| (33 intermediate revisions by 2 users not shown) | |||
| Line 25: | Line 25: | ||
== Define the test level == | == Define the test level == | ||
CS114 uses a | CS114 uses a limit that varies over the frequency range and therefor we have to configure a variable test level. | ||
These limits are configured with a correction file in {{RadiMation}}. | |||
=== Create a correction file for the test level by selecting the menu: === | === Create a correction file for the test level by selecting the menu: === | ||
{{Menu|File|New|Correction}} | {{Menu|File|New|Correction}} | ||
First create a correction file in {{RadiMation}} and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.'). | First create a correction file in {{RadiMation}} and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.' of the [[MIL-STD-461]].). | ||
In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured. | In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured. | ||
| Line 37: | Line 37: | ||
[[File:CS114_Test_level.png]] | [[File:CS114_Test_level.png]] | ||
=== Store the correction file for the test level by selecting the menu: === | |||
{{Menu|File|Save correction}} | {{Menu|File|Save correction}} | ||
Store the test level as correction file on disk. | Store the test level as correction file on disk. | ||
| Line 43: | Line 44: | ||
== Calibration procedure == | == Calibration procedure == | ||
The calibration is performed to characterize the losses and determine the correct power level in the used measurement setup. | |||
=== Calibration equipment === | === Calibration equipment === | ||
| Line 69: | Line 71: | ||
!colspan="4"|Cables | !colspan="4"|Cables | ||
|① || Cable current -> power meter || Cables || Cable | |① || Cable current -> power meter || Cables || Cable with a correction file specified for the cable loss | ||
|② || Cable current -> power meter || Cables || Cable | |② || Cable current -> power meter || Cables || Cable with the specified loss of the used attenuator | ||
|③ || Cable current -> power meter || Cables || Cable | |③ || Cable current -> power meter || Cables || Cable with a correction file specified for the cable loss | ||
=== Configure the calibration === | === Configure the calibration === | ||
There are three frequency ranges that have to be calibrated with the following settings (see TABLE III). | There are three frequency ranges that have to be calibrated with the following settings (see TABLE III in [[MIL-STD-461]].). | ||
! Band || Frequency range || Step size | ! Band || Frequency range || Step size | ||
| Line 91: | Line 93: | ||
{{Menu|Calibration|System calibration|Conducted immunity}} | {{Menu|Calibration|System calibration|Conducted immunity}} | ||
This example shows the | This example shows the configuration for band 1, a similar configuration can be created for the other bands. | ||
[[File:CS114_Calibration band 1.png]] | [[File:CS114_Calibration band 1.png]] | ||
{{ScreenElementDescription|Start|The start frequency of the | {{ScreenElementDescription|Start|The start frequency of the calibration. For example 10 kHz.}} | ||
{{ScreenElementDescription|End|The stop frequency of the | {{ScreenElementDescription|End|The stop frequency of the calibration. For example 1 MHz.}} | ||
{{ScreenElementDescription|Step|The frequency step, in this case 5%.}} | {{ScreenElementDescription|Step|The frequency step, in this case 5%.}} | ||
{{ScreenElementDescription|Test level|The test level, in this case current.}} | {{ScreenElementDescription|Test level|The test level, in this case {{ScreenElement|current}}.}} | ||
{{ScreenElementDescription|Variable|The variable test level, select the correction file created earlier.}} | {{ScreenElementDescription|Variable|The variable test level, select the correction file created earlier.}} | ||
{{ScreenElementDescription|Tolerance|The regulation tolerance.}} | {{ScreenElementDescription|Tolerance|The regulation tolerance.}} | ||
{{ScreenElementDescription|Calibration method|Forward power | {{ScreenElementDescription|Calibration method|The power measurement that should be the result of the calibration: {{ScreenElement|Forward power}}}} | ||
{{ScreenElementDescription|Test equipment|Calibration test equipment.}} | {{ScreenElementDescription|Test equipment|Calibration test equipment.}} | ||
{{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}} | {{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}} | ||
| Line 110: | Line 112: | ||
== Verification == | == Verification == | ||
CS114 requires to perform a verification of the setup with the current probe in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit. | |||
=== Verification equipment === | === Verification equipment === | ||
| Line 134: | Line 137: | ||
=== Configure the verification === | === Configure the verification === | ||
The verification can be performed in {{RadiMation}} by doing an actual conducted immunity test. This test can be started by selecting from the menu: | The verification can be performed in {{RadiMation}} by doing an actual conducted immunity test. This test can be started by selecting from the menu: | ||
{{Menu|Tests|Conducted Immunity|Multiband}} | {{Menu|Tests|Conducted Immunity|Multiband}} | ||
| Line 159: | Line 161: | ||
[[File:CS114 Verification Setup.png]] | [[File:CS114 Verification Setup.png]] | ||
