Report Generator[edit]
The Report Generator is a software module to automatically generate a test report from tests that are carried out by RadiMation®. Once RadiMation® carries out a test, all collected test data is available to the RadiMation® report generator and can be merged into a test report. Using the Report Generator module, a test engineer can save a huge amount of time. The Report Generator needs only a few minutes to generate a complete test report.
Test reports can be written in Microsoft Word 8.0 and in ASCII and HTML text formats. Other formats will be added on request.
In MS Word 8.0 text, tables and graphs can be added to the report.
Getting started[edit]
Before the Report Generator module can be used, it needs to be configured.
Select:
Select the tab on the screen with the label “Report Generator”
The Report Generator screen will be displayed. This screen contains only two items: The format of the template file, and the option to let the Report Generator check and alternate the extension of the report file. Choose your favourite format and select “close”. The Report Generator is now configured properly. Open an EUT and click the “Report” button on the EUT screen. This button shows the screen in which the settings can be selected.
To select a template file, select the folder button for the template. The red cross is for deleting the file selection, not the file itself. The magnifying glass is for viewing the template. If you want to print the report directly after it has been generated then check the check box for print report. If you want to edit the report then check the Edit report check box. By default all tests are included for generating a report. If you want to alter the selection press the Advanced Option button. An additional screen will be displayed like as shown above. All tests with the red checks in front will be included. Double clicking a test will alter the include/exclude selection. With the “Include All” button all tests will be included. With “Include None” all tests will be excluded from the report. “Invert Selection” will invert all the including/excluding. Selecting the “Generate” button will start the report generation process. The report will be finished in a few moments. The generation time varies between a couple of seconds for a text file to several minutes for an extensive report containing many tables and graphs in MS Word.
The report Generator uses two files to generate a report. These two files are the Template file and the Report file. The Template file is the complete Report containing keywords. The test Data is represented by unique keywords, which tell the report generator which data to insert at the position of the code. The Report Generator changes all codes into test data and thus creates the report. This Report file is the complete test report including all test data. The Template file is not altered. It is recommended to make this Template file read-only, so no one can accidentally delete or modify this file. The Template file must exist, the Report file will be created, if it doesn’t exist. The Template file and report file may not be the same file!
Keywords and Main Keywords[edit]
A template file contains one or more unique keywords. These keywords consist of a Main Keyword and a second Keyword. The codes are of the following type: ||Main keyword | second keyword|| The Main Keyword indicates the part of RadiMation® from which the data must be retrieved. The Keyword specifies the data to be merged into the report file. The Main Keyword can be:
- EUT
- “Configuration” - The name of a test specified in the list in the EUT Window
Every Main Keyword has a set of Keywords that can only be used in combination with the used Main Keyword.
The EUT Main Keyword can use the keyword “Serial Number”, but the “Serial Number” keyword is not supported by the “Configuration” Main Keyword. A list of Keywords is given in the succeeding paragraphs of in this manual. The keywords must be spelled correctly, including spaces and other symbols. To separate multiple words, only one space is used. Keywords and Main keywords are not case sensitive. When a Keyword or Main Keyword is not recognised, or when there is no test data available for this Keyword or Main keyword, the code will not be altered. The test engineer will be given a message in the settings screen of the Report Generator. One failure will be added to the total amount of failures. If a template file is altered, or the test-name has changed, the test engineer should search a generated report for codes from the template file for he may have used codes that contain no data or are not recognised by the Report Generator.
Available Report Generator Codes for EUT’s[edit]
The Main keyword for EUT data is “EUT”. All data entered in the EUT Window in RadiMation® can be merged into a report file. All data can be merged as single lines of text or in tables. Tables are supported for Client data, Manufacturer data and Test site data. In addition, a table with all information can be merged.
Available Report Generator Keywords[edit]
As different tests use different equipment and perform different actions each test has its own set of keywords. In many cases tests share the same keywords although this will refer to different data as the combination of main keyword and secondary keyword will always result in a unique combination. Errors are easily located by searching the report for the characters “||”. Once found these can be deleted, or changed into an existing code. If the Report Generator is executed once again, using the modified file as the template, the Report Generator will finish the report without errors.
D.A.R.E!! Consultancy Template and explanation[edit]
On you CD-ROM you will find a Template-file including several Report Generator Codes. This Template-file is used by D.A.R.E!! Consultancy, the “Sister company” of D.A.R.E!! Development. D.A.R.E!! Consultancy performs EMC tests on behalf of the CE regulations. Use this Template as an example and as a guide to writing your own Template files.
