To calculate the corresponding voltages based on the power levels the following formula can be used:</br>
<math>
<math>
\text{Voltage [dBuV]} = 20 \cdot \log_{10}\left(\frac{\sqrt{P \cdot R}}{1 \times 10^{-6}}\right)
\text{Voltage [dBµV]} = 20 \cdot \log_{10}\left(\frac{\sqrt{P \cdot R}}{1 \times 10^{-6}}\right)
</math>
</math>
Line 43: Line 43:
{{Note|Make sure to add the Vrms column with {{ScreenElement|Columns/Units}} and remove the {{ScreenElement|Correction column}}. In addition change the {{ScreenElement|Unit}} to dBµV.}}
Store the correction file for the test level by selecting the menu:
Store the correction file for the power limit by selecting the menu:
{{Menu|File|Save correction}}
Then specify a descriptive name for the test level correction file on disk, for example:
Then specify a descriptive name for the test level correction file on disk, for example: 'CS101-2 power limit.cor'.
== Calibration procedure ==
Line 75: Line 75:
[[File:CS101_Calibration_configuration.png]]
{{ScreenElementDescription|Start|The start frequency of the calibration. 30 Hz.}}
{{ScreenElementDescription|Start|The start frequency of the calibration: 30 Hz.}}
{{ScreenElementDescription|End|The stop frequency of the calibration. 150 kHz.}}
{{ScreenElementDescription|End|The stop frequency of the calibration: 150 kHz.}}
{{ScreenElementDescription|Step|The frequency step, 5%.}}
{{ScreenElementDescription|Step|The frequency step, for example: 5 %.}}
{{ScreenElementDescription|Test level|The test level, in this case {{ScreenElement|Voltage}}.}}
{{ScreenElementDescription|Variable|The variable test level, select the correction file created before 'CS101-2 power limit.cor'.}}
{{ScreenElementDescription|Calibration method|The power measurement that should be the result of the calibration: {{ScreenElement|Signal generator level}}}}
{{ScreenElementDescription|Tolerance|The regulation tolerance.}}
{{ScreenElementDescription|Voltage measuring device|The device that will be used to measure the voltage. In this example: {{ScreenElement|Oscilloscope}}.}}
{{ScreenElementDescription|Test equipment|Calibration test equipment.}}
{{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}}
Line 114: Line 115:
[[File:CS101 EUT Test Configuration.png]]
{{ScreenElementDescription|Start|The start frequency of the test. For example 30 Hz.}}
{{ScreenElementDescription|Start|The start frequency of the test: 30 Hz.}}
{{ScreenElementDescription|End|The stop frequency of the test. For example 150 kHz.}}
{{ScreenElementDescription|End|The stop frequency of the test: 150 kHz.}}
{{ScreenElementDescription|Frequency Step|The frequency step, see Table III.}}
{{ScreenElementDescription|Frequency Step|The frequency step, see Table III of the [[MIL-STD-461]].}}
{{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II of the [[MIL-STD-461]].}}
{{ScreenElementDescription|Frequency Change Mode|The method that is used to change from one to the next frequency: {{ScreenElement|Constant}}.}}
Line 127: Line 128:
{{Note|It is possible to change the voltage unit to dBµV instead of V, choose {{ScreenElement|Units}} and select Voltage and adjust the {{ScreenElement|Unit}} to dBµV.}}
At {{ScreenElement|Test Level}} click {{ScreenElement|Add}} to add a new Test level and select {{ScreenElement|Test Level - Voltage closed loop method}}
Line 134: Line 134:
{{ScreenElementDescription|Description|The description of the test level.}}
{{ScreenElementDescription|Calibration file|Select the calibration file to use.}}
{{ScreenElementDescription|Measuring device|Select the measuring device to use.}}
{{ScreenElementDescription|Test level|Variable voltage, and specify the correction file for the applicable test level, see CS101-1}}
{{ScreenElementDescription|Test level|{{ScreenElement|Variable voltage}}, and specify the correction file for the applicable test level, see CS101-1}}
{{ScreenElementDescription|Calibration method|Select signal power.}}
{{ScreenElementDescription|Tolerance|Specify the tolerance to use.}}
Line 145: Line 144:
[[File:CS101 Voltage limit.png]]
{{ScreenElementDescription|Description|The description of the limit}}
{{ScreenElementDescription|Description|The description of the limit.}}
{{ScreenElementDescription|Description|The calibration file}}
{{ScreenElementDescription|Calibration file|The calibration file, that was created during the calibration.}}
{{ScreenElementDescription|Variable voltage|The correction file for the power limit.}}
{{ScreenElementDescription|Power method|Signal generator level}}
{{ScreenElementDescription|Power method|{{ScreenElement|Signal Gen. Level (Pin)}}.}}
Latest revision as of 16:19, 26 February 2025
How to perform a MIL-STD-461 CS101, Conducted susceptibility, power leads[edit]
This application note explains how the MIL-STD-461 CS101, Conducted susceptibility, power leads test can be performed with RadiMation®
MIL-STD-461 CS101 is a test method used in military and aerospace environments to assess the conducted susceptibility of equipment to low-frequency interference, 30 Hz - 150 kHz, ensuring it operates reliably in the presence of conducted disturbances.
