RadiMation Application Note 164: Difference between revisions
(42 intermediate revisions by 2 users not shown) | |||
Line 11: | Line 11: | ||
* Signal generator | * Signal generator | ||
* Amplifier | * Amplifier | ||
* Oscilloscope or measurement receiver | * Oscilloscope or measurement receiver | ||
* 0.5 Ω resistor | * 0.5 Ω resistor | ||
* Coupling transformer | |||
== Define the calibration test level == | == Define the calibration test level == | ||
Line 22: | Line 21: | ||
To calculate the corresponding voltages based on the power levels the following formula can be used:</br> | To calculate the corresponding voltages based on the power levels the following formula can be used:</br> | ||
<math> | <math> | ||
\text{ | \text{Voltage [dBµV]} = 20 \cdot \log_{10}\left(\frac{\sqrt{P \cdot R}}{1 \times 10^{-6}}\right) | ||
</math> | </math> | ||
This | This results in the following voltages: | ||
{| class="wikitable" | {| class="wikitable" | ||
Line 37: | Line 36: | ||
A new correction file for the calibration test level can be created by selecting the menu: | |||
{{Menu|File|New|Correction}} | {{Menu|File|New|Correction}} | ||
Line 44: | Line 43: | ||
{{Note|Make sure to add the Vrms column with {{ScreenElement|Columns/Units}} and remove the {{ScreenElement|Correction column}}. In addition change the {{ScreenElement|Unit}} to dBµV.}} | {{Note|Make sure to add the Vrms column with {{ScreenElement|Columns/Units}} and remove the {{ScreenElement|Correction column}}. In addition change the {{ScreenElement|Unit}} to dBµV.}} | ||
Store the correction file for the power limit by selecting the menu: | |||
{{Menu|File|Save correction}} | {{Menu|File|Save correction}} | ||
Then specify a descriptive name for the test level correction file on disk, for example: 'CS101-2 power limit.cor'. | |||
== Calibration procedure == | == Calibration procedure == | ||
Line 65: | Line 65: | ||
| || Sensor power meter || Devices 2 || measurement receiver as power meter (in case no oscilloscope is used) | | || Sensor power meter || Devices 2 || measurement receiver as power meter (in case no oscilloscope is used) | ||
| || Injection device || Devices 2 || The coupling transformer to use | |||
=== Configure the calibration === | === Configure the calibration === | ||
The configuration of the calibration can be created by selecting the menu: | |||
{{Menu|Calibration|System calibration|Conducted susceptibility}} | {{Menu|Calibration|System calibration|Conducted susceptibility}} | ||
[[File:CS101_Calibration_configuration.png]] | |||
[[File: | |||
{{ScreenElementDescription|Start|The start frequency of the calibration | {{ScreenElementDescription|Start|The start frequency of the calibration: 30 Hz.}} | ||
{{ScreenElementDescription|End|The stop frequency of the calibration | {{ScreenElementDescription|End|The stop frequency of the calibration: 150 kHz.}} | ||
{{ScreenElementDescription|Step|The frequency step, | {{ScreenElementDescription|Step|The frequency step, for example: 5 %.}} | ||
{{ScreenElementDescription|Test level|The test level, in this case {{ScreenElement| | {{ScreenElementDescription|Test level|The test level, in this case {{ScreenElement|Voltage}}.}} | ||
{{ScreenElementDescription|Variable|The variable test level, select the correction file created | {{ScreenElementDescription|Variable|The variable test level, select the correction file created before 'CS101-2 power limit.cor'.}} | ||
{{ScreenElementDescription|Calibration method|The power measurement that should be the result of the calibration: {{ScreenElement|Signal generator level}}}} | |||
{{ScreenElementDescription|Tolerance|The regulation tolerance.}} | {{ScreenElementDescription|Tolerance|The regulation tolerance.}} | ||
{{ScreenElementDescription| | {{ScreenElementDescription|Voltage measuring device|The device that will be used to measure the voltage. In this example: {{ScreenElement|Oscilloscope}}.}} | ||
{{ScreenElementDescription|Test equipment|Calibration test equipment.}} | {{ScreenElementDescription|Test equipment|Calibration test equipment.}} | ||
{{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}} | {{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}} | ||
Line 94: | Line 88: | ||
=== Calibration result === | === Calibration result === | ||
When the calibration is | When the calibration is finished, the engineer is asked to save the results of that calibration to a calibration (.CAL) file. It is possible to give the .CAL file a clear and distinct name. | ||
== EUT Testing == | == EUT Testing == | ||
=== EUT Testing equipment === | === EUT Testing equipment === | ||
[[File: | [[File:CS101_EUT_test_setup.png]] | ||
The configuration of the eut test site should contain the following devices: | The configuration of the eut test site should contain the following devices: | ||
Line 192: | Line 103: | ||
| || Amplifier || Devices 1 || The amplifier to use | | || Amplifier || Devices 1 || The amplifier to use | ||
| || | | || Oscilloscope || Devices 1 || The oscilloscope to measure the voltage | ||
| || Sensor powermeter || Devices 2 || When a measurement receiver is used to measure voltage | |||
| || Sensor powermeter || Devices 2 || | |||
=== Configure the EUT test === | === Configure the EUT test === | ||
To perform the actual test on the EUT, create a conducted immunity multiband test | To perform the actual test on the EUT, create a conducted immunity multiband test. | ||
The conducted immunity multiband test | The conducted immunity multiband test configuration is created by selecting the menu: | ||
{{Menu|Tests|Conducted Immunity|Multiband}} | {{Menu|Tests|Conducted Immunity|Multiband}} | ||
[[File:CS101 EUT Test Configuration.png]] | |||
