m (→‎VISA Settings: Fix grammar)
(35 intermediate revisions by 3 users not shown)
Line 133: Line 133:
===EUT Controller===
===Auxiliary equipment===
When a Auxiliary equipment can be remotely controlled the specific window will appear. Please view chapter IEEE or RS232 setting for a complete description. When the Auxiliary equipment can not be remotely controlled the message This Device cannot be configured will appear.
== Communication interfaces and protocols ==
Different test and measurement instrumentation devices, can be controlled by different electronic communication buses and connectors, with on top of that different communication protocols. Which communication stream setting should be used to control a measurement instruments, cannot be easily based alone on the connector that is used to connect the measurement instrument. The following list shows the possible communication interfaces and protocols that are possible with a specific connector. This table can help to identify the actual communication protocol that is used. The manual of the test and measurement instrument also often lists the communication protocol that should be used, with similar terms, as is shown in the table. Based on the actual communication protocol the best communication stream to use is suggested.
{|class="wikitable"
! Connector || Communication protocol || Suggested communication stream || Alternative communication stream
| [[GPIB]] / [[wikipedia:IEEE-488|IEEE-488]] || GPIB || [[#GPIB_Setting|GPIB]] || [[#VISA_Setting|VISA]] (with GPIB sub option)
|rowspan="5"| Ethernet / [[wikipedia:Rj45|RJ45]] || VXI-11 || [[#VISA_Setting|VISA]] (with LAN sub option) ||
| [[wikipedia:High Speed LAN Instrument Protocol|HiSLIP]] || [[#VISA_Setting|VISA]] (with LAN sub option) ||
| [[wikipedia:Network_socket#Types|Raw socket]] || [[#TCPIP_Setting|TCPIP]] || [[#VISA_Setting|VISA]] (with Visa-ID sub option with <code>::SOCKET</code> identifier)
| HTTP || [[#HTTP_Setting|HTTP]] ||
| UDP || [[#UDP_Setting|UDP]] ||
|rowspan="4"| [[wikipedia:USB|USB]]|| [[wikipedia:Virtual COM port|Virtual COM Port (VCP)]] (RS-232) || [[#RS-232_Setting|RS-232]] || [[#VISA_Setting|VISA]] (with RS-232 sub option)
| USBTMC (Test and Measurement Class) || [[#VISA_Setting|VISA]] (with Visa-ID sub option) ||
| Raditeq USB devices || [[#USB_Setting|USB]] || [[#RS-232_Setting|RS-232]]<BR/>[[#VISA_Setting|VISA]] (with RS-232 sub option)
| Device specific USB ||  || should be implemented device specifically
| [[wikipedia:D-subminiature|Sub-D9]] / [[wikipedia:RS-232|RS-232]] || RS-232 || [[#RS-232_Setting|RS-232]] || [[#VISA_Setting|VISA]] (with RS-232 sub option)
|rowspan="2"| other software || [[wikipedia:Network_socket#Types|Raw socket]] || [[#TCPIP_Setting|TCPIP]] || [[#VISA_Setting|VISA]] (with Visa-ID sub option with <code>::SOCKET</code> identifier)
The details for the configuration of each possible communication stream is described in the next section.
== Communication settings ==
=== GPIB Setting ===
[[Image:GPIB DeviceStream Configuration.png]]
{{ScreenElementDescription|Primary Address|The primary address specifies the address of the device, so if the device you want to configure is on address 20 you enter 20. Please consult the manual of the device how to determine the GPIB address of the device.}}
{{ScreenElementDescription|Use advanced configuration|Allows the end-user to enable the more advanced configuration parameters. Those advanced option should only be used if the default settings are not sufficient.}}
{{ScreenElementDescription|GPIB Board|With GPIB Board you can specify the GPIB board that is used. The default value is 0 and can only be different when multiple GPIB boards are present.}}
{{ScreenElementDescription|Secondary Address|The second address is 0 by default, and should only be changed when needed. For further information please look in the help of National Instruments 488.2.}}
{{ScreenElementDescription|GPIB Delay|GPIB delay is the delay between GPIB reading and writing actions. Some older IEEE 488.2 machines have difficulty communicating with fast PCs (> 2.5 GHz). This is most of the time noticeable when a driver is sometimes working and sometimes gives a GPIB (EABO or TIMO) error. These errors are most of the time, generated randomly. Specifying a GPIB delay time of 3000 uSeconds can fix these random errors. Run the test again and see if the problem has disappeared. If the problem has disappeared then your problem was timing, if not please contact your reseller and report the problems you are having.}}
{{ScreenElementDescription|Clear device during initialisation|When checked, a low level GPIB command to reset the device will be sent to the device during initialisation.}}
{{ScreenElementDescription|Readdress device|When checked, a low level GPIB command to select the correct address will be transmitted every time a command is sent to the device.}}
=== HTTP Setting ===
The HTTP communication settings are used to specify the HTTP communication parameters, for the communication with a measurement device over HTTP. In this HTTP communication mode, the [[wikipedia:Hypertext Transfer Protocol|HyperText Transfer Protocol (HTTP)]] is used, and actual GET requests are performed.