The exact requirements and test methods for the CS101 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to run the calibration and EUT test:
  • Signal generator
  • Amplifier
  • Oscilloscope or measurement receiver
  • 0.5 Ω resistor
  • Coupling transformer
Define the calibration test level[edit]
The CS101 EUT test is limited by the power levels specified in CS101-2, applied through a 0.5 Ω resistor.
During the calibration RadiMation® is levelling on a voltage and the corresponding signal source is recorded which is used during EUT testing to limit on.
This results in the following voltages:
Frequency Power Voltage
30 Hz 80 Watts 136 dBµV
5 kHz 80 Watts 136 dBµV
150 kHz 0.09 Watts 106.5 dBµV
A new correction file for the calibration test level can be created by selecting the menu:
   Menu.svg File
      Menu.svg New
         Menu.svg Correction
Note: Make sure to add the Vrms column with Columns/Units and remove the Correction column. In addition change the Unit to dBµV.
   Menu.svg File
      Menu.svg Save correction
Calibration procedure[edit]
The calibration is performed to determine the signal source required for the corresponding power level through the 0.5 Ω resistor, which will be used during EUT testing as limit.
Calibration equipment[edit]
Oscilloscope Devices 1 Oscilloscope
Sensor power meter Devices 2 measurement receiver as power meter (in case no oscilloscope is used)
Injection device Devices 2 The coupling transformer to use
The configuration of the calibration can be created by selecting the menu:
   Menu.svg Calibration
      Menu.svg System calibration
         Menu.svg Conducted susceptibility
ScreenElementDescription.svg Start The start frequency of the calibration: 30 Hz.
ScreenElementDescription.svg End The stop frequency of the calibration: 150 kHz.
ScreenElementDescription.svg Step The frequency step, for example: 5 %.
ScreenElementDescription.svg Test level The test level, in this case Voltage.
ScreenElementDescription.svg Variable The variable test level, select the correction file created before 'CS101-2 power limit.cor'.
ScreenElementDescription.svg Calibration method The power measurement that should be the result of the calibration: Signal generator level
ScreenElementDescription.svg Voltage measuring device The device that will be used to measure the voltage. In this example: Oscilloscope.
ScreenElementDescription.svg Test equipment Calibration test equipment.
ScreenElementDescription.svg Test engineer The engineer that performed the calibration.
Calibration result[edit]
When the calibration is finished, the engineer is asked to save the results of that calibration to a calibration (.CAL) file. It is possible to give the .CAL file a clear and distinct name.
EUT Testing[edit]
EUT Testing equipment[edit]
Oscilloscope Devices 1 The oscilloscope to measure the voltage
Sensor powermeter Devices 2 When a measurement receiver is used to measure voltage
Configure the EUT test[edit]
To perform the actual test on the EUT, create a conducted immunity multiband test. The conducted immunity multiband test configuration is created by selecting the menu:
   Menu.svg Tests
      Menu.svg Conducted Immunity
         Menu.svg Multiband
ScreenElementDescription.svg Start The start frequency of the test: 30 Hz.
ScreenElementDescription.svg End The stop frequency of the test: 150 kHz.
ScreenElementDescription.svg Frequency Step The frequency step, see Table III of the MIL-STD-461.
ScreenElementDescription.svg Dwell Time The dwell time specified see TABLE II of the MIL-STD-461.
ScreenElementDescription.svg Frequency Change Mode The method that is used to change from one to the next frequency: Constant.
ScreenElementDescription.svg Modulation Modulation is not required.
ScreenElementDescription.svg Test equipment The equipment needed for the conducted susceptibility test.
ScreenElementDescription.svg Test Level Specify the Current voltage closed loop method test level and add a Calibrated voltage limit
Note: It is possible to change the voltage unit to dBµV instead of V, choose Units and select Voltage and adjust the Unit to dBµV.
At Test Level click Add to add a new Test level and select Test Level - Voltage closed loop method
ScreenElementDescription.svg Measuring device Select the measuring device to use.
ScreenElementDescription.svg Test level Variable voltage, and specify the correction file for the applicable test level, see CS101-1
Next to the test level add a limit, At TestLevel click Add to add a new Limit and select Limit - Calculated voltage limit With this limit the level will not be regulated higher than the calculated voltage based on the calibration file.
ScreenElementDescription.svg Description The description of the limit.
ScreenElementDescription.svg Calibration file The calibration file, that was created during the calibration.
ScreenElementDescription.svg Variable voltage The correction file for the power limit.
ScreenElementDescription.svg Power method Signal Gen. Level (Pin).
When all settings are configured press Start Test to start the EUT test.
EUT Test Result[edit]
The MIL-STD-461G CS101 calibration can be performed by using the Conducted susceptibility calibration. The RadiMation® Conducted immunity multiband test can then be used to perform the MIL-STD-461G CS101 EUT test with a voltage closed loop test level to regulate and a calculated voltage limit with a calibration file to limit on power during the EUT test.