== Overview == | == Overview == | ||
There are three possible setups of the used equipment to measure amplifier harmonics. Each | There are three possible setups of the used equipment to measure amplifier harmonics. Each method has its own advantages, and small differences in the used equipment. These setups and methods are explained in the following sections: | ||
# Measure the power of all harmonics using a spectrum analyser. | # [[#Measure the power of all harmonics using a spectrum analyser|Measure the power of all harmonics using a spectrum analyser]]. | ||
# Measure the broadband power with a powermeter and the harmonics using a spectrum analyser. | # [[#Measure the broadband power with a powermeter and the harmonics using a spectrum analyser|Measure the broadband power with a powermeter and the harmonics using a spectrum analyser]]. | ||
# Measure the actual field harmonics using a receive antenna. | # [[#Measure the actual field harmonics using a receive antenna|Measure the actual field harmonics using a receive antenna]]. | ||
Which equipment setup is used, is depending on the equipment that is selected in the testsite that is being used for the harmonics calibration. | |||
* If the selected testsite has a device driver selected for the {{ScreenElement|Sensor power meter}}, then the '[[#Measure the broadband power with a powermeter and the harmonics using a spectrum analyser|Measure the broadband power with a powermeter and the harmonics using a spectrum analyser]]' method will be used. | |||
* Otherwise, if the selected testsite has a device driver selected for the {{ScreenElement|Spectrum analyser}}, then the '[[#Measure the actual field harmonics using a receive antenna|Measure the actual field harmonics using a receive antenna]]' method will be used. | |||
* In all other cases the '[[#Measure the power of all harmonics using a spectrum analyser|Measure the power of all harmonics using a spectrum analyser]]' method will be used. | |||
=== Measure the power of all harmonics using a spectrum analyser === | === Measure the power of all harmonics using a spectrum analyser === | ||
Line 75: | Line 80: | ||
| || Sensor power meter || Devices 2 || yes || The powermeter device driver for the analyser that is being used to measure the harmonics. | | || Sensor power meter || Devices 2 || yes || The powermeter device driver for the analyser that is being used to measure the harmonics. | ||
!colspan="5"|Cables | |||
|① || Cable SG -> amplifier || Cables || no || Optional to correct for the cable loss between the signal generator and the amplifier | |||
|② || Cable amplifier -> coupler || Cables || no || Optional to correct for the cable loss between the amplifier and the coupler | |||
|③ || Cable coupler -> forward power meter || Cables || no || Optional to correct for the cable loss between the coupler and the forward power meter | |||
|④ || Cable current -> power meter || Cables || yes || Optional to correct for the cable loss between the coupler and the spectrum analyser (which is the sensor power meter) | |||
|⑤ || Cable current -> power meter || Cables || yes || Optional to correct for the attenuator between the coupler and the spectrum analyser (which is the sensor power meter) | |||
|⑥ || Cable current -> power meter || Cables || yes || Optional to correct for the cable loss between the coupler and the spectrum analyser (which is the sensor power meter) | |||
The signal path between the coupler and the spectrum analyser often consists of multiple cable segments. Each cable segment can have its own correction, and multiple corrections can be selected and added to the testsite configuration. The combined correction of all the selected cable segments will be used for the correction. | |||
=== Measure the actual field harmonics using a receive antenna === | === Measure the actual field harmonics using a receive antenna === | ||
Line 94: | Line 113: | ||
! Device name || Tab in testsite configuration window || suitable for harmonic frequencies || note | !# || Device name || Tab in testsite configuration window || suitable for harmonic frequencies || note | ||
| Signal Generator || Devices 1 || no | | || Signal Generator || Devices 1 || no | ||
| Amplifier || Devices 1 || no || This is the amplifier of which the harmonics are being measured. | | || Amplifier || Devices 1 || no || This is the amplifier of which the harmonics are being measured. | ||
| Coupler || Devices 1 || no | | || Coupler || Devices 1 || no | ||
| Forward power meter || Devices 1 || no || The powermeter device driver for the broadband power measurement of the forward power. | | || Forward power meter || Devices 1 || no || The powermeter device driver for the broadband power measurement of the forward power. | ||
| Calibration antenna || Devices 1 || yes || The antenna that is used for the receiving of the generated electrical fields on the harmonics. | | || Calibration antenna || Devices 1 || yes || The antenna that is used for the receiving of the generated electrical fields on the harmonics. | ||
| || Spectrum analyser || Devices 2 || yes || The device driver for the analyser that is being used to measure the harmonics. | |||
|④ || Cable preamp -> analyser || Cables || yes || Optional to correct for the attenuator between the receive antenna and the spectrum analyser | |||
|⑤ || Cable preamp -> analyser || Cables || yes || Optional to correct for the cable loss between the receive antenna and the spectrum analyser | |||
The signal path between the receive antenna and the spectrum analyser often consists of multiple cable segments. Each cable segment can have its own correction, and multiple corrections can be selected and added to the testsite configuration. The combined correction of all the selected cable segments will be used for the correction. | |||
