Line 659: Line 659:
===Turntable window===
{{ScreenElementDescription|Start position|Sets the start angle of the turntable.}}
Revision as of 11:35, 7 December 2017
Radiated emission[edit]
Loading a TSF file[edit]
   Menu.svg Tests
      Menu.svg Radiated Emission and then the desired test method.
  1. The presence of a valid software key is checked.
  2. The software settings are checked (i.e. is all test equipment used within its parameters?)
  3. The presence of all test equipment is checked.
  4. All relevant equipment will be initialised.
Test results screen[edit]
You can access the test results of the performed test by selecting the test in the EUT window and then press the Info button. For the Multiband test, the window that is appearing looks a lot like the manual mode screen, the reason that it looks the same is because it is. You can continue testing after you finished a test, just select the testsite you want to use and you can continue testing.
Note: The pass/fail criteria, shown in the tables do not take the measure uncertainty of the test setup into account.
Note: Continuing the test is only possible with the selected test site. RadiMation® will not select the correct test site for you. All measurements are done with the selected test site. If you want automatic changes of the test site, restart the configured TSF file.
Testsite information[edit]
An radiated emission test, requires that a testsite is used that can have the following components.
Device name Tab in RadiMation® configuration window
Spectrum analyser Devices 2
Pre amplifier Devices 2
Antenna Devices 1
Turn table Devices 1
Antennatower Devices 1
Cables
1 Cable preamp -> analyser Cables
2 Cable antenna -> preamp Cables
Note: Usage of cable correction factors is not mandatory. However, using cable correction factors will increase measurement accuracy during emission measurements. When in doubt contact your reseller.
Multiband[edit]
The radiated emission Multiband test, is a very flexible and powerful test to perform radiated emission tests. When the multiband test is performed, it will first perform a prescan measurement on all the specified locations. After this prescan is finished the temporary results are shown, and some peaks are automatically detected and shown. At this moment those peaks are NOT measured yet with the final measurement detectors. The end-user now has the possibility to add additional peaks, or do some other peak investigation/changes. When the end-user is satisfied with the list of detected peaks, he can continue with the final measurement, by pressing the Automatic button again. Then the peaks will be measured with the detector that are selected in the Final peak measurement box. Only after those measurements are finalized, those detector values will become available in the list of detected peaks.
The test engineer can configure a complete radiated emission test in the Radiated Emission configuration menu. The test engineer can change the frequency range from one frequency band or the RBW of several bands. It all depends on the selection of bands. When several bands are selected and some options are blank, this means that these settings in bands are different. For example if the test engineer selects both frequency band in the window above, the frequency range will be blank. If the engineer sets the start and stop frequency, all the selected bands will have the same frequency setting. One of the advantages of this feature is the quick settings of several settings like the RBW or the peak detection.
Frequency window[edit]
Turntable window[edit]
ScreenElementDescription.svg Start Angle Sets the start position of the turntable.
ScreenElementDescription.svg End Angle Sets the stop position of the turntable.
ScreenElementDescription.svg Steps Sets the number of steps for one turntable rotation.
ScreenElementDescription.svg Turn during measurements This will force the turntable to continuously rotate during the test and will disable steps setting.
ScreenElementDescription.svg EUT angle offset The offset of the EUT compared to the zero angle of the table.
ScreenElementDescription.svg Optimize Angle Optimized settings for searching the correct angle.
ScreenElementDescription.svg Area Range over which the peak search will be performed (0 to 360 degrees max.).
ScreenElementDescription.svg Step size The angle step size.
Antenna tower window[edit]
ScreenElementDescription.svg Max. height Sets maximum height of the antenna tower.
ScreenElementDescription.svg Min. height Sets the minimum height of the antenna tower.
ScreenElementDescription.svg Steps Sets the number of steps to go from min. height to max. height.
ScreenElementDescription.svg Move during measurements This will force the Antenna Tower to continuously move during the test and will disable steps setting.
ScreenElementDescription.svg Optimize height The software will automatically detect on which antenna height maximum emission will occur.
ScreenElementDescription.svg Area Range over which the peak search will be performed (0 to Max height).
ScreenElementDescription.svg Step size The height step size.
Antenna window[edit]
Polarisation[edit]
  1. Horizontal
  2. Vertical
  3. Both
Distance[edit]
Receiver settings[edit]
ScreenElementDescription.svg Reference level Sets the reference level of the receiver.
ScreenElementDescription.svg Attenuator Sets the receiver attenuator level.
ScreenElementDescription.svg RBW Sets the Resolution Band Width filter.
ScreenElementDescription.svg VBW Sets the Video Band Width filter.
