Line 884: Line 884:
{{ScreenElementDescription|Min. height|Sets the minimum height of the antenna tower.}}
{{ScreenElementDescription|Steps|Sets the number of steps to go from min. height to max. height.}}
{{ScreenElementDescription|Move during measurements|This will force the Antenna Tower to continuously move during the test and will disable step size.}}
{{ScreenElementDescription|Optimize height|The software will automatically detect on which antenna height maximum  emission will occur.}}
Revision as of 08:40, 9 December 2014
Radiated emission[edit]
The RadiMation® package support three main radiated emission methods:
  • Anechoic chamber method
  • Fixed height method
  • G-TEM method
Anechoic chamber[edit]
Frequency window[edit]
ScreenElementDescription.svg Start Sets start frequency
ScreenElementDescription.svg Stop Sets stop frequency
Turntable window[edit]
ScreenElementDescription.svg Start position Sets the start angle of the turntable.
ScreenElementDescription.svg Stop position Sets the stop angle of the turntable.
ScreenElementDescription.svg # Sets the number of steps for one turntable rotation.
ScreenElementDescription.svg Peak search The software will automatically detect on which angle maximum emission will occur.
  • If a turntable device is specified in the selected test-site, RadiMation will control the turntable to turn to the specified angle.
  • If no turntable device is specified in the selected test-site, RadiMation will assume that the EUT is already positioned on that specified angle. No actions on the turntable will be taken.
As long as a single angle is measured, and no turntable device is specified in the test-site, it is possible to use the Multiband Radiated Emission test without a turntable. So, if you specify the same 'Start Angle' and 'End Angle', and only specify a single 'Step' ('1'), then the test will assume that only that single specified angle should be measured.
ScreenElementDescription.svg Area Range over which the peak search will be performed (0 to 360 degrees max.).
ScreenElementDescription.svg Step size The angle step size.
ScreenElementDescription.svg Threshold This value determines the acceptance criteria for a peak.
ScreenElementDescription.svg Mode The mode setting determines the way a peak is detected.
  • Intelligent
  • Complete area
The Intelligent mode has the advantage that it stops with the scan as soon as a maximum has been determined. This could improve overall testtime as less movement of the clamp is possible. The Complete area mode has the advantage that the specified "Area" is measured, and that it ensures that the real maximum of the measurement value is measured.
Intelligent[edit]
Warning: This might cause a measurement error if the number of steps is set to a low value.
Complete area[edit]
Warning: This might cause a measurement error if the “Area” is set to a low value relative to the area given by the stop position minus start position dived by number of steps.
Antenna tower window[edit]
ScreenElementDescription.svg Max. height Sets maximum height of the antenna tower.
ScreenElementDescription.svg Min. height Sets the minimum height of the antenna tower.
ScreenElementDescription.svg # Sets the number of steps to go from min. height to max. height.
ScreenElementDescription.svg Peak search The software will automatically detect on which antenna height maximum emission will occur.
ScreenElementDescription.svg Area Range over which the peak search will be performed (0 to n meter).
ScreenElementDescription.svg Step size The distance for each step.
  • Intelligent
  • Complete area
Intelligent[edit]
Complete area[edit]
Warning: This might cause a measurement error if the “Area” is set to a low value relative to the area given by the max. height minus min. height dived by number of steps.
Antenna window[edit]
Polarisation[edit]
  1. Horizontal
  2. Vertical
  3. Both
Distance[edit]
Receiver settings[edit]
ScreenElementDescription.svg Reference level Sets the reference level of the receiver
ScreenElementDescription.svg Attenuator Sets the receiver attenuator level
ScreenElementDescription.svg RBW Sets the Resolution BandWidth filter
ScreenElementDescription.svg VBW Sets the Video BandWidth filter
ScreenElementDescription.svg Sweep Sets the sweep time of the receiver
ScreenElementDescription.svg Pre-amp Sets the pre-amp gain
ScreenElementDescription.svg Step frequency Sets the frequency step size of the receiver
ScreenElementDescription.svg Measure time Sets the measure time of the receiver
Warning: All these settings are generic. Therefore, always verify that the configured settings are also supported by the used hardware to ensure it is correctly used and no measurement faults are introduced. Use a fixed value to disabled interference of equipment intelligence.
Max. hold measurements[edit]
Traces window[edit]
Detectors window[edit]
Peak detection window[edit]
ScreenElementDescription.svg All peaks above “xx” below the limit lines In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 6 dB, the software will detect all peaks, which are higher than 6 dB below the limit line.
ScreenElementDescription.svg Max. Peaks With the max. peaks option, the user can limit the number of peak signals the software will detect and re-measure.