EUT[edit]
Common[edit]
| Keyword | Description |
| CONDITION RECEPTION | The condition of reception of the EUT specified in the EUT Window |
| DATE RECEIPT | The date of reception of the EUT specified in the EUT Window |
| DATE MEASURMENT | The date of measurment of the EUT specified in the EUT Window |
| RADIMATION VERSION | The Radimation version |
| SETUP TYPE | The test setup type of the EUT specified in the EUT Window |
Cable information[edit]
Standards Information[edit]
Keyword Description PRODUCT STANDARD … Specifications of the used Product Standard BASIC STANDARD … Specifications of the used Basic Standard
Where … is one of the following parameters:
| Parameter | Description |
| REFERRED STANDARDS | A string of referred standards, separated by a “,” |
Radiated Immunity[edit]
16 Point Calibration[edit]
| AD GRAPH nr | The graph representing the A/D graph of channel number [x], where x is a number between 1 and 8 |
| FIELD GRAPH nr | The graph representing the field graph of field strength sensor number [x], where x is a number between 1 and 8 |
| POINT 16 ERROR GRAPH nr | The graph representing the error graph of point number [x], where x is a number between 1 and 16 |
| POINT 16 FIELD GRAPH nr | The graph representing the field graph of point number [x], where x is a number between 1 and 16 |
| POINT 16 POWER GRAPH nr | The graph representing the power graph of point number [x], where x is a number between 1 and 16 |
| CURRENT UNIT | The selected current unit: [dBuA], [dBmA], [dBA], [uA],[mA] or [A] |
Equipment for 16 point calibration[edit]
The equipment keyword consists of two parts: the keyword followed by a parameter. For each keyword all parameters are valid.
| AMPLIFIER … | Specifications of the used Amplifier |
| ANTENNA … | Specifications of the used Antenna |
| COUPLER … | Specifications of the used Coupler |
| FIELD SENSOR nr … | Specifications of the used Field Sensor number [x], where x is a number between 1 and 8 |
| FORWARD POWER METER … | Specifications of the used Forward Power Meter |
| REFLECTED POWER METER … | Specifications of the used Reflected Power Meter |
| SIGNAL GENERATOR … | Specifications of the used Signal Generator |
| DEVICE DRIVER DLL | N/A |
| INI FILE NAME | N/A |
Radiated Immunity Tests[edit]
Keyword Description AD CONVERTOR … Specifications of the used A/D Converter AMPLIFIER … Specifications of the used Amplifier ANTENNA … Specifications of the used Antenna CABLE nr … Specifications of the used Cable number [x], where x is a number between 1 to 8 COUPLER … Specifications of the used Coupler CURRENT SENSOR … Specifications of the used Current Sensor FIELD SENSOR nr … Specifications of the used Field Sensor number [x], where x is a number between 1 and 8 FORWARD POWER METER … Specifications of the used Forward Power Meter REFLECTED POWER METER … Specifications of the used Reflected Power Meter SENSOR POWER METER … Specifications of the used Sensor Power Meter SIGNAL GENERATOR … Specifications of the used Signal Generator
Parameter Description BRAND Brand of Measurement equipment CALIBRATION EXPIRES Date when calibration expires of equipment, given by the user DESCRIPTION Description of Measurement equipment given by the user DEVICE DRIVER DLL N/A DEVICE DRIVER NAME The internal name of the device driver HARDWARE VERSION Hardware version of Measurement equipment ID Internal ID of Measurement equipment INI FILE NAME N/A NAME Name of Measurement equipment SERIAL NUMBER Serial number of Measurement equipment SOFTWARE VERSION Software version of Measurement equipment SUPPORTED DEVICE TYPES N/A TYPE Type of Measurement equipment
Conducted Immunity[edit]
Keyword Description NUMBER OF EVENTS The number of events that occur during a test TEST START TIME The date and time at which the test is started TEST STOP TIME The date and time at which the test is stopped TEST DATE The date at which the test is performed TEST ENGINEER The name of the test engineer TEST TIME The time at which the test is performed TEST SITE NAME The name of the test equipment list AMPLIFIER PROTECTION MODE The Amplifier Protection mode: Signal Power, Forward Power and / or Reflected Power. CALIBRATION METHOD The calibration method used: Signal Generator Level, Forward Power or Net-Power CURRENT MEASURED Indication whether or not the current is measured: Yes / No CURRENT TEST LEVEL The current test level in Ampere [A] CURRENT UNIT The selected current unit: [dBuA], [dBmA], [dBA], [uA], [mA] or [A] DESCRIPTION The description of the Test Setup File (TSF) DISTANCE The distance between the antenna and the EUT in meters [m] DWELL TIME The dwell time in seconds [s] FORWARD POWER MEASURED Indication whether or not the forward power is measured: Yes / No FREQUENCY CHANGE MODE The method used to change frequency: Stress Method, Ramp Method, Constant Method or Fast Constant