Release Notes/Solved items Version 5.8: Difference between revisions
(6 intermediate revisions by 2 users not shown) | |||
{{:Release Notes/Solved items Version 5.8.38}} | |||
{{:Release Notes/Solved items Version 5.8.37}} | |||
{{:Release Notes/Solved items Version 5.8.36}} | |||
{{:Release Notes/Solved items Version 5.8.35}} | |||
{{:Release Notes/Solved items Version 5.8.34}} | |||
{{:Release Notes/Solved items Version 5.8.33}} | |||
{{:Release Notes/Solved items Version 5.8.32}} | |||
{{:Release Notes/Solved items Version 5.8.31}} | |||
{{:Release Notes/Solved items Version 5.8.30}} | |||
{{:Release Notes/Solved items Version 5.8.29}} | |||
{{:Release Notes/Solved items Version 5.8.28}} | |||
{{:Release Notes/Solved items Version 5.8.27}} | {{:Release Notes/Solved items Version 5.8.27}} | ||
{{:Release Notes/Solved items Version 5.8.26}} | {{:Release Notes/Solved items Version 5.8.26}} |
Latest revision as of 15:04, 20 December 2012
RadiMation version 5.8.38[edit]
- #4409: Switchmatrix device driver is not notified when polarization is finished
- #4338: Communication with a Narda NBM550 is very slow
- #4235: Pulse modulation with R&S SMT
- #4221: Actions don't work in Test Equipments menu
- #4201: Driver for the 120S1G4
- #4200: Driver: EMControl Turntable
- #4196: Total testtime of immunity test with all RadiCentre equipment is too long
- #4194: Turntable INNCO CO1000 is not working anymore in 5.8.37
- #4187: Control of a 3 axis antenna system from a R&S receiver
- #4173: RadiCentre RSW1004A’s – Comport is already open error.
- #4156: Driver: AVL Turntable / EUT Controller
RadiMation version 5.8.37[edit]
- #4462 Error You are not licensed to use the Report Generator
- #4461 Driver updates from the Installation and Training visit
- #4452 Driver: default RS232 speed RadiCentre RPR1006A changed
- #4429 Is LLF information saved independently with test data
- #4420 Questions validation and calibration
- #4412 General questions about Report Generator
- #4401 Compatibility controlling RadiPower units with old RM
- #4400 Statements for sorting columns with a space
- #4391: Driver: ETS-Lindgren 7604
- #4307: R&S ESU40: set measurement bandwidth to 6dB
- #4305: Cause of the difference in noise in region below 1 kHz
- #4304: Driver: no communication PMM EP601
- #4302: Request for the latest version of RadiMation on CD
- #4300: Possibility to set amount of Reflected Power during Power Delivery Test
- #4283: Driver: Milmega 80RF1000-175 Amplifier
- #4268: Driver: wrong frequency range Agilent N5181A-503
- #4265: Implementing ESPC driver with option B3
- #4243: Wrong step size (rounding) in MB
- #4193: Issue auto perform CS114
- #4189: Show limit curve on graph during Immunity tests
- #4188: Performing RS101 susceptibility test described in MILSTD F
- #4183: Driver for R&S ESR
- #4176: Driver: AR PM2003
- #4168: RadiSwitch problem during initialisation
- #4154: Report generator crashes when a single band emission graph is included
- #4149: Report generator error 91
- #4148: Driver: AR 5000A225
- #4135: Wrong forward power measured with NRVD
- #4099: INNCO 2000 that only controls the polarization
- #4086: TDEMI driver with option for multisampling
- #4058 Driver: Luthi CDN L-801 AF4
- #544: Include RI and CI manual mode directly in main menu
RadiMation version 5.8.36[edit]
- #4141: Driver: TTi TGA1230
- #4138: Driver: Frankonia FC 05
- #4136: RadiSense device driver raises a 'no response received' error when controlled in a USB controlled RadiCentre 1
- #4134: Advantest TR4172 not working correctly in Multiband
- #4133: DLL for R&S SMF100 cannot be found (RSSMF_32.DLL)
- #4131: Agilent N9038A only activates the low frequency band preamplifier
- #4130: Configurable Signal generator device driver should turn carrier off during the deinitialization
- #4129: When turntable overshoots it's destination angle, RadiMation keeps waiting
- #4127: Selecting input2 of the Agilent N9038A raises a frequency error
- #4126: Driver: PMM 9060 and PMM 9180
- #4125: Errormessage "Com port is already open" for SOPM01 and SOPM03
- #4124: Unable to use the 10 MHz RBW on the FSP30
- #4123: Error messages: Unable to find device driver for ...