Test site window[edit]
Limit lines window[edit]
Test options window[edit]
ScreenElementDescription.svg Record data In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full tracebility, all test data should be stored.
ScreenElementDescription.svg Narrow band / broad band discrimination RadiMation® supports two types of narrow band / broad band determination.
Formula 7-1: Difference in a peak measurement, as a result of changing the used RBW
Formula 7-2: Difference in a peak amplitude
Limits Distortion type
Delta V <= 5 Delta RBW Narrow band
Delta V is > than 5 Delta RBW and delta V is <= 15 delta RBW Random noise
Delta V > 15 delta RBW Broad band
Loading a TSF file
  1. The presence of a valid software key is checked.
  2. The software settings are checked (i.e. is all test equipment used within its parameters?)
  3. The presence of all test equipment is checked.
  4. All relevant equipment will be initialised.
Test Procedure[edit]
  1. The receiver takes the desired sweeps of the spectrum.
  2. The antenna is moved to 4 meters.
  3. The receiver takes the desired sweeps of the spectrum.
  4. The turntable is moved to 45 degrees.
  5. The receiver takes the desired sweeps of the spectrum.
  6. The antenna tower is moved to 1 meter.
  7. The receiver takes the desired sweeps of the spectrum.
  8. The turntable is moved to 90 degrees.
  9. The receiver takes the desired sweeps of the spectrum.
  10. The antenna tower is moved to 4 meters
  11. The receiver takes the desired sweeps of the spectrum.
Info screen[edit]
Note: The pass/fail criteria, shown in the tables do not take the measure uncertainty of the test setup into account.
ScreenElementDescription.svg Settings In the settings window all the used settings are displayed.
ScreenElementDescription.svg Graph control In the graph control window the scale of the graph can be changed. You can show or hide the table by selecting or deselecting the show table checkbox.
ScreenElementDescription.svg Trace In the trace window you can control the traces that are shown in the graph by selecting or deselecting the checkbox. If during a measurement a trace is not used then the check box is grayed out.
ScreenElementDescription.svg Detector The detector window gives you the possibility to show and hide the values measured on the peaks. If during a measurement a type of detector is not used for determining the value on the peak the checkbox is grayed out.
ScreenElementDescription.svg Table In the table all the measured peaks are showed. The table also shows the limit values for that frequency point when a limit was given at the beginning of the test.
Test site information[edit]
Device name Tab in RadiMation® configuration window
Spectrum analyser Standard 2
Pre amplifier Standard 2
Antenna Standard 1
Cables
1 Pre amp->Spectrum Cables
2 Ant->Pre amp Cables
Note: Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during emission measurements.
Fixed height test method[edit]
  • Intelligent
  • Complete area
Intelligent[edit]
Complete area[edit]
ScreenElementDescription.svg Min. Height Sets the minimum height of the antenna tower.
  • Intelligent
  • Complete area
Intelligent[edit]
Complete area[edit]
Polarisation[edit]
  1. Horizontal
  2. Vertical
  3. Both
Distance[edit]
Traces window[edit]
ScreenElementDescription.svg Max. Peaks With the ‘max. peaks” option, the user can limit the number of peak signals the software will detect and re-measure.
Loading a TSF file
  1. The receiver takes the desired sweeps of the Spectrum.
  2. The turntable is moved to 45 degrees.
  3. The receiver takes the desired sweeps of the Spectrum.
  4. The turntable is moved to 90 degrees.
  5. The receiver takes the desired sweeps of the Spectrum.
Antenna Standard 1
Cables
measurements.
G-TEM Method[edit]
Correlation method[edit]
Measurement procedure[edit]
   Menu.svg Tests
      Menu.svg Radiated emission
         Menu.svg G-TEM
When a new test has to be configured please press the “Cancel” button. A blank test configuration screen will appear. The chapter below will explain all settings for this test configuration screen.
Test configuration[edit]
OATS settings[edit]
G-TEM settings[edit]
ScreenElementDescription.svg Reference level Sets the reference level of the receiver.
ScreenElementDescription.svg Attenuator Sets the receiver attenuator level.
ScreenElementDescription.svg RBW Sets the Resolution BandWidth filter.
ScreenElementDescription.svg VBW Sets the Video BandWidth filter.
ScreenElementDescription.svg Sweep Sets the sweep time of the receiver.
ScreenElementDescription.svg Pre-amp Sets the pre-amp gain.
ScreenElementDescription.svg Step frequency Sets the frequency step size of the receiver.
ScreenElementDescription.svg Measure time Sets the measure time of the receiver.