== Configuring the calibration == | == Configuring the calibration == | ||
Line 128: | Line 161: | ||
{{ScreenElementDescription|Type of harmonic|If only the even, odd or both (even and odd) harmonics should be measured.}} | {{ScreenElementDescription|Type of harmonic|If only the even, odd or both (even and odd) harmonics should be measured.}} | ||
{{ScreenElementDescription|Calibration method|On which power levelling point the desired testlevel should be regulated.}} | {{ScreenElementDescription|Calibration method|On which power levelling point the desired testlevel should be regulated.}} | ||
{{ScreenElementDescription|Test Equipment|In the test site window the test engineer can select which equipment list will be used during this test.}} | {{ScreenElementDescription|Test Equipment|In the test site window the test engineer can select which equipment list will be used during this test. This also determines which [[#Overview|method and equipment]] is being used.}} | ||
{{ScreenElementDescription|Test engineer|In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results}} | {{ScreenElementDescription|Test engineer|In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results}} | ||
{{ScreenElementDescription|Amount of samples|The amount of measurements that are measured on each harmonic to determine its power. The average of these measurements will be used as the power of the harmonic.}} | {{ScreenElementDescription|Amount of samples|The amount of measurements that are measured on each harmonic to determine its power. The average of these measurements will be used as the power of the harmonic.}} | ||
Line 140: | Line 173: | ||
{{ScreenElementDescription|Reporting|}} | {{ScreenElementDescription|Reporting|}} | ||
{{ScreenElementDescription|Photos|}} | {{ScreenElementDescription|Photos|}} | ||
{{ScreenElementDescription| | {{ScreenElementDescription|General info|}} | ||
Latest revision as of 11:25, 30 July 2024
The measurement of the harmonics is often needed to ensure that the used amplifier in the setup is not generating too much power on multiples of the carrier frequency.
Overview[edit]
- Measure the power of all harmonics using a spectrum analyser.
- Measure the broadband power with a powermeter and the harmonics using a spectrum analyser.
- Measure the actual field harmonics using a receive antenna.
- If the selected testsite has a device driver selected for the Sensor power meter, then the 'Measure the broadband power with a powermeter and the harmonics using a spectrum analyser' method will be used.
- Otherwise, if the selected testsite has a device driver selected for the Spectrum analyser, then the 'Measure the actual field harmonics using a receive antenna' method will be used.
- In all other cases the 'Measure the power of all harmonics using a spectrum analyser' method will be used.
# | Device name | Tab in testsite configuration window | suitable for harmonic frequencies | note |
Signal Generator | Devices 1 | no | ||
Amplifier | Devices 1 | no | This is the amplifier of which the harmonics are being measured. | |
Coupler | Devices 1 | yes | ||
Forward power meter | Devices 1 | yes | The powermeter device driver for the analyser that is being used to measure the harmonics. | |
Cables | ||||
① | Cable SG -> amplifier | Cables | no | Optional to correct for the cable loss between the signal generator and the amplifier |
② | Cable amplifier -> coupler | Cables | yes | Optional to correct for the cable loss between the amplifier and the coupler |
③ | Cable coupler -> forward power meter | Cables | yes | Optional to correct for the cable loss between the coupler and the forward power meter |
Coupler | Devices 1 | no | ||
Forward power meter | Devices 1 | no | The powermeter device driver for the broadband power measurement of the forward power. | |
Sensor power meter | Devices 2 | yes | The powermeter device driver for the analyser that is being used to measure the harmonics. | |
Cables | ||||
② | Cable amplifier -> coupler | Cables | no | Optional to correct for the cable loss between the amplifier and the coupler |
③ | Cable coupler -> forward power meter | Cables | no | Optional to correct for the cable loss between the coupler and the forward power meter |
④ | Cable current -> power meter | Cables | yes | Optional to correct for the cable loss between the coupler and the spectrum analyser (which is the sensor power meter) |
⑤ | Cable current -> power meter | Cables | yes | Optional to correct for the attenuator between the coupler and the spectrum analyser (which is the sensor power meter) |
⑥ | Cable current -> power meter | Cables | yes | Optional to correct for the cable loss between the coupler and the spectrum analyser (which is the sensor power meter) |
Calibration antenna | Devices 1 | yes | The antenna that is used for the receiving of the generated electrical fields on the harmonics. | |
Spectrum analyser | Devices 2 | yes | The device driver for the analyser that is being used to measure the harmonics. | |
Cables | ||||
④ | Cable preamp -> analyser | Cables | yes | Optional to correct for the attenuator between the receive antenna and the spectrum analyser |
⑤ | Cable preamp -> analyser | Cables | yes | Optional to correct for the cable loss between the receive antenna and the spectrum analyser |
To start the harmonics calibration first, open an EUT file. This EUT file represents the amplifier, which is the actual Equipment Under Test (EUT). By maintaining this EUT file, a yearly remeasurement of the harmonics of the amplifier will all be present in the same EUT file.
