(190 intermediate revisions by 4 users not shown)
Line 4: Line 4:
==Introduction==
==Introduction==
The RadiMation® package support three main radiated emission methods:
The {{RadiMation}} package supports several different radiated emission methods:
*Anechoic chamber method
*Fixed height method
* [[#Multiband|Multiband]]
*G-TEM method
* [[#Manual Mode|Manual Mode]]
* Burst
* [[#Anechoic chamber|Anechoic chamber]]
* [[#Fixed height|Fixed height]]
* [[#GTEM|GTEM]]
* Mil Std Magnetic
* [[#Manual mode (Singleband)|Manual mode (Singleband)]]
* [[#Merge|Merge]]
* [[#X + k * S|X + k * S]]
==Starting a radiated emission test==
===Loading a TSF file===
To start a radiated emission test, an EUT file must be defined first. From the EUT file, select: {{Menu|Tests|Radiated Emission}} and then the desired test method.
A list of Technical Setup Files (TSF files) will be displayed. The test engineer can select one of these TSF files to load all parameters from a previously defined emission test or press {{ScreenElement|New}} to define a new emission test.
[[File:SelectTSF REMultiband.png]]
When a TSF file is loaded, all test parameters will be already configured and pressing the {{ScreenElement|Start Test}} button on the right side of the screen will start the radiated emission test.
When no TSF file is loaded (by pressing {{ScreenElement|New}} in the TSF selection window) all parameters will be at zero and can be configured by the operator.
===Test initialisation and self test===
When the {{ScreenElement|Start Test}} button is pressed, the software first runs a number of checks.
# The presence of a valid software key is checked.
# The software settings are checked (i.e. is all test equipment used within its parameters?)
# The presence of all test equipment is checked.
# All relevant equipment will be initialised.
===Test results screen===
You can access the test results of the performed test by selecting the test in the EUT window and then press the {{ScreenElement|Info}} button.
For the Multiband test, the window that is appearing looks a lot like the manual mode screen, the reason that it looks the same is because it is. You can continue testing after you finished a test, just select the testsite you want to use and you can continue testing.
{{Note|The pass/fail criteria, shown in the tables do not take the measure uncertainty of the test setup into account.}}
{{Note|Continuing the test is only possible with the selected test site. {{RadiMation}} will not select the correct test site for you. All measurements are done with the selected test site. If you want automatic changes of the test site, restart the configured TSF file.}}
===Testsite information===
An radiated emission test, requires that a testsite is used that can have the following components.
[[Image:Radiated Emission Test Setup Chamber.png]]
{|class="wikitable"
!# || Device name || Tab in {{RadiMation}} configuration window
|  ||Spectrum analyser || Devices 2
|  ||Pre amplifier || Devices 2
|  ||Antenna || Devices 1
|  ||Turn table || Devices 1
|  ||Antennatower || Devices 1
!colspan="3" |'''Cables'''
|1 || Cable preamp -> analyser || Cables
|2 || Cable antenna -> preamp|| Cables
{{Note|Usage of cable correction factors is not mandatory. However, using cable correction factors will increase measurement accuracy during emission measurements. When in doubt contact your reseller.}}
== Multiband ==
The radiated emission Multiband test, is a very flexible and powerful test to perform radiated emission tests.
When the multiband test is performed, it will first perform a prescan measurement on all the specified locations. After this prescan is finished the temporary results are shown, and some peaks are automatically detected and shown. At this moment those peaks are NOT measured yet with the final measurement detectors and are also not optimized yet. The end-user now has the possibility to add additional peaks, or do some other peak investigation/changes.
When the end-user is satisfied with the list of detected peaks, he can continue with the final measurement, by pressing the {{ScreenElement|Automatic}} button again.
When the end-user has configured peak optimization to be active, the peak optimization will now also be performed.
Now the peaks will be measured with the detector that are selected in the {{ScreenElement|Final peak measurement}} box. Only after those measurements are finalized, those detector values will become available in the list of detected peaks.
[[Image:Radiated Emission Multi Band Configuration Window.png]]
The test engineer can configure a complete radiated emission test in the Radiated Emission configuration menu. The test engineer can change the frequency range from one frequency band or the RBW of several bands. It all depends on the selection of bands. When several bands are selected and some options are blank, this means that these settings in bands are different. For example if the test engineer selects both frequency band in the window above, the frequency range will be blank. If the engineer sets the start and stop frequency, all the selected bands will have the same frequency setting. One of the advantages of this feature is the quick settings of several settings like the RBW or the peak detection.
===Frequency window===
[[Image:Radiated Emission Multi Band Bands Window.png]]
In the frequency window, the test engineer can enter a frequency range to be measured.
Pressing {{ScreenElement|Add}} will add a new frequency band to the test, {{ScreenElement|Remove}} will remove all the selected bands and {{ScreenElement|Duplicate}} will duplicate all selected bands.
===Turntable window===
[[Image:Radiated Emission Multi Band Turntable Angle Window.png]]
{{ScreenElementDescription|Start Angle|Sets the start position of the turntable.}}
{{ScreenElementDescription|End Angle|Sets the stop position of the turntable.}}
{{ScreenElementDescription|Steps|Sets the number of steps for one turntable rotation.}}
{{ScreenElementDescription|Turn during measurements|This will force the turntable to continuously rotate during the test and will disable steps setting.}}
{{ScreenElementDescription|EUT angle offset|The offset of the EUT compared to the zero angle of the table.}}
{{ScreenElementDescription|Optimize Angle|Optimized settings for searching the correct angle.}}
===Turntable window optimize angle settings===
[[Image:Radiated Emission Multi Band Optimized Angle Settings Window.png]]
When optimize angle is selected, the user can configure the way a peak search is performed.
In the optimize angle configuration window, the following configurations can be made:
{{ScreenElementDescription|Area|Range over which the peak search will be performed (0 to 360 degrees max.).}}
{{ScreenElementDescription|Step size|The angle step size.}}
The area is defined 50% left of the starting position and 50% right. So if the area is 30 degrees, {{RadiMation}} will scan between starting position plus and minus 15 degrees.
In the detector window the test engineer can select which detectors to use for measuring the optimal angle.
After selecting the detectors the measure time and observation time of the selected detector are enabled. It is possible to use other settings for finding the peak then the settings on the main configuration window.
===Antenna tower window===
[[Image:Radiated Emission Multi Band Antenna Height Window.png]]
{{ScreenElementDescription|Max. height|Sets maximum height of the antenna tower.}}
{{ScreenElementDescription|Min. height|Sets the minimum height of the antenna tower.}}
{{ScreenElementDescription|Steps|Sets the number of steps to go from min. height to max. height.}}
{{ScreenElementDescription|Move during measurements|This will force the Antenna Tower to continuously move during the test and will disable steps setting.}}
{{ScreenElementDescription|Optimize height|The software will automatically detect on which antenna height maximum  emission will occur.}}
===Optimize height settings===
[[Image:Radiated Emission Multi Band Optimized Height Settings Window.png]]
When optimize height is selected, the user can configure the way a peak search is performed.
In the optimize height configuration window, the following configurations can be made:
{{ScreenElementDescription|Area|Range over which the peak search will be performed (0 to Max height).}}
{{ScreenElementDescription|Step size|The height step size.}}
The area is defined 50% above of the starting position and 50% below. So if the area is 1 meter, {{RadiMation}} will scan between starting position plus and minus 50 cm.
===Antenna window===
[[Image:Radiated Emission Multi Band Antenna Settings Window.png]]
====Polarisation====
# Horizontal
# Vertical
# Both
====Distance====
{{RadiMation}} uses the distance information for reference only.