[[File:HTTPDeviceStreamConfiguration.png]]
{{ScreenElementDescription|Address|Allows to specify the IP-address or the (FQDN) hostname and the socket port number of the measurement device. This is normally done with a string like: "tcpip://<address-or-name>:<port-number>", where '<address-or-name>' is the IP address or hostname, and '<port-number>' is the socket port number on which the connection should be initiated. By default the socket port number is 80, but this can be changed to another port if desired.}}
If for example in a configurable device driver this HTTP communication stream is used, and the specified command is for example "SETA=1" (which set relay A of a Mini-Circuits relay to position 1), an actual "GET /SETA=1" HTTP request is transmitted to the specified hostname and socket port that is configured for the {{ScreenElement|Address}}.
=== Manual Setting ===
The manual communication stream can be used to show message boxes on the screen for the end-user, to manually perform some operation on a device. The commands that are specified in the device driver that is using the manual communication stream, are thus the messages that are shown to the end-user in a message box, for example with the instruction to change a setting on a test and measurement device.
No further configuration in the manual communication stream is needed.
[[File:ManualDeviceStreamConfiguration.png]]
===RS-232 Setting===
The RS-232 Communication settings are used to specify the RS-232 communication parameters, for the communication with a measurement device over RS-232. A lot of modern measurement devices are connected by USB, however very often then a Virtual RS-232 COMPort (VCP) is generated over the USB connection. In that situation also the RS-232 communication settings should be configured.
The RS-232 communication settings specify the serial port parameters when communicating with a measurement device over RS-232. Although many modern devices connect via USB, they often appear as a virtual COM port (VCP). In that case you still configure the RS-232 settings as below.
[[Image:RS-232 DeviceStream Configuration.png]]
Line 144: Line 215:
{{ScreenElementDescription|COM Port|The COM port which {{RadiMation}} has to be used to communicate with the measurement device.}}
{{ScreenElementDescription|Baudrate|Allows to configure the baudrate that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
{{ScreenElementDescription|Baud rate|Allows to configure the baud rate that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
{{ScreenElementDescription|Data bits|Allows to configure the number of data bits that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
{{ScreenElementDescription|Parity|Allows to configure the parity bit that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
Line 153: Line 224:
=== RSIB Setting ===
The RSIB communication settings are used to communicate with an ethernet connected measurement device through the [[Rohde & Schwarz]] RSISB32.DLL, which implements a VISA VXI-11 compatible layer.
[[File:RSIBDeviceStreamConfiguration.png]]
{{ScreenElementDescription|RSIB Address|Allows to specify RSIB identifier of the device that should be controlled.}}
=== TCPIP Setting ===
The TCPIP communication settings are used to specify that a (raw-)socket based [[wikipedia:TCP/IP|TCP/IP]] connection to a measurement device should be used.
[[Image:TCPIP DeviceStream Configuration.png]]
{{ScreenElementDescription|Address|Allows to specify the IP-address or the (FQDN) hostname and the socket port number of the measurement device. This is normally done with a string like: "tcpip://<address-or-name>:<port-number>", where '<address-or-name>' is the IP address or hostname, and '<port-number>' is the socket port number on which the connection should be initiated. If the socket port number is a fixed port number, it is already shown as the default value, and it will be automatically added when no socket port number is specified.}}
=== VISA Settings ===
=== UDP Setting ===
The VISA Communication settings are used to specify the VISA communication parameters, for the communication with a measurement device. The VISA library is a higher level communication library that supports different kinds of communication methods. {{RadiMation}} doesn't provide a VISA library itself, and thus requires that a VISA library from another supplier like [[National Instruments]] or [[Keysight Technologies]] is installed. The VISA manager that is provided by that VISA library can be used to determine the correct VISA resource. The selected VISA configuration thus also has an influence if a measurement device is controlled by GPIB, RS-232, LAN, USB-TMC, VXI11, network-socket or another communication method .
The UPD communication settings are used to specify that an UDP based communication over ethernet should be used to control the measurement.