Method. HUMIDITY Relative Humidity during the test in percentage rel. Humidity [%RH] LOGARITHMIC STEP The logarithmic frequency step in a percentage of the previous step LOGARITHMIC STEP FROM The frequency the test was started from to calculate the proper logarithmic step size MAIN VOLTAGE The mains voltage during the test in Volt [V] NET POWER MEASURED Indication whether or not the net-power power is measured: Yes / No NOTE Note or remark made during the test POLARIZATION Antenna polarisation: Horizontal or Vertical PRESSURE Air Pressure during the test in milliBar [mbar] REFLECTED POWER MEASURED Indication whether or not the reflected power is measured: Yes / No START FREQUENCY The start frequency in megahertz [MHz] STEP FREQUENCY The linear frequency step in megahertz [MHz] STOP FREQUENCY The stop frequency in megahertz [MHz] TEMPERATURE Temperature during the test in degrees Celsius [oC] TSF FILENAME The filename of the TSF file USE CURRENT TEST LEVEL Indication the current test level is used: Yes / No USE VOLTAGE TEST LEVEL Indication the voltage test level is used: Yes / No VOLTAGE TEST LEVEL The voltage test level in the unit selected by the user EQUIPMENT TABLE A table containing all used equipment EVENTS Table of events which occur during a test POWER UNIT The selected unit for power: [dBuW], [dBm], [dBmW], [dBW], [uW], [mW] or [W] VOLTAGE UNIT The selected unit for voltage: [dBuVrms], [dBmVrms], [dBVrms], [uVrms], [mVrms] or [Vrms]
Keyword Description AMPLIFIER … Specifications of the used Amplifier ANTENNA … Specifications of the used Antenna CABLE nr … Specifications of the used Cable number [x], where x is a number between 1 to 8 CLAMP … Specifications of the used Clamp CLAMP RESISTANCE The resistance in Ohm [Ω] of the clamp COUPLER … Specifications of the used Coupler CURRENT SENSOR … Specifications of the used Current Sensor FORWARD POWER METER … Specifications of the used Forward Power Meter JIG … Specifications of the used Calibration Jig REFLECTED POWER METER … Specifications of the used Reflected Power Meter SENSOR POWER METER … Specifications of the used Sensor Power Meter SIGNAL GENERATOR … Specifications of the used Signal Generator
Keyword Description “keyword or name of photo” The name or keyword of the photo NUMBER OF EVENTS The number of events that occur during a test TEST START TIME The date and time at which the test is started TEST STOP TIME The date and time at which the test is stopped TEST DATE The date at which the test is performed TEST ENGINEER The name of the test engineer TEST TIME The time at which the test is performed TEST SITE NAME The name of the test equipment list AD GRAPH nr The graph representing the A/D graph of channel number [x], where x is a number between 1 and 8 CALCULATED CURRENT GRAPH (1) The graph representing the current graph CALCULATED VOLTAGE GRAPH (1) The graph representing the calculated voltage graph CURRENT GRAPH The graph representing the current graph FIELD GRAPH nr The graph representing the field graph of field strength sensor number [x], where x is a number between 1 and 8 FORWARD POWER GRAPH The graph representing the forward power graph NET POWER GRAPH The graph representing the net-power power graph REFLECTED POWER GRAPH The graph representing the reflected power graph SIGNAL POWER GRAPH The graph representing the signal power graph VOLTAGE GRAPH The graph representing the voltage graph AMPLIFIER PROTECTION MODE The Amplifier Protection mode: Signal Power, Forward Power and / or Reflected Power. CALIBRATION FILE (1) The name of the calibration file used for the test CALIBRATION METHOD The calibration method used: Signal Generator Level, Forward Power or Net-Power CURRENT LIMIT The Current limit in Ampere [A] CURRENT MEASURED Indication whether or not the current is measured: Yes / No CURRENT TEST LEVEL The current test level in Ampere [A] CURRENT UNIT The selected current unit: [dBuA],[dBmA], [dBA], [uA], [mA] or [A] DESCRIPTION The description of the Test Setup File (TSF) DISTANCE The distance between the antenna and the EUT in meters [m] DWELL TIME The dwell time in seconds [s] FORWARD POWER MEASURED Indication whether or not the forward power is measured: Yes / No FREQUENCY CHANGE MODE The method used to change frequency: Stress Method, Ramp Method, Constant Method or Fast Constant Method. HUMIDITY Relative Humidity during the test in percentage relative Humidity [%RH] LOGARITHMIC STEP The logarithmic frequency step in a percentage of the previous step LOGARITHMIC STEP FROM The frequency the test was started from to calculate the proper logarithmic step size MAIN VOLTAGE The mains voltage during the test in Volt [V]