===Receiver settings===
[[Image:Radiated Emission Multi Band Receiver Settings Window.png]]
{{ScreenElementDescription|Reference level|Sets the reference level of the receiver.}}
{{ScreenElementDescription|Attenuator|Sets the receiver attenuator level.}}
{{ScreenElementDescription|RBW|Sets the Resolution Band Width filter.}}
{{ScreenElementDescription|VBW|Sets the Video Band Width filter.}}
{{ScreenElementDescription|Sweep|Sets the sweep time of the receiver.}}
{{ScreenElementDescription|Step frequency|Sets the frequency step size of the receiver.}}
{{ScreenElementDescription|Measure time|Sets the measure time of the receiver.}}
{{ScreenElementDescription|Pre-amp|Sets the pre-amp gain.}}
When a parameter is set to auto, {{RadiMation}} will determine the most optimum settings for this parameter. If the parameter is set to coupled, the parameter will be set to automatic internally in the receiver.
===Traces window===
[[Image:Radiated Emission Multi Band Traces Settings Window.png]]
{{ScreenElementDescription|Detectors|The detector settings (Peak, Average, Quasi-peak, RMS) used during sweeptime. This setting is generic. So it is possible to select multiple detectors, but only one detector can be selected due to the device limitations. In that case for each enabled detector the amount of sweeps will be performed by {{RadiMation}}.}}
{{ScreenElementDescription|Number of sweeps|The number op peak hold sweeps for each measurement can be selected. If this value is set to, for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweep speed. For spectrum analysers and CW signals you can use a relatively low amount of peak sweeps, for example 10. But when you have discontinues signals like a spark bridge, you may need 100 peak sweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope.}}
===Final peak measurement===
[[Image:Radiated Emission Multi Band Final Peak Measurement Settings Window.png]]
In the final peak measurement window the test engineer can configure the detectors used during the final emission measurements of the detected peaks. The first value is the time constant of the detector while the second value is used for the measure time at each frequency point for this detector. The time constant is send to the receiver, to set the sample time of the detector. To select a detector, check the desired button.
===Peak detection window===
[[Image:Radiated Emission Multi Band Peak Detection Settings Window.png]]
{{ScreenElementDescription|No peaks|When no final peak measurements are required select the option no peaks.}}
{{ScreenElementDescription|Maximum Peaks|With the max. peaks option, the user can limit the number of peak signals the software will detect and re-measure.}}
===Test site window===
[[Image:Radiated Emission Multi Band Test Site Settings Window.png]]
{{ScreenElementDescription|Equipment icon|By clicking on the equipment icon, the equipment list can be viewed and edited.}}
===Limit lines window===
[[Image:Radiated Emission Multi Band Limit Lines Settings Window.png]]
{{ScreenElementDescription|All peaks above “xx” below the limit lines|In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 5 dB, the software will detect all peaks, which are higher than 5 dB below the limit line.}}
===Change Order window===
[[Image:Radiated Emission Multi Band Change Order Settings Window.png]]
In some case you want to switch for example the antenna height scan before the turn table scan because this will decrease the total measuring time. With the change order the test engineer can control the way {{RadiMation}} performs the testing sequence.  Read the list from top to bottom to see what changes first. The best way explain the change order window is by example.
[[Image:Radiated Emission Multi Band Band A Band B.png]]
You need to perform a test as shown but you have 2 test sites. One with an antenna with the frequency range that is equal to the total band. The second test site you have 2 antennas, one for every frequency band. For test site one you want the RBW filter to change as little as possible to prevent the old analyser from breaking down.
So for Site 1 you want:
*Band A RBW 9 kHz
*Band B RBW 9 kHz
*Band A RBW 120 kHz
*Band B RBW 120 kHz
And for test site 2:
*Band A RBW 9 kHz
*Band A RBW 120 kHz
*Band B RBW 9 kHz
*Band B RBW 120 kHz
*Band A RBW 9 kHz
*Band B RBW 9 kHz
*Band B RBW 120 kHz
*Band A RBW 120 kHz
This is even better then we had hoped for, the amount of switching has been decrees due to the fact that {{RadiMation}} detected that it had to measure the same frequency twice.
*RBW
*Testsite
*Frequency band
*Band A RBW 9 kHz
*Band A RBW 120 kHz
*Band B RBW 120 kHz
*Band B RBW 9 kHz
This is also better then we had hoped for, the amount of switching has been decrees due to the fact that {{RadiMation}} detected that it had to measure the same RBW twice.
{{Note|If you have any questions regarding the setting of the change order please contact your reseller.}}
== Manual Mode ==
[[Image:Radiated Emission Manual Mode Multi Band Test.png]]
[[Image:Radiated Emission Multi Band Manual Mode Receiver Window.png]]
===Turn table===
[[Image:Radiated Emission Multi Band Manual Mode Turn Table Window.png]]
The turn table settings are directly linked to the turn table. Meaning that when a test engineer changes one of the settings then {{RadiMation}} will apply this setting directly.
===Antenna tower===
[[Image:Radiated Emission Multi Band Manual Mode Antenna Tower Window.png]]
The antenna tower settings are directly linked to the antenna tower. Meaning that when a test engineer changes one of the settings then {{RadiMation}} will apply this setting directly.
===Test site===
[[Image:Radiated Emission Multi Band Manual Mode Test Site Window.png]]
Before manual mode can be used the test site needs to be selected so that {{RadiMation}} knows which devices to use.
===Final measurement===
[[image:Radiated Emission Multi Band Final Peak Measurement Settings Window.png]]
===Peaks window===
[[Image:Radiated Emission Multi Band Manual Mode Peaks Window.png]]
{{ScreenElementDescription|Max Peaks|Maximum allowed amount of peaks to be automatically determined.}}
{{ScreenElementDescription|Detect Peaks|Detects the configured amount of sweeps in the measured spectrum.}}
{{ScreenElementDescription|Measure Peaks|Measures all selected peaks with the configured settings in the final measurement window.}}
{{ScreenElementDescription|Peak Actions|Allows several actions to be performed on the detected peaks, like select all and deselect all.
    {{ScreenElementDescription|Peaks to maximum|Start performing measurement to redetermine on which turntable angle and antenna tower height the highest emission value was measured.
This angle and height will be updated in the selected peak, together with the corresponding measured emission values. If multiple peaks are selected, this search for the highest emission value is done individually for every selected peak.
This 'peaks to maximum' function can be very helpful during the angle or height optimization of a peak. If after manual changes of the antenna height or turntable angle, no other maximum is found, the peak can be reset to the angle and height where the maximum was determined. 'Peaks to maximum', will thus select that location on which the highest emission value was measured.