[[Image:VISA DeviceStream Configuration.png]]
[[Image:UDPDeviceStreamConfiguration.png]]
{{ScreenElementDescription|Alias|Allows to specify a VISA Alias that should be used for the communication with the measurement device. Any Alias that is supported by VISA is accepted. The correct Alias can be determined by using the VISA manager (eg. [[National Instruments]] MAX) that is installed on the PC.}}
{{ScreenElementDescription|GPIB|Allows to specify that GPIB should be used for the communication with the measurement device. The GPIB address of the measurement device should be specified.}}
{{ScreenElementDescription|LAN|Allows to specify that a VXI11 or LXI connection should be used for the communication with the measurement device. It is possible to specify the IP-address or the (FQDN) hostname of the measurement device.}}
{{ScreenElementDescription|RS-232|Allows to specify that a RS232 (ASRL in VISA terms) connection should be used for the communication with the measurement device. The COM port of the measurement device should be specified.}}
{{ScreenElementDescription|Visa-ID|Allows to specify a VISA Identifier that should be used for the communication with the measurement device. Any VISA Identifier that is supported by VISA is accepted. If one of the other communication methods is selected the corresponding VISA Identifier is also shown in this setting. Often used VISA identifiers are:
* <code>TCPIP[board]::<IP-address>::INSTR</code>: VXI11 or LXI ethernet communication with the device with '<IP-address>'.
* <code>TCPIP[board]::<IP-address>::<port-number>::SOCKET</code>: Raw socket based ethernet communication on the port '<port-number>' with the the device '<IP-address>'
* <code>GPIB[board]::<primary address>::INSTR</code>: GPIB communication to the GPIB device with the address '<primary address>'
* <code>ASRL<port>::INSTR</code>: Serial communcation on COM-port '<port>'
=== USB Settings ===
=== USB Setting ===
The USB Communication settings are used to specify that an USB connection to a DARE!! Instruments measurement device is used. It is not possible to use this USB communication setting for devices that simulate a Virtual COMPort (VCP) over an USB connection. The RS232 Settings should be used for such a kind of measurement device.
The USB Communication settings are used to specify that a USB connection to a [[Raditeq]] or [[DARE!! Instruments]] measurement device is used. For devices that appear as a Virtual COM port (VCP) over USB, configure the [[Chapter_15#RS-232_Setting|RS-232 setting]] rather than the USB setting.
{{ScreenElementDescription|Device Identifier|The device identifier (which is an unique identifier of 8 groups of digits) that identifies the measurement device that is connected over USB.}}
{{ScreenElementDescription|Device Identifier|The device identifier (which is a unique identifier of 8 groups of digits) that identifies the measurement device that is connected over USB.}}
{{ScreenElementDescription|Detect|Will automatically determine the correct {{ScreenElement|Device Identifier}} for the measurement device that is connected.}}
=== TCPIP Settings ===
=== VISA Setting ===
The TCPIP Communication settings are used to specify that a socket based TCPIP connection to a measurement device should be used.
The VISA Communication settings are used to specify the VISA communication parameters, for the communication with a measurement device. The VISA library is a higher level communication library that supports different kinds of communication methods. {{RadiMation}} doesn't provide a VISA library itself, and thus requires that a VISA library from another supplier like [[National Instruments]] or [[Keysight Technologies]] is installed. The VISA manager that is provided by that VISA library can be used to determine the correct VISA resource. The selected VISA configuration thus also has an influence if a measurement device is controlled by GPIB, RS-232, LAN, USBTMC, VXI-11, network-socket or another communication method .
{{ScreenElementDescription|Alias|Allows to specify a VISA Alias that should be used for the communication with the measurement device. Any Alias that is supported by VISA is accepted. The correct Alias can be determined by using the VISA manager (eg. [[National Instruments]] MAX or [[Keysight Technologies]] connection expert) that is installed on the PC.}}
{{ScreenElementDescription|LAN|Allows to specify that a VXI-11 or LXI connection should be used for the communication with the measurement device. It is possible to specify the IP-address or the (FQDN) hostname of the measurement device.}}
* <code>TCPIP[board]::<IP-address>::INSTR</code>: VXI-11 or LXI ethernet communication with the device with '<IP-address>'.
==Configurable device drivers==
{{RadiMation}} allows the user to create its own device drivers for test and measurement equipment, which is (not yet) implemented in the standard device driver list.
{{RadiMation}} allows the user to create their own device drivers for test and measurement equipment, which is (not yet) implemented in the standard device driver list.
Line 215: Line 289:
# Select the driver called “Configurable xx” (i.e. if you want to make a signal generator device driver you would select the “Configurable signal generator” device driver).
