m (→Polarisation: Grammar fixes) |
m (→Zoom function: Fix typo) |
(29 intermediate revisions by the same user not shown) | |||
Line 175: | Line 175: | ||
When this option is selected, all data will be displayed in one graph thus losing some polarisation information. However the peak list table will show if a peak was measured in horizontal or vertical polarisation. | When this option is selected, all data will be displayed in one graph thus losing some polarisation information. However the peak list table will show if a peak was measured in horizontal or vertical polarisation. | ||
If both emission graphs of horizontal and vertical polarisation are required, two separate tests must be configured. These tests can be put in a sequence, eliminating the need for | If both emission graphs of horizontal and vertical polarisation are required, two separate tests must be configured. These tests can be put in a sequence, eliminating the need for user intervention. | ||
====Distance==== | ====Distance==== | ||
Line 194: | Line 194: | ||
{{ScreenElementDescription|Pre-amp|Sets the pre-amp gain.}} | {{ScreenElementDescription|Pre-amp|Sets the pre-amp gain.}} | ||
When a parameter is set to auto, {{RadiMation}} will determine the most | When a parameter is set to auto, {{RadiMation}} will determine the most optimal settings for this parameter. If the parameter is set to coupled, the parameter will be set to automatic internally in the receiver. | ||
{{Warning| When older analyzers and receivers are used that switch between measuring in a linear or logarithmic scale during the pre-scan and final-measurement phases, it is advised to not use a fixed reference level}} | {{Warning| When older analyzers and receivers are used that switch between measuring in a linear or logarithmic scale during the pre-scan and final-measurement phases, it is advised to not use a fixed reference level}} | ||
Line 205: | Line 205: | ||
{{ScreenElementDescription|Detectors|The detector settings (Peak, Average, Quasi-peak, RMS) used during sweeptime. This setting is generic. So it is possible to select multiple detectors, but only one detector can be selected due to the device limitations. In that case for each enabled detector the amount of sweeps will be performed by {{RadiMation}}.}} | {{ScreenElementDescription|Detectors|The detector settings (Peak, Average, Quasi-peak, RMS) used during sweeptime. This setting is generic. So it is possible to select multiple detectors, but only one detector can be selected due to the device limitations. In that case for each enabled detector the amount of sweeps will be performed by {{RadiMation}}.}} | ||
{{ScreenElementDescription|Number of sweeps|The number | {{ScreenElementDescription|Number of sweeps|The number of peak hold sweeps for each measurement can be selected. If this value is set to for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps depends on the type of signal and the setting of the sweep speed. For spectrum analysers and CW signals you can use a relatively low amount of peak sweeps, for example 10. But when you have discontinuous signals like a spark bridge, you may need 100 peak sweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope.}} | ||
===Final peak measurement=== | ===Final peak measurement=== | ||
Line 211: | Line 211: | ||
[[Image:Radiated Emission Multi Band Final Peak Measurement Settings Window.png]] | [[Image:Radiated Emission Multi Band Final Peak Measurement Settings Window.png]] | ||
In the final peak measurement window the test engineer can configure the detectors used during the final emission measurements of the detected peaks. The first value is the time constant of the detector while the second value is used for the measure time at each frequency point for this detector. The time constant is | In the final peak measurement window the test engineer can configure the detectors used during the final emission measurements of the detected peaks. The first value is the time constant of the detector while the second value is used for the measure time at each frequency point for this detector. The time constant is sent to the receiver, to set the sample time of the detector. To select a detector, check the desired button. | ||
===Peak detection window=== | ===Peak detection window=== | ||
Line 232: | Line 232: | ||
[[Image:Radiated Emission Multi Band Limit Lines Settings Window.png]] | [[Image:Radiated Emission Multi Band Limit Lines Settings Window.png]] | ||
In the limit line window the operator | In the limit line window the operator selects one or more limit lines to be used during the test. | ||