    {{ScreenElementDescription|Measure Peaks|Start performing a final measurement on the selected peaks.}}
    {{ScreenElementDescription|Optimize Angle|Start performing a measurement as configured for the Optimize Angle.}}
    {{ScreenElementDescription|Optimize Height|Start performing a measurement as configured for the Optimize Height.}}
    {{ScreenElementDescription|Go To Peak Location|If possible it will adjust the Antenna Tower and Turntable to the location of the selected peak.}}
    {{ScreenElementDescription|Set receiver to Peak settings|Set the current receiver settings to the used settings of the selected peak.}}
    {{ScreenElementDescription|Configure -> Optimize Angle|Opens a new window to configure the Optimize Angle settings. (See also Optimize Angle settings.)}}
    {{ScreenElementDescription|Configure -> Optimize Height|Opens a new window to configure the Optimize Height settings. (See also Optimize height settings.)}}
    {{ScreenElementDescription|Delete peaks|Delete selected peak(s).}}
{{ScreenElementDescription|Delete Peaks|Delete the selected peaks from the list.}}
{{ScreenElementDescription|Sweeps|Amount of sweeps to be measured.}}
{{ScreenElementDescription|Take Sweeps|Triggers the receiver of analyser to take the configured amount of sweeps.}}
{{ScreenElementDescription|Continuous Measure|Start performing of taking measurements. This is a heavy process and it is normal when more CPU is used by the computer.}}
{{ScreenElementDescription|Continuous Measure Write|Each new measurement will overwrite the previous measurement data.}}
{{ScreenElementDescription|Continuous Measure Max-hold|Only higher new measurement values will overwrite the previous measurement data.}}
{{ScreenElementDescription|Clear measurement data|Removes all previous measurement data.}}
==Anechoic chamber==
Line 29: Line 364:
{{ScreenElementDescription|Start position|Sets the start angle of the turntable.}}
{{ScreenElementDescription|Stop position|Sets the stop angle of the turntable.}}
{{ScreenElementDescription|#|Sets the number of steps for one turntable rotation.}}
{{ScreenElementDescription|#|Number of locations from {{ScreenElement|Start postion}} to {{ScreenElement|Stop position}} that should be measured.}}
{{ScreenElementDescription|Peak search|The software will automatically detect on which angle maximum emission will occur.}}
{{ScreenElementDescription|Peak search|The software will automatically detect on which turntable angle the maximum measurement value of each peak is present.}}
*If a turntable device is specified in the selected test-site, {{RadiMation}} will control the turntable to turn to the specified angle.
*If no turntable device is specified in the selected test-site, {{RadiMation}} will assume that the EUT is already positioned on that specified angle. No actions on the turntable will be taken.
As long as a single angle is measured, and no turntable device is specified in the test-site, it is possible to use the Multiband Radiated Emission test without a turntable.
So, if you specify the same {{ScreenElement|Start position}} and {{ScreenElement|Stop position}}, and specify '1' to the number of steps ({{ScreenElement|#}}), then the test will assume that only that single specified angle should be measured.
When {{ScreenElement|Peak search}} is selected, the user can configure the method that is used during the peak search. In the peak search configuration window, the following configurations can be made:
{{ScreenElementDescription|Area|Range over which the peak search will be performed. The initially detected turntable angle of the peak will be the middle of this {{ScreenElement|Area}}, and the total search area for this peak will be half of this area clockwise and half of this area counter-clockwise.}}
{{ScreenElementDescription|Step size|The step size for each turntable movement.}}
{{ScreenElementDescription|Threshold|This value is only used during the {{ScreenElement|Intelligent}} mode. It influences how long the movement is continued. The movement will stop when the measurement value is lower than the specified {{ScreenElement|Threshold}} below the highest known measurement value of the searched peak.}}
{{ScreenElementDescription|Mode|The mode setting determines the way a peak is detected. The following modes are available:
* {{ScreenElement|Intelligent}}
* {{ScreenElement|Complete area}}
When peak search is selected, the user can configure the way a peak search is performed.
The Intelligent mode has the advantage that it stops with the scan as soon as a maximum measurement value has been determined. This could improve overall testtime as less movement of the turntable angle is possible. The Complete area mode has the advantage that the complete specified {{ScreenElement|Area}} is measured.
{{ScreenElementDescription|Threshold|This value determines the acceptance criteria for a peak.}}
{{ScreenElementDescription|Mode|The mode setting determines the way a peak is detected.}} The following modes are available:
*Intelligent
*Complete area
====Intelligent====
====Intelligent====
In this mode the software increases the angle of the turntable from the initial point with a step equal to {{ScreenElement|Step size}} as long as the measurement value is not more than {{ScreenElement|Threshold}} below the already known maximum measurement value. As soon as measurement value is detected that is {{ScreenElement|Threshold}} below the already known meaximum value, the turntable movement is reversed until the initial point and from this initial point the angle of the turntable is decreased with a step equal to {{ScreenElement|Step size}} until a measurement value is measured that is {{ScreenElement|Threshold}} below the already known maximum measurement value. The step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak.
{{Warning|This might cause a measurement error if the intial number of steps ({{ScreenElement|#}}) is set to a low value.}}
{{Note|This might cause a measurement error if the number of steps is set to a low value.}}
====Complete area====
In this mode the complete area around the initial point and defined by {{ScreenElement|Area}} is scanned for a peak. The turntable angle is moved to one end of the to be measured area, and is then moved with the specified {{ScreenElement|Step size}} to the other end of the to be measured area. On each step the measurement value is determined, and the step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak.
{{Warning|This might cause a measurement error if the {{ScreenElement|Area}} is set to a smaller value than the original stepsize (given by the stop position minus start position divided by the number of steps).}}
{{Note|This might cause a measurement error if the “Area” is set to a low value relative to the area given by the stop position minus start position dived by number of steps.}}
{{ScreenElementDescription|#|Number op locations from {{ScreenElement|Max. height}} to {{ScreenElement|Min. height}} that should be measured.}}
{{ScreenElementDescription|Peak search|The software will automatically detect on which antenna height the maximum measurement value of each peak is present.}}
Max. height : Sets maximum height of the antenna tower.
Min. height : Sets the minimum height of the antenna tower.
# : Sets the number of steps to go from min. height to max. height.
Peak search : The software will automatically detect on which antenna height maximum emission will occur.
{{ScreenElementDescription|Area|Range over which the peak search will be performed. The initially detected antenna height of the peak will be the middle of this {{ScreenElement|Area}}, and the total search area for this peak will be half of this area up and half of this area down.}}
{{ScreenElementDescription|Step size|The step size for each antenna height movement.}}
Area : Range over which the peak search will be performed (0 to n meter).
The Intelligent mode has the advantage that it stops with the scan as soon as a maximum measurement value has been determined. This could improve overall testtime as less movement of the antenna tower is possible. The Complete area mode has the advantage that the complete specified {{ScreenElement|Area}} is measured.
Step size : The distance for each step.
Threshold : This value determines the acceptance criteria for a peak
Mode : The mode setting determines the way a peak is detected. The following modes are available
- Intelligent
- Complete area
Intelligent
====Intelligent====
In this mode the software increases the height of the antenna from the initial point with a step equal to {{ScreenElement|Step size}} as long as the measurement value is not more than {{ScreenElement|Threshold}} below the already known maximum measurement value. As soon as measurement value is detected that is {{ScreenElement|Threshold}} below the already known meaximum value, the antenna movement is reversed until the initial point and from this initial point the distance of the clamp is decreased with a step equal to {{ScreenElement|Step size}} until a measurement value is measured that is {{ScreenElement|Threshold}} below the already known maximum measurement value. The step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak.
Remark: This might cause a measurement error if the number of steps is set to a low value.
Complete area
In this mode the complete area around the initial point and defined by {{ScreenElement|Area}} is scanned for a peak. The antenna is moved to one end of the to be measured area, and is then moved with the specified {{ScreenElement|Step size}} to the other end of the to be measured area. On each step the measurement value is determined, and the step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak.
Remark: This might cause a measurement error if the “Area” is set to a low value relative to the area given by the max. height minus min. height dived by number of steps.
Polarisation
Line 104: Line 446:
Distance
====Distance====
RadiMation uses the distance information for reference only.
