Tag: Manual revert
(7 intermediate revisions by 2 users not shown)
Line 33: Line 33:
=== Configuration of the loop sensor antenna ===
The loop antenna needs to have the correction data applied, which is normally known by the manufacturer. The correction data is expressed in dBpT/µV.
The loop antenna needs to have the correction data applied, which is normally provided by the manufacturer, or a calibration institute. The correction data is expressed in dBpT/µV.
In {{RadiMation}} this can be configured by creating a correction file, which contains a {{ScreenElement|Frequency}} and a {{ScreenElement|Correction}} column. The frequencies and the correction values of the actual correction of the loop antenna can then be specified in the correction file. The values in the {{ScreenElement|Correction}} column in the correction should be expressed in dBpT/µV.
==== Convert k-factor from dB(S/m) or dB(/Ohm m) to dBpT/µV ====
==== Convert k-factor from dB(S/m) or dB/(Ohm m) to dBpT/µV ====
Some calibration certificates of loop antennas contain the magnetic field factor of the loop antenna as a k-factor with a dB(S/m) unit (which is the same as a dB(/Ohm m) unit). {{RadiMation}} expects however expect that the values in the correction file are expressed in the dBpt/µV unit.
Some calibration certificates of loop antennas contain the magnetic field factor of the loop antenna as a k-factor with a dB(S/m) unit or a dB/(Ohm m) unit (which are exactly the same to each other). {{RadiMation}} expects that the values in the correction file are expressed in the dBpt/µV unit.
The conversion is: 'MagneticFieldFactor [dBpT/µV] = k [dB(S/m)] + 1.984'. The value 1.984 is often simplified to be 2.0, but 1.984 is the more accurate value.
The conversion from dB(S/m) is: 'MagneticFieldFactor [dBpT/µV] = k [dB(S/m)] + 1.984'.
The conversion for dB/Ohm/m is: 'MagneticFieldFactor [dBpT/µV] = dB/Ohm/m + 1.984'.
The conversion from dB/(Ohm m) is: 'MagneticFieldFactor [dBpT/µV] = k [dB/(Ohm m)] + 1.984'.
{{Note|1=The relation between dBpT and A/m is: Magnetic Flux Density [dBpT] = Magnetic field [dBµA/m] + 1.984
Line 57: Line 59:
Also see: https://www.tf.uni-kiel.de/matwis/amat/elmat_en/kap_4/backbone/r4_1_1.html
{{Note|1=As {{RadiMation}} only has a correction file for the magnetic field factor that contains the frequency and actual factor value, it is not known for {{RadiMation}} at which distance this magnetic field factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct magnetic field factor is specified, corresponding to the antenna distance that is actually used during the measurement.
If you would like to convert the magnetic field antenna factor from one distance to another distance, it is possible to use a 18 dB correction if the distance is doubled.
An example correction file for this 'Magnetic Field factor' thus looks like:
Line 88: Line 97:
[[File:RE101_Test_configuration.png]]
{{ScreenElementDescription|Start|The start frequency of the test. For example 30 kHz.}}
Line 100: Line 108:
{{ScreenElementDescription|Traces|The type of trace set in the receiver. For MIL-STD-461 RE101 only a peak detector is used.}}
{{ScreenElementDescription|Test Equipment|The equipment used for the measurement.}}
{{ScreenElementDescription|Location type|The type of measurement that should be used. Set this to {{ScreenElement|Magnetic Field antenna}}.}}
{{ScreenElementDescription|Location type|The type of measurement that should be used. Set this to {{ScreenElement|Magnetic Field antenna}}, as this will present the final result in dBpT or dBuA/m (depending on these selected unit).}}
{{ScreenElementDescription|Max Height|The maximum height for the antenna tower. This configuration is depending on how many positions should be measured.}}
{{ScreenElementDescription|Min Height|The minimal height for the antenna tower. This configuration is depending on how many positions should be measured.}}
Latest revision as of 09:56, 15 April 2025
How to perform a MIL-STD-461 RE101 Radiated Emission test[edit]
Necessary equipment[edit]
  • Receiver
  • Loop sensor (Antenna)
  • LISN
  • Cable
Configuration of the testsite[edit]
Antenna Devices 1 Loop sensor antenna for the required frequency range
Cable preamp -> analyser Cables Cable with a correction file specified for the loss
Note: The LISN is only used to decouple the EUT and it is not needed to specify and configure this in RadiMation®
Configuration of the loop sensor antenna[edit]
In RadiMation® this can be configured by creating a correction file, which contains a Frequency and a Correction column. The frequencies and the correction values of the actual correction of the loop antenna can then be specified in the correction file. The values in the Correction column in the correction should be expressed in dBpT/µV.