- #4121: Version number of the device driver DLL's are confusing
- #4120: FSEK 30 only has a single input
- #4118: Driver: Powermeter for IFI SMC 100
- #4117: Driver: EMCO1085 (as a turntable controller)
- #4116: R&S ESMI spectrum analyser shows strange emission shapes
- #4115: Driver: Maturo MCU as clamp positioner
- #4113: Problem with identifier of Agilent powermeter
- #4110: Report for a small frequency span from a large scan
- #4107: PMM timeout when cancelling multiband test
- #4106: Error 113 on the display of SMA 100A
- #4105: Agilent N9038A with multiple detectors
- #4104: Agilent 9038 in receiver mode in GTEM-test
- #4097: Keyword for test level in EFT tests
- #4085: Driver problem with Agilent N9038A
- #4082: Driver: Rohde & Schwarz ESL6
- #4075: Turntable from Frankonia uses unit cm
- #4047: First point from final measurement is not measured correctly
- #4041: Antenna mast Frankonia FC 02 not working
- #4030: TDEMI controlled LISNs
- #3954: PMM uses wrong RBW and Measurement time in conducted emission tests
- #3953: Error when input from PMM is switched from low to high (9 kHz RBW in high input)
- #3935: Driver: PMM Controlled LISN
- #3775: wrong spelling in advanced settings for Maturo MC32 antenna tower
- #3723: Configurable CDN
- #3613: driver for R&S ESL 3
RadiMation version 5.8.35[edit]
- #4102: LISN Emission graph shows a large amount of zero's for the frequency on the x-axis
- #4093: Command error (4198) during start of report generation
- #4083: Driver: SMC 100A via usb/ethernet
- #4077: the PULSE modulation is not working with R&S SMP02 generator
- #4073: Keithley2000 driver as multimeter
- #4070: System calibration for attenuation/gain (Cable calibration) is complaining of amplifier protection
- #4054: Zerospan command is not allowed with Agilent N9000A
- #4024: The canoe driver is giving zero back instead of signal value
- #3993: Support for more then 8 channels in the configurable ad convertor
RadiMation version 5.8.34[edit]
- #4091: 40 GHz end freqency is too low for measurement.
- #4076: RadiSense III not working
RadiMation version 5.8.33[edit]
- #4065: Limit line is not shown between 5 and 30 MHz in a CE LISN test
- #4055: Older embedded software of the R&S SMX needs the space removed for the frequency
- #4048: Driver: Prana DT25DC RF
- #4045: ESU 8 Input 2 not working as sensor powermeter
- #4044: lower start frequency for the vector/signal generator Agilent E8267D which supports downrange
- #4043: PMM 9010/30P is giving error back for setting the RBW while by hand this is possible
- #4040: Driver for signal generator Agilent 33220A
- #4039: Driver for the AR 350AH1
- #2728: Driver: Switch Matrix VSE Electronic ISF – 1HE 2/4-2
- #1565: Possibility to specify attenuation of coupler as absolute value and not as a correction of the factory default value
RadiMation version 5.8.32[edit]
- #4211: Error report generator
- #4037: Power substitution emission measurements in dBm cannot use turntable optimization
- #4035: Save EUT file as on a different location does not work on Windows 7
- #4034: Excel Exporter crashes when some specific EUT data is exported
- #4029: N9000A driver is crashing at about 70% of the sweep when using over lan
- #4025: Using N9020A with configurable number of points equal to 10001 and higher result in complete crash of Radimation.