Reference level : Sets the reference level of the receiver
{{ScreenElementDescription|Reference level|Sets the reference level of the receiver}}
Attenuator : Sets the receiver attenuator level
{{ScreenElementDescription|Attenuator|Sets the receiver attenuator level}}
RBW : Sets the Resolution BandWidth filter
{{ScreenElementDescription|RBW|Sets the Resolution BandWidth filter}}
VBW : Sets the Video BandWidth filter
{{ScreenElementDescription|VBW|Sets the Video BandWidth filter}}
Sweep : Sets the sweep time of the receiver
{{ScreenElementDescription|Sweep|Sets the sweep time of the receiver}}
Pre-amp : Sets the pre-amp gain
{{ScreenElementDescription|Pre-amp|Sets the pre-amp gain}}
Step frequency : Sets the frequency step size of the receiver
{{ScreenElementDescription|Step frequency|Sets the frequency step size of the receiver}}
Measure time : Sets the measure time of the receiver
{{ScreenElementDescription|Measure time|Sets the measure time of the receiver}}
However, using the 'Auto' setting for the attenuator can let the analyser decides to use a different attenuator setting during (re-)measuring of a peak, which as a result can lead to measuring a higher Quasi-Peak value than the Peak value. This is related to the used analyser when auto attenuator is used. If it is not desired that the analyser can decide to switch to a different attenuator setting, then this should be disabled by the test engineer in {{RadiMation}} by using a fixed attenuator value.
{{Warning| All these settings are generic. Therefore, always verify that the configured settings are also supported by the used hardware to ensure it is correctly used and no measurement faults are introduced. Use a fixed value to disabled interference of equipment intelligence.}}
===Max. hold measurements===
In this window the number op peak hold sweeps for each measurement can be selected. If this value is set to, for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweepspeed. For spectrum analysers and CW signals you can use a relatively low amount of  peaksweeps, for example 10. But when you have discontinues signals like a spark bridge, you may need 100 peaksweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope.
In this window the number of peak hold sweeps for each measurement can be selected. If this value is set to, for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweepspeed. For spectrum analysers and CW signals you can use a relatively low amount of  peaksweeps, for example 10. But when you have discontinues signals like a spark bridge, you may need 100 peaksweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope.
===Traces window===
===Traces window===
Line 129: Line 475:
===Detectors window===
{{ScreenElementDescription|All peaks above “xx” below the limit lines|In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 6 dB, the software will detect all peaks, which are higher than 6 dB below the limit line.}}
===Max. Peaks===
{{ScreenElementDescription|Max. Peaks|With the max. peaks option, the user can limit the number of peak signals the software will detect and re-measure.}}
Test equipment
Test engineer
In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results.
Equipment icon
{{ScreenElementDescription|Test engineer|In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results.}}
Line 162: Line 497:
===Test options window===
Record data
{{ScreenElementDescription|Record data|In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full tracebility, all test data should be stored.}}
{{ScreenElementDescription|Narrow band / broad band discrimination|{{RadiMation}} supports two types of narrow band / broad band determination.}}
The first one is called 'Generic'. During the determination of the peak, the so called ‘zooming’. {{RadiMation}} verifies if the peak is bigger or equal to twice the RBW filter. If so, the peak is determined broad band. Otherwise it is called narrow band.
The second method is called 'Mil Standard'. This method is also based on the RBW filter. But now the filter is changed to determine the type of peak. First the peak is measured with the current RBW, then it is made 10 times bigger. If multiplying the current RBW gives a RBW that is bigger then 1 MHz the current RBW is divided by 10. A delta RBW is calculated according to the formula below.
<math>Delta RBW = 10 * log(\frac{RBW_1}{RBW_2})</math>
Radimation supports two types of narrow band / broad band determination.
{{Formula|7|1|Difference in a peak measurement, as a result of changing the used RBW}}
The first one is called “Generic”. During the determination of the peak, the so called ‘zooming’. Radimation verifies if the peak is bigger or equal to twice the RBW filter. If so, the peak is determined broad band. Otherwise it is called narrow band.
The second method is called “Mil Standard”. This method is also based on the RBW filter. But now the filter is changed to determine the type of peak. First the peak is measured with the current RBW, then it is made 10 times bigger. If multiplying the current RBW gives a RBW that is bigger then 1 MHz the current RBW is divided by 10. A delta RBW is calculated according to the formula below.
Delta RBW = 10log(RBW1 / RBW2)
Delta V = V1 – V2
<math>DeltaV={V_1}\ -\ {V_2}</math>
{{Formula|7|2|Difference in a peak amplitude}}
Limits Distortion type
Delta V <= 5 Delta RBW Narrow band
!Limits || Distortion type
Delta V is > than 5 Delta RBW and delta V is <= 15 delta RBW Random noise
Delta V > 15 delta RBW Broad band
|Delta V <= 5 Delta RBW || Narrow band
|Delta V is > than 5 Delta RBW and delta V is <= 15 delta RBW || Random noise
|Delta V > 15 delta RBW || Broad band
===Starting a Anechoic Chamber test===
Line 200: Line 538:
[[Image:Radiated Emission Chamber Configuration Window.png]]
Line 205: Line 545:
====Test initialisation and self test====
Line 216: Line 556:
[[Image:Radiated Emission Chamber Initialization Window.png]]
===Test Procedure===
Line 237: Line 578:
#The receiver takes the desired sweeps of the spectrum.
Radimation now performs a maxhold on all the sweeps and determines the peaks.
{{RadiMation}} now performs a maxhold on all the sweeps and determines the peaks.
Then it determines the position a peak was found.
And changes the turntable and antenna to that current position to re-measure the peak.
Line 244: Line 585:
===Info screen===
===Info screen===
Line 252: Line 592:
A window like below here will appear.
Settings
[[Image:Radiated Emission Chamber Information Window.png]]
In the settings window all the used settings are displayed.
{{ScreenElementDescription|Settings|In the settings window all the used settings are displayed.}}
{{ScreenElementDescription|Graph control|In the graph control window the scale of the graph can be changed. You can show or hide the table by selecting or deselecting the show table checkbox.}}
Graph control
{{ScreenElementDescription|Trace|In the trace window you can control the traces that are shown in the graph by selecting or deselecting the checkbox. If during a measurement a trace is not used then the check box is grayed out.}}
{{ScreenElementDescription|Detector|The detector window gives you the possibility to show and hide the values measured on the peaks. If during a measurement a type of detector is not used for determining the value on the peak the checkbox is grayed out.}}
Trace
{{ScreenElementDescription|Table|In the table all the measured peaks are showed. The table also shows the limit values for that frequency point when a limit was given at the beginning of the test.}}
Detector
Table
===Test site information===
{|class="wikitable"
!# ||Device name || Tab in {{RadiMation}} configuration window
!colspan="3"|'''Cables'''
|1 || cable Preamp -> analyser || Cables
|2 || cable Antenna -> Preamp || Cables
Device name Tab in Radimation configuration window
{{Note|Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during emission measurements.
Spectrum analyser Standard 2
Pre amplifier Standard 2
Antenna Standard 1
Cables
# Device name Tab in Radimation configuration window
1 Pre amp->Spectrum Cables
2 Ant->Pre amp Cables
{{Note|Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuraccy during emission measurements.