Convert k-factor from dB(S/m) or dB/(Ohm m) to dBpT/µV[edit]
Some calibration certificates of loop antennas contain the magnetic field factor of the loop antenna as a k-factor with a dB(S/m) unit or a dB/(Ohm m) unit (which are exactly the same to each other). RadiMation® expects that the values in the correction file are expressed in the dBpt/µV unit.
Note: The relation between dBpT and A/m is: Magnetic Flux Density [dBpT] = Magnetic field [dBµA/m] + 1.984
In this equation, the constant 1.984 is derived by using: The magnetic flux density (B) in Tesla (T), is related to the magnetic field strength (H) in A/m, by the permeability of the medium (Failed to parse (SVG (MathML can be enabled via browser plugin): Invalid response ("Math extension cannot connect to Restbase.") from server "https://wikimedia.org/api/rest_v1/":): {\displaystyle \mu_0} ) in Henry's per meter (H/m). For free space, the permeability is given as:
Failed to parse (SVG (MathML can be enabled via browser plugin): Invalid response ("Math extension cannot connect to Restbase.") from server "https://wikimedia.org/api/rest_v1/":): {\displaystyle \mu_0 [H/m] = 4 \pi \cdot 10^{-7}}
Failed to parse (SVG (MathML can be enabled via browser plugin): Invalid response ("Math extension cannot connect to Restbase.") from server "https://wikimedia.org/api/rest_v1/":): {\displaystyle B [T] = \mu_0 * H [A/m]}
Converting from T to pT and from A/m to µA/m, and converting it to decibels, using the Log, the constant becomes:
Failed to parse (SVG (MathML can be enabled via browser plugin): Invalid response ("Math extension cannot connect to Restbase.") from server "https://wikimedia.org/api/rest_v1/":): {\displaystyle 240 - 120 + 20 log^{10}(4 \pi \cdot 10^{-7}) = 1.984} (to be exact: 1.9841972804419248955749215547977)
Note: As RadiMation® only has a correction file for the magnetic field factor that contains the frequency and actual factor value, it is not known for RadiMation® at which distance this magnetic field factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct magnetic field factor is specified, corresponding to the antenna distance that is actually used during the measurement.
Once this correction is stored to disk, this correction file can be selected in the loop antenna driver configuration as 'Magnetic Field Factor':
Perform the test[edit]
   Menu.svg Tests
      Menu.svg Radiated Emission
         Menu.svg Multiband
Band Frequency range RBW Step size Measure time receiver Measure time FFT receiver
Band 1 30 Hz - 1 kHz 10 Hz 5 Hz 0.15 s 1 s
Band 2 1 kHz - 10 kHz 100 Hz 50 Hz 0.015 s 1 s
Band 3 10 kHz - 150 kHz 1 kHz 500 Hz 0.015 s 1 s
In the following screenshot a configuration is shown to perform the measurement on an EUT, applicable for the Navy Application.
The Location type setting for every band should be set to Magnetic Field antenna, as that determines that the measurement result will be shown as magnetic field values.
ScreenElementDescription.svg Start The start frequency of the test. For example 30 kHz.
ScreenElementDescription.svg Stop The stop frequency of the test. For example 100 kHz.
ScreenElementDescription.svg Reference Level The reference level set in an analyser. This setting is not used when a receiver is used.
ScreenElementDescription.svg Attenuation The attenuation set in the receiver.
ScreenElementDescription.svg RBW The RBW set in the receiver. See the table with different bands for the specific values for the RE101 measurement.
