(69 intermediate revisions by the same user not shown)
__NOTOC__
= How to measure amplifier harmonics using RadiMation =
In this application note we explain how you can use RadiMation to perform accurate measurement of the harmonics that are generated by an amplifier.
This application note explains how {{RadiMation}} can be used to perform accurate measurement of the harmonics that are generated by an amplifier.
The measurement of the harmonics is often needed to ensure that the used amplifier in the setup is not generating too much power on multiples of the carrier frequency.
As the harmonics are thus present on higher frequencies than the CW signal, all the equipment that is connected to the output of the [[amplifier]] up to and including the used [[spectrum analyser]] should be able to be used on those higher frequencies. For an amplifier of 80 MHz - 1000 MHz, measuring up to the 6th harmonic, it is thus necessary that the measurement equipment is usable up to 6 GHz. The measurement equipment that should be able to measure those higher frequencies are shown in the blue areas in the equipment drawings below.
== Overview ==
== Overview ==
There are two ways to measure amplifier harmonics.
There are three possible setups of the used equipment to measure amplifier harmonics. Each method has its own advantages, and small differences in the used equipment. These setups and methods are explained in the following sections:
#Measure the initial power on the first harmonic and the rest of the harmonics with a spectrum analyser.
# [[#Measure the power of all harmonics using a spectrum analyser|Measure the power of all harmonics using a spectrum analyser]].
#Measure the initial power with a broadband power meter and measure the harmonics with a spectrum analyser.
# [[#Measure the broadband power with a powermeter and the harmonics using a spectrum analyser|Measure the broadband power with a powermeter and the harmonics using a spectrum analyser]].
# [[#Measure the actual field harmonics using a receive antenna|Measure the actual field harmonics using a receive antenna]].
Which equipment setup is used, is depending on the equipment that is selected in the testsite that is being used for the harmonics calibration.
* If the selected testsite has a device driver selected for the {{ScreenElement|Sensor power meter}}, then the '[[#Measure the broadband power with a powermeter and the harmonics using a spectrum analyser|Measure the broadband power with a powermeter and the harmonics using a spectrum analyser]]' method will be used.
* Otherwise, if the selected testsite has a device driver selected for the {{ScreenElement|Spectrum analyser}}, then the '[[#Measure the actual field harmonics using a receive antenna|Measure the actual field harmonics using a receive antenna]]' method will be used.
* In all other cases the '[[#Measure the power of all harmonics using a spectrum analyser|Measure the power of all harmonics using a spectrum analyser]]' method will be used.
=== Measure the power of all harmonics using a spectrum analyser ===
[[File:HarmonicsSpectrumAnalyserEquipment.png]]
A schema of the test setup:
To use only the spectrum analyser the Forward power meter in the testsite has to be selected to be the spectrum analyser device driver.
Keep in mind to not select any device for the sensor power meter on the 'Devices 2' tab.
[[File:Harmonics schematic.png]]
{|class="wikitable"
!# || Device name || Tab in testsite configuration window || suitable for harmonic frequencies || note
| || Signal Generator || Devices 1 || no
| || Amplifier || Devices 1 || no || This is the amplifier of which the harmonics are being measured.
| || Coupler || Devices 1 || yes ||
| || Forward power meter || Devices 1 || yes || The powermeter device driver for the analyser that is being used to measure the harmonics.
!colspan="5"|Cables
|① || Cable SG -> amplifier || Cables || no || Optional to correct for the cable loss between the signal generator and the amplifier
|② || Cable amplifier -> coupler || Cables || yes || Optional to correct for the cable loss between the amplifier and the coupler
|③ || Cable coupler -> forward power meter || Cables || yes || Optional to correct for the cable loss between the coupler and the forward power meter
== Configuring the testsite ==
=== Measure the broadband power with a powermeter and the harmonics using a spectrum analyser ===
=== Using a spectrum analyser ===
To use only the spectrum analyser we need to configure the the Forward power meter to be the spectrum analyser device driver in the testsite.
Keep in mind to not select any device for the sensor power meter on the second tab.