Keyword Description MODULATION The type of modulation: none, Internal AM, External AM or Pulse MODULATION CONSERVATION Indication whether peak conservation is used: Power, Voltage, none MODULATION DEPTH Modulation depth in percentage of the carrier [%] MODULATION DUTY CYCLE The duty cycle of the pulse modulation as a percentage [%] MODULATION FREQUENCY The modulation frequency in Hertz [Hz] MODULATION SIGNAL The type of modulation signal: Sine Wave or Square Wave NET POWER MEASURED Indication whether or not the net-power power is measured: Yes / No NOTE Note or remark made during the test POLARIZATION Antenna polarisation: Horizontal or Vertical PRESSURE Air Pressure during the test in milliBar [mbar] REFLECTED POWER MEASURED Indication whether or not the reflected power is measured: Yes / No START FREQUENCY The start frequency in megahertz [MHz] STEP FREQUENCY The linear frequency step in megahertz [MHz] STOP FREQUENCY The stop frequency in megahertz [MHz] TAKE START PICTURE Indication whether or not a photo is taken at the start of the test: Yes/No TEMPERATURE Temperature during the test in degrees Celsius [oC] TSF FILENAME The filename of the TSF file USE CURRENT LIMIT Is the current limit used: Yes / No USE CURRENT TEST LEVEL (2) Indication the current test level is used: Yes / No USE VOLTAGE TEST LEVEL (6) Indication the voltage test level is used: Yes / No VOLTAGE TEST LEVEL (6) The voltage test level in the unit selected by the user EQUIPMENT TABLE A table containing all used equipment EVENTS Table of events which occur during a test MANUAL MODE Table containing the information collected during a manual mode session POWER UNIT The selected unit for power: [dBuW], [dBm], [dBmW], [dBW], [uW], [mW] or [W] VOLTAGE UNIT The selected unit for voltage: [dBuVrms], [dBmVrms], [dBVrms], [uVrms], [mVrms] or [Vrms] POWER TEST LEVEL (3) The power test level in the unit selected by the user STEP DOWN (4) The number of dB's the test level is stepped down during a minimum performance test TOLERANCE (5) The maximum allowable tolerance of the test-level
1) Fixed Generator method and Minimum Performance Level Method Excluded 2) Substitution method and Fixed Current Method only 3) Fixed Generator level method only 4) Minimum Performance test method only 5) Fixed Current method and Minimum Performance test method only 6) Substitution method only
Keyword Description AD CONVERTOR … Specifications of the used A/D Converter AMPLIFIER … Specifications of the used Amplifier ANTENNA … Specifications of the used Antenna CABLE nr … Specifications of the used Cable number [x], where x is a number between 1 to 8 CLAMP … Specifications of the used Clamp CLAMP RESISTANCE The resistance in Ohm [Ω] of the clamp COUPLER … Specifications of the used Coupler CURRENT SENSOR … Specifications of the used Current Sensor FORWARD POWER METER … Specifications of the used Forward Power Meter REFLECTED POWER METER … Specifications of the used Reflected Power Meter SENSOR POWER METER … Specifications of the used Sensor Power Meter SIGNAL GENERATOR … Specifications of the used Signal Generator
Pulsed Immunity[edit]
ESD Test[edit]
Keyword Description NEG AIR MEASURED POINT nr Indication a Negative Air Discharge is performed at point number [x]: Yes / No NEG AIR RESULT POINT nr Results of Negative Air Discharge at point number [x]: pass or fail NEG CONTACT MEASURED POINT nr Indication a Negative Contact Discharge is performed at point number [x]: Yes / No NEG CONTACT RESULT POINT nr Results of Negative Contact Discharge at point number [x]: pass or fail NUMBER OF EVENTS The number of events that occur during a test POS AIR MEASURED POINT nr Indication a Positive Air Discharge is performed at point number [x]: Yes / No POS AIR RESULT POINT nr Results of Positive Air Discharge at point number [x]: pass or fail POS CONTACT MEASURED POINT nr Indication a Positive Contact Discharge is performed at point number [x]: Yes / No POS CONTACT RESULT POINT nr Results of Positive Contact Discharge at point number [x]: pass or fail TEST START TIME The time at which the test is stopped TEST STOP TIME The time at which the test is started TEST SITE NAME The name of the test equipment list AMPLITUDE POINT nr Amplitude of pulse performed at point number [x] in [V] CONFIGURATION TIME The date and time when the TSF file is saved DESCRIPTION Description of the Test Setup File (TSF) DISCHARGE INTERVAL The interval of the discharge in seconds. HUMIDITY Relative Humidity during the test in percentage relative Humidity [%RH] MAIN VOLTAGE Mains Voltage during the test in Volt [V] NEG AIR LEVEL POINT nr Level of the Negative Air Discharge at point number [x] in [V] NEG CONTACT LEVEL POINT nr Level of the Negative Contact Discharge at point number [x] in [V] NOTE Note or remark made during the test NOTE POINT nr Note or remark for point number [x] NUMBER OF DISCHARGES The number of discharges PERFORMANCE CRITERIA CUSTOMER The performance criteria according to the customer PERFORMANCE CRITERIA PRODUCT STANDARD The performance criteria according to the product standard POS AIR LEVEL POINT nr Level of the Positive Air Discharge at point number [x] in [V] POS CONTACT LEVEL POINT nr Level of the Positive Contact Discharge at point number [x] in [V] PRESSURE Air Pressure during the test in MilliBar [mbar] TDURATION POINT nr Duration of pulse at point number [x] in milliseconds [ms] TEMPERATURE Temperature during the test in degrees Celsius [oC] TEST ENGINEER The name of the test engineer TESTPOINT DESCRIPTION POINT nr Position at EUT where discharge was applied TRISE POINT nr Rise time of pulse at point number [x] in milliseconds [ms] EQUIPMENT TABLE A table containing all used equipment EVENTS Table of events which occur during a test