ScreenElementDescription.svg Step size The step size set in the receiver. See the table with different bands for the specific values for the RE101 measurement.
ScreenElementDescription.svg Measure time The measure time set in the receiver. See the table with different bands for the specific values for the RE101 measurement.
ScreenElementDescription.svg Preamplifier The preamplifier setting set in the receiver.
ScreenElementDescription.svg Traces The type of trace set in the receiver. For MIL-STD-461 RE101 only a peak detector is used.
ScreenElementDescription.svg Test Equipment The equipment used for the measurement.
ScreenElementDescription.svg Location type The type of measurement that should be used. Set this to Magnetic Field antenna, as this will present the final result in dBpT or dBuA/m (depending on these selected unit).
ScreenElementDescription.svg Max Height The maximum height for the antenna tower. This configuration is depending on how many positions should be measured.
ScreenElementDescription.svg Min Height The minimal height for the antenna tower. This configuration is depending on how many positions should be measured.
ScreenElementDescription.svg Steps The number of height positions to measure. This configuration is depending on how many positions should be measured.
ScreenElementDescription.svg Move during measurement If the antenna tower can move during the measurement. This is normally not used during MIL-STD-461 RE101 measurements, and can thus be unchecked.
ScreenElementDescription.svg Optimize Height Find the optimal emission around a certain height. This is normally not used during MIL-STD-461 RE101 measurements, and can thus be unchecked.
ScreenElementDescription.svg Distance The distance between the antenna and the EUT. This is normally 7cm for a MIL-STD-461 RE101 measurement.
ScreenElementDescription.svg Antenna Polarization Which polarization is used, Vertical, Horizontal or Both.
ScreenElementDescription.svg Start Angle The minimal turn table angle.
ScreenElementDescription.svg End Angle The maximum turn table angle.
ScreenElementDescription.svg Steps The number of angles to measure.
ScreenElementDescription.svg Turn during measurement If the turn table can move during the measurement. This is normally not used during MIL-STD-461 RE101 measurements, and can thus be unchecked.
ScreenElementDescription.svg Optimize Angle Find the optimal emission around a certain angle. This is normally not used during MIL-STD-461 RE101 measurements, and can thus be unchecked.
ScreenElementDescription.svg EUT Angle Offset The offset of the EUT on the table. This is normally not used during MIL-STD-461 RE101 measurements, and can thus be set to 0 degrees.
ScreenElementDescription.svg Limit Lines The applicable limit line can be added, in this example the "MIL-STD-461 RE101-2 Navy Applications" limit line file.
ScreenElementDescription.svg All peaks above x dB below the limit line Which level above x dB below the limit line the peaks should be detected.
ScreenElementDescription.svg Peak detection How many peaks should be detected automatically, for example, use 10 peaks to be detected
ScreenElementDescription.svg Final peak measurement It is possible to perform a final measurement on a peak with the selected detectors. Normally no final measurement is needed for the MIl-STD-461 RE101 measurement, and all detectors can thus be unselected.
When all parameters are configured for the RE101 test, press Start Test to perform the actual emission measurement.
Review the initial scan and measure frequencies for maximum emission[edit]
With the the frequencies of maximum emission located, RadiMation® will detect any peaks above the limit line, but it is also possible to add additional peaks by clicking in the graph. A list of peaks with frequencies should appear below in the Peaks tab. As described in section 5.17.3.4c(4) and 5.17.3.4c(5) it is needed to find the maximum radiation, the following steps describe how to perform this within RadiMation.
To measure one single frequency and find the maximum radiation:
  1. Select one frequency by checking one single frequency in the peaks list.
  2. Click Peak Actions and select Set receiver to Peak settings to set the correct measure time and RBW as configured.
  3. Enable Continuous Measure and move the loop antenna to find the maximum radiation.
  4. Click Continuous Measure again to stop the measurement.
  5. Repeat this step for each frequency.
The MIL-STD-461 RE101 test is using different receiver settings depending on the frequency band to cover the measurement of the magnetic field for the complete frequency range. All these settings can be easily configured in a multiband "Test Setup File" and tested with RadiMation®. Also the optimization of the maximum emission can be performed, and all this information will be present in the peak table in the test results.