== Tips ==
== Tips ==
{{Note|If the current unit is shown as μA, it can be changed to dBμA by clicking {{ScreenElement|Units}}, select {{ScreenElement|Current}} and click {{ScreenElement|Edit}} to change this.}}
{{Note|If the current unit is shown as mA, it can be changed to dBmA by clicking {{ScreenElement|Units}}, select {{ScreenElement|Current}} and click {{ScreenElement|Edit}} to change this.}}
== Define the test level ==
Line 46: Line 46:
=== Calibration equipment ===
{{Note|When calibrating at high levels ensure that enough adequate attenuation is used to shield the current sensing device from damage.}}
[[File:ISO11452-4-Calibration.png]]
{{Note|When calibrating using amplifiers rated with more power than the calibration setup can tolerate, ensure that enough adequate attenuation is used after the amplifier to prevent damaging levels from reaching the setup.}}
The configuration of the calibration test setup should include the following equipment:
Line 88: Line 92:
| Band 3 || 3 MHz - 200 MHz
| Band 4 || 400 MHz - 400 MHz
| Band 4 || 200 MHz - 400 MHz
Line 151: Line 155:
[[File:ISO-11452-4_Test_config.png]]
{{ScreenElementDescription|Start|The start frequency of the test. For example 0.1 kMHz.}}
{{ScreenElementDescription|Start|The start frequency of the test. For example 0.1 MHz.}}
{{ScreenElementDescription|End|The stop frequency of the test. For example 1 MHz.}}
{{ScreenElementDescription|Frequency Step|The frequency step, see Table III.}}
{{ScreenElementDescription|Frequency Step|The frequency step.}}
{{ScreenElementDescription|Dwell Time|The dwell time.}}
{{ScreenElementDescription|Frequency Change Mode|Constant.}}
Line 164: Line 168:
To configure any required modulation settings click {{ScreenElement|Config}} next to the modulation settings, click the applicable tab, and enter the modulation settings as specified in ISO 11452-1.
To configure any required modulation settings click {{ScreenElement|Config}} next to the modulation settings, click the applicable tab, and enter the modulation settings as specified in [[ISO 11452-1]].
[[File:ISO11452-4-Modulation_settings.png]]
{{ScreenElementDescription|Activate modulation|Check this checkbox to enable the pulse modulation.}}
{{ScreenElementDescription|Activate modulation|Check this checkbox to enable the required modulation.}}
{{ScreenElementDescription|Frequency|The PM frequency 1 kHz.}}
{{ScreenElementDescription|Frequency|The frequency.}}
{{ScreenElementDescription|Duty Cycle|A duty cycle of 50%.}}
{{ScreenElementDescription|Depth|The modulation depth.}}
{{ScreenElementDescription|Also apply modulation out of the dwell-time period|Check this checkbox to enable modulation outside the dwell-time.}}
Line 185: Line 189:
{{ScreenElementDescription|Tolerance|Specify the tolerance to use.}}
4
When all bands are configured press {{ScreenElement|Start Test}} to run the EUT test.
Latest revision as of 08:37, 21 October 2024
How to perform an ISO 11452-4 BCI (Bulk Current Injection) substitution test[edit]
This application note explains how the ISO 11452-4 BCI (Bulk Current Injection) substitution test can be performed with RadiMation®.
The ISO 11452-4 standard is used to test the immunity of automotive electronic components to conducted RF (radio frequency) disturbances injected into the wiring harness. The test frequency typically ranges from 1 MHz to 400 MHz, using both continuous and amplitude-modulated signals. The test ensures that the electronic devices can withstand the RF disturbances without performance degradation or failure.
The exact requirements and test methods for the BCI test are specified in the ISO 11452-4.
Necessary equipment[edit]
The following devices are required to run the calibration, verification, and EUT (Equipment Under Test) test:
  • Signal generator
  • Amplifier
  • Coupler
  • Forward power meter
  • Current sensor
  • 50 Ohm load
  • 50 Ohm attenuator
  • Current injection device
  • Current injection calibration jig
Tips[edit]
Note: If the current unit is shown as mA, it can be changed to dBmA by clicking Units, select Current and click Edit to change this.
Define the test level[edit]
The ISO 11452-4 BCI test requires test levels to be defined for different frequency ranges. The test level may vary depending on the application class or specific test plan, and these levels are set up as a correction file in RadiMation®.