Equipment for ESD Tests[edit]
Keyword Description ESDGUN … Specifications of the used ESD Gun
EFT Test[edit]
Keyword Description (L1-L2)/PE RESULT Results on (L1-L2)/PE: Pass or Fail CAPCLAMP RESULT Results on Capacitive Clamp: Pass or Fail L1 RESULT Results on L1: Pass or Fail L1/L2 RESULT Results on L1/L2: Pass or Fail L1/PE RESULT Results on L1/PE: Pass or Fail L2 RESULT Results on L2: Pass or Fail L2/PE RESULT Results on L2/PE: Pass or Fail NUMBER OF EVENTS The number of events that occur during a test PE RESULT Results on PE (Protective Earth): Pass or Fail TEST START TIME The time at which the test is stopped TEST STOP TIME The time at which the test is started TEST SITE NAME The name of the test equipment list BRAND Brand of Measurement equipment CALIBRATION EXPIRES Date when calibration expires of equipment, given by the user DESCRIPTION Description of Measurement equipment given by the user DEVICE DRIVER DLL N/A DEVICE DRIVER NAME The internal name of the device driver HARDWARE VERSION Hardware version of Measurement equipment ID Internal ID of Measurement equipment INI FILE NAME N/A NAME Name of Measurement equipment SERIAL NUMBER Serial number of Measurement equipment SOFTWARE VERSION Software version of Measurement equipment SUPPORTED DEVICE TYPES N/A TYPE Type of Measurement equipment (L1-L2)/PE MEASURED (L1-L2)/PE Measured: Yes / No AMPLITUDE Amplitude of burst in Volt [V] CAPCLAMP MEASURED Capacitive Clamp Measured: Yes / No CONFIGURATION TIME The date and time when the TSF file is saved DESCRIPTION Description of the Test Setup File (TSF) DURATION Duration of test in milliseconds [ms] FREQUENCY Frequency of burst in Hertz [Hz] HUMIDITY Relative Humidity during the test in percentage relative Humidity [%RH] L1 MEASURED L1 Measured: Yes / No L1/L2 MEASURED L1/L2 Measured: Yes / No L1/PE MEASURED L1/PE Measured: Yes / No L2 MEASURED L2 Measured: Yes / No L2/PE MEASURED L2/PE Measured: Yes / No MAIN VOLTAGE Mains Voltage during the test in Volt [V]
Keyword Description NOTE Note or remark made during the test NOTE (L1-L2)/PE Notes or remarks for (L1-L2)/PE NOTE CAPCLAMP Notes or remarks for Capacitive Clamp NOTE L1 Notes or remarks for L1 NOTE L1/L2 Notes or remarks for L1/L2 NOTE L1/PE Notes or remarks for L1/PE NOTE L2 Notes or remarks for L2 NOTE L2/PE Notes or remarks for L2/PE NOTE PE Notes or remarks for PE PE MEASURED PE Measured: Yes / No PRESSURE Air Pressure during the test in milliBar [mbar] TBURST Duration of burst in milliseconds [ms] TEMPERATURE Temperature during the test in degrees Celsius [oC] TEST ENGINEER The name of the test engineer TPERIOD Duration of period in milliseconds [ms] EQUIPMENT TABLE A table containing all used equipment EVENTS Table of events which occur during a test
Keyword Description MEASURED If the line was measured RESULT The result on this line AMPLITUDE The amplitude on this line NOTE The note that is related to this line
Equipment for EFT Tests[edit]
Keyword Description EFTGENERATOR … Specifications of the used EFT Generator
Surge Test[edit]
Keyword Description (L1-L2)/PE MEASURED (L1-L2)/PE measured: Yes / No (L1-L2)/PE RESULT Results of Surge Test at (L1-L2)/PE: pass or fail L1/L2 MEASURED Indication whether or not L1/L2 is measured: Yes / No L1/L2 RESULT Results of Surge Test at L1/L2: pass or fail L1/PE MEASURED Indication whether or not L1/PE is measured: Yes / No L1/PE RESULT Results of Surge Test at L1/PE: pass or fail L2/PE MEASURED Indication whether or not L2/PE is measured: Yes / No L2/PE RESULT Results of Surge Test at L2/PE: pass or fail NUMBER OF EVENTS The number of events that occur during a test TEST START TIME The time at which the test is stopped TEST STOP TIME The time at which the test is started TEST SITE NAME The name of the test equipment list CINTSETTINGS (L1-L2)/PE Internal Capacitance for (L1-L2)/PE CINTSETTINGS L1/L2 Internal Capacitance for L1/L2 CINTSETTINGS L1/PE Internal Capacitance for L1/PE CINTSETTINGS L2/PE Internal Capacitance for L2/PE CONFIGURATION TIME The date and time when the TSF file is saved DESCRIPTION Description of the Test Setup File (TSF) HUMIDITY Relative Humidity during the test in