[[File: | |||
{{ScreenElementDescription|Start|The start frequency of the test | {{ScreenElementDescription|Start|The start frequency of the test: 30 Hz.}} | ||
{{ScreenElementDescription|End|The stop frequency of the test | {{ScreenElementDescription|End|The stop frequency of the test: 150 kHz.}} | ||
{{ScreenElementDescription|Frequency Step|The frequency step, see Table III.}} | {{ScreenElementDescription|Frequency Step|The frequency step, see Table III of the [[MIL-STD-461]].}} | ||
{{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II of the [[MIL-STD-461]].}} | {{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II of the [[MIL-STD-461]].}} | ||
{{ScreenElementDescription|Frequency Change Mode|The method that is used to change from one to the next frequency: {{ScreenElement|Constant}}.}} | {{ScreenElementDescription|Frequency Change Mode|The method that is used to change from one to the next frequency: {{ScreenElement|Constant}}.}} | ||
{{ScreenElementDescription|Modulation| | {{ScreenElementDescription|Modulation|Modulation is not required.}} | ||
{{ScreenElementDescription|Test equipment|The equipment needed for the conducted | {{ScreenElementDescription|Test equipment|The equipment needed for the conducted susceptibility test.}} | ||
{{ScreenElementDescription|Location Type|Injection clamp.}} | {{ScreenElementDescription|Location Type|Injection clamp.}} | ||
{{ScreenElementDescription|Test Level|Specify the {{ScreenElement|Current | {{ScreenElementDescription|Test Level|Specify the {{ScreenElement|Current voltage closed loop method}} test level and add a {{ScreenElement|Calibrated voltage limit}}}} | ||
{{ScreenElementDescription|Inputs|No inputs are needed.}} | {{ScreenElementDescription|Inputs|No inputs are needed.}} | ||
{{Note|It is possible to change the voltage unit to dBµV instead of V, choose {{ScreenElement|Units}} and select Voltage and adjust the {{ScreenElement|Unit}} to dBµV.}} | |||
At {{ScreenElement| | At {{ScreenElement|Test Level}} click {{ScreenElement|Add}} to add a new Test level and select {{ScreenElement|Test Level - Voltage closed loop method}} | ||
[[File: | [[File:CS101 Voltage Testlevel.png]] | ||
{{ScreenElementDescription|Description|The description of the test level.}} | {{ScreenElementDescription|Description|The description of the test level.}} | ||
{{ScreenElementDescription| | {{ScreenElementDescription|Measuring device|Select the measuring device to use.}} | ||
{{ScreenElementDescription|Test level|Variable | {{ScreenElementDescription|Test level|{{ScreenElement|Variable voltage}}, and specify the correction file for the applicable test level, see CS101-1}} | ||
{{ScreenElementDescription|Tolerance|Specify the tolerance to use.}} | {{ScreenElementDescription|Tolerance|Specify the tolerance to use.}} | ||
Next to the test level add a limit, At {{ScreenElement|TestLevel}} click {{ScreenElement|Add}} to add a new Limit and select {{ScreenElement|Limit - | Next to the test level add a limit, At {{ScreenElement|TestLevel}} click {{ScreenElement|Add}} to add a new Limit and select {{ScreenElement|Limit - Calculated voltage limit}} | ||
With this limit the level will not be regulated higher than the calculated voltage based on the calibration file. | |||
With this limit the level will not be regulated higher than the | |||
[[File: | [[File:CS101 Voltage limit.png]] | ||
{{ScreenElementDescription|Description|The description of the limit}} | {{ScreenElementDescription|Description|The description of the limit.}} | ||
{{ScreenElementDescription| | {{ScreenElementDescription|Calibration file|The calibration file, that was created during the calibration.}} | ||
{{ScreenElementDescription|Variable voltage|The correction file for the power limit.}} | |||
{{ScreenElementDescription|Power method|{{ScreenElement|Signal Gen. Level (Pin)}}.}} | |||
When all | When all settings are configured press {{ScreenElement|Start Test}} to start the EUT test. | ||
=== EUT Test Result === | === EUT Test Result === | ||
Once the test is finished, the results of this test are stored in the EUT file and available as one of the performed {{ScreenElement|Tests}} in the EUT file. Selecting the corresponding test result and pressing on {{ScreenElement|Info}} will show the test results again. | Once the test is finished, the results of this test are stored in the EUT file and available as one of the performed {{ScreenElement|Tests}} in the EUT file. Selecting the corresponding test result and pressing on {{ScreenElement|Info}} will show the test results again. | ||
[[File: | [[File:CS101 EUT Test Result.png]] | ||
The MIL-STD-461G | The MIL-STD-461G CS101 calibration can be performed by using the Conducted susceptibility calibration. | ||
The {{RadiMation}} Conducted immunity multiband test can then be used to perform the MIL-STD-461G | The {{RadiMation}} Conducted immunity multiband test can then be used to perform the MIL-STD-461G CS101 EUT test with a voltage closed loop test level to regulate and a calculated voltage limit with a calibration file to limit on power during the EUT test. | ||
[[Category:RadiMation Application Note]] | [[Category:RadiMation Application Note]] |
Latest revision as of 16:19, 26 February 2025
How to perform a MIL-STD-461 CS101, Conducted susceptibility, power leads[edit]
This application note explains how the MIL-STD-461 CS101, Conducted susceptibility, power leads test can be performed with RadiMation®
MIL-STD-461 CS101 is a test method used in military and aerospace environments to assess the conducted susceptibility of equipment to low-frequency interference, 30 Hz - 150 kHz, ensuring it operates reliably in the presence of conducted disturbances.