# Enter a description for the device driver (for example the type number of the generator) and press OK.
# The name of the device driver will be added in the available device driver’s list.
# The name of the device driver will be added in the available device drivers list.
# Select the new device driver from the available device device’s list.
# Select the new device driver from the available device drivers list.
# Press the “Edit” button.
# A configuration screen for the device driver will appear. For active controlled equipment all required control commands for the device must be entered. Refer to the operating manual of the equipment for these codes.
Line 224: Line 298:
===Generic settings===
{{ScreenElementDescription|Start and stop frequency|The default value for the start frequency is 1 Hz and for the stop frequency is 120 GHz. The creator of the device is encouraged to set the correct start and stop frequency. If the start and stop frequency are set correctly, {{RadiMation}} can warn the test engineer when he/she want to use the device out of its valid frequency range.}}
{{ScreenElementDescription|Start and stop frequency|The default value for the start frequency is 1 Hz and for the stop frequency is 120 GHz. The creator of the device is encouraged to set the correct start and stop frequency. If the start and stop frequency are set correctly, {{RadiMation}} can warn the test engineer when they want to use the device out of its valid frequency range.}}
{{ScreenElementDescription|Reset|In the reset window you need to specify the string that the software needs to send when it want to reset the device. For example *RST is commonly used reset string. If you don’t know the string then leave this window blank, and make sure that the device in a neutral state.}}
{{ScreenElementDescription|Init|In the init window you need to specify the string that the software needs to send when it want to Init  the device. For example *RST is commonly used reset string.}}
{{ScreenElementDescription|Get ID|In the Get ID window you need to specify the string that the software needs to send when it want to get the ID string of the device. For example *IDN? is commonly used Get ID string.}}
{{ScreenElementDescription|Get ID|In the Get ID window you need to specify the string that the software needs to send when it wants to get the ID string of the device. For example *IDN? is commonly used Get ID string.}}
{{ScreenElementDescription|Returned ID|In the Returned ID window you need to specify the string that the software will receive so that it know that it has the right device. For example "Hewlett_Packard,8643A," can be used for the Hewlett Packard 8643A. If you leave this window empty then all returned strings are accepted.}}
{{ScreenElementDescription|Returned ID|In the Returned ID window you need to specify the string that the software will receive so that it knows that it has the right device. For example "Hewlett_Packard,8643A," can be used for the Hewlett Packard 8643A. If you leave this window empty then all returned strings are accepted.}}
Line 234: Line 308:
[[Image:Configurable Signal Generator Window.png]]
{{ScreenElementDescription|Set Frequency|Set Frequency is the string that needs to send to set the signal generator frequency. The unit is in MHZ, so the string should be made for MHz. }}
{{ScreenElementDescription|Set Frequency|Set Frequency is the string that needs to be sent to set the signal generator frequency. The unit is in MHZ, so the string should be made for MHz. }}
'''Example:''' “FRQ__freq__MHZ”.
Line 248: Line 322:
__freq__ will be replaced by the value that {{RadiMation}} want to set the signal generator to.  The “__” of  “__freq__” are two “_”.
{{ScreenElementDescription|Set Carrier|Set Carrier is the string that needs to send to set the signal generator carrierlevel. The unit is in dBm, so the string should be made for dBM.}}
{{ScreenElementDescription|Set Carrier|Set Carrier is the string that needs to be sent to set the signal generator carrierlevel. The unit is in dBm, so the string should be made for dBM.}}
'''Example:''' “:SOURCE:POWER __carrier__ DBM”.
__carrier__ will be replaced by the value that {{RadiMation}} want to set the signal generator to.  The “__” of  “__carrier__ ” are two “_”.
__carrier__ will be replaced by the value that {{RadiMation}} wants to set the signal generator to.  The “__” of  “__carrier__ ” are two “_”.
{{ScreenElementDescription|Carrier on|Carrier on is the string that needs to send to set the signal generator carrier on.}}
{{ScreenElementDescription|Carrier on|Carrier on is the string that needs to be sent to set the signal generator carrier on.}}
'''Example:''' “OUTPUT ON”.
{{ScreenElementDescription|Carrier off|Carrier off is the string that needs to send to set the signal generator carrier off.}}
{{ScreenElementDescription|Carrier off|Carrier off is the string that needs to be sent to set the signal generator carrier off.}}
'''Example:''' “OUTPUT OFF”.