percentage relative Humidity [%RH] MAIN VOLTAGE Mains Voltage during the test in Volt [V] NOTE Note or remark made during the test NOTE (L1-L2)/PE Note or remark for Surge Test at (L1-L2)/PE NOTE L1/L2 Note or remark for Surge Test at L1/L2 NOTE L1/PE Note or remark for Surge Test at L1/PE NOTE L2/PE Note or remark for Surge Test at L2/PE PHASE (L1-L2)/PE Phase angle between L1 - L2 and PE in degrees: 0, 90, 270, random PHASE L1/L2 Phase angle between L1 and L2 in degrees: 0, 90, 270, random PHASE L1/PE Phase angle between L1 and PE in degrees: 0, 90, 270, random PHASE L2/PE Phase angle between L2 and PE in degrees: 0, 90, 270, random POLARISATION OF (L1-L2)/PE TESTLEVEL Polarisation of test voltage at (L1-L2)/PE: Positive, Negative or both POLARISATION OF L1/L2 TESTLEVEL Polarisation of test voltage at L1/L2: Positive, Negative or both POLARISATION OF L1/PE TESTLEVEL Polarisation of test voltage at L1/PE: Positive, Negative or both POLARISATION OF L2/PE TESTLEVEL Polarisation of test voltage at L2/PE: Positive, Negative or both PRESSURE Air Pressure during the test in milliBar [mbar] RINTSETTINGS (L1-L2)/PE Internal Resistance for (L1-L2)/PE RINTSETTINGS L1/L2 Internal Resistance for L1/L2 RINTSETTINGS L1/PE Internal Resistance for L1/PE RINTSETTINGS L2/PE Internal Resistance for L2/PE TEMPERATURE Temperature during the test in degrees Celsius [oC] TEST ENGINEER The name of the test engineer
Keyword Description TESTLEVEL (L1-L2)/PE Test level at (L1-L2)/PE in kilo Volt [kV] TESTLEVEL L1/L2 Test level at L1/L2 in kilo Volt [kV] TESTLEVEL L1/PE Test level at L1/PE in kilo Volt [kV] TESTLEVEL L2/PE Test level at L2/PE in kilo Volt [kV] EQUIPMENT TABLE A table containing all used equipment EVENTS Table of events which occur during a test
For every available line the following codes can be applied: Keyword Description MEASURED If the selected line was measured RESULT The result of the measurement POSITIVE TEST LEVEL The positive test level(Yes / No) NEGATIVE TEST LEVEL The negative testlevel(Yes / No) CI The capacity used for this line configuration RI The resistance used for this line configuration Test Level The test level used for this line configuration Note The note that applies to this line configuration PHASE 0 Phase 0 measured (Yes / No) PHASE 90 Phase 90 measured (Yes / No) PHASE 270 Phase 270 measured (Yes / No) PHASE RANDOM Phase random measured (Yes / No)
Keyword Description SURGEGENERATOR … Specifications of the used Surge Generator
Emission[edit]
Note : The pass/fail criteria, shown in the tables of the report generator do not take the measure uncertainty of the test setup into account.
Radiated Emission OATS[edit]
Keyword Description TEST SITE NAME The name of the test equipment list EQUIPMENT TABLE A table containing all used equipment
Equipment for OATS Emission Tests[edit]
Keyword Description ANTENNA … Specifications of the used Antenna ANTENNA TOWER … Specifications of the used Antenna Tower CABLE nr … Specifications of the used Cable number [x], where x is a number between 1 to 8 NUMBER OF DETECTORS The number of detectors the Receiver is able to handle simultaneously PRE AMPLIFIER … Specifications of the used Pre Amplifier RECEIVER TYPE The type of receiver: Spectrum Analyser, Measurement Receiver or Scanning Receiver SPECTRUM ANALYZER … Specifications of the used Spectrum Analyser or Measurement Receiver TEST SITE NAME The name of the test equipment list TURN TABLE … Specifications of the used Turntable
Radiated Emission Tests[edit]
Keyword Description “keyword or name of photo” The name or keyword of the photo NUMBER OF EVENTS The number of events that occur during a test PEAK DETECTION COUNT The number of detected peaks TEST START TIME The time at which the test is stopped TEST STOP TIME The time at which the test is started MAX HOLD GRAPH The max hold graph of the all the traces MAX HOLD PEAKS GRAPH The max hold graph of all the traces and markers MAX PEAK GRAPH The max hold graph of the Peak trace MAX AVERAGE GRAPH The max hold graph of the Average trace MAX QUASI PEAK GRAPH The max hold graph of the Quasi Peak trace MAX RMS GRAPH The max hold graph of the RMS trace ANGLE STEPS The number of steps between the start position and the stop position ATTENUATOR The attenuation of the attenuator in dB ATTENUATOR MODE The Attenuator Mode: Auto, Coupled or Manual AVERAGE DETECTOR Indication whether or not the Average detector is measured: Yes / No AVERAGE MEASURE TIME The measurement time of the average peak detector in seconds [s] AVERAGE OBSERVATION TIME The observation time per average peak in seconds [s] BELOW LIMITLINE OFFSET The level of peaks to detect in dB below the limit line CONFIGURATION TIME The date and time when the TSF file is saved CURRENT UNIT The selected unit for current: [dBuA], [dBmA], [dBA], [uA], [mA] or [A] DB/DIV N/A DB/DIV MODE N/A DESCRIPTION