[[File:Harmonics_Configure_Spectrum_analyser_as_power_meter.PNG]]
Some standards like the [[IEC 61000-4-3]] describe that the harmonic content in the field strength should be specified. By using this setup, only the harmonic content in the power is measured, and the antenna factor of the used antenna should still be used to calculate the harmonic content in the field strength.
=== Using a broadband powermeter and a spectrum analyser ===
When using the powermeter to setup the required testlevel and a spectrum analyser to measure the harmonics we need to setup the forward power meter to the power meter and on the devices 2 tab we select the spectrum analyser.
[[File:Harmonics_Configure_Broadband_power_meter.PNG]]
[[File:HarmonicsSpectrumAnalyserAndPowermeterEquipment.png]]
[[File:Harmonics_Configure_Spectrum_Analyser_as_Sensor_Power_meter.PNG]]
== Configuring the test ==
To start the harmonics test open a EUT file and choose the menu:
{{Menu|Calibration|EUT Calibration|Harmonics|Element}}
{|class="wikitable"
| || Coupler || Devices 1 || no ||
| || Forward power meter || Devices 1 || no || The powermeter device driver for the broadband power measurement of the forward power.
| || Sensor power meter || Devices 2 || yes || The powermeter device driver for the analyser that is being used to measure the harmonics.
!colspan="5"|Cables
|② || Cable amplifier -> coupler || Cables || no || Optional to correct for the cable loss between the amplifier and the coupler
|③ || Cable coupler -> forward power meter || Cables || no || Optional to correct for the cable loss between the coupler and the forward power meter
|④ || Cable current -> power meter || Cables || yes || Optional to correct for the cable loss between the coupler and the spectrum analyser (which is the sensor power meter)
|⑤ || Cable current -> power meter || Cables || yes || Optional to correct for the attenuator between the coupler and the spectrum analyser (which is the sensor power meter)
|⑥ || Cable current -> power meter || Cables || yes || Optional to correct for the cable loss between the coupler and the spectrum analyser (which is the sensor power meter)
The signal path between the coupler and the spectrum analyser often consists of multiple cable segments. Each cable segment can have its own correction, and multiple corrections can be selected and added to the testsite configuration. The combined correction of all the selected cable segments will be used for the correction.
=== Measure the actual field harmonics using a receive antenna ===
Some standards like the [[IEC 61000-4-3]] describe that the harmonic content in the field strength should be measured and verified. By using this setup, the harmonic content in the field strength is directly measured, and the actual gain of the transmitting antenna is present in the measurement results.
[[File:HarmonicsSpectrumAnalyserAndAntennasEquipment.png]]
{|class="wikitable"
| || Calibration antenna || Devices 1 || yes || The antenna that is used for the receiving of the generated electrical fields on the harmonics.
| || Spectrum analyser || Devices 2 || yes || The device driver for the analyser that is being used to measure the harmonics.
!colspan="5"|Cables
|④ || Cable preamp -> analyser || Cables || yes || Optional to correct for the attenuator between the receive antenna and the spectrum analyser
|⑤ || Cable preamp -> analyser || Cables || yes || Optional to correct for the cable loss between the receive antenna and the spectrum analyser
The signal path between the receive antenna and the spectrum analyser often consists of multiple cable segments. Each cable segment can have its own correction, and multiple corrections can be selected and added to the testsite configuration. The combined correction of all the selected cable segments will be used for the correction.
== Configuring the calibration ==
To start the harmonics calibration first, open an EUT file. This EUT file represents the amplifier, which is the actual Equipment Under Test (EUT). By maintaining this EUT file, a yearly remeasurement of the harmonics of the amplifier will all be present in the same EUT file.