{{ScreenElementDescription|All peaks above “xx” below the limit lines|In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 5 dB, the software will detect all peaks | {{ScreenElementDescription|All peaks above “xx” below the limit lines|In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 5 dB, the software will detect all peaks which are higher than 5 dB below the limit line.}} | ||
===Change Order window=== | ===Change Order window=== | ||
Line 244: | Line 244: | ||
[[Image:Radiated Emission Multi Band Band A Band B.png]] | [[Image:Radiated Emission Multi Band Band A Band B.png]] | ||
You need to perform a test as shown but you have 2 test sites. One with an antenna with the frequency range that is equal to the total band. | You need to perform a test as shown but you have 2 test sites. One with an antenna with the frequency range that is equal to the total band. At the second test site you have 2 antennas, one for every frequency band. For test site one you want the RBW filter to change as little as possible to prevent the old analyser from breaking down. | ||
So for Site 1 you want: | So for Site 1 you want: | ||
Line 258: | Line 258: | ||
*Band B RBW 120 kHz | *Band B RBW 120 kHz | ||
For test site one, we do not have to change the order. Why? you may wonder. The explanation is quite easy. Look at the frequency change list, first change frequency band then change RBW. That is exactly what you want. Let check what the software is going to do: | For test site one, we do not have to change the order. Why? you may wonder. The explanation is quite easy. Look at the frequency change list, first change frequency band then change RBW. That is exactly what you want. Let's check what the software is going to do: | ||
*Band A RBW 9 kHz | *Band A RBW 9 kHz | ||
Line 265: | Line 265: | ||
*Band A RBW 120 kHz | *Band A RBW 120 kHz | ||
This is even better | This is even better than we had hoped for, the amount of switching has been decreased due to the fact that {{RadiMation}} detected that it had to measure the same frequency twice. | ||
For test site two, it is a bit more complicated. You might wonder why? It looks easy just put the RBW in front the Frequency band and you are done. So far you are correct, but a different antenna also means a different test site. So changing the test site should also be above frequency band. When doing so, lets check it again. The top three settings: | For test site two, it is a bit more complicated. You might wonder why? It looks easy just put the RBW in front the Frequency band and you are done. So far you are correct, but a different antenna also means a different test site. So changing the test site should also be above the frequency band. When doing so, lets check it again. The top three settings: | ||
*RBW | *RBW | ||
*Testsite | *Testsite | ||
Line 279: | Line 279: | ||
*Band B RBW 9 kHz | *Band B RBW 9 kHz | ||
This is also better | This is also better than we had hoped for, the amount of switching has been decreased due to the fact that {{RadiMation}} detected that it had to measure the same RBW twice. | ||
{{Note|If you have any questions regarding the setting of the change order please contact your reseller.}} | {{Note|If you have any questions regarding the setting of the change order please contact your reseller.}} | ||
Line 315: | Line 315: | ||
[[image:Radiated Emission Multi Band Final Peak Measurement Settings Window.png]] | [[image:Radiated Emission Multi Band Final Peak Measurement Settings Window.png]] | ||
In the final peak measurement window the test engineer can configure the detectors used during the final emission measurements of the peaks. The first value is the time constant of the detector while the second value is used for the measure time at each frequency point for this detector. The time constant is | In the final peak measurement window the test engineer can configure the detectors used during the final emission measurements of the peaks. The first value is the time constant of the detector while the second value is used for the measure time at each frequency point for this detector. The time constant is sent to the receiver, to set the sample time of the detector. To select a detector, check the desired button. | ||
===Peaks window=== | ===Peaks window=== | ||
Line 325: | Line 325: | ||
{{ScreenElementDescription|Measure Peaks|Measures all selected peaks with the configured settings in the final measurement window.}} | {{ScreenElementDescription|Measure Peaks|Measures all selected peaks with the configured settings in the final measurement window.}} | ||