Line 291: Line 629:
==Measurement of peak amplitude and peak frequency==
When using a spectrum analyser for emission measurements, {{RadiMation}} first performs a full span max. hold measurement with the configured trace detector. The number of max. hold peak sweeps can be defined by the user. From the obtained max hold graph, {{RadiMation}} determines a number of peaks, which will be further investigated. The number of peak signals to be measured can be defined by the user. For each peak from the graph, {{RadiMation}} will zoom in, by reducing the span in several steps. This must be done while span inaccuracy of the analyser in large frequency spans will give slightly incorrect frequency information. After the exact peak frequency is determined, {{RadiMation}} will perform the configured Peak,Quasi peak, Average and / or RMS measurements. Due to the above, the finally measured frequency and amplitude of a peak can slightly differ from the max hold graph. This is not an error, but merely an indication that the final measurements are more accurate. The peak table always shows the final measured values (most accurate).
== Fixed height test method==
== Fixed height ==
===Introduction===
===Introduction===
Line 299: Line 637:
This method decreases measurement time dramatically. A trade-off for this method is a slight decrease in measurement accuracy.
[[Image:Radiated Emission Fixed Height Configuration Window.png]]
Line 306: Line 645:
Start : Sets start frequency
{{ScreenElementDescription|Start|Sets start frequency}}
Stop : Sets stop frequency
{{ScreenElementDescription|Stop|Sets stop frequency}}
In the turn table window, the following parameters can be configured:
Start position : Sets the start angle of the turntable.
Stop position : Sets the stop angle of the turntable.
# : Sets the number of steps for one turntable rotation.
Peak search : The software will automatically detect on which angle maximum emission will occur.
Area : Range over which the peak search will be performed (0 to 360 degrees max.).
Step size : The angle step size.
Threshold : This value determines the acceptance criteria for a peak.
Mode : The mode setting determines the way a peak is detected. The following modes are available:
- Intelligent
- Complete area
Intelligent
====Intelligent====
Complete area
Min. Height : Sets the minimum height of the antenna tower.
====Intelligent====
- Intelligent
- Complete area
Intelligent
Complete area
Polarisation
Line 395: Line 734:
Distance
====Distance====
In this box, the operator can enter the distance between antenna and EUT. RadiMation uses the distance information for reference only.
In this box, the operator can enter the distance between antenna and EUT. {{RadiMation}} uses the distance information for reference only.
Line 422: Line 761:
Max. Peaks
With the ‘max. peaks” option, the user can limit the number of peak signals the software will detect and re-measure.
{{ScreenElementDescription|Max. Peaks|With the ‘max. peaks” option, the user can limit the number of peak signals the software will detect and re-measure.}}
Test equipment
Test engineer
Equipment icon
Line 483: Line 810:
[[Image:Radiated Emission Fixed Height Initialization Window.png]]
Line 500: Line 826:
#The receiver takes the desired sweeps of the Spectrum.
The turntable position is changes to that position to re-measure the peak.
Line 509: Line 835:
[[Image:Radiated Emission Test Setup Fixed Height.png]]
Antenna Standard 1
Cables
{|class="wikitable"
|  ||Pre amplifier || Devices2
|1 || cable amplifier ->Spectrum || Cables
|2 || cable antenna ->Preamp || Cables
measurements.
measurements.
Line 528: Line 860:
==G-TEM Method==
== GTEM ==
==Correlation method==
===Correlation method===
Line 538: Line 870:
Ptr2 = X2 + Y2 + Z2
<math>{{P_{tr}}^2} = X^2+Y^2+Z^2</math>
[[Image:Radiated Emission GTEM Configuration Window.png]]
{{RadiMation}} uses the above method to correlate the radiated emission measurements performed in a G-TEM to the emission levels on an OATS.
===Measurement procedure===
“Tests> Radiated emission > G-TEM”.
{{Menu|Tests|Radiated emission|G-TEM}}
Line 558: Line 892:
===Test configuration===
Frequency range selection
====Frequency range selection====
Line 568: Line 901:
Frequency list selection
====Frequency list selection====
OATS settings
====OATS settings====
Line 583: Line 916:
G-TEM settings
====G-TEM settings====
Receiver settings
====Receiver settings====
{{ScreenElementDescription|RBW|Sets the Resolution BandWidth filter.}}
{{ScreenElementDescription|VBW|Sets the Video BandWidth filter.}}
====Max. hold measurements====
====Traces window====
Traces window
Detectors window
====Detectors window====
Line 627: Line 957:
Test equipment
Test engineer
Equipment icon
Line 650: Line 971:
===Starting a G-TEM test===
Loading a TSF file
====Loading a TSF file====
To start a radiated emission test in the G-TEM cell, an EUT file must be defined first. From the EUT file, select “Tests > Radiated emission > G-TEM.
Line 663: Line 984:
Line 677: Line 998:
[[Image:Radiated Emission Test Setup GTEM.png]]
{|class="wikitable"
Antenna Standard 1
|  || Antenna || Devices 1
|1 ||cable Preamp->analyser || Cables
|2 || cable Antenna->Preamp || Cables
Cables
measurements.
measurements.
Line 700: Line 1,027:
==Manual mode==
==Manual mode (Singleband)==
For product improvements, it quit often is required to operate the test set-up in manual mode. This can be done either by operating the measurement receiver (or spectrum analyser) manually or by the {{RadiMation}} software. In our opinion it is no use to take over the buttons of the receiver without adding any extra functionality. For this reason a number of handsome futures are included in manual mode.
Line 709: Line 1,036:
The following parameters can be configured:
Reference : Sets the reference level of the receiver
{{ScreenElementDescription|Reference|Sets the reference level of the receiver}}
RBW : Sets the Resoution Band Width filter
{{ScreenElementDescription|RBW|Sets the Resoution Band Width filter}}
The figure on the next page shows the receiver settings menu.
[[Image:Radiated Emission Manual Mode Test.png]]
Data correction
====Data correction====
The semi- manual mode of {{RadiMation}} uses the PC screen to display the emission data real time on the computer screen. However the measured data is corrected for antenna factors and cable losses.
===Peak menu===
===Peak menu===
Search peak
{{ScreenElementDescription|Search peak|{{RadiMation}} can detect peaks automatically. From the “peaks” TAB, choose “Search peak” to detect the maximum signal. The software will measure the detected peak. }}
{{ScreenElementDescription|Add peak|The measured values can be added to the peak list by pressing the “Add Peak” button.}}
RadiMation can detect peaks automatically. From the “peaks” TAB, choose “Search peak” to detect the maximum signal. The software will measure the detected peak.
Add peak
{{ScreenElementDescription|Search angle|By pressing the “Search Angle” button, the software starts a turntable scan to determine at which turn table angle maximum emission takes place.}}
Search angle
{{ScreenElementDescription|Search Height|By pressing the “Search Height” button, the software starts an antenna height scan to determine at which antenna height maximum emission takes place.}}
Search Height
[[Image:Radiated Emission Manual Mode Test Peak Search.png]]
===Traces Menu===
===Traces Menu===
Traces
====Traces====
Line 758: Line 1,078:
[[Image:Radiated Emission Manual Mode Test Traces.png]]
====Detectors====
Detectors
- Peak detector
*Peak detector
- Average detector
*Average detector
- Quasi peak detector
*Quasi peak detector
- RMS detector
*RMS detector
Turntable and antenna tower position
====Turntable and antenna tower position====
====Zoom function====
Zoom function
{{RadiMation}} features a live zoom function. When the operator wants to “zoom in” on a certain signal just use the mouse to drag a box around the signal of interest. The {{RadiMation}} software will fully automatic zoom in to this signal. The screen, as well as the measurement receiver will be zoomed. This reduces the number of actions of the operator significantly.
Line 787: Line 1,102:
{{ScreenElementDescription|X-axis|When this option is switched on, the frequency settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph.}}
{{ScreenElementDescription|Y-axis|When this option is switched on, the amplitude settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph.}}
===Limit lines===
===Limit lines===
[[Image:Radiated Emission Manual Mode Test Limit Lines.png]]
===Export menu===
===Export menu===
[[Image:Radiated Emission Manual Mode Test Export.png]]
== Merge ==
==Merge function==
Line 825: Line 1,132:
[[Image:Radiated Emission Merge Configuration Window.png]]
== X + k * S==
==X+k*S Calculations==
==Anechoic chamber (Multi band)==
Pressing Add a new frequency band to the test, Remove will remove all selected bands and Duplicate will duplicate all selected bands.
Steps : Sets the number of steps for one turntable rotation.
EUT angle offset : The offset of the EUT compared to the zero angle of the table
Optimize Angle : Optimized settings for searching the correct angle.
The area is defined 50% left of the starting position and 50% right. So if the area is 30 degrees, RadiMation will scan between starting position plus and minus 15 degrees.
Steps : Sets the number of steps to go from min. height to max. height.
Optimize height : The software will automatically detect on which antenna height maximum  emission will occur.
Area : Range over which the peak search will be performed (0 to Max height).
Step size : The height step size.
The area is defined 50% above of the starting position and 50% below. So if the area is 1 meter, RadiMation will scan between starting position plus and minus 50 cm.
Polarisation
# Horizontal
# Vertical
# Both
Distance
RBW : Sets the Resolution Band Width filter
VBW : Sets the Video Band Width filter
===Traces window===
Number of sweeps
No peaks
Max. Peaks
Test equipment
Equipment icon
In some case you want to switch for example the antenna height scan before the turn table scan because this will decrease the total measuring time. With the change order the test engineer can control the way RadiMation® performs the testing sequence.  Read the list from top to bottom to see what changes first. The best way explain the change order window is by example.
*Band A RBW 9 kHz
*Band B RBW 9 kHz
*Band A RBW 120 kHz
*Band B RBW 120 kHz
*Band A RBW 9 kHz
*Band A RBW 120 kHz
*Band B RBW 9 kHz
*Band B RBW 120 kHz
*Band A RBW 9 kHz
*Band B RBW 9 kHz
*Band B RBW 120 kHz
*Band A RBW 120 kHz
*RBW
*Testsite
*Frequency band
*Band A RBW 9 kHz
*Band A RBW 120 kHz
*Band B RBW 120 kHz
*Band B RBW 9 kHz
Loading a TSF file
To start a radiated emission test, an EUT file must be defined first. From the EUT file, select “Tests > Radiated emission > Anechoic chamber.
# The presence of a valid software key is checked.
# The software settings are checked (i.e. is all test equipment used within its parameters?)
# The presence of all test equipment is checked.
# All relevant equipment will be initialised.
===Info screen===
{{Note|The pass/fail criteria, shown in the tables do not take the measure uncertainty of the test setup into account.}}
You can access the info screen by selecting the test you whish to see and then press the info button.
The window appearing looks a lot like the manual mode screen, the reason that it looks the same is because it is. You can continue testing after you finished a test, just select the test site you want to use and you can continue testing.
{{Note|Continuing the test is only possible with the selected test site. RadiMation will not select the correct test site for you. All measurements are done with the selected test site. If you want automatic changes of the test site, restart the configured TSF file.}}
Antenna Standard 1
Cables
measurements.
When in doubt contact your reseller.}}
Manual Mode (Multi band)
===Turn table===
===Antenna tower===
===Test site===
Before manual mode can be used the test site needs to be selected so that Radimation® knows which devices to use.
===Peaks window===
Take sweeps
Triggers the receiver of analyzer to take the configured amount of sweeps
Detect peaks
Measure Peaks
Delete Peaks
Peak actions
[[Category:Manual]]
Latest revision as of 12:36, 4 May 2021
Radiated emission[edit]
Loading a TSF file[edit]
   Menu.svg Tests
      Menu.svg Radiated Emission and then the desired test method.
  1. The presence of a valid software key is checked.
  2. The software settings are checked (i.e. is all test equipment used within its parameters?)
  3. The presence of all test equipment is checked.
  4. All relevant equipment will be initialised.
Test results screen[edit]
You can access the test results of the performed test by selecting the test in the EUT window and then press the Info button. For the Multiband test, the window that is appearing looks a lot like the manual mode screen, the reason that it looks the same is because it is. You can continue testing after you finished a test, just select the testsite you want to use and you can continue testing.
Note: The pass/fail criteria, shown in the tables do not take the measure uncertainty of the test setup into account.
Note: Continuing the test is only possible with the selected test site. RadiMation® will not select the correct test site for you. All measurements are done with the selected test site. If you want automatic changes of the test site, restart the configured TSF file.
Testsite information[edit]
An radiated emission test, requires that a testsite is used that can have the following components.
Spectrum analyser Devices 2
Pre amplifier Devices 2
Antenna Devices 1
Turn table Devices 1
Antennatower Devices 1
Cables
1 Cable preamp -> analyser Cables
2 Cable antenna -> preamp Cables
Note: Usage of cable correction factors is not mandatory. However, using cable correction factors will increase measurement accuracy during emission measurements. When in doubt contact your reseller.
Multiband[edit]
The radiated emission Multiband test, is a very flexible and powerful test to perform radiated emission tests. When the multiband test is performed, it will first perform a prescan measurement on all the specified locations. After this prescan is finished the temporary results are shown, and some peaks are automatically detected and shown. At this moment those peaks are NOT measured yet with the final measurement detectors and are also not optimized yet. The end-user now has the possibility to add additional peaks, or do some other peak investigation/changes. When the end-user is satisfied with the list of detected peaks, he can continue with the final measurement, by pressing the Automatic button again. When the end-user has configured peak optimization to be active, the peak optimization will now also be performed. Now the peaks will be measured with the detector that are selected in the Final peak measurement box. Only after those measurements are finalized, those detector values will become available in the list of detected peaks.
The test engineer can configure a complete radiated emission test in the Radiated Emission configuration menu. The test engineer can change the frequency range from one frequency band or the RBW of several bands. It all depends on the selection of bands. When several bands are selected and some options are blank, this means that these settings in bands are different. For example if the test engineer selects both frequency band in the window above, the frequency range will be blank. If the engineer sets the start and stop frequency, all the selected bands will have the same frequency setting. One of the advantages of this feature is the quick settings of several settings like the RBW or the peak detection.
Frequency window[edit]
Turntable window[edit]
ScreenElementDescription.svg Start Angle Sets the start position of the turntable.
ScreenElementDescription.svg End Angle Sets the stop position of the turntable.
ScreenElementDescription.svg Steps Sets the number of steps for one turntable rotation.
ScreenElementDescription.svg Turn during measurements This will force the turntable to continuously rotate during the test and will disable steps setting.
ScreenElementDescription.svg EUT angle offset The offset of the EUT compared to the zero angle of the table.
ScreenElementDescription.svg Optimize Angle Optimized settings for searching the correct angle.
ScreenElementDescription.svg Area Range over which the peak search will be performed (0 to 360 degrees max.).
ScreenElementDescription.svg Step size The angle step size.
Antenna tower window[edit]
ScreenElementDescription.svg Max. height Sets maximum height of the antenna tower.
ScreenElementDescription.svg Min. height Sets the minimum height of the antenna tower.
ScreenElementDescription.svg Steps Sets the number of steps to go from min. height to max. height.
ScreenElementDescription.svg Move during measurements This will force the Antenna Tower to continuously move during the test and will disable steps setting.
ScreenElementDescription.svg Optimize height The software will automatically detect on which antenna height maximum emission will occur.
ScreenElementDescription.svg Area Range over which the peak search will be performed (0 to Max height).
ScreenElementDescription.svg Step size The height step size.
Antenna window[edit]
Polarisation[edit]
  1. Horizontal
  2. Vertical
  3. Both
Distance[edit]
Receiver settings[edit]
ScreenElementDescription.svg Reference level Sets the reference level of the receiver.
ScreenElementDescription.svg Attenuator Sets the receiver attenuator level.
ScreenElementDescription.svg RBW Sets the Resolution Band Width filter.
ScreenElementDescription.svg VBW Sets the Video Band Width filter.
ScreenElementDescription.svg Sweep Sets the sweep time of the receiver.
ScreenElementDescription.svg Step frequency Sets the frequency step size of the receiver.
ScreenElementDescription.svg Measure time Sets the measure time of the receiver.
ScreenElementDescription.svg Pre-amp Sets the pre-amp gain.
Traces window[edit]
ScreenElementDescription.svg Detectors The detector settings (Peak, Average, Quasi-peak, RMS) used during sweeptime. This setting is generic. So it is possible to select multiple detectors, but only one detector can be selected due to the device limitations. In that case for each enabled detector the amount of sweeps will be performed by RadiMation®.
ScreenElementDescription.svg Number of sweeps The number op peak hold sweeps for each measurement can be selected. If this value is set to, for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweep speed. For spectrum analysers and CW signals you can use a relatively low amount of peak sweeps, for example 10. But when you have discontinues signals like a spark bridge, you may need 100 peak sweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope.
Final peak measurement[edit]
In the final peak measurement window the test engineer can configure the detectors used during the final emission measurements of the detected peaks. The first value is the time constant of the detector while the second value is used for the measure time at each frequency point for this detector. The time constant is send to the receiver, to set the sample time of the detector. To select a detector, check the desired button.
Peak detection window[edit]
ScreenElementDescription.svg No peaks When no final peak measurements are required select the option no peaks.
ScreenElementDescription.svg Maximum Peaks With the max. peaks option, the user can limit the number of peak signals the software will detect and re-measure.
Test site window[edit]
Limit lines window[edit]
ScreenElementDescription.svg All peaks above “xx” below the limit lines In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 5 dB, the software will detect all peaks, which are higher than 5 dB below the limit line.
Change Order window[edit]
You need to perform a test as shown but you have 2 test sites. One with an antenna with the frequency range that is equal to the total band. The second test site you have 2 antennas, one for every frequency band. For test site one you want the RBW filter to change as little as possible to prevent the old analyser from breaking down.
So for Site 1 you want:
  • Band A RBW 9 kHz
  • Band B RBW 9 kHz
  • Band A RBW 120 kHz
  • Band B RBW 120 kHz
And for test site 2:
  • Band A RBW 9 kHz
  • Band B RBW 9 kHz
  • Band A RBW 9 kHz
  • Band B RBW 9 kHz
  • RBW
  • Testsite
  • Frequency band
  • Band A RBW 9 kHz
  • Band B RBW 9 kHz
Note: If you have any questions regarding the setting of the change order please contact your reseller.
Manual Mode[edit]
Turn table[edit]
Antenna tower[edit]
Test site[edit]
Final measurement[edit]
Peaks window[edit]
ScreenElementDescription.svg Max Peaks Maximum allowed amount of peaks to be automatically determined.
ScreenElementDescription.svg Detect Peaks Detects the configured amount of sweeps in the measured spectrum.
ScreenElementDescription.svg Measure Peaks Measures all selected peaks with the configured settings in the final measurement window.
ScreenElementDescription.svg Peak Actions Allows several actions to be performed on the detected peaks, like select all and deselect all.
ScreenElementDescription.svg Peaks to maximum Start performing measurement to redetermine on which turntable angle and antenna tower height the highest emission value was measured.
This angle and height will be updated in the selected peak, together with the corresponding measured emission values. If multiple peaks are selected, this search for the highest emission value is done individually for every selected peak.
This 'peaks to maximum' function can be very helpful during the angle or height optimization of a peak. If after manual changes of the antenna height or turntable angle, no other maximum is found, the peak can be reset to the angle and height where the maximum was determined. 'Peaks to maximum', will thus select that location on which the highest emission value was measured.
ScreenElementDescription.svg Measure Peaks Start performing a final measurement on the selected peaks.
ScreenElementDescription.svg Optimize Angle Start performing a measurement as configured for the Optimize Angle.
ScreenElementDescription.svg Optimize Height Start performing a measurement as configured for the Optimize Height.
ScreenElementDescription.svg Go To Peak Location If possible it will adjust the Antenna Tower and Turntable to the location of the selected peak.
ScreenElementDescription.svg Set receiver to Peak settings Set the current receiver settings to the used settings of the selected peak.
ScreenElementDescription.svg Configure -> Optimize Angle Opens a new window to configure the Optimize Angle settings. (See also Optimize Angle settings.)
ScreenElementDescription.svg Configure -> Optimize Height Opens a new window to configure the Optimize Height settings. (See also Optimize height settings.)
ScreenElementDescription.svg Delete peaks Delete selected peak(s).
ScreenElementDescription.svg Delete Peaks Delete the selected peaks from the list.
ScreenElementDescription.svg Sweeps Amount of sweeps to be measured.
ScreenElementDescription.svg Take Sweeps Triggers the receiver of analyser to take the configured amount of sweeps.
ScreenElementDescription.svg Continuous Measure Start performing of taking measurements. This is a heavy process and it is normal when more CPU is used by the computer.
ScreenElementDescription.svg Continuous Measure Write Each new measurement will overwrite the previous measurement data.
ScreenElementDescription.svg Continuous Measure Max-hold Only higher new measurement values will overwrite the previous measurement data.
ScreenElementDescription.svg Clear measurement data Removes all previous measurement data.
Anechoic chamber[edit]
ScreenElementDescription.svg Start Sets start frequency
ScreenElementDescription.svg Stop Sets stop frequency
ScreenElementDescription.svg Start position Sets the start angle of the turntable.
ScreenElementDescription.svg Stop position Sets the stop angle of the turntable.
ScreenElementDescription.svg # Number of locations from Start postion to Stop position that should be measured.
ScreenElementDescription.svg Peak search The software will automatically detect on which turntable angle the maximum measurement value of each peak is present.
  • If a turntable device is specified in the selected test-site, RadiMation® will control the turntable to turn to the specified angle.
  • If no turntable device is specified in the selected test-site, RadiMation® will assume that the EUT is already positioned on that specified angle. No actions on the turntable will be taken.
ScreenElementDescription.svg Area Range over which the peak search will be performed. The initially detected turntable angle of the peak will be the middle of this Area, and the total search area for this peak will be half of this area clockwise and half of this area counter-clockwise.
ScreenElementDescription.svg Step size The step size for each turntable movement.
ScreenElementDescription.svg Threshold This value is only used during the Intelligent mode. It influences how long the movement is continued. The movement will stop when the measurement value is lower than the specified Threshold below the highest known measurement value of the searched peak.
ScreenElementDescription.svg Mode The mode setting determines the way a peak is detected. The following modes are available:
  • Intelligent
  • Complete area
The Intelligent mode has the advantage that it stops with the scan as soon as a maximum measurement value has been determined. This could improve overall testtime as less movement of the turntable angle is possible. The Complete area mode has the advantage that the complete specified Area is measured.
Intelligent[edit]
In this mode the software increases the angle of the turntable from the initial point with a step equal to Step size as long as the measurement value is not more than Threshold below the already known maximum measurement value. As soon as measurement value is detected that is Threshold below the already known meaximum value, the turntable movement is reversed until the initial point and from this initial point the angle of the turntable is decreased with a step equal to Step size until a measurement value is measured that is Threshold below the already known maximum measurement value. The step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak.
Warning: This might cause a measurement error if the intial number of steps (#) is set to a low value.
Complete area[edit]
Warning: This might cause a measurement error if the Area is set to a smaller value than the original stepsize (given by the stop position minus start position divided by the number of steps).
ScreenElementDescription.svg # Number op locations from Max. height to Min. height that should be measured.
ScreenElementDescription.svg Peak search The software will automatically detect on which antenna height the maximum measurement value of each peak is present.
ScreenElementDescription.svg Area Range over which the peak search will be performed. The initially detected antenna height of the peak will be the middle of this Area, and the total search area for this peak will be half of this area up and half of this area down.
ScreenElementDescription.svg Step size The step size for each antenna height movement.
  • Intelligent
  • Complete area
The Intelligent mode has the advantage that it stops with the scan as soon as a maximum measurement value has been determined. This could improve overall testtime as less movement of the antenna tower is possible. The Complete area mode has the advantage that the complete specified Area is measured.
Intelligent[edit]
In this mode the software increases the height of the antenna from the initial point with a step equal to Step size as long as the measurement value is not more than Threshold below the already known maximum measurement value. As soon as measurement value is detected that is Threshold below the already known meaximum value, the antenna movement is reversed until the initial point and from this initial point the distance of the clamp is decreased with a step equal to Step size until a measurement value is measured that is Threshold below the already known maximum measurement value. The step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak.
Complete area[edit]
Polarisation[edit]
  1. Horizontal
  2. Vertical
  3. Both
Distance[edit]
ScreenElementDescription.svg RBW Sets the Resolution BandWidth filter
ScreenElementDescription.svg VBW Sets the Video BandWidth filter
Warning: All these settings are generic. Therefore, always verify that the configured settings are also supported by the used hardware to ensure it is correctly used and no measurement faults are introduced. Use a fixed value to disabled interference of equipment intelligence.
Max. hold measurements[edit]
Traces window[edit]
Detectors window[edit]
ScreenElementDescription.svg All peaks above “xx” below the limit lines In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 6 dB, the software will detect all peaks, which are higher than 6 dB below the limit line.
ScreenElementDescription.svg Max. Peaks With the max. peaks option, the user can limit the number of peak signals the software will detect and re-measure.
Test options window[edit]
ScreenElementDescription.svg Record data In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full tracebility, all test data should be stored.
ScreenElementDescription.svg Narrow band / broad band discrimination RadiMation® supports two types of narrow band / broad band determination.
Formula 7-1: Difference in a peak measurement, as a result of changing the used RBW
Formula 7-2: Difference in a peak amplitude
Loading a TSF file
Test Procedure[edit]
  1. The receiver takes the desired sweeps of the spectrum.
  2. The antenna is moved to 4 meters.
  3. The receiver takes the desired sweeps of the spectrum.
  4. The turntable is moved to 45 degrees.
  5. The receiver takes the desired sweeps of the spectrum.
  6. The antenna tower is moved to 1 meter.
  7. The receiver takes the desired sweeps of the spectrum.
  8. The turntable is moved to 90 degrees.
  9. The receiver takes the desired sweeps of the spectrum.
  10. The antenna tower is moved to 4 meters
  11. The receiver takes the desired sweeps of the spectrum.
Info screen[edit]
ScreenElementDescription.svg Settings In the settings window all the used settings are displayed.
ScreenElementDescription.svg Graph control In the graph control window the scale of the graph can be changed. You can show or hide the table by selecting or deselecting the show table checkbox.
ScreenElementDescription.svg Trace In the trace window you can control the traces that are shown in the graph by selecting or deselecting the checkbox. If during a measurement a trace is not used then the check box is grayed out.
ScreenElementDescription.svg Detector The detector window gives you the possibility to show and hide the values measured on the peaks. If during a measurement a type of detector is not used for determining the value on the peak the checkbox is grayed out.
ScreenElementDescription.svg Table In the table all the measured peaks are showed. The table also shows the limit values for that frequency point when a limit was given at the beginning of the test.
Test site information[edit]
Antenna Devices 1
Cables
1 cable Preamp -> analyser Cables
2 cable Antenna -> Preamp Cables
Note: Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during emission measurements.
Fixed height[edit]
  • Intelligent
  • Complete area
Intelligent[edit]
Complete area[edit]
  • Intelligent
  • Complete area
Intelligent[edit]
Complete area[edit]
Polarisation[edit]
  1. Horizontal
  2. Vertical
  3. Both
Distance[edit]
Traces window[edit]
ScreenElementDescription.svg Max. Peaks With the ‘max. peaks” option, the user can limit the number of peak signals the software will detect and re-measure.
Loading a TSF file
  1. The receiver takes the desired sweeps of the Spectrum.
  2. The turntable is moved to 45 degrees.
  3. The receiver takes the desired sweeps of the Spectrum.
  4. The turntable is moved to 90 degrees.
  5. The receiver takes the desired sweeps of the Spectrum.
Pre amplifier Devices2
Antenna Devices 1
Cables
1 cable amplifier ->Spectrum Cables
2 cable antenna ->Preamp Cables
measurements.
GTEM[edit]
Correlation method[edit]
Measurement procedure[edit]
   Menu.svg Tests
      Menu.svg Radiated emission
         Menu.svg G-TEM
When a new test has to be configured please press the “Cancel” button. A blank test configuration screen will appear. The chapter below will explain all settings for this test configuration screen.
Test configuration[edit]
OATS settings[edit]
G-TEM settings[edit]
ScreenElementDescription.svg RBW Sets the Resolution BandWidth filter.
ScreenElementDescription.svg VBW Sets the Video BandWidth filter.
Traces window[edit]
Max. Peaks
Starting a G-TEM test[edit]
Antenna Devices 1
Cables
1 cable Preamp->analyser Cables
2 cable Antenna->Preamp Cables
measurements.
ScreenElementDescription.svg Reference Sets the reference level of the receiver
ScreenElementDescription.svg RBW Sets the Resoution Band Width filter
Data correction[edit]
Peak menu[edit]
ScreenElementDescription.svg Search peak RadiMation® can detect peaks automatically. From the “peaks” TAB, choose “Search peak” to detect the maximum signal. The software will measure the detected peak.
ScreenElementDescription.svg Add peak The measured values can be added to the peak list by pressing the “Add Peak” button.
ScreenElementDescription.svg Search angle By pressing the “Search Angle” button, the software starts a turntable scan to determine at which turn table angle maximum emission takes place.
ScreenElementDescription.svg Search Height By pressing the “Search Height” button, the software starts an antenna height scan to determine at which antenna height maximum emission takes place.
Traces Menu[edit]
Traces[edit]
Detectors[edit]
  • Peak detector
  • Average detector
  • Quasi peak detector
  • RMS detector
Zoom function[edit]
ScreenElementDescription.svg X-axis When this option is switched on, the frequency settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph.
ScreenElementDescription.svg Y-axis When this option is switched on, the amplitude settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph.
Limit lines[edit]
Export menu[edit]
Merge[edit]
X + k * S[edit]