{{ScreenElementDescription|Start|The start frequency of the | {{ScreenElementDescription|Start|The start frequency of the verification. For example 10 kHz.}} | ||
{{ScreenElementDescription|End|The stop frequency of the | {{ScreenElementDescription|End|The stop frequency of the verification. For example 1 MHz.}} | ||
{{ScreenElementDescription|Frequency Step|The frequency step, see 5.12.3.4.c.3.}} | {{ScreenElementDescription|Frequency Step|The frequency step, see 5.12.3.4.c.3.}} | ||
{{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II.}} | {{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II of the [[MIL-STD-461]].}} | ||
{{ScreenElementDescription|Frequency Change Mode|Constant.}} | {{ScreenElementDescription|Frequency Change Mode|The method that is used to change from one to the next frequency: {{ScreenElement|Constant}}.}} | ||
{{ScreenElementDescription|Modulation|Modulation is optional during the verification.}} | {{ScreenElementDescription|Modulation|Modulation is optional during the verification.}} | ||
{{ScreenElementDescription|Test equipment|The equipment needed for the verification.}} | {{ScreenElementDescription|Test equipment|The equipment needed for the verification.}} | ||
{{ScreenElementDescription|Location Type|Injection clamp.}} | {{ScreenElementDescription|Location Type|Injection clamp.}} | ||
{{ScreenElementDescription|Test Level|Specify the | {{ScreenElementDescription|Test Level|Specify the test level {{ScreenElement|Current substitution method}}.}} | ||
{{ScreenElementDescription|Inputs|Add the | {{ScreenElementDescription|Inputs|Add the {{ScreenElement|Current sensor}} as an input.}} | ||
| Line 176: | Line 178: | ||
{{ScreenElementDescription|Description|The description of the test level.}} | {{ScreenElementDescription|Description|The description of the test level.}} | ||
{{ScreenElementDescription|Calibration file|Select the calibration file | {{ScreenElementDescription|Calibration file|Select the calibration file created during the calibration procedure.}} | ||
{{ScreenElementDescription|Test level|Variable current | {{ScreenElementDescription|Test level|{{ScreenElement|Variable current}} and specify the correction file for the applicable test level created during the calibration procedure.}} | ||
{{ScreenElementDescription|Calibration method|Select Forward Power.}} | {{ScreenElementDescription|Calibration method|Select Forward Power.}} | ||
{{ScreenElementDescription|Tolerance|Specify the tolerance to use.}} | {{ScreenElementDescription|Tolerance|Specify the tolerance to use.}} | ||
Add an input {{ScreenElement|Current sensor}}, this will ensure that the current is measured during the test. | |||
=== Run the verification test === | === Run the verification test === | ||
| Line 217: | Line 219: | ||
| Line 245: | Line 247: | ||
{{ScreenElementDescription|End|The stop frequency of the test. For example 1 MHz.}} | {{ScreenElementDescription|End|The stop frequency of the test. For example 1 MHz.}} | ||
{{ScreenElementDescription|Frequency Step|The frequency step, see Table III.}} | {{ScreenElementDescription|Frequency Step|The frequency step, see Table III.}} | ||
{{ScreenElementDescription|Dwell Time|The dwell time.}} | {{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II of the [[MIL-STD-461]].}} | ||
{{ScreenElementDescription|Modulation|Configure Pulse Modulation with 1 kHz and 50% duty cycle.}} | {{ScreenElementDescription|Modulation|Configure Pulse Modulation with 1 kHz and 50% duty cycle.}} | ||
{{ScreenElementDescription|Test equipment|The equipment needed for the conducted immunity test.}} | {{ScreenElementDescription|Test equipment|The equipment needed for the conducted immunity test.}} | ||
{{ScreenElementDescription|Test Level|Specify the | {{ScreenElementDescription|Test Level|Specify the {{ScreenElement|Current substition method}} and add a limit {{ScreenElement|Measured Current Limit}}}} | ||
{{ScreenElementDescription|Inputs|No inputs are needed.}} | {{ScreenElementDescription|Inputs|No inputs are needed.}} | ||
To configure the pulse modulation settings click {{ScreenElement|Config}} next to the modulation settings, click the {{ScreenElement|PM}} tab, and enter the pulse modulation settings as defined in 5.12.3.4.d.1. | To configure the pulse modulation settings click {{ScreenElement|Config}} next to the modulation settings, click the {{ScreenElement|PM}} tab, and enter the pulse modulation settings as defined in 5.12.3.4.d.1. | ||
[[File:CS114 Modulation settings.png]] | [[File:CS114 Modulation settings.png]] | ||
| Line 292: | Line 295: | ||
The {{RadiMation}} Conducted | The MIL-STD-461G CS114 calibration can be performed by using the Conducted immunity calibration. | ||
The {{RadiMation}} Conducted immunity multiband test can then be used to perform the MIL-STD-461G CS114 verification and the EUT test with a calibration file and the current sensor used as a limit in the EUT test. | |||
Latest revision as of 12:10, 7 November 2024
How to perform a MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test[edit]
CS114 is a test method used in military and aerospace environments to assess the susceptibility of electronic equipment to conducted RF (radio frequency) currents over a frequency range of 10 kHz to 200 MHz. The test uses pulse modulation to ensure that devices can withstand conducted interference without experiencing performance degradation or failure.