ScreenElementDescription.svg Sweep Sets the sweep time of the receiver.
ScreenElementDescription.svg Step frequency Sets the frequency step size of the receiver.
ScreenElementDescription.svg Measure time Sets the measure time of the receiver.
ScreenElementDescription.svg Pre-amp Sets the pre-amp gain.
Traces window[edit]
ScreenElementDescription.svg Detectors The detector settings (Peak, Average, Quasi-peak, RMS) used during sweeptime. This setting is generic. So it is possible to select multiple detectors, but only one detector can be selected due to the device limitations. In that case for each enabled detector the amount of sweeps will be performed by RadiMation®.
ScreenElementDescription.svg Number of sweeps The number op peak hold sweeps for each measurement can be selected. If this value is set to, for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweep speed. For spectrum analysers and CW signals you can use a relatively low amount of peak sweeps, for example 10. But when you have discontinues signals like a spark bridge, you may need 100 peak sweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope.
Final peak measurement[edit]
Peak detection window[edit]
ScreenElementDescription.svg No peaks When no final peak measurements are required select the option no peaks.
ScreenElementDescription.svg Maximum Peaks With the max. peaks option, the user can limit the number of peak signals the software will detect and re-measure.
Test site window[edit]
Limit lines window[edit]
ScreenElementDescription.svg All peaks above “xx” below the limit lines In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 5 dB, the software will detect all peaks, which are higher than 5 dB below the limit line.
Change Order window[edit]
So for Site 1 you want:
  • Band A RBW 9 kHz
  • Band B RBW 9 kHz
  • Band A RBW 120 kHz
  • Band B RBW 120 kHz
And for test site 2:
  • Band A RBW 9 kHz
  • Band B RBW 9 kHz
  • Band A RBW 9 kHz
  • Band B RBW 9 kHz
  • RBW
  • Testsite
  • Frequency band
  • Band A RBW 9 kHz
  • Band B RBW 9 kHz
Note: If you have any questions regarding the setting of the change order please contact your reseller.
Manual Mode[edit]
Turn table[edit]
Antenna tower[edit]
Test site[edit]
Final measurement[edit]
Peaks window[edit]
ScreenElementDescription.svg Max Peaks Maximum allowed amount of peaks to be automatically determined.
ScreenElementDescription.svg Detect Peaks Detects the configured amount of sweeps in the measured spectrum.
ScreenElementDescription.svg Measure Peaks Measures all selected peaks with the configured settings in the final measurement window.
ScreenElementDescription.svg Peak Actions Allows several actions to be performed on the detected peaks, like select all and deselect all.
ScreenElementDescription.svg Peaks to maximum Start performing measurement to redetermine on which turntable angle and antenna tower height the highest emission value was measured.
This angle and height will be updated in the selected peak, together with the corresponding measured emission values. If multiple peaks are selected, this search for the highest emission value is done individually for every selected peak.
This 'peaks to maximum' function can be very helpful during the angle or height optimization of a peak. If after manual changes of the antenna height or turntable angle, no other maximum is found, the peak can be reset to the angle and height where the maximum was determined. 'Peaks to maximum', will thus select that location on which the highest emission value was measured.
ScreenElementDescription.svg Measure Peaks Start performing a final measurement on the selected peaks.
ScreenElementDescription.svg Optimize Angle Start performing a measurement as configured for the Optimize Angle.
ScreenElementDescription.svg Optimize Height Start performing a measurement as configured for the Optimize Height.
ScreenElementDescription.svg Go To Peak Location If possible it will adjust the Antenna Tower and Turntable to the location of the selected peak.
ScreenElementDescription.svg Set receiver to Peak settings Set the current receiver settings to the used settings of the selected peak.
ScreenElementDescription.svg Configure -> Optimize Angle Opens a new window to configure the Optimize Angle settings. (See also Optimize Angle settings.)
ScreenElementDescription.svg Configure -> Optimize Height Opens a new window to configure the Optimize Height settings. (See also Optimize height settings.)
ScreenElementDescription.svg Delete peaks Delete selected peak(s).
ScreenElementDescription.svg Delete Peaks Delete the selected peaks from the list.
ScreenElementDescription.svg Sweeps Amount of sweeps to be measured.
ScreenElementDescription.svg Take Sweeps Triggers the receiver of analyzer to take the configured amount of sweeps.
ScreenElementDescription.svg Continuous Measure Start performing of taking measurements. This is a heavy process and it is normal when more CPU is used by the computer.
ScreenElementDescription.svg Continuous Measure Write Each new measurement will overwrite the previous measurement data.
ScreenElementDescription.svg Continuous Measure Max-hold Only higher new measurement values will overwrite the previous measurement data.