{{ScreenElementDescription|Peak Actions|Allows several actions to be performed on the detected peaks, like select all and deselect all. | {{ScreenElementDescription|Peak Actions|Allows several actions to be performed on the detected peaks, like select all and deselect all. | ||
{{ScreenElementDescription|Peaks to maximum|Start performing measurement to redetermine on which turntable angle and antenna tower height the highest emission value was measured. | {{ScreenElementDescription|Peaks to maximum|Start performing a measurement to redetermine on which turntable angle and antenna tower height the highest emission value was measured. | ||
This angle and height will be updated in the selected peak, together with the corresponding measured emission values. If multiple peaks are selected, this search for the highest emission value is done individually for every selected peak. | This angle and height will be updated in the selected peak, together with the corresponding measured emission values. If multiple peaks are selected, this search for the highest emission value is done individually for every selected peak. | ||
Line 393: | Line 393: | ||
In this mode the software increases the angle of the turntable from the initial point with a step equal to {{ScreenElement|Step size}} as long as the measurement value is not more than {{ScreenElement|Threshold}} below the already known maximum measurement value. As soon as measurement value is detected that is {{ScreenElement|Threshold}} below the already known | In this mode the software increases the angle of the turntable from the initial point with a step equal to {{ScreenElement|Step size}} as long as the measurement value is not more than {{ScreenElement|Threshold}} below the already known maximum measurement value. As soon as a measurement value is detected that is {{ScreenElement|Threshold}} below the already known maximum value, the turntable movement is reversed until the initial point and from this initial point the angle of the turntable is decreased with a step equal to {{ScreenElement|Step size}} until a measurement value is measured that is {{ScreenElement|Threshold}} below the already known maximum measurement value. The step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak. | ||
{{Warning|This might cause a measurement error if the | {{Warning|This might cause a measurement error if the initial number of steps ({{ScreenElement|#}}) is set to a low value.}} | ||
====Complete area==== | ====Complete area==== | ||
Line 407: | Line 407: | ||
{{ScreenElementDescription|Max. height|Sets maximum height of the antenna tower.}} | {{ScreenElementDescription|Max. height|Sets maximum height of the antenna tower.}} | ||
{{ScreenElementDescription|Min. height|Sets the minimum height of the antenna tower.}} | {{ScreenElementDescription|Min. height|Sets the minimum height of the antenna tower.}} | ||
{{ScreenElementDescription|#|Number | {{ScreenElementDescription|#|Number of locations from {{ScreenElement|Max. height}} to {{ScreenElement|Min. height}} that should be measured.}} | ||
{{ScreenElementDescription|Peak search|The software will automatically detect on which antenna height the maximum measurement value of each peak is present.}} | {{ScreenElementDescription|Peak search|The software will automatically detect on which antenna height the maximum measurement value of each peak is present.}} | ||
Line 423: | Line 423: | ||
In this mode the software increases the height of the antenna from the initial point with a step equal to {{ScreenElement|Step size}} as long as the measurement value is not more than {{ScreenElement|Threshold}} below the already known maximum measurement value. As soon as measurement value is detected that is {{ScreenElement|Threshold}} below the already known | In this mode the software increases the height of the antenna from the initial point with a step equal to {{ScreenElement|Step size}} as long as the measurement value is not more than {{ScreenElement|Threshold}} below the already known maximum measurement value. As soon as a measurement value is detected that is {{ScreenElement|Threshold}} below the already known maximum value, the antenna movement is reversed until the initial point and from this initial point the distance of the clamp is decreased with a step equal to {{ScreenElement|Step size}} until a measurement value is measured that is {{ScreenElement|Threshold}} below the already known maximum measurement value. The step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak. | ||
Line 446: | Line 446: | ||