Once the EUT file is opened, select from the menu:
{{Menu|Calibration|EUT Calibration|Harmonics}}
[[File:Harmonics_configure_test.png]]
{{ScreenElementDescription|Description|The name of the selected TSF file.}}
{{Frequency window immunity}}
{{ScreenElementDescription|Constant|The test level can be set to a constant level to which the power should be regulated.}}
{{ScreenElementDescription|Variable|A .COR file can also be used. This can be the same file as for the Frequency List.}}
{{ScreenElementDescription|Amount|The number of harmonics that should be measured.}}
{{ScreenElementDescription|Type of harmonic|If only the even, odd or both (even and odd) harmonics should be measured.}}
{{ScreenElementDescription|Calibration method|On which power levelling point the desired testlevel should be regulated.}}
{{ScreenElementDescription|Test Equipment|In the test site window the test engineer can select which equipment list will be used during this test. This also determines which [[#Overview|method and equipment]] is being used.}}
{{ScreenElementDescription|Amount of samples|The amount of measurements that are measured on each harmonic to determine its power. The average of these measurements will be used as the power of the harmonic.}}
{{ScreenElementDescription|Start Test|Starts the configured calibration.}}
{{ScreenElementDescription|Cancel|Cancel the configured calibration.}}
{{ScreenElementDescription|Environment| Display the environment window.}}
{{ScreenElementDescription|Amplifier|}}
{{ScreenElementDescription|Inputs|}}
{{ScreenElementDescription|Note|Display the note window.}}
{{ScreenElementDescription|Units|Display the Units configuration window}}
{{ScreenElementDescription|Photos|}}
{{ScreenElementDescription|General info|}}
== Calibration procedure ==
# Set the signal generator to the first frequency and a low power.
# Measure the power using the forward powermeter. If the power is within the desired tolerance, continue at step 4.
# Adjust the signal generator carrier level. If the forward power was too high, reduce the carrier level, otherwise increase the carrier level.
# Use the spectrum analyser to measure the first harmonic
# Use the spectrum analyser to measure the other harmonics that should be measured
# Increase the signal generator carrier frequency to the next frequency. continue at step 2.
{|class="wikitable"
! Carrier frequency
| 100 MHz || 200 MHz || 300 MHz || 400 MHz || 500 Mhz
! Forward power
| 56.0 dBm || 56.1 dBm || 56.2 dBm || 56.1 dBm || 56.0 dBm
! 1st harmonic
| 55.9 dBm @ 100 MHz || 55.8 dBm @ 200 MHz || 55.9 dBm @ 300 MHz  || 55.8 dBm @ 400 MHz || 55.9 dBm @ 500 MHz
! 3rd harmonic
| 37.8 dBm @ 300 MHz || 42.2 dBm @ 600 MHz || 52.7 dBm @ 900 MHz  || 43.1 dBm @ 1200 MHz || 39.2 dBm @ 1500 MHz
! 5th harmonic
| 33.1 dBm @ 500 MHz || 37.4 dBm @ 1000 MHz || 37.9 dBm @ 1500 MHz  || 38.4 dBm @ 2000 MHz || --- @ 2500 MHz
It is possible that one or more of the higher harmonics have not been measured due to limitations of the allowed frequency range of the used equipment (which can be coupler, spectrum analyzer, antenna and cables). For a CW frequency of 500 MHz, the 5th harmonic is generated at 2500 MHz. If however the used spectrum analyzer is only able to measure up to 2000 MHz, it is not possible to measure the 5th harmonic of 500 MHz, and thus the value in the table will be empty. Also the graphline of the 5th harmonic will then only be drawn up to 400 MHz, which corresponds to the 5th harmonic at 2000 MHz, which is the highest measurable frequency of that spectrum analyzer.
== Test results ==
== Test results ==
In the test results it is possible to show the forward power measured.
When only using a spectrum analyser this will be the same as the first harmonic and when also using the broadband power meter this of course will be the broadband power measured.
When only a spectrum analyser is used, the forward power will be the same as the first harmonic. When also a broadband power meter is used, the forward power measurement will be the measured broadband power, and it thus will be higher than the power of the first harmonic alone.
[[File:Harmonics_test_results.PNG]]
{{ScreenElementDescription|Signal power|The power of the signal generator that is driving the amplifier.}}
{{ScreenElementDescription|Forward power|The (broadband) output power that is generated by the amplifier.}}
{{ScreenElementDescription|Power (1st harmonic)|The power of the first harmonic (the actual CW frequency) that is generated by the amplifier.}}
{{ScreenElementDescription|Power (...th harmonic)|The power of specified harmonic that is generated by the amplifier.}}
{{ScreenElementDescription|Gain (...th harmonic)|The dBc difference between the specified harmonic and the first harmonic.}}
{{ScreenElementDescription|Std. Dev. (...th harmonic)|The standard deviation of the measured power samples of the specified harmonic.}}
== Conclusion ==
The harmonics calibration in RadiMation can thus be used in different measurement configurations to measure the actually generated harmonics. Depending on the requirements and the used standard, the best suitable measurement configuration can be used.