{{ScreenElementDescription|AM on|AM on is the string that needs to send to set the internal AM generator of the signal generator. The unit of frequency is KHz and the unit of Modulation Depth is %.}}
{{ScreenElementDescription|AM on|AM on is the string that needs to be sent to set the internal AM generator of the signal generator. The unit of frequency is KHz and the unit of Modulation Depth is %.}}
'''Example:''' “SOURCE2:FREQ  __freq__  KHZ;SOURCE:AM:DEPTH  __depth__ PCT;SOURCE:AM:STATE ON”.
Line 269: Line 343:
“__freq__  ” are two “_”.
{{ScreenElementDescription|AM off|AM off is the string that needs to send to set the internal am modulation of the signal generator off.}}
{{ScreenElementDescription|AM off|AM off is the string that needs to be sent to set the internal am modulation of the signal generator off.}}
'''Example:''' “SOURCE:AM:STATE OFF”.
{{ScreenElementDescription|External on|External on is the string that needs to send to set the modulator to the external input.}}
{{ScreenElementDescription|External on|External on is the string that needs to be sent to set the modulator to the external input.}}
'''Example:''' “SOURCE:AM:SOURCE EXT”.
{{ScreenElementDescription|External off|External off is the string that needs to send to set the modulator to the internal input.}}
{{ScreenElementDescription|External off|External off is the string that needs to be sent to set the modulator to the internal input.}}
'''Example:''' “SOURCE:AM:SOURCE INT”.
{{ScreenElementDescription|PM on|PM on is the string that needs to send to set the internal PM generator of the signal generator. The unit of frequency is KHz and the unit of duty cycle is %.}}
{{ScreenElementDescription|PM on|PM on is the string that needs to be sent to set the internal PM generator of the signal generator. The unit of frequency is KHz and the unit of duty cycle is %.}}
'''Example:''' “SOURCE2:FREQ  __freq__  KHZ;SOURCE:PM:DUTY  __duty__ PCT;SOURCE:PM:STATE ON”.
Line 288: Line 362:
{{ScreenElementDescription|PM off|PM off is the string that needs to send to set the internal PM modulation of the signal generator off.}}
{{ScreenElementDescription|PM off|PM off is the string that needs to be sent to set the internal PM modulation of the signal generator off.}}
'''Example:''' “SOURCE:PM:STATE OFF”.
{{ScreenElementDescription|Sine Wave|Sine wave is the string that needs to be send when {{RadiMation}} wants to set the wave form of the internal source to sine.}}
{{ScreenElementDescription|Sine Wave|Sine wave is the string that needs to be sent when {{RadiMation}} wants to set the wave form of the internal source to sine.}}
'''Example:''' “SOURCE2:FUNC SIN”
{{ScreenElementDescription|Square Wave|Square wave is the string that needs to be send when {{RadiMation}} wants to set the wave form of the internal source to square.}}
{{ScreenElementDescription|Square Wave|Square wave is the string that needs to be sent when {{RadiMation}} wants to set the wave form of the internal source to square.}}
'''Example:''' “SOURCE2:FUNC SQU”
Line 310: Line 384:
[[Image:Configurable Power Meter Window.png]]
{{ScreenElementDescription|Set frequency|Set Frequency is the string that needs to send to set the powermeter frequency. The unit is in MHz, so the string should be made for MHz.}}
{{ScreenElementDescription|Set frequency|Set Frequency is the string that needs to be sent to set the powermeter frequency. The unit is in MHz, so the string should be made for MHz.}}
Line 324: Line 398:
'''Example:''' “X1”
'''Example:''' “X1”
{{ScreenElementDescription|Result Format|Result Format is the string that {{RadiMation}} needs to use to decode the value from the string send by the powermeter. The return value is interpreted in dBm.}}
{{ScreenElementDescription|Result Format|Result Format is the string that {{RadiMation}} needs to use to decode the value from the string sent by the powermeter. The return value is interpreted in dBm.}}
'''Example:''' “__result__”
Line 335: Line 409:
'''Example:''' “O1”
'''Example:''' “O1”
{{ScreenElementDescription|Duration|Duration is the time that {{RadiMation}} waits so that the powermeter can zero properly. Make sure that this time is big enough, an incorrect value may result in unpredictable result.}}
{{ScreenElementDescription|Duration|Duration is the time that {{RadiMation}} waits so that the powermeter can zero properly. Make sure that this time is big enough, an incorrect value may result in unpredictable results.}}
=== Configurable AD converter ===
Line 375: Line 449:
{{:Configurable EUT Controller}}
{{:Configurable equipment}}
==Messages ==
==Messages ==
Line 381: Line 458:
[[Image:Amplifier Device Driver Can Not Be Configured Window.png]]
===Unknown Device Driver===
Line 401: Line 478:
==How to Report an Error==
When encountering a problem with the software you might would like to report it to the RadiMation support.