When this option is selected, all data will be displayed in one graph thus loosing some polarisation information. However the peak list table will show if a peak was measured in horizontal or vertical polarisation. | When this option is selected, all data will be displayed in one graph thus loosing some polarisation information. However the peak list table will show if a peak was measured in horizontal or vertical polarisation. | ||
If the both emission graphs of horizontal and vertical polarisation are required, two separate tests must be configured. These tests can be put in a sequence, eliminating the need for | If the both emission graphs of horizontal and vertical polarisation are required, two separate tests must be configured. These tests can be put in a sequence, eliminating the need for user intervention. | ||
Line 463: | Line 463: | ||
{{ScreenElementDescription|Measure time|Sets the measure time of the receiver}} | {{ScreenElementDescription|Measure time|Sets the measure time of the receiver}} | ||
However, using the 'Auto' setting for the attenuator can let the analyser | However, using the 'Auto' setting for the attenuator can let the analyser decide to use a different attenuator setting during (re-)measuring of a peak, which as a result can lead to measuring a higher Quasi-Peak value than the Peak value. This is related to the used analyser when auto attenuator is used. If it is not desired that the analyser can decide to switch to a different attenuator setting, then this should be disabled by the test engineer in {{RadiMation}} by using a fixed attenuator value. | ||
{{Warning| All these settings are generic. Therefore, always verify that the configured settings are also supported by the used hardware to ensure it is correctly used and no measurement faults are introduced. Use a fixed value to | {{Warning| All these settings are generic. Therefore, always verify that the configured settings are also supported by the used hardware to ensure it is correctly used and no measurement faults are introduced. Use a fixed value to disable interference of equipment intelligence.}} | ||
Line 473: | Line 473: | ||
===Max. hold measurements=== | ===Max. hold measurements=== | ||
In this window the number of peak hold sweeps for each measurement can be selected. If this value is set to | In this window the number of peak hold sweeps for each measurement can be selected. If this value is set to for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweepspeed. For spectrum analysers and CW signals you can use a relatively low amount of peaksweeps, for example 10. But when you have discontinuous signals like a spark bridge, you may need 100 peaksweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope. | ||
===Traces window=== | ===Traces window=== | ||
Line 502: | Line 502: | ||
===Test options window=== | ===Test options window=== | ||
{{ScreenElementDescription|Record data|In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full | {{ScreenElementDescription|Record data|In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full traceability, all test data should be stored.}} | ||
{{ScreenElementDescription|Narrow band / broad band discrimination|{{RadiMation}} supports two types of narrow band / broad band determination.}} | {{ScreenElementDescription|Narrow band / broad band discrimination|{{RadiMation}} supports two types of narrow band / broad band determination.}} | ||
Line 602: | Line 602: | ||
{{ScreenElementDescription|Graph control|In the graph control window the scale of the graph can be changed. You can show or hide the table by selecting or deselecting the show table checkbox.}} | {{ScreenElementDescription|Graph control|In the graph control window the scale of the graph can be changed. You can show or hide the table by selecting or deselecting the show table checkbox.}} | ||
{{ScreenElementDescription|Trace|In the trace window you can control the traces that are shown in the graph by selecting or deselecting the checkbox. If during a measurement a trace is not used then the | {{ScreenElementDescription|Trace|In the trace window you can control the traces that are shown in the graph by selecting or deselecting the checkbox. If during a measurement a trace is not used then the checkbox is grayed out.}} | ||
{{ScreenElementDescription|Detector|The detector window gives you the possibility to show and hide the values measured on the peaks. If during a measurement a type of detector is not used for determining the value on the peak the checkbox is grayed out.}} | {{ScreenElementDescription|Detector|The detector window gives you the possibility to show and hide the values measured on the peaks. If during a measurement a type of detector is not used for determining the value on the peak the checkbox is grayed out.}} | ||
