Difference between revisions of "RadiMation Application Note 154"
Jump to: navigation, search
(Configure the calibration)
(Configure the verification)
Line 161: Line 161:
[[File:CS114 Verification Setup.png]]
{{ScreenElementDescription|Start|The start frequency of the test. For example 10 kHz.}}
{{ScreenElementDescription|Start|The start frequency of the verification. For example 10 kHz.}}
{{ScreenElementDescription|End|The stop frequency of the test. For example 1 MHz.}}
{{ScreenElementDescription|End|The stop frequency of the verification. For example 1 MHz.}}
{{ScreenElementDescription|Frequency Step|The frequency step, see 5.12.3.4.c.3.}}
{{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II.}}
Revision as of 13:42, 29 August 2024
How to perform a MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test
CS114 is a test method used in military and aerospace environments to assess the susceptibility of electronic equipment to conducted RF (radio frequency) currents over a frequency range of 10 kHz to 200 MHz. The test uses pulse modulation to ensure that devices can withstand conducted interference without experiencing performance degradation or failure.
Necessary equipment
The following devices are necessary to run the calibration, verification and EUT test:
  • Signal generator
  • Amplifier
  • Coupler
  • Forward power meter
  • Current sensor
  • 50 Ohm load
  • 50 Ohm attenuator
  • Current injection device
  • Current injection calibration jig (Calibration fixture)
Tips
Note: If the current unit is shown as μA, it can be changed to dBμA by clicking Units, select Current and click Edit to change this.
Define the test level
CS114 uses a limit that varies over the frequency range and therefor we have to configure a variable test level. These limits are configured with a correction file in RadiMation®.
Create a correction file for the test level by selecting the menu:
   Menu.png File
      Menu.png New
         Menu.png Correction
First create a correction file in RadiMation® and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.' in MIL-STD-461.). In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured.
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBμA.
Store the correction file for the test level by selecting the menu:
   Menu.png File
      Menu.png Save correction
This correction file with the test level will be used during the calibration, verification and the EUT testing.
Calibration procedure
The calibration is performed to characterize the losses and determine the correct power level in the used measurement setup.
Calibration equipment
Forward power meter Devices 1 The forward power meter to use
Sensor powermeter Devices 2 The power meter or analyser to use for measuring the current
Current Sensor Devices 2 The current sensor to use with transfer factor attached to the driver
Injection device Devices 2 The injection clamp to use
Jig Devices 2 The jig to use
Cables
Cable current -> power meter Cables Cable with a correction file specified for the cable loss
Cable current -> power meter Cables Cable with the specified loss of the used attenuator
There are three frequency ranges that have to be calibrated with the following settings (see TABLE III in MIL-STD-461.).
Band 1 10 kHz - 1 MHz 5 %
Band 2 1 MHz - 30 MHz 1 %
Band 2 30 MHz - 200 MHz 0.5 %
   Menu.png Calibration
      Menu.png System calibration
         Menu.png Conducted immunity
This example shows the configuration for band 1, a similar configuration can be created for the other bands.
ScreenElementDescription.png Start The start frequency of the calibration. For example 10 kHz.
ScreenElementDescription.png End The stop frequency of the calibration. For example 1 MHz.
ScreenElementDescription.png Step The frequency step, in this case 5%.
ScreenElementDescription.png Test level The test level, in this case current.
ScreenElementDescription.png Variable The variable test level, select the correction file created earlier.
ScreenElementDescription.png Tolerance The regulation tolerance.
ScreenElementDescription.png Calibration method Forward power.
ScreenElementDescription.png Test equipment Calibration test equipment.
ScreenElementDescription.png Test engineer The engineer that performed the calibration.
Calibration result
Verification
Verification equipment
Cables
Cable current -> power meter Cables Cable (1) with a correction file specified for the cable loss
Cable current -> power meter Cables Cable (2) with the specified loss of the used attenuator
Cable current -> power meter Cables Cable (3) with a correction file specified for the cable loss
CS114 requires to perform a verification of the setup with the probes in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit. The verification can be performed in RadiMation® by doing an actual conducted immunity test. This test can be started by selecting from the menu:
   Menu.png Tests
      Menu.png Conducted Immunity
         Menu.png Multiband
Band 1 10 kHz - 1 MHz 10 %
Band 2 1 MHz - 30 MHz 2 %
Band 2 30 MHz - 200 MHz 1 %
ScreenElementDescription.png Start The start frequency of the verification. For example 10 kHz.
ScreenElementDescription.png End The stop frequency of the verification. For example 1 MHz.
ScreenElementDescription.png Frequency Step The frequency step, see 5.12.3.4.c.3.
ScreenElementDescription.png Dwell Time The dwell time specified see TABLE II.
ScreenElementDescription.png Frequency Change Mode Constant.
ScreenElementDescription.png Modulation Modulation is optional during the verification.
ScreenElementDescription.png Test equipment The equipment needed for the verification.
ScreenElementDescription.png Location Type Injection clamp.
ScreenElementDescription.png Test Level Specify the current substitution method.
ScreenElementDescription.png Inputs Add the current sensor as an input.
ScreenElementDescription.png Description The description of the test level.
ScreenElementDescription.png Calibration file Select the calibration file to use.
ScreenElementDescription.png Test level Variable current, and specify the correction file for the applicable test level.
ScreenElementDescription.png Calibration method Select Forward Power.
ScreenElementDescription.png Tolerance Specify the tolerance to use.
Run the verification test
Verification result
EUT Testing
EUT Testing equipment
Cables
Configure the EUT test
   Menu.png Tests
         Menu.png Multiband
ScreenElementDescription.png Start The start frequency of the test. For example 10 kHz.
ScreenElementDescription.png End The stop frequency of the test. For example 1 MHz.
ScreenElementDescription.png Frequency Step The frequency step, see Table III.
ScreenElementDescription.png Dwell Time The dwell time.
ScreenElementDescription.png Modulation Configure Pulse Modulation with 1 kHz and 50% duty cycle.
ScreenElementDescription.png Test equipment The equipment needed for the conducted immunity test.
ScreenElementDescription.png Test Level Specify the current substitution method. Also add a "Measured Current Limit"
ScreenElementDescription.png Inputs No inputs are needed.
To configure the pulse modulation settings click Config next to the modulation settings, click the PM tab, and enter the pulse modulation settings as defined in 5.12.3.4.d.1. CS114 Modulation settings.png
ScreenElementDescription.png Activate modulation Check this checkbox to enable the pulse modulation.
ScreenElementDescription.png Frequency The PM frequency 1 kHz.
ScreenElementDescription.png Duty Cycle A duty cycle of 50%.
ScreenElementDescription.png Also apply modulation out of the dwell-time period Check this checkbox to enable modulation outside the dwell-time.
ScreenElementDescription.png Calibration file Select the calibration file to use for this band.
ScreenElementDescription.png Description The description of the limit
ScreenElementDescription.png Constant current The current to limit on.
EUT Test Result
Conclusion
