RadiMation Application Note 165: Difference between revisions
Line 41: Line 41:
|① || Cable current -> power meter || Cables || Cable with the specified loss of the used attenuator
=== Configure the calibration ===
The calibration can be configured and started from the menu by selecting:
{{Menu|Calibration|System calibration|Conducted immunity}}
[[File:CI Calibration configuration.png]]
{{ScreenElementDescription|Start|The start frequency of the calibration. For example 150 kHz.}}
{{ScreenElementDescription|End|The stop frequency of the calibration. For example 80 MHz.}}
{{ScreenElementDescription|Step|The frequency step, in this case 1%.}}
{{ScreenElementDescription|Test level|The test level, in this case {{ScreenElement|Voltage}}.}}
{{ScreenElementDescription|Voltage|The voltage test level, 10 V.}}
{{ScreenElementDescription|Tolerance|The regulation tolerance.}}
{{ScreenElementDescription|Calibration method|The power measurement that should be the result of the calibration: {{ScreenElement|Forward power}}}}
{{ScreenElementDescription|Test equipment|Calibration test equipment.}}
{{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}}
Revision as of 14:10, 9 September 2025
How to perform an IEC 61000-4-6, Conducted immunity test[edit]
This application note explains how the IEC 61000-4-6 conducted immunity test can be performed with RadiMation®. The IEC 61000-4-6 conducted immunity test method is used for electrical and electronic equipment that may be subjected to radio-frequency disturbances conducted along cables coming from radio transmitters in the frequency range from 150 kHz up to 80 MHz. The exact requirements, calibration and EUT test methods are described in the IEC 61000-4-6 standard.
Necessary equipment[edit]
  • Signal generator
  • Amplifier
  • Coupler
  • Forward power meter
  • Injection device
  • Calibration jig
  • Attenuator (optional)
  • Sensor power meter
Calibration procedure[edit]
The calibration is performed to determine the power required to match the corresponding voltage test level seen at the output of the injection device.
Coupler Devices 1 Coupler
Forward power meter Devices 1 Forward power meter
Injection device Devices 2 The injection clamp or CDN
Calibration Jig Devices 2 The calibration jig to use
Sensor power meter Devices 2 Power meter
Cables
Cable current -> power meter Cables Cable with the specified loss of the used attenuator
   Menu.svg Calibration
      Menu.svg System calibration
         Menu.svg Conducted immunity
ScreenElementDescription.svg Start The start frequency of the calibration. For example 150 kHz.
ScreenElementDescription.svg End The stop frequency of the calibration. For example 80 MHz.
ScreenElementDescription.svg Step The frequency step, in this case 1%.
ScreenElementDescription.svg Test level The test level, in this case Voltage.
ScreenElementDescription.svg Voltage The voltage test level, 10 V.
ScreenElementDescription.svg Calibration method The power measurement that should be the result of the calibration: Forward power
ScreenElementDescription.svg Test equipment Calibration test equipment.
ScreenElementDescription.svg Test engineer The engineer that performed the calibration.