When encountering a problem with the software you might like to report it to the RadiMation support.
The RadiMation error popups, contains detailed section. This can be expanded with details button on the error popup.
The RadiMation error popups, contains a detailed section. This can be expanded with the details button on the error popup.
In the expanded detailed error popup, a Report Error button is present. This allows Error Reporting to the RadiMation support within RadiMation.
More information about Error Reporting can be found here:
Device drivers[edit]
Virtual device drivers[edit]
  • Usable frequency range
  • Maximum input power
Signal generator[edit]
Amplifier[edit]
Antenna[edit]
Coupler[edit]
Powermeter[edit]
Measuring: Powermeters which do not have a RMS detector, should not be used in Fast constant mode.
Field Sensor[edit]
Axis setting panel[edit]
AD converter[edit]
Injection device[edit]
Calibration Jigs[edit]
Current Sensor[edit]
Pre Amplifiers[edit]
LISN[edit]
Turn Table[edit]
Antenna Tower[edit]
Absorbing Clamps[edit]
Clamp positioner[edit]
Cables[edit]
Switch matrix[edit]
EUT Controller[edit]
Auxiliary equipment[edit]
Connector Communication protocol Suggested communication stream Alternative communication stream
GPIB / IEEE-488 GPIB GPIB VISA (with GPIB sub option)
Ethernet / RJ45 VXI-11 VISA (with LAN sub option)
HiSLIP VISA (with LAN sub option)
Raw socket TCPIP VISA (with Visa-ID sub option with ::SOCKET identifier)
HTTP HTTP
UDP UDP
USB Virtual COM Port (VCP) (RS-232) RS-232 VISA (with RS-232 sub option)
USBTMC (Test and Measurement Class) VISA (with Visa-ID sub option)
Raditeq USB devices USB RS-232
Device specific USB should be implemented device specifically
Sub-D9 / RS-232 RS-232 RS-232 VISA (with RS-232 sub option)
other software Raw socket TCPIP VISA (with Visa-ID sub option with ::SOCKET identifier)
Communication settings[edit]
GPIB Setting[edit]
ScreenElementDescription.svg Primary Address The primary address specifies the address of the device, so if the device you want to configure is on address 20 you enter 20. Please consult the manual of the device how to determine the GPIB address of the device.
ScreenElementDescription.svg Use advanced configuration Allows the end-user to enable the more advanced configuration parameters. Those advanced option should only be used if the default settings are not sufficient.
ScreenElementDescription.svg Secondary Address The second address is 0 by default, and should only be changed when needed. For further information please look in the help of National Instruments 488.2.
ScreenElementDescription.svg GPIB Delay GPIB delay is the delay between GPIB reading and writing actions. Some older IEEE 488.2 machines have difficulty communicating with fast PCs (> 2.5 GHz). This is most of the time noticeable when a driver is sometimes working and sometimes gives a GPIB (EABO or TIMO) error. These errors are most of the time, generated randomly. Specifying a GPIB delay time of 3000 uSeconds can fix these random errors. Run the test again and see if the problem has disappeared. If the problem has disappeared then your problem was timing, if not please contact your reseller and report the problems you are having.
ScreenElementDescription.svg Clear device during initialisation When checked, a low level GPIB command to reset the device will be sent to the device during initialisation.
ScreenElementDescription.svg Readdress device When checked, a low level GPIB command to select the correct address will be transmitted every time a command is sent to the device.
HTTP Setting[edit]
ScreenElementDescription.svg Address Allows to specify the IP-address or the (FQDN) hostname and the socket port number of the measurement device. This is normally done with a string like: "tcpip://<address-or-name>:<port-number>", where '<address-or-name>' is the IP address or hostname, and '<port-number>' is the socket port number on which the connection should be initiated. By default the socket port number is 80, but this can be changed to another port if desired.
Manual Setting[edit]
RS-232 Setting[edit]
ScreenElementDescription.svg COM Port The COM port which RadiMation® has to be used to communicate with the measurement device.
ScreenElementDescription.svg Baud rate Allows to configure the baud rate that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.
RSIB Setting[edit]
ScreenElementDescription.svg RSIB Address Allows to specify RSIB identifier of the device that should be controlled.
TCPIP Setting[edit]
UDP Setting[edit]
USB Setting[edit]
ScreenElementDescription.svg Device Identifier The device identifier (which is a unique identifier of 8 groups of digits) that identifies the measurement device that is connected over USB.