- #2505: Bonn Elektronik BLWA 2010-50 and BSA 0122-100 driver is not working
RadiMation version 5.8.31[edit]
- #3098: Agilent 4417 device driver is not working correctly anymore
- #3099: Driver for the AR 25S1G4A doesn't work correctly anymore
- #3102: Using the HP 33250 as an external modulation states that wrong GPIB commands are used
- #3407: Driver: SC1000
- #3733: Driver: Milmega AS0104-200/200
- #4000: Driver for spectrum analyser Agilent N9320A
- #4010: Limit lines are not displayed correctly in the graph
- #4016: ESU error: More points received from the receiver/analyser than calculated
- #4021: RE multi band 360 degrees area does not work all the time
RadiMation version 5.8.30[edit]
- #3863: report generator code horizontal and vertical max peak are not working correctly as the complete frequency range is not displayed.
- #3884: The configuration window of the configurable switch is not closing when pressing ok
- #3886: CANoe Analyser CLSID cannot be found
- #4007: Final measurement IF overload is not detected correctly
- #4013: Configurable amplifier is crashing when settings the frequency range, driverversion 2011-6-7
RadiMation version 5.8.29[edit]
- #2616: First point of a calibration is too low
- #3887: Agilent E4446A analyser should use the correct RBW setting for MIL-STD
- #4002: Power meter device driver Advantest 3361B is giving event Time out
- #4011: Reviewing Harmonic test is crashing RadiMation
RadiMation version 5.8.28[edit]
- #3867: Net power and calculated current on the CAN-bus
- #3999: Driver Spectrum Analyser HP E7405A does not work any more after return from agilent service point
- #4006: No prescan is done with FSV40
- #4009: No final measurement is done with FSP 3 and LISN in multiband
RadiMation version 5.8.27[edit]
- #2601: Test initialization window should also show the name of the test-site that is being initialized
- #3976: Agilent N9020A: Configurable number of points
- #3998: Driver for AR500A250A
- #4004: Report generator code for emission tests in multiband
- #4008: Measurement with HF-probe
RadiMation version 5.8.26[edit]
- #4003: Closing the Visa window is crashing RadiMation with new gcc compiler
- #3990: Driver: FSEA20
- #3986: Configurable power meter driver can not find expression value while the value is send over the GPIB bus
- #1942: No check for connected device is included in device driver of Marconi 6960
RadiMation version 5.8.25[edit]
- #3543: Unclear message when the RadiPower is not connected to the RadiCentre
- #3783: Advantest Analyser is not measuring CE peaks correctly with version 5.7.4
- #3903: Advantest R3361B analyser raises timeout errors in a Multiband Emission test
- #3937: Scope driver Multi band is decreasing the scale when the returned value is zero and the scale is still maximum
- #3988: Report Generator crashes when a ) is included in the template filename
RadiMation version 5.8.24[edit]
- #3920: Driver: Marconi 2390
- #3965: Is not possible to choose between EXT-AC and EXT-DC using the siggen HP8648A.
- #3979: Compression test fails when the amplifier has correction files attached
- #3980: device driver for amplifier AR 60S1G4
- #3982: Engineer system in RadiMation allows impossible selections and removals.
