RadiMation Application Note 165: Difference between revisions
| Line 94: | Line 94: | ||
{{ScreenElementDescription|Test equipment|The equipment needed for the conducted immunity test.}} | {{ScreenElementDescription|Test equipment|The equipment needed for the conducted immunity test.}} | ||
{{ScreenElementDescription|Location Type|Injection clamp.}} | {{ScreenElementDescription|Location Type|Injection clamp.}} | ||
{{ScreenElementDescription|Test Level|Specify the {{ScreenElement|Current substition method | {{ScreenElementDescription|Test Level|Specify the {{ScreenElement|Current substition method}}}} | ||
{{ScreenElementDescription|Inputs|No inputs are needed.}} | {{ScreenElementDescription|Inputs|No inputs are needed.}} | ||
Revision as of 14:58, 9 September 2025
How to perform an IEC 61000-4-6, Conducted immunity test[edit]
This application note explains how the IEC 61000-4-6 conducted immunity test can be performed with RadiMation®. The IEC 61000-4-6 conducted immunity test method is used for electrical and electronic equipment that may be subjected to radio-frequency disturbances conducted along cables coming from radio transmitters in the frequency range from 150 kHz up to 80 MHz. The exact requirements, calibration and EUT test methods are described in the IEC 61000-4-6 standard.
Necessary equipment[edit]
- Signal generator
- Amplifier
- Coupler
- Forward power meter
- Injection device
- Calibration jig
- Attenuator (optional)
- Sensor power meter
Calibration procedure[edit]
The calibration is performed to determine the power required to match the corresponding voltage test level seen at the output of the injection device.
| # | Device name | Tab in testsite configuration window | note |
| Signal Generator | Devices 1 | The signal generator to use | |
| Amplifier | Devices 1 | The amplifier to use | |
| Coupler | Devices 1 | Coupler | |
| Forward power meter | Devices 1 | Forward power meter | |
| Injection device | Devices 2 | The injection clamp or CDN | |
| Calibration Jig | Devices 2 | The calibration jig to use | |
| Sensor power meter | Devices 2 | Power meter | |
| Cables | |||
| ① | Cable current -> power meter | Cables | Cable with the specified loss of the used attenuator |
The calibration can be configured and started from the menu by selecting:
Calibration
System calibration
Conducted immunity
| The start frequency of the calibration. For example 150 kHz. |
| The stop frequency of the calibration. For example 80 MHz. |
| The frequency step, in this case 1%. |
| The test level, in this case Voltage. |
| The voltage test level, 10 V. |
| The regulation tolerance. |
| The power measurement that should be the result of the calibration: Forward power |
| Calibration test equipment. |
| The engineer that performed the calibration. |
When the calibration has finished, RadiMation® will ask to store the calibration file.
EUT Test[edit]
EUT testing equipment[edit]
| Coupler | Devices 1 | The coupler to use | |
| Forward power meter | Devices 1 | The forward power meter to use | |
| Injection device | Devices 2 | The injection clamp or CDN to use |
Configure the EUT test[edit]
To perform the actual test on the EUT, create a conducted immunity multiband test. The conducted immunity multiband test, can be started by selecting from the menu:
Tests
Conducted Immunity
Multiband
| The start frequency of the test. For example 150 kHz. |
| The stop frequency of the test. For example 80 MHz. |
| The frequency step, 1%. |
| The dwell time. |
| The method that is used to change from one to the next frequency: Constant. |
| Configure Amplitude Modulation with 1 kHz and 80 % duty cycle. |
| The equipment needed for the conducted immunity test. |
| Injection clamp. |
| Specify the Current substition method |
| No inputs are needed. |