{{ScreenElementDescription|Table|In the table all the measured peaks are | {{ScreenElementDescription|Table|In the table all the measured peaks are shown. The table also shows the limit values for that frequency point when a limit was given at the beginning of the test.}} | ||
===Test site information=== | ===Test site information=== | ||
Line 683: | Line 683: | ||
Line 697: | Line 697: | ||
Line 713: | Line 713: | ||
Line 735: | Line 735: | ||
When Horizontal or Vertical is selected, the software will measure only one antenna polarisation. When the option “Both” is selected, the software will measure both Horizontal and Vertical polarisation (an antenna positioner is required for this option). | When Horizontal or Vertical is selected, the software will measure only one antenna polarisation. When the option “Both” is selected, the software will measure both Horizontal and Vertical polarisation (an antenna positioner is required for this option). | ||
When this option is selected, all data will be displayed in one graph thus | When this option is selected, all data will be displayed in one graph thus losing some polarisation information. However the peak list table will show if a peak was measured in horizontal or vertical polarisation. | ||
If | If both emission graphs of horizontal and vertical polarisation are required, two separate tests must be configured. These tests can be put in a sequence, eliminating the need for a user intervention. | ||
Line 755: | Line 755: | ||
Line 761: | Line 761: | ||
In this window the number | In this window the number of peak hold sweeps for each measurement can be selected. If this value is set to, for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweepspeed. For spectrum analysers and CW signals you can use a relatively low amount of peaksweeps, for example 10. But when you have discontinuous signals like a spark bridge, you may need 100 peaksweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope. | ||
Line 790: | Line 790: | ||
In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full | In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full traceability, all test data should be stored. | ||
===Starting a fixed height test=== | ===Starting a fixed height test=== | ||
Line 835: | Line 835: | ||
{{RadiMation}} now performs a maxhold on all the sweeps and determines the peaks. | {{RadiMation}} now performs a maxhold on all the sweeps and determines the peaks. | ||
Then it determines the position a peak was found. | Then it determines the position a peak was found. | ||
The turntable position is | The turntable position is changed to that position to re-measure the peak. | ||
The antenna tower is changed in height according to the settings. | The antenna tower is changed in height according to the settings. | ||
Line 939: | Line 939: | ||
Line 945: | Line 945: | ||
====Max. hold measurements==== | ====Max. hold measurements==== | ||
====Traces window==== | ====Traces window==== | ||
Line 976: | Line 976: | ||
===Starting a G-TEM test=== | ===Starting a G-TEM test=== | ||
Line 1,038: | Line 1,038: | ||
==Manual mode (Singleband)== | ==Manual mode (Singleband)== | ||
For product improvements, it quit often is required to operate the test set-up in manual mode. This can be done either by operating the measurement receiver (or spectrum analyser) manually or by the {{RadiMation}} software. In our opinion it is no use to take over the buttons of the receiver without adding any extra functionality. For this reason a number of | For product improvements, it quit often is required to operate the test set-up in manual mode. This can be done either by operating the measurement receiver (or spectrum analyser) manually or by the {{RadiMation}} software. In our opinion it is no use to take over the buttons of the receiver without adding any extra functionality. For this reason a number of handy features are included in manual mode. | ||
Line 1,049: | Line 1,049: | ||
{{ScreenElementDescription|Reference|Sets the reference level of the receiver}} | {{ScreenElementDescription|Reference|Sets the reference level of the receiver}} | ||
{{ScreenElementDescription|Attenuator|Sets the receiver attenuator level}} | {{ScreenElementDescription|Attenuator|Sets the receiver attenuator level}} | ||
{{ScreenElementDescription|RBW|Sets the | {{ScreenElementDescription|RBW|Sets the Resolution Band Width filter}} | ||
