RadiMation Application Note 106: Difference between revisions
| Line 28: | Line 28: | ||
[[File:MeasureLoop.png]] | [[File:MeasureLoop.png]] | ||
See the measured power graph above. Each dot, is 1 | See the measured power graph above. Each dot, is 1 measurement. In this case, you could set the measure time to 5, and maximum measure time to 10. The diference can be set to the highest(green) point and one previous point(orange). | ||
In this case, RadiMation will use the first point, where the difference less than the allowed difference. The value of the green point will be used. | In this case, RadiMation will use the first point, where the difference less than the allowed difference. The value of the green point will be used. | ||
Revision as of 15:14, 29 January 2019
Powermeter parameter configuration[edit]
This application note describes how the measurement parameters of a Powermeter can be configured to get a fast and stable power measurement.
Each powermeter in RadiMation® has a powermeter configuration tab in the device drivers. In this tab, the accuracy of power measurement can be optimized. This function allows to make multiple measurements, and wait until the value is stable. Using a measurement-loop.
| Wait time before starting the measurement-loop. Can be used to wait for a signal generator, and or amplifier to stabalize. |
| The minimum ammount of measurements in the measure loop. Continues to loop even if power level is stable. |
| The wait time before each measurement is taken, during the measurement-loop. |
| If the measured power between each measurement in the measurement-loop, is less than this value. The measured power is stable. The loop will stop. |
| The maximum ammount of measure times, for the measurement-loop |
As can be seen on the screenshot. Both leveling, and monitoring settings can be set.
The leveling is used, whenever RadiMation is measuring the power, to regulate the signal generator.
For example, if you have a immunity power closed loop method. This will regulate the signal generator until certain power is measured.
In this case, the leveling settings are used.
If the powermeter is used as input. For example in the immunity multiband test, in the inputs table. Then the monitoring settings are used.
Example[edit]
See the measured power graph above. Each dot, is 1 measurement. In this case, you could set the measure time to 5, and maximum measure time to 10. The diference can be set to the highest(green) point and one previous point(orange). In this case, RadiMation will use the first point, where the difference less than the allowed difference. The value of the green point will be used.
| Warning: | With the available settings, you can make measurements very slow. In some cases, you can end up in measuring noise during test. Then the maximum measure times will be used. You may need to decrease the Max. Measure value in that case. |