| (7 intermediate revisions by the same user not shown) | |||
| Line 170: | Line 170: | ||
{{ScreenElementDescription|Location Type|Injection clamp.}} | {{ScreenElementDescription|Location Type|Injection clamp.}} | ||
{{ScreenElementDescription|Test Level|Specify the test level {{ScreenElement|Current substitution method}}.}} | {{ScreenElementDescription|Test Level|Specify the test level {{ScreenElement|Current substitution method}}.}} | ||
{{ScreenElementDescription|Inputs|Add the | {{ScreenElementDescription|Inputs|Add the {{ScreenElement|Current sensor}} as an input.}} | ||
| Line 179: | Line 179: | ||
{{ScreenElementDescription|Description|The description of the test level.}} | {{ScreenElementDescription|Description|The description of the test level.}} | ||
{{ScreenElementDescription|Calibration file|Select the calibration file created during the calibration procedure.}} | {{ScreenElementDescription|Calibration file|Select the calibration file created during the calibration procedure.}} | ||
{{ScreenElementDescription|Test level|Variable current | {{ScreenElementDescription|Test level|{{ScreenElement|Variable current}} and specify the correction file for the applicable test level created during the calibration procedure.}} | ||
{{ScreenElementDescription|Calibration method|Select Forward Power.}} | {{ScreenElementDescription|Calibration method|Select Forward Power.}} | ||
{{ScreenElementDescription|Tolerance|Specify the tolerance to use.}} | {{ScreenElementDescription|Tolerance|Specify the tolerance to use.}} | ||
| Line 247: | Line 247: | ||
{{ScreenElementDescription|End|The stop frequency of the test. For example 1 MHz.}} | {{ScreenElementDescription|End|The stop frequency of the test. For example 1 MHz.}} | ||
{{ScreenElementDescription|Frequency Step|The frequency step, see Table III.}} | {{ScreenElementDescription|Frequency Step|The frequency step, see Table III.}} | ||
{{ScreenElementDescription|Dwell Time|The dwell time.}} | {{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II of the [[MIL-STD-461]].}} | ||
{{ScreenElementDescription|Frequency Change Mode|Constant.}} | {{ScreenElementDescription|Frequency Change Mode|The method that is used to change from one to the next frequency: {{ScreenElement|Constant}}.}} | ||
{{ScreenElementDescription|Modulation|Configure Pulse Modulation with 1 kHz and 50% duty cycle.}} | {{ScreenElementDescription|Modulation|Configure Pulse Modulation with 1 kHz and 50% duty cycle.}} | ||
{{ScreenElementDescription|Test equipment|The equipment needed for the conducted immunity test.}} | {{ScreenElementDescription|Test equipment|The equipment needed for the conducted immunity test.}} | ||
{{ScreenElementDescription|Test Level|Specify the | {{ScreenElementDescription|Test Level|Specify the {{ScreenElement|Current substition method}} and add a limit {{ScreenElement|Measured Current Limit}}}} | ||
{{ScreenElementDescription|Inputs|No inputs are needed.}} | {{ScreenElementDescription|Inputs|No inputs are needed.}} | ||
| Line 295: | Line 295: | ||
The {{RadiMation}} Conducted | The MIL-STD-461G CS114 calibration can be performed by using the Conducted immunity calibration. | ||
The {{RadiMation}} Conducted immunity multiband test can then be used to perform the MIL-STD-461G CS114 verification and the EUT test with a calibration file and the current sensor used as a limit in the EUT test. | |||
Latest revision as of 12:10, 7 November 2024
How to perform a MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test[edit]
CS114 is a test method used in military and aerospace environments to assess the susceptibility of electronic equipment to conducted RF (radio frequency) currents over a frequency range of 10 kHz to 200 MHz. The test uses pulse modulation to ensure that devices can withstand conducted interference without experiencing performance degradation or failure.
Necessary equipment[edit]
The following devices are necessary to run the calibration, verification and EUT test:
- Signal generator
- Amplifier
- Coupler
- Forward power meter
- Current sensor
- 50 Ohm load
- 50 Ohm attenuator
- Current injection device
Tips[edit]
| Note: | If the current unit is shown as μA, it can be changed to dBμA by clicking Units, select Current and click Edit to change this. |
Define the test level[edit]
CS114 uses a limit that varies over the frequency range and therefor we have to configure a variable test level. These limits are configured with a correction file in RadiMation®.