Description of the Test Setup File (TSF) DISTANCE The distance between the antenna and the EUT in meters [m] EMISSION OPTIMIZING METHOD N/A HEIGHT STEPS The number of steps between the minimum height and the maximum height HIGHEST HEIGHT The maximum height of the antenna tower in meters [m] HUMIDITY Relative Humidity during the test in percentage relative Humidity [%RH] INTERNAL PREAMPLIFIER The amplification of the internal pre amplifier in [dB] INTERNAL PREAMPLIFIER MODE The Internal Pre Amplifier Mode: Auto, Coupled or Manual LOWEST HEIGHT The minimum height of the antenna tower in meters [m] MAIN VOLTAGE Mains Voltage during the test in Volt [V] MAXIMUM PEAK COUNT The maximum number of peaks to be detected MEASURE TIME The measure time in milliseconds [ms] MEASURE TIME MODE The Mode: Auto, Coupled or Manual NOTE Note or remark made during the test PEAK DETECTOR Indication whether or not the Peak detector is measured: Yes / No PEAK MEASURE TIME The measurement time of the peak detector in seconds [s] PEAK OBSERVATION TIME The observation time per peak in seconds [s] PEAK SEARCH Indication whether or not a peak search is performed: Yes / No PEAK SEARCH Indication a peak search is performed: Yes / No PEAK TRACE Indication whether or not the Peak trace is measured: Yes / No PEAKSWEEPS Number of peak sweeps performed Keyword Description POINT X FREQUENCY The frequency of point X in the peak list ( X starting from 1) POINT X PEAK The peak value of point X in the peak list (X starting from 1) POINT X QUASI PEAK The quasi peak value of point X in the peak list (X starting from 1) POINT X AVERAGE The average value of point X in the peak list (X starting from 1) POLARIZATION The polarisation of the antenna: horizontal, vertical or both PRESSURE Air Pressure during the test in milliBar [mbar] QUASIPEAK DETECTOR Indication whether or not the Quasi peak detector is measured: Yes/No QUASIPEAK MEASURE TIME The measurement time of the quasi peak detector in seconds [s] QUASIPEAK OBSERVATION TIME The observation time per quasi peak in seconds [s] QUASIPEAK TRACE Indication whether or not the Quasi peak trace is measured: Yes / No RBW The Radio Bandwidth in kilo Hertz [kHz] RBW MODE The RBW Mode: Auto, Coupled or Manual RECEIVER DATA SAVE MODE Indication whether all data is saved or only the test results REFERENCE LEVEL The reference level in the unit selected by the user REFERENCE LEVEL MODE The Reference Level Mode: Auto, Coupled or Manual RMS DETECTOR Indication whether or not the RMS detector is measured: Yes / No RMS MEASURE TIME The measurement time of the RMS peak detector in seconds [s] RMS OBSERVATION TIME The observation time per RMS peak in seconds [s] RMS TRACE Indication whether or not the RMS trace is measured: Yes / No START ANGLE The start position in degrees START FREQUENCY The start frequency in megahertz [MHz] STEP FREQUENCY The step frequency in kilo Hertz [kHz] STEP FREQUENCY MODE The Step frequency Mode: Auto, Coupled or Manual STOP ANGLE The stop position in degrees STOP FREQUENCY The stop frequency in megahertz [MHz] SWEEP TIME The sweep time in milliseconds [ms] SWEEP TIME MODE The Sweep Time Mode: Auto, Coupled or Manual TAKE START PICTURE Indication whether or not a picture is taken at the start of the test: Yes / No TEMPERATURE Temperature during the test in degrees Celsius [oC] TEST ENGINEER The name of the test engineer TEST SITE NAME The name of the test equipment list USE INTERNAL PREAMP N/A USE START STOP FREQUENCY Yes / No, if Yes: Start/Stop if No: Frequency list is used. VBW The Video Bandwidth in kilo Hertz [kHz] VBW MODE The Video Bandwidth Mode: Auto, Coupled or Manual EMISSION TABLE Table of detected peaks and per peak: frequency, level(s), limit value, pass/fail, location EQUIPMENT TABLE A table containing all used equipment EVENTS Table of events which occur during a test EUT HEIGHT (1) The height of the EUT in meters [m] SEPTUM HEIGHT (1) The height of the septum in meters [m]
1) GTEM Method only
Keyword Description ANTENNA … Specifications of the used Antenna ANTENNA TOWER … Specifications of the used Antenna Tower CABLE nr … Specifications of the used Cable number [x], where x is a number between 1 to 8 NUMBER OF DETECTORS The number of detectors the Receiver is able to handle simultaneously PRE AMPLIFIER … Specifications of the used Pre Amplifier RECEIVER TYPE The type of receiver: Spectrum Analyser, Measurement Receiver or Scanning Receiver SPECTRUM ANALYZER … Specifications of the used Spectrum Analyser or Measurement Receiver TURN TABLE … Specifications of the used Turntable
