • #3428: Driver: HP 59307
  • #3426: Driver: ESV-Z1 VHF Current Probe
  • #3430: Driver: LISN PMM L3-64
  • #3431: Driver: ENI 603L and 411LA
  • #3432: Driver: Solar 9142-1N and 9719-1N
  • #3433: Driver: EATON 93686-8L and EM test HFK
  • #3437: LISN ESU User port
  • #3438: TDS220 is only available as an AD convertor not as an oscilloscope
  • #3448: Conducted Immunity Fixed Current with activated performance threshold is crashing
  • #3460: Format string does not work correctly with %lf and a double.
Known Problems
R&S spectrum analysers/receivers communication problems
  1. Open the configuration window of RadiMation®
  2. Select the Device Drivers tab
  3. Select Receivers/Spectrum Analysers
  4. Select the R&S receiver
  5. Press Edit to open the device driver configuration window
  6. Press the Advanced button
    • The GPIB configuration dialog of the R&S device driver will be shown
  7. Activate the Show advanced Settings checkbox
    • The GPIB Configuration dialog will show a Wait time box
  8. Specify 3000 us in the Wait time box
  9. Close all windows, and repeat these steps for each and every R&S receiver.
Previous RadiMation® version raises a run-time error
Windows 8, Windows 7, XP, Vista and older are not supported