Necessary equipment[edit]
The following devices are necessary to run the calibration, verification and EUT test:
- Signal generator
- Amplifier
- Coupler
- Forward power meter
- Current sensor
- 50 Ohm load
- 50 Ohm attenuator
- Current injection device
Tips[edit]
| Note: | If the current unit is shown as μA, it can be changed to dBμA by clicking Units, select Current and click Edit to change this. |
Define the test level[edit]
CS114 uses a limit that varies over the frequency range and therefor we have to configure a variable test level. These limits are configured with a correction file in RadiMation®.
[edit]
File
New
Correction
First create a correction file in RadiMation® and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.' of the MIL-STD-461.). In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured.
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBμA.
[edit]
File
Save correction
This correction file with the test level will be used during the calibration, verification and the EUT testing.
Calibration procedure[edit]
Calibration equipment[edit]
| # | Device name | Tab in testsite configuration window | note |
| Signal Generator | Devices 1 | The signal generator to use | |
| Amplifier | Devices 1 | The amplifier to use | |
| Coupler | Devices 1 | The coupler to use | |
| Forward power meter | Devices 1 | The forward power meter to use | |
| Sensor powermeter | Devices 2 | The power meter or analyser to use for measuring the current | |
| Current Sensor | Devices 2 | The current sensor to use with transfer factor attached to the driver | |
| Injection device | Devices 2 | The injection clamp to use | |
| Jig | Devices 2 | The jig to use | |
| Cables | |||
| ① | Cable current -> power meter | Cables | Cable with a correction file specified for the cable loss |
| ② | Cable current -> power meter | Cables | Cable with the specified loss of the used attenuator |
| ③ | Cable current -> power meter | Cables | Cable with a correction file specified for the cable loss |
There are three frequency ranges that have to be calibrated with the following settings (see TABLE III in MIL-STD-461.).
| Band | Frequency range | Step size |
| Band 1 | 10 kHz - 1 MHz | 5 % |
| Band 2 | 1 MHz - 30 MHz | 1 % |
| Band 2 | 30 MHz - 200 MHz | 0.5 % |
Calibration
System calibration
Conducted immunity
| The start frequency of the calibration. For example 10 kHz. |
| The stop frequency of the calibration. For example 1 MHz. |
| The frequency step, in this case 5%. |
| The test level, in this case current. |
| The variable test level, select the correction file created earlier. |
| The regulation tolerance. |
| The power measurement that should be the result of the calibration: Forward power |
| Calibration test equipment. |
| The engineer that performed the calibration. |
Calibration result[edit]
Verification[edit]
Verification equipment[edit]
| Cables | |||
Configure the verification[edit]
Tests
Conducted Immunity
Multiband
| Band 1 | 10 kHz - 1 MHz | 10 % |
| Band 2 | 1 MHz - 30 MHz | 2 % |
| Band 2 | 30 MHz - 200 MHz | 1 % |
| The start frequency of the verification. For example 10 kHz. |
| The stop frequency of the verification. For example 1 MHz. |
| The frequency step, see 5.12.3.4.c.3. |
| The dwell time specified see TABLE II of the MIL-STD-461. |
| The method that is used to change from one to the next frequency: Constant. |
| Modulation is optional during the verification. |
| The equipment needed for the verification. |
| Injection clamp. |
| Specify the test level Current substitution method. |
| Add the Current sensor as an input. |
| The description of the test level. |
| Select the calibration file created during the calibration procedure. |
| Variable current and specify the correction file for the applicable test level created during the calibration procedure. |
| Select Forward Power. |
| Specify the tolerance to use. |
Add an input Current sensor, this will ensure that the current is measured during the test.
Verification result[edit]
EUT Testing[edit]
EUT Testing equipment[edit]
| Cables | |||
Configure the EUT test[edit]
Tests
Conducted Immunity
Multiband
| The start frequency of the test. For example 10 kHz. |
| The stop frequency of the test. For example 1 MHz. |
| The frequency step, see Table III. |
| Configure Pulse Modulation with 1 kHz and 50% duty cycle. |
| The equipment needed for the conducted immunity test. |
| Injection clamp. |
| Specify the Current substition method and add a limit Measured Current Limit |
| No inputs are needed. |
| Check this checkbox to enable the pulse modulation. |
| The PM frequency 1 kHz. |
| A duty cycle of 50%. |
| Check this checkbox to enable modulation outside the dwell-time. |
| Select the calibration file to use for this band. |
| Variable current, and specify the correction file for the applicable test level. |
| The description of the limit |
| The current to limit on. |
EUT Test Result[edit]
The MIL-STD-461G CS114 calibration can be performed by using the Conducted immunity calibration. The RadiMation® Conducted immunity multiband test can then be used to perform the MIL-STD-461G CS114 verification and the EUT test with a calibration file and the current sensor used as a limit in the EUT test.