In this mode the software increases the angle of the turntable from the initial point with a step equal to {{ScreenElement|Step size}} as long as the measurement value is not more than {{ScreenElement|Threshold}} below the already known maximum measurement value. As soon as measurement value is detected that is {{ScreenElement|Threshold}} below the already known | In this mode the software increases the angle of the turntable from the initial point with a step equal to {{ScreenElement|Step size}} as long as the measurement value is not more than {{ScreenElement|Threshold}} below the already known maximum measurement value. As soon as a measurement value is detected that is {{ScreenElement|Threshold}} below the already known maximum value, the turntable movement is reversed until the initial point and from this initial point the angle of the turntable is decreased with a step equal to {{ScreenElement|Step size}} until a measurement value is measured that is {{ScreenElement|Threshold}} below the already known maximum measurement value. The step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak. | ||
{{Warning|This might cause a measurement error if the intial number of steps ({{ScreenElement|#}}) is set to a low value.}} | {{Warning|This might cause a measurement error if the intial number of steps ({{ScreenElement|#}}) is set to a low value.}} | ||
Revision as of 12:13, 6 February 2025
Radiated emission[edit]
- Multiband
- Manual Mode
- Burst
- Anechoic chamber
- Fixed height
- GTEM
- Mil Std Magnetic
- Manual mode (Singleband)
- Merge
- X + k * S
Loading a TSF file[edit]
Tests
Test results screen[edit]
| Note: | The pass/fail criteria, shown in the tables do not take the measurement uncertainty of the test setup into account. |
Testsite information[edit]
| Antenna | Devices 1 | |
| Turn table | Devices 1 | |
| Cables | ||
Multiband[edit]
Frequency window[edit]
Turntable window[edit]
| The angle step size. |
Antenna tower window[edit]
| The height step size. |
Antenna window[edit]
Polarisation[edit]
- Horizontal
- Vertical
- Both
When this option is selected, all data will be displayed in one graph thus losing some polarisation information. However the peak list table will show if a peak was measured in horizontal or vertical polarisation.
If both emission graphs of horizontal and vertical polarisation are required, two separate tests must be configured. These tests can be put in a sequence, eliminating the need for a user intervention.
Distance[edit]
Receiver settings[edit]
| Sets the pre-amp gain. |
| Warning: | When older analyzers and receivers are used that switch between measuring in a linear or logarithmic scale during the pre-scan and final-measurement phases, it is advised to not use a fixed reference level |
Traces window[edit]
| The number of peak hold sweeps for each measurement can be selected. If this value is set to for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps depends on the type of signal and the setting of the sweep speed. For spectrum analysers and CW signals you can use a relatively low amount of peak sweeps, for example 10. But when you have discontinuous signals like a spark bridge, you may need 100 peak sweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope. |
Final peak measurement[edit]
Peak detection window[edit]
Test site window[edit]
Limit lines window[edit]
In the limit line window the operator selects one or more limit lines to be used during the test.
| In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 5 dB, the software will detect all peaks which are higher than 5 dB below the limit line. |
Change Order window[edit]
You need to perform a test as shown but you have 2 test sites. One with an antenna with the frequency range that is equal to the total band. At the second test site you have 2 antennas, one for every frequency band. For test site one you want the RBW filter to change as little as possible to prevent the old analyser from breaking down.
So for Site 1 you want:
- Band A RBW 9 kHz
- Band B RBW 9 kHz
- Band A RBW 120 kHz
- Band B RBW 120 kHz
And for test site 2:
- Band A RBW 9 kHz
- Band B RBW 9 kHz
For test site one, we do not have to change the order. Why? you may wonder. The explanation is quite easy. Look at the frequency change list, first change frequency band then change RBW. That is exactly what you want. Let's check what the software is going to do:
- Band A RBW 9 kHz
- Band B RBW 9 kHz
For test site two, it is a bit more complicated. You might wonder why? It looks easy just put the RBW in front the Frequency band and you are done. So far you are correct, but a different antenna also means a different test site. So changing the test site should also be above the frequency band. When doing so, lets check it again. The top three settings:
- RBW
- Testsite
- Frequency band
- Band A RBW 9 kHz
- Band B RBW 9 kHz
Manual Mode[edit]
Turn table[edit]
Antenna tower[edit]
Test site[edit]
Final measurement[edit]
In the final peak measurement window the test engineer can configure the detectors used during the final emission measurements of the peaks. The first value is the time constant of the detector while the second value is used for the measure time at each frequency point for this detector. The time constant is sent to the receiver, to set the sample time of the detector. To select a detector, check the desired button.
Peaks window[edit]
| Triggers the receiver of analyser to take the configured amount of sweeps. |
Anechoic chamber[edit]
| Sets start frequency |
| Sets stop frequency |
Intelligent[edit]
| Warning: | This might cause a measurement error if the intial number of steps (#) is set to a low value. |
Complete area[edit]
| Number op locations from Max. height to Min. height that should be measured. |
Intelligent[edit]
In this mode the software increases the height of the antenna from the initial point with a step equal to Step size as long as the measurement value is not more than Threshold below the already known maximum measurement value. As soon as measurement value is detected that is Threshold below the already known meaximum value, the antenna movement is reversed until the initial point and from this initial point the distance of the clamp is decreased with a step equal to Step size until a measurement value is measured that is Threshold below the already known maximum measurement value. The step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak.
Complete area[edit]
Polarisation[edit]
- Horizontal
- Vertical
- Both
Distance[edit]
| Sets the pre-amp gain |
| Warning: | All these settings are generic. Therefore, always verify that the configured settings are also supported by the used hardware to ensure it is correctly used and no measurement faults are introduced. Use a fixed value to disabled interference of equipment intelligence. |
Max. hold measurements[edit]
Traces window[edit]
Detectors window[edit]
Test options window[edit]
| In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full tracebility, all test data should be stored. |
| Limits | Distortion type |
Loading a TSF file
Test Procedure[edit]
Info screen[edit]
| In the trace window you can control the traces that are shown in the graph by selecting or deselecting the checkbox. If during a measurement a trace is not used then the check box is grayed out. |
| In the table all the measured peaks are showed. The table also shows the limit values for that frequency point when a limit was given at the beginning of the test. |
Test site information[edit]
| Cables | ||
Fixed height[edit]
Intelligent[edit]
In this mode the software increases the angle of the turntable from the initial point with a step equal to Step size as long as the measurement value is not more than Threshold below the already known maximum measurement value. As soon as measurement value is detected that is Threshold below the already known meaximum value, the turntable movement is reversed until the initial point and from this initial point the angle of the turntable is decreased with a step equal to Step size until a measurement value is measured that is Threshold below the already known maximum measurement value. The step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak.
Complete area[edit]
Intelligent[edit]
Complete area[edit]
Polarisation[edit]
- Horizontal
- Vertical
- Both
Distance[edit]
Traces window[edit]
Loading a TSF file
| Cables | ||
GTEM[edit]
Correlation method[edit]
Measurement procedure[edit]
Tests
Radiated emission
G-TEM
Test configuration[edit]
OATS settings[edit]
G-TEM settings[edit]
Traces window[edit]
Max. Peaks
Starting a G-TEM test[edit]
| Cables | ||
| Sets the Resoution Band Width filter |
Data correction[edit]
[edit]
Traces Menu[edit]
Traces[edit]
Detectors[edit]
- Peak detector
- Average detector
- Quasi peak detector
- RMS detector
Zoom function[edit]
Limit lines[edit]
[edit]
Merge[edit]
X + k * S[edit]