VISA Setting[edit]
ScreenElementDescription.svg Alias Allows to specify a VISA Alias that should be used for the communication with the measurement device. Any Alias that is supported by VISA is accepted. The correct Alias can be determined by using the VISA manager (eg. National Instruments MAX or Keysight Technologies connection expert) that is installed on the PC.
ScreenElementDescription.svg LAN Allows to specify that a VXI-11 or LXI connection should be used for the communication with the measurement device. It is possible to specify the IP-address or the (FQDN) hostname of the measurement device.
  • TCPIP[board]::<IP-address>::INSTR: VXI-11 or LXI ethernet communication with the device with '<IP-address>'.
  • TCPIP[board]::<IP-address>::<port-number>::SOCKET: Raw socket based ethernet communication on the port '<port-number>' with the the device '<IP-address>'
  2. Press the “New” button.
  5. The name of the device driver will be added in the available device drivers list.
  6. Select the new device driver from the available device drivers list.
  7. Press the “Edit” button.
Generic settings[edit]
ScreenElementDescription.svg Start and stop frequency The default value for the start frequency is 1 Hz and for the stop frequency is 120 GHz. The creator of the device is encouraged to set the correct start and stop frequency. If the start and stop frequency are set correctly, RadiMation® can warn the test engineer when they want to use the device out of its valid frequency range.
ScreenElementDescription.svg Get ID In the Get ID window you need to specify the string that the software needs to send when it wants to get the ID string of the device. For example *IDN? is commonly used Get ID string.
ScreenElementDescription.svg Returned ID In the Returned ID window you need to specify the string that the software will receive so that it knows that it has the right device. For example "Hewlett_Packard,8643A," can be used for the Hewlett Packard 8643A. If you leave this window empty then all returned strings are accepted.
ScreenElementDescription.svg Set Frequency Set Frequency is the string that needs to be sent to set the signal generator frequency. The unit is in MHZ, so the string should be made for MHz.
Example: “FRQ__freq__MHZ”.
Example: “FRQ__freqHz__HZ”.
Example: “FRQ__freqkHz__KHZ”.
Example: “FRQ__freqMHz__MHZ”.
Example: “FRQ__freqGHz__GHZ”.
ScreenElementDescription.svg Set Carrier Set Carrier is the string that needs to be sent to set the signal generator carrierlevel. The unit is in dBm, so the string should be made for dBM.
ScreenElementDescription.svg Carrier on Carrier on is the string that needs to be sent to set the signal generator carrier on.
Example: “OUTPUT ON”.
ScreenElementDescription.svg Carrier off Carrier off is the string that needs to be sent to set the signal generator carrier off.
Example: “OUTPUT OFF”.
ScreenElementDescription.svg AM on AM on is the string that needs to be sent to set the internal AM generator of the signal generator. The unit of frequency is KHz and the unit of Modulation Depth is %.
ScreenElementDescription.svg AM off AM off is the string that needs to be sent to set the internal am modulation of the signal generator off.
ScreenElementDescription.svg External on External on is the string that needs to be sent to set the modulator to the external input.
ScreenElementDescription.svg External off External off is the string that needs to be sent to set the modulator to the internal input.
ScreenElementDescription.svg PM on PM on is the string that needs to be sent to set the internal PM generator of the signal generator. The unit of frequency is KHz and the unit of duty cycle is %.
ScreenElementDescription.svg PM off PM off is the string that needs to be sent to set the internal PM modulation of the signal generator off.
ScreenElementDescription.svg Sine Wave Sine wave is the string that needs to be sent when RadiMation® wants to set the wave form of the internal source to sine.
Example: “SOURCE2:FUNC SIN”
ScreenElementDescription.svg Square Wave Square wave is the string that needs to be sent when RadiMation® wants to set the wave form of the internal source to square.
Example: “SOURCE2:FUNC SQU”
Configurable Amplifier[edit]
Main[edit]
Operation[edit]
Example[edit]
1 = RF ON
1 = BAND 2
1 = LINE ON
Configurable Antenna[edit]
Configurable Powermeter[edit]
ScreenElementDescription.svg Set frequency Set Frequency is the string that needs to be sent to set the powermeter frequency. The unit is in MHz, so the string should be made for MHz.
Example: “P1,U1”.
Example: “X1”
ScreenElementDescription.svg Result Format Result Format is the string that RadiMation® needs to use to decode the value from the string sent by the powermeter. The return value is interpreted in dBm.
Example: “__result__”
Example: “O1”
ScreenElementDescription.svg Duration Duration is the time that RadiMation® waits so that the powermeter can zero properly. Make sure that this time is big enough, an incorrect value may result in unpredictable results.