Once the EUT file is opened, select from the menu:
Calibration
EUT Calibration
Harmonics
The name of the selected TSF file. |
Sets the frequency specification to a frequency range. |
Sets start frequency. |
Sets end frequency. |
Sets a fixed step size from start to end frequency. |
Sets a logarithmic step size from start to stop frequency |
Sets the start frequency for a logarithmic step from which the actual measurement frequencies will be calculated. The starting from frequency should be lower than the actual start frequency. The test will start at the first calculated frequency that is equal to or higher than the specified start frequency. |
With frequency list the operator can select a frequency list file by browsing to the file location. The frequencies inside the frequency list file will be measured. |
The test level can be set to a constant level to which the power should be regulated. |
A .COR file can also be used. This can be the same file as for the Frequency List. |
The number of harmonics that should be measured. |
If only the even, odd or both (even and odd) harmonics should be measured. |
On which power levelling point the desired testlevel should be regulated. |
In the test site window the test engineer can select which equipment list will be used during this test. This also determines which method and equipment is being used. |
In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results |
The amount of measurements that are measured on each harmonic to determine its power. The average of these measurements will be used as the power of the harmonic. |
Starts the configured calibration. |
Cancel the configured calibration. |
Display the environment window. |
Display the note window. |
Display the Units configuration window |
Calibration procedure[edit]
- Set the signal generator to the first frequency and a low power.
- Measure the power using the forward powermeter. If the power is within the desired tolerance, continue at step 4.
- Adjust the signal generator carrier level. If the forward power was too high, reduce the carrier level, otherwise increase the carrier level.
- Use the spectrum analyser to measure the first harmonic
- Use the spectrum analyser to measure the other harmonics that should be measured
- Increase the signal generator carrier frequency to the next frequency. continue at step 2.
Carrier frequency | 100 MHz | 200 MHz | 300 MHz | 400 MHz | 500 Mhz |
Forward power | 56.0 dBm | 56.1 dBm | 56.2 dBm | 56.1 dBm | 56.0 dBm |
1st harmonic | 55.9 dBm @ 100 MHz | 55.8 dBm @ 200 MHz | 55.9 dBm @ 300 MHz | 55.8 dBm @ 400 MHz | 55.9 dBm @ 500 MHz |
3rd harmonic | 37.8 dBm @ 300 MHz | 42.2 dBm @ 600 MHz | 52.7 dBm @ 900 MHz | 43.1 dBm @ 1200 MHz | 39.2 dBm @ 1500 MHz |
5th harmonic | 33.1 dBm @ 500 MHz | 37.4 dBm @ 1000 MHz | 37.9 dBm @ 1500 MHz | 38.4 dBm @ 2000 MHz | --- @ 2500 MHz |
It is possible that one or more of the higher harmonics have not been measured due to limitations of the allowed frequency range of the used equipment (which can be coupler, spectrum analyzer, antenna and cables). For a CW frequency of 500 MHz, the 5th harmonic is generated at 2500 MHz. If however the used spectrum analyzer is only able to measure up to 2000 MHz, it is not possible to measure the 5th harmonic of 500 MHz, and thus the value in the table will be empty. Also the graphline of the 5th harmonic will then only be drawn up to 400 MHz, which corresponds to the 5th harmonic at 2000 MHz, which is the highest measurable frequency of that spectrum analyzer.
Test results[edit]
The power of the signal generator that is driving the amplifier. |
The (broadband) output power that is generated by the amplifier. |
The power of the first harmonic (the actual CW frequency) that is generated by the amplifier. |
The power of specified harmonic that is generated by the amplifier. |
The dBc difference between the specified harmonic and the first harmonic. |
The standard deviation of the measured power samples of the specified harmonic. |
The harmonics calibration in RadiMation can thus be used in different measurement configurations to measure the actually generated harmonics. Depending on the requirements and the used standard, the best suitable measurement configuration can be used.