Create a correction file for the test level by selecting the menu:[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg Correction
First, create a correction file in RadiMation® and specify the applicable test level for the BCI test based on the test requirements. The levels are typically specified in terms of current in mA for different frequency bands (see 'ISO 11452-4 BCI test severity levels '). In this example the levels for test severity level 1 are configured.
Ensure that the correction file has a Frequency column and a Current column with the unit set to mA.
Store the correction file for the test level by selecting the menu:[edit]
   Menu.svg File
      Menu.svg Save correction Save the test level as a correction file on disk.
This correction file will be used during the calibration and EUT testing.
Calibration procedure[edit]
The calibration procedure is necessary to characterize losses in the setup and determine the correct power level for injection.
Calibration equipment[edit]
Note: When calibrating at high levels ensure that enough adequate attenuation is used to shield the current sensing device from damage.
Note: When calibrating using amplifiers rated with more power than the calibration setup can tolerate, ensure that enough adequate attenuation is used after the amplifier to prevent damaging levels from reaching the setup.
# Device name Tab in test site configuration window note
Coupler Devices 1 The directional coupler to use
Forward power meter Devices 1 The forward power sensor to use
Power meter Devices 2 The power meter or analyzer for measuring current
Injection clamp Devices 2 The bulk current injection (BCI) clamp
Calibration jig Devices 2 The jig used for the current injection calibration
Cables
Cable current -> power meter Cables Cable with a correction file specified for cable loss
Cable current -> power meter Cables Cable with specified loss of the used attenuator
The BCI test covers four primary frequency bands. The following settings can be configured according to ISO 11452-4 requirements: The minimal requirements related to the step sizes are found in ISO 11452-1, 5.6 Table 2.
Band Frequency range
Band 1 0.1 MHz - 1 MHz
Band 2 1 MHz - 3 MHz
Band 3 3 MHz - 200 MHz
Band 4 200 MHz - 400 MHz
These four bands can be configured and performed in one calibration. The calibration can be started by selecting:
   Menu.svg Calibration
      Menu.svg System calibration
         Menu.svg Conducted immunity
This example shows an example of the configuration for required for the calibration.
ScreenElementDescription.svg Start The start frequency of the calibration (e.g., 1 MHz).
ScreenElementDescription.svg End The stop frequency of the calibration (e.g., 30 MHz).
ScreenElementDescription.svg Step The frequency step, in this case, 1%.
ScreenElementDescription.svg Test level The test level, in this case current.
ScreenElementDescription.svg Variable The variable test level, select the correction file created earlier.
ScreenElementDescription.svg Calibration method The power measurement that should result from the calibration: Forward power.
ScreenElementDescription.svg Test equipment The calibration test equipment.
ScreenElementDescription.svg Test engineer The engineer performing the calibration.
Calibration result[edit]
Once calibration is complete, the engineer will save the results in a calibration (.CAL) file.
EUT Testing[edit]
EUT Testing equipment[edit]
The configuration for EUT testing should contain the following devices:
Injection clamp Devices 2 The bulk current injection clamp
Configure the EUT test[edit]
To perform the EUT test, configure the conducted immunity multiband test by selecting from the menu:
   Menu.svg Tests
      Menu.svg Conducted Immunity
         Menu.svg Multiband
The following bands should be configured:
ScreenElementDescription.svg Start The start frequency of the test. For example 0.1 MHz.
ScreenElementDescription.svg Frequency Step The frequency step.
ScreenElementDescription.svg Dwell Time The dwell time.
ScreenElementDescription.svg Frequency Change Mode Constant.
ScreenElementDescription.svg Modulation Configure the required modulation.
ScreenElementDescription.svg Test Level Specify the current substitution method.
To configure any required modulation settings click Config next to the modulation settings, click the applicable tab, and enter the modulation settings as specified in ISO 11452-1.
ScreenElementDescription.svg Activate modulation Check this checkbox to enable the required modulation.
ScreenElementDescription.svg Frequency The frequency.
ScreenElementDescription.svg Depth The modulation depth.
ScreenElementDescription.svg Also apply modulation out of the dwell-time period Check this checkbox to enable modulation outside the dwell-time.
4 When all bands are configured press Start Test to run the EUT test.
EUT Test Result[edit]
The RadiMation® Conducted Immunity multiband test can be used to perform the ISO 11452-4 BCI substitution test.