Channels[edit]
Read Back examples[edit]
Communication[edit]
Configurable LISN[edit]
Configurable Turn Table[edit]
Configurable commands[edit]
Check device[edit]
Device is initialised[edit]
Device is deinitialised[edit]
Some examples are:
Communication[edit]
Identification[edit]
Antenna Tower[edit]
Antenna Tower Advanced[edit]
Currently unavailable.
Configurable Cable[edit]
sending the next command.
keyword inserted value
__calculatedvoltage__ The calculated voltage expressed in V.
Auxiliary equipment[edit]
Messages[edit]
Unknown Device Driver[edit]
Device not connected[edit]
How to Report an Error[edit]
AD convertors[edit]
Type of communication: IEEE.
Channels 1:
Hewlett Packard 34401A[edit]
Channel 1
Type of measurement: AC Volt
Minimum value: 0Volt
Maximum value: 1 kVolt.
Channel 2:
Minimum value: 0 Amp.
Maximum value: 3 Amp.
Channel 3
Type of measurement: DC Volt
Channel 4:
Maximum value: 3 Amp. (rms)
Channel 5
Minimum value: 0 Ω
Maximum value: 100 MΩ
Channel 6
Minimum value: 0 Hz.
Maximum value: 300 kHz
Channel 7 and 8
Not used
Hewlett Packard 54600[edit]
Channel 1:
Type of measurement: V Max.
Maximum value: 1000 Volt.
Channel 2:
Type of measurement: V Min
Minimum value: 0 Volt.
Channel 3
Channel 4:
Channel 5
Minimum value: 0 Hz
Maximum value: 100 kHz
Channel 6
Type of measurement: Period
Minimum value: 0 ms
Maximum value: 10000 ms
Channel 7
Minimum value: 0 ms
Channel 8
Minimum value: 0 ms
Hewlett Packard 3562A[edit]
Type of communication: GPIB
Channels
Minimum value: 0 dB
Maximum value: 100 dB
Hewlett Packard 59313[edit]
Channel 1
Minimum value: 0
Maximum value: 2048
Channel 2:
Minimum value: 0
Maximum value: 2048
Channel 3
Minimum value: 0
Maximum value: 2048
Channel 4:
Minimum value: 0
Maximum value: 2048
Channel 5 to 8
Not used.
Channels
Marconi 2305[edit]
Channel 1
Minimum value: 0
Maximum value: 1000 MHz
Channel 2:
Minimum value: 0
Maximum value: 1000
Channel 3 to 8
Not used.
Fluke 45 AC Current[edit]
Channels
Minimum value: 0 mA
Maximum value: 10.000 mA
Fluke 45 DC Current[edit]
Channels
Minimum value: 0 mA
Fluke 45 AC Voltage[edit]
Channels
Minimum value: 0 mV
Maximum value: 1.000.000 mV
Fluke 45 DC Voltage[edit]
Channels
Minimum value: 0 mV
Fluke 45 Frequency[edit]
Channels
Minimum value: 0 Hz
Maximum value: 1.000.000 Hz
Fluke 45 Resistance[edit]
Channels
Minimum value: 0 Ω
Maximum value: 100.000.000 Ω
Channel 1
Maximum value: 353.55 Volt.
Channel 2:
Channel 3
Channel 4:
Maximum value: 707.1 Volt
Channel 5
Type of measurement: RMS
Maximum value: 250 Volt
Channel 6
Maximum value: 200 MHz
Channel 7 and 8
Not used
Fluke 8840A AC Current[edit]
Channels
Minimum value: 0 mA
Fluke 8840A DC Current[edit]
Channels
Minimum value: 0 mA
Fluke 8840A AC Voltage[edit]
Channels
Minimum value: 0 mV
Fluke 8840A DC Voltage[edit]
Channels
Minimum value: 0 mV
Channels
Minimum value: 0 mA
Channels
Minimum value: 0 mA
Channels
Minimum value: 0 mV
Maximum value: 300.000 mV
Channels
Minimum value: 0 mV
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 8
Minimum value: 0
Maximum value: 16383
EIP 575[edit]
Channel 1
Minimum value: 0 Hz
Maximum value: 10 kHz
Channel 2:
Channel 3
Minimum value: 0 Hz
Maximum value: 1 MHz.
Channel 4:
Maximum value: 10 MHz.
Channel 5
Minimum value: 0 Hz
Maximum value: 100 MHz
Channel 6
Minimum value: 0 Hz
Maximum value: 1 GHz
Channel 7
Minimum value: 0 Hz
Maximum value: 10 GHz
Channel 8
Minimum value: 0 Hz
Maximum value: 100 GHz
Channel 1 to 8