- #3983: Modulation is always marked red after stopping and continuing RI multiband test
- #3984: Allow more than 8 AD-channels
RadiMation version 5.8.23[edit]
- #3969: Selecting the lowest location type in a multiband Lisn test configuration results in a crash of the software
- #3975: Configured sweeptime of a band is not used, due to sweeptime/MHz calculation
- #3977: Changing reference level is not always allowed
- #3978: Agilent 4446A: Sweeptime is changing when detector is selected
RadiMation version 5.8.22[edit]
- #3964: Pauses in a sequence are crashing the software
- #3966: Peak not detected in conducted emission current probe test
RadiMation version 5.8.21[edit]
- #2162: Create base class for the RadiCentre drivers
- #3795: Adding the device driver version to the about screen of RadiMation
- #3921: Extend driver of RadiTower to control a seperate RadiPol
- #3967: Pause test can not be used; problem during creating it
- #3968: Multiband emission test files cannot be opened anymore.
RadiMation version 5.8.20[edit]
- #3960: HD 100 Antenna tower driver is always in max speed.
- #3948: Agilent E4446A is not keeping 50 sec sweeptime in multiband test
RadiMation version 5.8.19[edit]
- #3910: External AM sine and square is not working with the Agilent PSG E8257D
- #3932: Driver: AR FP4080 controlled as an AR FP3000
- #3949: Printout is not generating the same plot twice, and the printout during the test is a combination of both
- #3652: Conducted emission - current probe TSF is saved in conducted immunity folder
- #3956: Agilent E4440A as a powermeter
- #3957: Registration of STDOLE2.TLB Error message during installation on Windows 7
- #3958: Some device drivers are missing
- #3959: Report generator code AVERAGE DETECTOR uses information of Peak Detector
RadiMation version 5.8.18[edit]
- #3925: R&S FSV13 does supports scans from 10Hz
- #3926: RBW is not correctly set on the R&S FSV13
- #3927: Raditurn is giving already in progress after motor error
- #3934: Several Multiband Conducted Emission Current probe problems
- #3938: Incorrect limit line when using Merge function
- #3947: AD/dc selection in power meter driver advantest is giving problems
- #3950: Zoom-out an emission graph by drawing a reverse zoom-in box
RadiMation version 5.8.17[edit]
- #3922: Copyright is 2010 instead of 2011
- #3941: Agilent E4417A power meter is limited to 18 GHz
- #3942: Driver: Agilent N9000A
- #3943: Driver: Agilent N9030A
- #3944: FAQ is not explaining why the measured value of the field probe can be higher with the am modulation on.
- #3945: How does the cable calibration works in the wiki page
- #3946: explain what is the difference between a printout during the test and a print out in the eut window
RadiMation version 5.8.16[edit]
- #3905: Test sites are lost in from time to time without any leads.
RadiMation version 5.8.15[edit]
- #3924: Sweeptime is not set in the Agilent N9020A
RadiMation version 5.8.14[edit]
- #3876: Gaps in emisson graph after final measurement
- #3894: Function "Continuous Measure" can block RadiMation with an error warning
- #3895: Test-site from a band can not be chosen after pre-scan
- #3904: Setting of the sweep time of the HP 4395A is not possible
- #3909: Pre amplifier is not yet turned on while sweeping
- #3917: Nonexistent peak and wrong final measurement with PMM during conducted emission multiband tests
- #3918: FSH 3 is setting wrong detector
- #3928: Driver: HP 4396A
RadiMation version 5.8.13[edit]
- #1856: No possibility to specify the modulation depth and frequency for external AM modulation
- #3196: Driver: Agilent N9020A
- #3875: Question: Radiated Emission Multiband: Order of measured peaks in the final measurement
- #3896: Older test results show a low resolution Multiband Emission graph
- #3902: The command ||LISN LINE|| is not recognized anymore by the report generator
- #3906: Scaling of the graphs can make the AD convertor graphs unusable
- #3907: Angle and height of detected peaks are always the same (in singleband emission)
- #3908: Agilent 4446A cannot be used below 9 kHz
- #3911: Driver: Agilent 83640A
- #3914: Report of MB emission tests contains a large list of detected peaks, if no peaks are detected
- #3915: Use a Microsoft SQL Server database to retrieve customer information
- #3916: External PM is not allowing the selection of modulation out side the dwell time
- #3919: Possibility to draw graphs via report generator with the full dataset
RadiMation version 5.8.12[edit]
- #3857: PMM 9030 driver is giving time out during measurement of peak
