How to perform a MIL-STD-461 CS109, Conducted susceptibility, structure current test[edit]
This application note explains how the MIL-STD-461 CS109, Conducted susceptibility, structure current test can be performed with RadiMation®
The exact requirements and test methods for the CS109 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to perform the CS109 test:
- Signal generator
- Amplifier
- Coupling Transformer
- Current sensor
- 0.5 Ohm resistor
- Isolation Transformer
Note:
If the current unit is shown as μA, it can be changed to dBμA by clicking Units, select Current and click Edit to change this.
Define the test level[edit]
Create a correction file for the test level[edit]
File
New
Correction
First create a correction file in RadiMation® and specify the test level (See 'FIGURE CS109-1. CS109 limit.').
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBμA.
File
Save correction
This correction file with the test level will be used during EUT testing.
EUT Testing[edit]
EUT Testing equipment[edit]
Device name |
Tab in testsite configuration window |
note
Signal Generator |
Devices 1 |
The signal generator to use
Amplifier |
Devices 1 |
The amplifier to use
Sensor powermeter |
Devices 2 |
The power meter or analyser to use for measuring the current
Current Sensor |
Devices 2 |
The current sensor to use with transfer factor attached to the driver
Injection device |
Devices 2 |
The injection device to use
| Cables
| ① |
Cable current -> power meter |
Cables |
Cable with a correction file specified for the cable loss
Configure the EUT test[edit]
To perform the test on the EUT, create a conducted immunity multiband test.
The conducted immunity multiband test, can be started by selecting from the menu:
Tests
Conducted Immunity
Multiband
Start
The start frequency of the test. For example 30 Hz.
End
The stop frequency of the test. For example 100 kHz.
Frequency Step
The frequency step, see Table III.
Dwell Time
The dwell time.
Frequency Change Mode
Constant.
Modulation
CS109 does not specify a modulation to use.
Test equipment
The equipment needed for the conducted immunity test.
Location Type
Injection clamp.
Test Level
Specify the current substitution method. Also add a "Measured Current Limit"
Inputs
No inputs are needed.
At TestLevel click Add to add a new Test level and select TestLevel - Current closed loop method
Description
The description of the test level.
Test level
Variable current, and specify the correction file for the applicable test level.
Tolerance
Specify the tolerance to use.
When all settings are configured press Start Test to run the test.
EUT Test Result[edit]
The RadiMation® Conducted Immunity multiband test can be used to perform the CS109 test with a current closed loop method test level and a variable test level specified in a correction file.