ScreenElementDescription.svg Clear measurement data Removes all previous measurement data.
Anechoic chamber[edit]
ScreenElementDescription.svg Start Sets start frequency
ScreenElementDescription.svg Stop Sets stop frequency
ScreenElementDescription.svg Start position Sets the start angle of the turntable.
ScreenElementDescription.svg Stop position Sets the stop angle of the turntable.
ScreenElementDescription.svg # Number of locations from Start postion to Stop position that should be measured.
ScreenElementDescription.svg Peak search The software will automatically detect on which turntable angle the maximum measurement value of each peak is present.
  • If a turntable device is specified in the selected test-site, RadiMation® will control the turntable to turn to the specified angle.
  • If no turntable device is specified in the selected test-site, RadiMation® will assume that the EUT is already positioned on that specified angle. No actions on the turntable will be taken.
ScreenElementDescription.svg Area Range over which the peak search will be performed. The initially detected turntable angle of the peak will be the middle of this Area, and the total search area for this peak will be half of this area clockwise and half of this area counter-clockwise.
ScreenElementDescription.svg Step size The step size for each turntable movement.
ScreenElementDescription.svg Threshold This value is only used during the Intelligent mode. It influences how long the movement is continued. The movement will stop when the measurement value is lower than the specified Threshold below the highest known measurement value of the searched peak.
ScreenElementDescription.svg Mode The mode setting determines the way a peak is detected. The following modes are available:
  • Intelligent
  • Complete area
The Intelligent mode has the advantage that it stops with the scan as soon as a maximum measurement value has been determined. This could improve overall testtime as less movement of the turntable angle is possible. The Complete area mode has the advantage that the complete specified Area is measured.
Intelligent[edit]
In this mode the software increases the angle of the turntable from the initial point with a step equal to Step size as long as the measurement value is not more than Threshold below the already known maximum measurement value. As soon as measurement value is detected that is Threshold below the already known meaximum value, the turntable movement is reversed until the initial point and from this initial point the angle of the turntable is decreased with a step equal to Step size until a measurement value is measured that is Threshold below the already known maximum measurement value. The step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak.
Warning: This might cause a measurement error if the intial number of steps (#) is set to a low value.
Complete area[edit]
Warning: This might cause a measurement error if the Area is set to a smaller value than the original stepsize (given by the stop position minus start position divided by the number of steps).
ScreenElementDescription.svg # Number op locations from Max. height to Min. height that should be measured.
ScreenElementDescription.svg Peak search The software will automatically detect on which antenna height the maximum measurement value of each peak is present.
ScreenElementDescription.svg Area Range over which the peak search will be performed. The initially detected antenna height of the peak will be the middle of this Area, and the total search area for this peak will be half of this area up and half of this area down.
ScreenElementDescription.svg Step size The step size for each antenna height movement.
  • Intelligent
  • Complete area
The Intelligent mode has the advantage that it stops with the scan as soon as a maximum measurement value has been determined. This could improve overall testtime as less movement of the antenna tower is possible. The Complete area mode has the advantage that the complete specified Area is measured.
Intelligent[edit]
In this mode the software increases the height of the antenna from the initial point with a step equal to Step size as long as the measurement value is not more than Threshold below the already known maximum measurement value. As soon as measurement value is detected that is Threshold below the already known meaximum value, the antenna movement is reversed until the initial point and from this initial point the distance of the clamp is decreased with a step equal to Step size until a measurement value is measured that is Threshold below the already known maximum measurement value. The step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak.
Complete area[edit]
Polarisation[edit]
  1. Horizontal
  2. Vertical
  3. Both
Distance[edit]
ScreenElementDescription.svg RBW Sets the Resolution BandWidth filter
ScreenElementDescription.svg VBW Sets the Video BandWidth filter
Warning: All these settings are generic. Therefore, always verify that the configured settings are also supported by the used hardware to ensure it is correctly used and no measurement faults are introduced. Use a fixed value to disabled interference of equipment intelligence.
Max. hold measurements[edit]
Traces window[edit]
Detectors window[edit]
ScreenElementDescription.svg All peaks above “xx” below the limit lines In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 6 dB, the software will detect all peaks, which are higher than 6 dB below the limit line.
ScreenElementDescription.svg Max. Peaks With the max. peaks option, the user can limit the number of peak signals the software will detect and re-measure.
Test options window[edit]
ScreenElementDescription.svg Record data In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full tracebility, all test data should be stored.
ScreenElementDescription.svg Narrow band / broad band discrimination RadiMation® supports two types of narrow band / broad band determination.
