Difference between revisions of "RadiMation Application Note 157"
(→Configure the EUT test) |
(7 intermediate revisions by the same user not shown) | |||
Line 6: | Line 6: | ||
The exact requirements and test methods for the RS103 are specified in the [[MIL-STD-461]]. | The exact requirements and test methods for the RS103 are specified in the [[MIL-STD-461]]. | ||
+ | This application note demonstrates how to setup a closed loop test. When testing closed loop, no calibration is required. | ||
== Necessary equipment == | == Necessary equipment == | ||
Line 62: | Line 63: | ||
− | The radiated immunity multiband test | + | The radiated immunity multiband test can be configured by selecting the menu: |
{{Menu|Tests|Radiated Immunity|Multiband|New}} | {{Menu|Tests|Radiated Immunity|Multiband|New}} | ||
Line 69: | Line 70: | ||
{{ScreenElementDescription|Start|The start frequency of the test. For example 2 MHz.}} | {{ScreenElementDescription|Start|The start frequency of the test. For example 2 MHz.}} | ||
{{ScreenElementDescription|Stop|The stop frequency of the test. For example 30 MHz.}} | {{ScreenElementDescription|Stop|The stop frequency of the test. For example 30 MHz.}} | ||
− | {{ScreenElementDescription|Frequency Step|The frequency step, see Table III.}} | + | {{ScreenElementDescription|Frequency Step|The frequency step, for the applicable steps see "Table III susceptibility scanning" in MIL-STD-461G.}} |
{{ScreenElementDescription|Dwell Time|The dwell time.}} | {{ScreenElementDescription|Dwell Time|The dwell time.}} | ||
{{ScreenElementDescription|Frequency Change Mode|Constant.}} | {{ScreenElementDescription|Frequency Change Mode|Constant.}} | ||
Line 75: | Line 76: | ||
{{ScreenElementDescription|Test equipment|The equipment needed for the radiated susceptibility test.}} | {{ScreenElementDescription|Test equipment|The equipment needed for the radiated susceptibility test.}} | ||
{{ScreenElementDescription|Location Type|Antenna height, distance, polarization and angle.}} | {{ScreenElementDescription|Location Type|Antenna height, distance, polarization and angle.}} | ||
− | {{ScreenElementDescription|Test Level|Specify the {{ScreenElement|Electrical field closed loop method}} test level | + | {{ScreenElementDescription|Test Level|Specify the {{ScreenElement|Electrical field closed loop method}} test level and configure a {{ScreenElement|Constant field}} or {{ScreenElement|Variable field}}}} |
{{ScreenElementDescription|Inputs|No inputs are needed.}} | {{ScreenElementDescription|Inputs|No inputs are needed.}} | ||
Line 84: | Line 85: | ||
{{ScreenElementDescription|Description|The description of the test level.}} | {{ScreenElementDescription|Description|The description of the test level.}} | ||
− | {{ScreenElementDescription|Test level|Variable field | + | {{ScreenElementDescription|Test level|Check {{ScreenElement|Variable field}} and specify the correction file for the applicable test level. When the level is one fixed value check {{ScreenElement|Constant field}} and specify the test level}} |
{{ScreenElementDescription|Tolerance|Specify the tolerance to use.}} | {{ScreenElementDescription|Tolerance|Specify the tolerance to use.}} | ||
− | {{ScreenElementDescription|Probes|Check Probe 1 to measure with the first probe.}} | + | {{ScreenElementDescription|Probes|Check {{ScreenElement|Probe 1}} to measure with the first probe.}} |
Latest revision as of 11:39, 29 August 2024
How to perform a MIL-STD-461 RS103, radiated susceptibility, electric field
This application note explains how the MIL-STD-461 RS103, radiated susceptibility, electric field test can be performed with RadiMation®
RS103 is a test method used in military and aerospace environments to evaluate the susceptibility of electronic equipment to radiated electromagnetic fields across a frequency range of 2 MHz to 40 GHz. It primarily uses pulse modulation to simulate high-intensity electromagnetic interference, ensuring that devices can withstand these conditions without malfunctioning.
The exact requirements and test methods for the RS103 are specified in the MIL-STD-461. This application note demonstrates how to setup a closed loop test. When testing closed loop, no calibration is required.
Necessary equipment
The following devices are necessary to perform the RS103 test:
- Signal generator
- Amplifier
- Antenna
- Field probe
Define the test level
If the limit in "Table XI RS103" is the same value for the whole frequency range the creation and usage of the correction file can be skipped. Usage of the correction file is only applicable to variable limits.
Create a correction file for the test level
First create a correction file in RadiMation® and specify the test level (See 'Table XI RS103 limits.').
Store the correction file with the RS103 limits as file on disk.
This correction file with the test level will be used during EUT testing.
EUT Testing
EUT Testing equipment
The configuration of the EUT test site should contain the following devices:
Antenna | Devices 1 | The antenna to use | |
Field Sensor 1 | Field probes | The field sensor to use |
To perform the test on the EUT, create a radiated immunity multiband test.
Create a test setup configuration file for the following settings:
Frequency range | Frequency step | Polarization |
2 MHz to 30 MHz | 1% | Vertical |
30 MHz - 18 GHz | 0.5% | Both (Horizontal and vertical) |
18 GHz - 40 GHz | 0.25% | Both (Horizontal and vertical) |
Start | The start frequency of the test. For example 2 MHz. |
Stop | The stop frequency of the test. For example 30 MHz. |
Frequency Step | The frequency step, for the applicable steps see "Table III susceptibility scanning" in MIL-STD-461G. |
Modulation | The modulation to use see below. |
Test equipment | The equipment needed for the radiated susceptibility test. |
Location Type | Antenna height, distance, polarization and angle. |
Test Level | Specify the Electrical field closed loop method test level and configure a Constant field or Variable field |
At TestLevel click Add to add a new Test level and select TestLevel - Electrical field closed loop method
Test level | Check Variable field and specify the correction file for the applicable test level. When the level is one fixed value check Constant field and specify the test level |
Probes | Check Probe 1 to measure with the first probe. |
To configure the pulse modulation settings click Config next to the modulation settings, click the PM tab, and enter the pulse modulation settings as defined in 5.21.3.4.c.1.a
Activate modulation | Check this checkbox to enable the pulse modulation. |
Frequency | The PM frequency 1 kHz. |
Duty Cycle | A duty cycle of 50%. |
Also apply modulation out of the dwell-time period | Check this checkbox to enable modulation outside the dwell-time. |
When all settings are configured press Start Test to run the test.
EUT test result
Conclusion
The RadiMation® Radiated Immunity multiband test can be used to perform the RS103 EUT test.