Keyword Description "keyword or name of photo" The name or keyword of the photo NUMBER OF EVENTS The number of events that occur during a test PEAK DETECTION COUNT The number of detected peaks TEST START TIME The time at which the test is stopped TEST STOP TIME The time at which the test is started TEST SITE NAME The name of the test equipment list MAX HOLD GRAPH The max hold graph of the all the traces MAX HOLD PEAKS GRAPH The max hold graph of all the traces and markers MAX PEAK GRAPH The max hold graph of the Peak trace MAX AVERAGE GRAPH The max hold graph of the Average trace MAX QUASI PEAK GRAPH The max hold graph of the Quasi Peak trace MAX RMS GRAPH The max hold graph of the RMS trace ATTENUATOR The attenuation of the attenuator in dB ATTENUATOR MODE The Attenuator Mode: Auto, Coupled or Manual AVERAGE DETECTOR Indication whether or not the Average detector is measured: Yes / No AVERAGE MEASURE TIME The measurement time of the average peak detector in seconds [s] AVERAGE OBSERVATION TIME The observation time per average peak in seconds [s] BELOW LIMITLINE OFFSET The level of peaks to detect in dB below the limit line CONFIGURATION TIME The date and time when the TSF file is saved CURRENT UNIT The selected unit for current: [dBuA], [dBmA], [dBA], [uA], [mA] or [A] DB/DIV N/A DB/DIV MODE N/A DESCRIPTION Description of the Test Setup File (TSF) EMISSION OPTIMIZING METHOD N/A HUMIDITY Relative Humidity during the test in percentage relative Humidity [%RH] INTERNAL PREAMPLIFIER The amplification of the internal pre amplifier in [dB] INTERNAL PREAMPLIFIER MODE The Internal Pre Amplifier Mode: Auto, Coupled or Manual LISN LINE Which LISN line is used: Neutral, Phase 1, Phase 2 or Phase 3 MAIN VOLTAGE Mains Voltage during the test in Volt [V] MAXIMUM PEAK COUNT The maximum number of peaks to be detected MEASURE TIME The measure time in milliseconds [ms] MEASURE TIME MODE The Mode: Auto, Coupled or Manual NOTE Note or remark made during the test PEAK DETECTOR Indication whether or not the Peak detector is measured: Yes / No PEAK MEASURE TIME The measurement time of the peak detector in seconds [s] PEAK OBSERVATION TIME The observation time per peak in seconds [s] PEAK TRACE Indication whether or not the Peak trace is measured: Yes / No PEAKSWEEPS Number of peak sweeps performed PRESSURE Air Pressure during the test in milliBar [mbar] QUASIPEAK DETECTOR Indication whether or not the Quasi peak detector is measured: Yes / No QUASIPEAK MEASURE TIME The measurement time of the quasi peak detector in seconds [s] QUASIPEAK OBSERVATION TIME The observation time per quasi peak in seconds [s] QUASIPEAK TRACE Indication whether or not the Quasi peak trace is measured: Yes / No Keyword Description RBW The Radio Bandwidth in kilo Hertz [kHz] RBW MODE The RBW Mode: Auto, Coupled or Manual RECEIVER DATA SAVE MODE Indication whether all data is saved or only the test results REFERENCE LEVEL The reference level in the unit selected by the user REFERENCE LEVEL MODE The Reference Level Mode: Auto, Coupled or Manual RMS DETECTOR Indication whether or not the RMS detector is measured: Yes / No RMS MEASURE TIME The measurement time of the RMS peak detector in seconds [s] RMS OBSERVATION TIME The observation time per RMS peak in seconds [s] RMS TRACE Indication whether or not the RMS trace is measured: Yes / No START FREQUENCY The start frequency in megahertz [MHz] STEP FREQUENCY The step frequency in kilo Hertz [kHz] STEP FREQUENCY MODE The Step frequency Mode: Auto, Coupled or Manual STOP FREQUENCY The stop frequency in megahertz [MHz] SWEEP TIME The sweep time in milliseconds [ms] SWEEP TIME MODE The Sweep Time Mode: Auto, Coupled or Manual TAKE START PICTURE Indication whether or not a picture is taken at the start of the test: Yes / No TEMPERATURE Temperature during the test in degrees Celsius [oC] TEST ENGINEER The name of the test engineer USE INTERNAL PREAMP N/A USE START STOP FREQUENCY Yes / No, if Yes: Start/Stop if No: Frequency list is used. VBW The Video Bandwidth in kilo Hertz [kHz] VBW MODE The Video Bandwidth Mode: Auto, Coupled or Manual EMISSION TABLE Table of detected peaks and per peak: frequency, level(s), limit value, pass/fail, location EQUIPMENT TABLE A table containing all used equipment EVENTS Table of events which occur during a test ABSORBING CLAMP … (1) Specifications of the used Absorbing Clamp CLAMP MOVER … (1) Specifications of the used Clamp Mover LARGEST DISTANCE The maximum distance of the clamp in meters [m] SMALLEST DISTANCE The minimum distance of the clamp in meters [m] STEPS The number of steps between the minimum distance and the maximum distance
Keyword Description CABLE nr [x] … Specifications of the used Cable number [x], where x is a number between 1 to 8 LISN … Specifications of the used LISN NUMBER OF DETECTORS The number of detectors the Receiver is able to handle simultaneously PRE AMPLIFIER … Specifications of the used Pre Amplifier RECEIVER TYPE The type of receiver: Spectrum Analyser, Measurement Receiver or Scanning Receiver SPECTRUM ANALYZER … Specifications of the used Spectrum Analyser or Measurement Receiver