The exact requirements and test methods for the CS101 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to run the calibration and EUT test:
- Signal generator
- Amplifier
- Oscilloscope or measurement receiver
- 0.5 Ω resistor
- Coupling transformer
Define the calibration test level[edit]
The CS101 EUT test is limited by the power levels specified in CS101-2, applied through a 0.5 Ω resistor.
During the calibration RadiMation® is levelling on a voltage and the corresponding signal source is recorded which is used during EUT testing to limit on.
Frequency | Power | Voltage |
30 Hz | 80 Watts | 136 dBµV |
5 kHz | 80 Watts | 136 dBµV |
150 kHz | 0.09 Watts | 106.5 dBµV |
File
New
Correction
Note: | Make sure to add the Vrms column with Columns/Units and remove the Correction column. In addition change the Unit to dBµV. |
File
Save correction
Calibration procedure[edit]
The calibration is performed to determine the signal source required for the corresponding power level through the 0.5 Ω resistor, which will be used during EUT testing as limit.
Calibration equipment[edit]
Oscilloscope | Devices 1 | Oscilloscope | |
Sensor power meter | Devices 2 | measurement receiver as power meter (in case no oscilloscope is used) | |
Injection device | Devices 2 | The coupling transformer to use |
Calibration
System calibration
The start frequency of the calibration: 30 Hz. |
The stop frequency of the calibration: 150 kHz. |
The frequency step, for example: 5 %. |
The test level, in this case Voltage. |
The variable test level, select the correction file created before 'CS101-2 power limit.cor'. |
The power measurement that should be the result of the calibration: Signal generator level |
The regulation tolerance. |
The device that will be used to measure the voltage. In this example: Oscilloscope. |
Calibration test equipment. |
The engineer that performed the calibration. |
Calibration result[edit]
EUT Testing[edit]
EUT Testing equipment[edit]
Oscilloscope | Devices 1 | The oscilloscope to measure the voltage | |
Sensor powermeter | Devices 2 | When a measurement receiver is used to measure voltage |
Configure the EUT test[edit]
To perform the actual test on the EUT, create a conducted immunity multiband test. The conducted immunity multiband test configuration is created by selecting the menu:
Tests
Conducted Immunity
Multiband
The start frequency of the test: 30 Hz. |
The stop frequency of the test: 150 kHz. |
The frequency step, see Table III of the MIL-STD-461. |
The dwell time specified see TABLE II of the MIL-STD-461. |
The method that is used to change from one to the next frequency: Constant. |
Modulation is not required. |
The equipment needed for the conducted susceptibility test. |
Injection clamp. |
Specify the Current voltage closed loop method test level and add a Calibrated voltage limit |
No inputs are needed. |
Note: | It is possible to change the voltage unit to dBµV instead of V, choose Units and select Voltage and adjust the Unit to dBµV. |
At Test Level click Add to add a new Test level and select Test Level - Voltage closed loop method
The description of the test level. |
Select the measuring device to use. |
Variable voltage, and specify the correction file for the applicable test level, see CS101-1 |
Specify the tolerance to use. |
Next to the test level add a limit, At TestLevel click Add to add a new Limit and select Limit - Calculated voltage limit
The description of the limit. |
The calibration file, that was created during the calibration. |
The correction file for the power limit. |
Signal Gen. Level (Pin). |
When all settings are configured press Start Test to start the EUT test.
EUT Test Result[edit]
The MIL-STD-461G CS101 calibration can be performed by using the Conducted susceptibility calibration. The RadiMation® Conducted immunity multiband test can then be used to perform the MIL-STD-461G CS101 EUT test with a voltage closed loop test level to regulate and a calculated voltage limit with a calibration file to limit on power during the EUT test.