Formula 7-1: Difference in a peak measurement, as a result of changing the used RBW
Formula 7-2: Difference in a peak amplitude
Limits Distortion type
Delta V <= 5 Delta RBW Narrow band
Delta V is > than 5 Delta RBW and delta V is <= 15 delta RBW Random noise
Delta V > 15 delta RBW Broad band
Loading a TSF file
Test Procedure[edit]
  1. The receiver takes the desired sweeps of the spectrum.
  2. The antenna is moved to 4 meters.
  3. The receiver takes the desired sweeps of the spectrum.
  4. The turntable is moved to 45 degrees.
  5. The receiver takes the desired sweeps of the spectrum.
  6. The antenna tower is moved to 1 meter.
  7. The receiver takes the desired sweeps of the spectrum.
  8. The turntable is moved to 90 degrees.
  9. The receiver takes the desired sweeps of the spectrum.
  10. The antenna tower is moved to 4 meters
  11. The receiver takes the desired sweeps of the spectrum.
Info screen[edit]
ScreenElementDescription.svg Settings In the settings window all the used settings are displayed.
ScreenElementDescription.svg Graph control In the graph control window the scale of the graph can be changed. You can show or hide the table by selecting or deselecting the show table checkbox.
ScreenElementDescription.svg Trace In the trace window you can control the traces that are shown in the graph by selecting or deselecting the checkbox. If during a measurement a trace is not used then the check box is grayed out.
ScreenElementDescription.svg Detector The detector window gives you the possibility to show and hide the values measured on the peaks. If during a measurement a type of detector is not used for determining the value on the peak the checkbox is grayed out.
ScreenElementDescription.svg Table In the table all the measured peaks are showed. The table also shows the limit values for that frequency point when a limit was given at the beginning of the test.
Test site information[edit]
Spectrum analyser Standard 2
Pre amplifier Standard 2
Antenna Standard 1
Cables
1 Pre amp->Spectrum Cables
2 Ant->Pre amp Cables
Note: Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during emission measurements.
Fixed height[edit]
  • Intelligent
  • Complete area
Intelligent[edit]
Complete area[edit]
  • Intelligent
  • Complete area
Intelligent[edit]
Complete area[edit]
Polarisation[edit]
  1. Horizontal
  2. Vertical
  3. Both
Distance[edit]
Traces window[edit]
ScreenElementDescription.svg Max. Peaks With the ‘max. peaks” option, the user can limit the number of peak signals the software will detect and re-measure.
Loading a TSF file
  1. The receiver takes the desired sweeps of the Spectrum.
  2. The turntable is moved to 45 degrees.
  3. The receiver takes the desired sweeps of the Spectrum.
  4. The turntable is moved to 90 degrees.
  5. The receiver takes the desired sweeps of the Spectrum.
Antenna Standard 1
Cables
measurements.
GTEM[edit]
Correlation method[edit]
Measurement procedure[edit]
   Menu.svg Tests
      Menu.svg Radiated emission
         Menu.svg G-TEM
When a new test has to be configured please press the “Cancel” button. A blank test configuration screen will appear. The chapter below will explain all settings for this test configuration screen.
Test configuration[edit]
OATS settings[edit]
G-TEM settings[edit]
ScreenElementDescription.svg RBW Sets the Resolution BandWidth filter.
ScreenElementDescription.svg VBW Sets the Video BandWidth filter.
Traces window[edit]
Max. Peaks
Starting a G-TEM test[edit]
Antenna Standard 1
Cables
measurements.
ScreenElementDescription.svg Reference Sets the reference level of the receiver
ScreenElementDescription.svg RBW Sets the Resoution Band Width filter
Data correction[edit]
Peak menu[edit]
ScreenElementDescription.svg Search peak RadiMation® can detect peaks automatically. From the “peaks” TAB, choose “Search peak” to detect the maximum signal. The software will measure the detected peak.
ScreenElementDescription.svg Add peak The measured values can be added to the peak list by pressing the “Add Peak” button.
ScreenElementDescription.svg Search angle By pressing the “Search Angle” button, the software starts a turntable scan to determine at which turn table angle maximum emission takes place.
ScreenElementDescription.svg Search Height By pressing the “Search Height” button, the software starts an antenna height scan to determine at which antenna height maximum emission takes place.
Traces Menu[edit]
Traces[edit]
Detectors[edit]
  • Peak detector
  • Average detector
  • Quasi peak detector
  • RMS detector
Zoom function[edit]
ScreenElementDescription.svg X-axis When this option is switched on, the frequency settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph.
ScreenElementDescription.svg Y-axis When this option is switched on, the amplitude settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph.
Limit lines[edit]
Export menu[edit]
Merge[edit]
X + k * S[edit]