- #3877: Slow final measurement with PMM 9010/30P via RS232
- #3878: PMM 9010/30P hangs up in final measurement
- #3885: Peak and Average measurement at the same time with PMM not possible
- #3891: TSF does not save measure- and observation time from "optimize" - settings
- #3892: Wrong detector is used in final measurement
- #3899: Y-axis unit is not shown anymore in singleband emission manual mode
- #3900: Observation time of a singleband emission test is always 1 second
- #3901: Opening Multiband Emission tests is slower
RadiMation version 5.8.11[edit]
- #3308: Measure with ESVS10/ESHS10 with PK and AVG detectors together: average trace is not visible until the sweep is finished and the method is saved, closed and open again
- #3868: Creating new 16 point verification on a chinese pc is displaying a message box and not crashing
- #3888: Graphic of correction file not displayed
- #3889: RadiMation manual contains a lot of unrelated text
- #3890: High peaks in the emission graphs of the R&S ESHS
RadiMation version 5.8.10[edit]
- #3869: Saving a cor file is giving problems with ) at the end with a chinese pc
- #3882: CE Current probe test is crashing
- #3883: Multiband Emission graph is showing Bars
- #3913: list of detected peaks does not have all the Average values specified
RadiMation version 5.8.9[edit]
- #3791: Peak detection wrong in CE-MB (LISN) test
- #3843: Multiband emsion data file cannot be loaded anymore. It seems to be corrupt
- #3872: Reselecting the report is giving the .doc instead of .txt in the result file
RadiMation version 5.8.8[edit]
- #2915: Report generator template should be selectable in the file selection window
- #3757: Report Generator to ASCII
RadiMation version 5.8.7[edit]
- #457: Possibility to perform MIL-STD emission measurements in dBuA
- #469: Not possible to perform broadband measurements (dBuV/MHz)
- #475: Vertical scale of emission graph is not possible in dBuA
- #552: Configuration of the emission graph lines
- #784: Radiated emission graph dBuV should be dBuV/m
- #1300: Merge function - Scaling is wrong in Report Generator
- #1688: Sorting in unit configuration window is not working as expected
- #1704: IMaster unit transformer
- #1834: Possibility of measuring a magnetic field with a loop antenna (Unit dBµA/m)
- #2063: Changing the unit in a CE Current clamp test, doesn't change the graph
- #2205: Is fast scan mode supported in the multiband emission
- #2248: RE magnetic field MIL STD test reference value is always zero.
- #2274: Multiband Conducted Emission current probe
- #2327: RE multiband test can not be started from restart last test button
- #2413: Radiated emission substitution method test
- #2656: A multiband graphline is included multiple times on the graph
- #2966: Possibility to see the internal attenuation of a coupler device driver
- #3184: The unit of the reference level and the Emission is the same, so the correct reference level can not be reported
- #3381: Traces are not converted to the correct amplitude unit
- #3395: Limit lines hidden during pre-scan
- #3415: The sequence of the Multiband is according to the way it's been entered
- #3420: AdvantestTR4172 device driver is missing for input 1 50Hz - 1800MHz
- #3596: Limit Lines not shown during CE measurement in Multiband
- #3607: CE Multi band can only be performed with a LISN
- #3699: Possibility to set a default unit for radiated Emission and conducted Emission
- #3718: Limit line with unit
- #3719: Emission Test: Results are not converted to chosen unit - just the display is upgraded
- #3731: sx32w.dll is causing problems with the dongle in version 5.7.2
- #3784: RI multiband test is not turning on the amplifiers in version 5.7.4 while this was working in version 5.6.13
- #3830: Physical properties do not match with the units in the menus when using German language
- #3831: Support for the mil std dbuV/Mhz/M according to the 462C standard
- #3832: changing dbuV to dbuV/m does not change the reading in the display
- #3833: Narrow band / broad band detection according to the Mil std 462C standard
- #3835: Specification of the limit line in the configuration of the limit line
- #3844: dbuV/m/MHz axis is not displayed is not displayed in the mil std test
- #3845: BCI multiband is not displaying the lines correctly when the levels are identical
- #3852: Saving a file doesn't show a progress bar
- #3855: ESHS10 is scanning the frequency range 3 times
- #3856: Opening older (< 5.8) multi band lisn test is causing a crash
- #3859: AC coupling on input 1 decreases the start frequency of the ESU26.
- #3861: Communication Problem with configurable AD converter via TCP/IP
RadiMation version 5.8.6[edit]
- #3418: Driver: Advantest R3261A as powermeter
- #3604: RadiSwitch is generating sometimes is generating an error during the test.
- #3653: Driver: IFI 404
- #3656: Driver: OPHIR 5019E
- #3681: RadiMation tries to put FSP 3 in Receiver Mode
- #3803: CARAD - Timeout
- #3804: CARAD - debug window
- #3805: CARAD - clear buffer after timeout
- #3806: CARAD - error message when using wrong IP-address
- #3809: Not possible to configure stepsize in RE-MB RC test
- #3818: several questions regarding CAPL
- #3824: Configurable alarm detection is always reporting alarm, even when the correct response is returned.
- #3829: Driver for Amplifier Bonn BSA1050-250/100
- #3840: Copyright is 2009 instead of 2010
- #3841: Closed loop band before substitution band is checking the calibration information of the substitution band
- #3842: end-of-line character for AD Converter via TCP/IP
- #3846: Cannot select multiple cables for a specific part in a testsite
- #3849: Am modulation is not turned off when selecting no modulation in RI multiband test.
RadiMation version 5.8.5[edit]
- #2600: Undocumented error 0x300000C9 when Prana amplifier was already switched off
- #2775: Measurement of current by using the current sensor in combination with a multimeter or oscilloscope is wrong
- #3190: Driver: Anritsu MS2621B
- #3447: Conducted Immunity Tests are slow in 5.5.7 in comparison to 5.3.18
- #3636: Driver: Logimetrics AN8200G
- #3637: Driver: Logimetrics AN 8100Q
- #3651: Driver: Kalmus 122FC
- #3652: Driver: Kalmus LA200U
- #3732: Driver: Milmega Rf1000
- #3734: Driver: schwarzbeck antenna's FESP 5132 + FESP 5133-7/41
- #3735: Driver: SCHWARZBECK TEMZ 5231
- #3819: Network Analyser EUT Calibration, using a R&S ZVK, crashes with an overflow
- #3828: Unit is single band graph do not change after changing the unit in the unit configuration of the result file
- #3838: Why is it required to specify a magnetic flux density file for a MIL STD Calibration?