Traces window[edit]
Max. Peaks
Starting a G-TEM test[edit]
Loading a TSF file[edit]
Antenna Standard 1
Cables
measurements.
Manual mode[edit]
ScreenElementDescription.svg Reference Sets the reference level of the receiver
ScreenElementDescription.svg RBW Sets the Resoution Band Width filter
Data correction[edit]
Peak menu[edit]
ScreenElementDescription.svg Search peak RadiMation® can detect peaks automatically. From the “peaks” TAB, choose “Search peak” to detect the maximum signal. The software will measure the detected peak.
ScreenElementDescription.svg Add peak The measured values can be added to the peak list by pressing the “Add Peak” button.
ScreenElementDescription.svg Search angle By pressing the “Search Angle” button, the software starts a turntable scan to determine at which turn table angle maximum emission takes place.
ScreenElementDescription.svg Search Height By pressing the “Search Height” button, the software starts an antenna height scan to determine at which antenna height maximum emission takes place.
Traces Menu[edit]
Traces[edit]
Detectors[edit]
  • Peak detector
  • Average detector
  • Quasi peak detector
  • RMS detector
Zoom function[edit]
ScreenElementDescription.svg X-axis When this option is switched on, the frequency settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph.
ScreenElementDescription.svg Y-axis When this option is switched on, the amplitude settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph.
Limit lines[edit]
Export menu[edit]
Merge function[edit]
X+k*S Calculations[edit]
Anechoic chamber (Multi band)[edit]
ScreenElementDescription.svg Start Angle Sets the start position of the turntable.
ScreenElementDescription.svg End Angle Sets the stop position of the turntable.
ScreenElementDescription.svg Steps Sets the number of steps for one turntable rotation.
ScreenElementDescription.svg Turn during measurements This will force the turntable to continuously rotate during the test and will disable step size.
ScreenElementDescription.svg EUT angle offset The offset of the EUT compared to the zero angle of the table.
ScreenElementDescription.svg Optimize Angle Optimized settings for searching the correct angle.
ScreenElementDescription.svg Steps Sets the number of steps to go from min. height to max. height.
ScreenElementDescription.svg Move during measurements This will force the Antenna Tower to continuously move during the test and will disable step size.
ScreenElementDescription.svg Optimize height The software will automatically detect on which antenna height maximum emission will occur.
ScreenElementDescription.svg Area Range over which the peak search will be performed (0 to Max height).
ScreenElementDescription.svg Step size The height step size.
Polarisation[edit]
  1. Horizontal
  2. Vertical
  3. Both
Distance[edit]
ScreenElementDescription.svg RBW Sets the Resolution Band Width filter.
ScreenElementDescription.svg VBW Sets the Video Band Width filter.
Traces window[edit]
ScreenElementDescription.svg Number of sweeps The number op peak hold sweeps for each measurement can be selected. If this value is set to, for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweep speed. For spectrum analysers and CW signals you can use a relatively low amount of peak sweeps, for example 10. But when you have discontinues signals like a spark bridge, you may need 100 peak sweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope.
Final peak measurement[edit]
ScreenElementDescription.svg No peaks When no final peak measurements are required select the option no peaks.
Change Order window[edit]
So for Site 1 you want:
  • Band A RBW 9 kHz
  • Band B RBW 9 kHz
  • Band A RBW 120 kHz
  • Band B RBW 120 kHz
And for test site 2:
  • Band A RBW 9 kHz
  • Band B RBW 9 kHz
  • Band A RBW 9 kHz
  • Band B RBW 9 kHz
  • RBW
  • Testsite
  • Frequency band
  • Band A RBW 9 kHz
  • Band B RBW 9 kHz
Note: If you have any questions regarding the setting of the change order please contact your reseller.
Info screen[edit]
Note: Continuing the test is only possible with the selected test site. RadiMation® will not select the correct test site for you. All measurements are done with the selected test site. If you want automatic changes of the test site, restart the configured TSF file.
Antenna Standard 1
Cables
Manual Mode (Multi band)[edit]
Turn table[edit]
Antenna tower[edit]
Test site[edit]
Final measurement[edit]
Peaks window[edit]
ScreenElementDescription.svg Take sweeps Triggers the receiver of analyzer to take the configured amount of sweeps.
ScreenElementDescription.svg Detect peaks Detects the configured amount of sweeps in the measured spectrum.
ScreenElementDescription.svg Measure Peaks Measures all selected peaks with the configured settings in the final measurement window.
ScreenElementDescription.svg Delete Peaks Delete the selected peaks from the list.
ScreenElementDescription.svg Peak actions Allows several actions to be performed on the detected peaks, like select all and deselect all.