[edit]
File
New
Correction
First create a correction file in RadiMation® and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.' of the MIL-STD-461.). In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured.
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBμA.
[edit]
File
Save correction
This correction file with the test level will be used during the calibration, verification and the EUT testing.
Calibration procedure[edit]
The calibration is performed to characterize the losses and determine the correct power level in the used measurement setup.
Calibration equipment[edit]
| # | Device name | Tab in testsite configuration window | note |
| Signal Generator | Devices 1 | The signal generator to use | |
| Amplifier | Devices 1 | The amplifier to use | |
| Coupler | Devices 1 | The coupler to use | |
| Forward power meter | Devices 1 | The forward power meter to use | |
| Sensor powermeter | Devices 2 | The power meter or analyser to use for measuring the current | |
| Current Sensor | Devices 2 | The current sensor to use with transfer factor attached to the driver | |
| Injection device | Devices 2 | The injection clamp to use | |
| Jig | Devices 2 | The jig to use | |
| Cables | |||
| ① | Cable current -> power meter | Cables | Cable with a correction file specified for the cable loss |
| ② | Cable current -> power meter | Cables | Cable with the specified loss of the used attenuator |
| ③ | Cable current -> power meter | Cables | Cable with a correction file specified for the cable loss |
There are three frequency ranges that have to be calibrated with the following settings (see TABLE III in MIL-STD-461.).
| Band | Frequency range | Step size |
| Band 1 | 10 kHz - 1 MHz | 5 % |
| Band 2 | 1 MHz - 30 MHz | 1 % |
| Band 2 | 30 MHz - 200 MHz | 0.5 % |
Calibration
System calibration
Conducted immunity
This example shows the configuration for band 1, a similar configuration can be created for the other bands.
| The start frequency of the calibration. For example 10 kHz. |
| The stop frequency of the calibration. For example 1 MHz. |
| The frequency step, in this case 5%. |
| The test level, in this case current. |
| The variable test level, select the correction file created earlier. |
| The regulation tolerance. |
| The power measurement that should be the result of the calibration: Forward power |
| Calibration test equipment. |
| The engineer that performed the calibration. |
Calibration result[edit]
Verification[edit]
CS114 requires to perform a verification of the setup with the current probe in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit.
Verification equipment[edit]
| Cables | |||
Configure the verification[edit]
Tests
Conducted Immunity
Multiband
| Band 1 | 10 kHz - 1 MHz | 10 % |
| Band 2 | 1 MHz - 30 MHz | 2 % |
| Band 2 | 30 MHz - 200 MHz | 1 % |
| The start frequency of the verification. For example 10 kHz. |
| The stop frequency of the verification. For example 1 MHz. |
| The frequency step, see 5.12.3.4.c.3. |
| The dwell time specified see TABLE II of the MIL-STD-461. |
| The method that is used to change from one to the next frequency: Constant. |
| Modulation is optional during the verification. |
| The equipment needed for the verification. |
| Injection clamp. |
| Specify the test level Current substitution method. |
| Add the Current sensor as an input. |
| The description of the test level. |
| Select the calibration file created during the calibration procedure. |
| Variable current and specify the correction file for the applicable test level created during the calibration procedure. |
| Select Forward Power. |
| Specify the tolerance to use. |
Add an input Current sensor, this will ensure that the current is measured during the test.
Verification result[edit]
EUT Testing[edit]
EUT Testing equipment[edit]
| Cables | |||
Configure the EUT test[edit]
Tests
Conducted Immunity
Multiband
| The start frequency of the test. For example 10 kHz. |
| The stop frequency of the test. For example 1 MHz. |
| The frequency step, see Table III. |
| Configure Pulse Modulation with 1 kHz and 50% duty cycle. |
| The equipment needed for the conducted immunity test. |
| Injection clamp. |
| Specify the Current substition method and add a limit Measured Current Limit |
| No inputs are needed. |
| Check this checkbox to enable the pulse modulation. |
| The PM frequency 1 kHz. |
| A duty cycle of 50%. |
| Check this checkbox to enable modulation outside the dwell-time. |
| Select the calibration file to use for this band. |
| Variable current, and specify the correction file for the applicable test level. |
| The description of the limit |
| The current to limit on. |
EUT Test Result[edit]
The MIL-STD-461G CS114 calibration can be performed by using the Conducted immunity calibration. The RadiMation® Conducted immunity multiband test can then be used to perform the MIL-STD-461G CS114 verification and the EUT test with a calibration file and the current sensor used as a limit in the EUT test.