{{ScreenElementDescription|VBW|Sets the Video BandWidth filter}} | {{ScreenElementDescription|VBW|Sets the Video BandWidth filter}} | ||
{{ScreenElementDescription|Sweep|Sets the sweep time of the receiver}} | {{ScreenElementDescription|Sweep|Sets the sweep time of the receiver}} | ||
{{ScreenElementDescription|Pre-amp|Sets the pre-amp gain}} | {{ScreenElementDescription|Pre-amp|Sets the pre-amp gain}} | ||
Line 1,066: | Line 1,066: | ||
====Data correction==== | ====Data correction==== | ||
The semi- manual mode of {{RadiMation}} uses the PC screen to display the emission data real time on the computer screen. However the measured data is corrected for antenna factors and cable losses. | The semi- manual mode of {{RadiMation}} uses the PC screen to display the emission data real-time on the computer screen. However the measured data is corrected for antenna factors and cable losses. | ||
===Peak menu=== | ===Peak menu=== | ||
Line 1,085: | Line 1,085: | ||
====Traces==== | ====Traces==== | ||
In the “traces” menu, a maximum | In the “traces” menu, a maximum of four traces can be switched on. Each trace can be appointed to be a Max hold trace, Average trace, Write trace or be switched either to View mode or just left Blank. | ||
The figure below shows the traces menu: | The figure below shows the traces menu: | ||
[[Image:Radiated Emission Manual Mode Test Traces.png]] | [[Image:Radiated Emission Manual Mode Test Traces.png]] | ||
====Detectors==== | ====Detectors==== | ||
Line 1,107: | Line 1,106: | ||
====Zoom function==== | ====Zoom function==== | ||
{{RadiMation}} features a live zoom function. When the operator wants to “zoom in” on a certain signal just use the mouse to drag a box around the signal of interest. The {{RadiMation}} software will fully | {{RadiMation}} features a live zoom function. When the operator wants to “zoom in” on a certain signal just use the mouse to drag a box around the signal of interest. The {{RadiMation}} software will fully automatically zoom in to this signal. The screen, as well as the measurement receiver will be zoomed. This reduces the number of actions of the operator significantly. | ||
The figures on the next page show how to drag a box to zoom into a graph, while the second graph shows the zoomed graph. | The figures on the next page show how to drag a box to zoom into a graph, while the second graph shows the zoomed graph. |
Latest revision as of 15:24, 6 February 2025
Radiated emission[edit]
- Multiband
- Manual Mode
- Burst
- Anechoic chamber
- Fixed height
- GTEM
- Mil Std Magnetic
- Manual mode (Singleband)
- Merge
- X + k * S
Loading a TSF file[edit]
Tests
Test results screen[edit]
Note: | The pass/fail criteria, shown in the tables do not take the measurement uncertainty of the test setup into account. |
Note: | Continuing the test is only possible with the selected test site. RadiMation® will not select the correct test site for you. All measurements are done with the selected test site. If you want automatic changes of the test site, restart the configured TSF file. |
Testsite information[edit]
Spectrum analyser | Devices 2 | |
Pre amplifier | Devices 2 | |
Antenna | Devices 1 | |
Turn table | Devices 1 | |
Antennatower | Devices 1 | |
Cables | ||
1 | Cable preamp -> analyser | Cables |
2 | Cable antenna -> preamp | Cables |
Note: | Usage of cable correction factors is not mandatory. However, using cable correction factors will increase measurement accuracy during emission measurements. When in doubt contact your reseller. |
Multiband[edit]
Frequency window[edit]
Turntable window[edit]
Sets the start position of the turntable. |
Sets the stop position of the turntable. |
Sets the number of steps for one turntable rotation. |
This will force the turntable to continuously rotate during the test and will disable steps setting. |
The offset of the EUT compared to the zero angle of the table. |
Optimized settings for searching the correct angle. |
Range over which the peak search will be performed (0 to 360 degrees max.). |
The angle step size. |
Antenna tower window[edit]
Sets maximum height of the antenna tower. |
Sets the minimum height of the antenna tower. |
Sets the number of steps to go from min. height to max. height. |
This will force the Antenna Tower to continuously move during the test and will disable steps setting. |
The software will automatically detect on which antenna height maximum emission will occur. |
Range over which the peak search will be performed (0 to Max height). |
The height step size. |
Antenna window[edit]
Polarisation[edit]
- Horizontal
- Vertical
- