Difference between revisions of "RadiMation Application Note 154"
(→Create calibration test setup files) |
(→Calibration procedure) |
Line 154: | Line 154: | ||
+ | === Calibration result === | ||
When the three calibrations are performed there should be three calibration files created. | When the three calibrations are performed there should be three calibration files created. | ||
Revision as of 15:40, 16 October 2023
How to perform a MIL-STD CS114, Conducted susceptibility, bulk cable injection test
This application note explains how the MIL-STD CS114, Conducted susceptibility, bulk cable injection test can be performed with RadiMation®
Necessary equipment
- Signal generator
- Amplifier
- Coupler
- Forward power meter
- Reflected power meter (optionally)
- Current sensor
- 50 Ohm load
- 50 Ohm attenuator
- LISN
- Current injection device
Configuration of the test sites
Calibration
Device name | Tab in testsite configuration window | note |
Signal generator | Devices 1 | The signal generator to use |
Amplifier | Devices 1 | The amplifier to use |
Coupler | Devices 1 | The coupler to use |
Forward power meter | Devices 1 | The forward power meter to use |
Reflected power meter | Devices 1 | The reflected power meter to use (optionally) |
Sensor power meter | Devices 2 | The power meter or analyser to use for measuring the current |
Current sensor | Devices 2 | The current sensor to use with transfer factor attached to the driver |
Injection device | Devices 2 | The injection clamp to use |
Jig | Devices 2 | The jig to use |
Cable current -> power meter | Cables | Cable (1) with a correction file specified for the cable loss |
Cable current -> power meter | Cables | Cable (2) with the specified loss of the used attenuator |
Cable current -> power meter | Cables | Cable (3) with a correction file specified for the cable loss |
Verification
EUT Testing
Create correction file for the test level
First create a correction file in RadiMation® and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.'). In this example the limits for 'Aircraft (External or safety critcal) / Curve 5 are configured.
Make sure to remove the Correction column and add the Current column with the unit set to dBuV.
Band | Frequency range | Step size |
Band 1 | 10 kHz - 1 MHz | 5 % |
Band 2 | 1 MHz - 30 MHz | 1 % |
Band 2 | 30 MHz - 200 MHz | 0.5 % |
Create three test setup files, this example shows the settings for band 1.
Start | The start frequency of the test. For example 10 kHz. |
End | The stop frequency of the test. For example 1 MHz. |
Step | The frequency step, in this case 5%. |
Test level | The test level, in this case 5%. |
Variable | The variable test level, select the correction file created earlier. |
Tolerance | The regulation tolerance. |
Calibration method | Forward power. |
Test equipment | Calibration test equipment. |
Test engineer | The engineer that performed the calibration. |
Calibration result