- #3839: Radiated Immunity MIL STD Calibration is always using a 70 dBuV Reference level
RadiMation version 5.8.4[edit]
- #1290: Radiated Immunity not completely usable for H-Field measurements
- #2604: Full support for Magnetic field tests in Radiated Immunity module
- #3095: RadiTurn controller does not allow a destination angle below 0 or above 360 degrees
- #3147: Memory leak in opening/closing RE MB TSF
- #3283: Modulation command is not retransmitted by configurable device driver if modulation parameter is changed
- #3686: AD Converter is read out before modulation is switched on
- #3720: During emission test: crosses instead of line
- #3779: Calculated field is not send to switch matrix in multiband test
- #3781: Driver: Farnell PSG 1000B
- #3789: Memory leak in reviewing Multiband Emission test
- #3802: Memory leakage in Multiband Emission test, when test is opened and closed
- #3823: TDS 600 Series scope driver reports unknown device driver when pressing advanced in the device driver settings window
RadiMation version 5.8.3[edit]
- #507: Overload indication of an analyser/receiver is not detected
- #2764: Optimize IDataIdentifier comparing
- #3183: Crash of RadiMation Version 5.3.18 when CARAD driver is included
- #3354: CARAD interface: two commands get combined to one
- #3372: ESU8 returned an error message during Radiated Emission MB measurement
- #3535: *OPC for ZVK with antenna measurement
- #3544: Device driver for Narda NBM550 uses the wrong baudrate
- #3615: Multiband-Conducted Immunity: data is not displayed
- #3780: Driver: Pre amplifier Schwarzbeck BBV 9718
- #3782: Driver: Antenna Schwarzbeck STLP 9128D and Elektromectrics EM 6954
- #3794: Driver for amplifier Bonn BLWA-0820-160/75/50D
- #3814: Driver: AR 15S1G4
- #3816: *OPC problems for ZVK (as signal generator) when setting the power level
RadiMation version 5.8.2[edit]
- #2152: Peak search for turntable angle should be configurable to include another angle range than the configured turntable
- #2153: Peak search for antenna tower height should be configurable to include another height range than the configured antennatower
- #2843: Incrementing the turntable angle, should wrap to the corresponding value if the maximum angle is reached
- #2849: When the turntable angle is changed, the direction with the shortest angle difference should be used to turn the turntable
- #3545: Automatic Angle and Height optimization should use the physical limitations of the turntable and antenna tower
- #3592: Turn Table Action in Manual Mode not possible with EMCO 1050
- #3593: Antenna Tower Action in Manual Mode not possible with EMCO 1050
- #3755: RadiMation is setting up the ESPI with incorrect RBW during CI-testing
RadiMation version 5.8.1[edit]
- #3201: Dificulties with OATS measurements
- #3563: Limit lines should be shown in polar plot and height graph
- #3701: Multiband Radiated Emission: Default value in dropdown list for Turntable Angle specifies 235°
- #3744: When 'Detect Peaks' is pressed, already detected peaks should be placed on the angle/height with the maximum emission
- #3745: Turntable doesn't turn a complete turn when it starts on a negative angle
- #3746: Peak zoom-in doesn't seem to work in multiband emission
- #3747: Improve peak detection by splitting the span in chunks
- #3748: Improve EMCO1060 turntable controller
- #3749: Click on the graph to add peak, should allow multiple peaks on the same frequency but with a different polarization
- #3751: Frequency marker is not moved on the graph when it is dragged
- #3752: Clicking on the turntable CW button to release it again doesn't respond
- #3753: HP4395A is always using 10 Hz RBW when it is used for a voltage measurement
- #3754: Climate chamber is not controlled correctly by RadiMation.
RadiMation version 5.8.0[edit]
- #1713: Continuous scanning is not possible in Multiband Emission Manual mode
- #2512: Zooming into emission graph while measurement still running
- #2597: In multiband immunity the cancel button only reacts on the end of the dwelltime
- #3203: Manual mast control does not work correctly, Up is not always Up
- #3386: the Radisense ctr1001B driver does not allow you to set the filter when using in gpib mode.
- #3490: Multi band emission manual mode is generating error when selecting the test site
- #3456: Manual mode multiband emission test does not have an initial graph which results in a crash
- #3547: First graph in a multiband emission measurements shows a bottom y-axis value of -1000 dBuV/m
- #3552: Spurious peaks in polar plot as a result of the initial measurement
- #3566: Change the frequency/angle/height of a detected peak by moving the marker in the graph
- #3611: Turning range of the antenna tower when changing polarization
- #3709: zero-span mode in radiated emission multiband
- #3714: Emission Multiband: Scan analysis while scan is running
- #3724: Radisense 4 over RS232 does not set the filter setting
- #3729: CE multi band measurement is not showing the information below 1 MHz
- #3736: I can not select apply modulation outside dwell time when selecting pulse