For more information on how to perform the measurement of amplifier harmonics, please see the [[Chapter_12#EUT_Harmonics|  EUT Harmonics section in Chapter 12]] of the {{RadiMation}} manual.
[[Category:RadiMation]]
Latest revision as of 11:25, 30 July 2024
Overview[edit]
  1. Measure the power of all harmonics using a spectrum analyser.
  2. Measure the broadband power with a powermeter and the harmonics using a spectrum analyser.
  3. Measure the actual field harmonics using a receive antenna.
# Device name Tab in testsite configuration window suitable for harmonic frequencies note
Signal Generator Devices 1 no
Amplifier Devices 1 no This is the amplifier of which the harmonics are being measured.
Coupler Devices 1 yes
Forward power meter Devices 1 yes The powermeter device driver for the analyser that is being used to measure the harmonics.
Cables
Cable SG -> amplifier Cables no Optional to correct for the cable loss between the signal generator and the amplifier
Cable amplifier -> coupler Cables yes Optional to correct for the cable loss between the amplifier and the coupler
Cable coupler -> forward power meter Cables yes Optional to correct for the cable loss between the coupler and the forward power meter
Coupler Devices 1 no
Forward power meter Devices 1 no The powermeter device driver for the broadband power measurement of the forward power.
Sensor power meter Devices 2 yes The powermeter device driver for the analyser that is being used to measure the harmonics.
Cables
Cable amplifier -> coupler Cables no Optional to correct for the cable loss between the amplifier and the coupler
Cable coupler -> forward power meter Cables no Optional to correct for the cable loss between the coupler and the forward power meter
Cable current -> power meter Cables yes Optional to correct for the cable loss between the coupler and the spectrum analyser (which is the sensor power meter)
Cable current -> power meter Cables yes Optional to correct for the attenuator between the coupler and the spectrum analyser (which is the sensor power meter)
Calibration antenna Devices 1 yes The antenna that is used for the receiving of the generated electrical fields on the harmonics.
Spectrum analyser Devices 2 yes The device driver for the analyser that is being used to measure the harmonics.
Cables
Cable preamp -> analyser Cables yes Optional to correct for the attenuator between the receive antenna and the spectrum analyser
Cable preamp -> analyser Cables yes Optional to correct for the cable loss between the receive antenna and the spectrum analyser
   Menu.svg Calibration
      Menu.svg EUT Calibration
         Menu.svg Harmonics
ScreenElementDescription.svg Test Equipment In the test site window the test engineer can select which equipment list will be used during this test. This also determines which method and equipment is being used.
Calibration procedure[edit]
  1. Set the signal generator to the first frequency and a low power.
  2. Measure the power using the forward powermeter. If the power is within the desired tolerance, continue at step 4.
  3. Adjust the signal generator carrier level. If the forward power was too high, reduce the carrier level, otherwise increase the carrier level.
  4. Use the spectrum analyser to measure the first harmonic
  5. Use the spectrum analyser to measure the other harmonics that should be measured
  6. Increase the signal generator carrier frequency to the next frequency. continue at step 2.
Carrier frequency 100 MHz 200 MHz 300 MHz 400 MHz 500 Mhz
Forward power 56.0 dBm 56.1 dBm 56.2 dBm 56.1 dBm 56.0 dBm
1st harmonic 55.9 dBm @ 100 MHz 55.8 dBm @ 200 MHz 55.9 dBm @ 300 MHz 55.8 dBm @ 400 MHz 55.9 dBm @ 500 MHz
3rd harmonic 37.8 dBm @ 300 MHz 42.2 dBm @ 600 MHz 52.7 dBm @ 900 MHz 43.1 dBm @ 1200 MHz 39.2 dBm @ 1500 MHz
5th harmonic 33.1 dBm @ 500 MHz 37.4 dBm @ 1000 MHz 37.9 dBm @ 1500 MHz 38.4 dBm @ 2000 MHz --- @ 2500 MHz
Test results[edit]