RadiMation Application Note 157: Difference between revisions
Line 34: Line 34:
| || Signal Generator || Devices 1 || The signal generator to use
| || Antenna || Devices 1 || The antenna to use
| || Amplifier || Devices 1 || The amplifier to use
Revision as of 08:58, 5 December 2023
How to perform a MIL-STD-461 RS103, radiated susceptibility, electric field[edit]
This application note explains how the MIL-STD-461 RS103, radiated susceptibility, electric field test can be performed with RadiMation®
The exact requirements and test methods for the RS103 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to perform the CS109 test:
  • Signal generator
  • Amplifier
  • Antenna
  • Field probe
Define the test level[edit]
Create a correction file for the test level[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg Correction
First create a correction file in RadiMation® and specify the test level (See 'Table XI RS103 limits.').
   Menu.svg File
      Menu.svg Save correction
This correction file with the test level will be used during EUT testing.
EUT Testing[edit]
EUT Testing equipment[edit]
Antenna Devices 1 The antenna to use
Field Sensor 1 Field probes The field sensor to use
Configure the EUT test[edit]
To perform the test on the EUT, create a radiated immunity multiband test.
Create a test setup configuration file for the following frequency ranges:
Frequency range Frequency step Polarization
2 MHz to 30 MHz 1% Vertical
30 MHz - 18 GHz 0.5% Both (Horizontal and vertical)
18 GHz - 40 GHz 0.25% Both (Horizontal and vertical)
The radiated immunity multiband test, can be configured by selecting from the menu:
   Menu.svg Tests
      Menu.svg Radiated Immunity
         Menu.svg Multiband
ScreenElementDescription.svg Start The start frequency of the test. For example 2 MHz.
ScreenElementDescription.svg End The stop frequency of the test. For example 30 MHz.
ScreenElementDescription.svg Dwell Time The dwell time.
ScreenElementDescription.svg Frequency Change Mode Constant.
ScreenElementDescription.svg Modulation The modulation to use see below.
ScreenElementDescription.svg Test equipment The equipment needed for the radiated susceptibility test.
ScreenElementDescription.svg Location Type Antenna height, distance, polarization and angle.
ScreenElementDescription.svg Test Level Specify the Electrical field closed loop method test level. Also configured a "Variable field"
At TestLevel click Add to add a new Test level and select TestLevel - Electrical field closed loop method
ScreenElementDescription.svg Test level Variable field, and specify the correction file for the applicable test level. When the level is one value this value can be specified with Constant field
ScreenElementDescription.svg Probes Check Probe 1 to measure with the first probe.
To configure the pulse modulation settings click Config next to the modulation settings, click the PM tab, and enter the pulse modulation settings as defined in 5.21.3.4.c.1.a RS103 Modulation settings.png
ScreenElementDescription.svg Activate modulation Check this checkbox to enable the pulse modulation.
ScreenElementDescription.svg Frequency The PM frequency 1 kHz.
ScreenElementDescription.svg Duty Cycle A duty cycle of 50%.
ScreenElementDescription.svg Also apply modulation out of the dwell-time period Check this checkbox to enable modulation outside the dwell-time.
When all settings are configured press Start Test to run the test.
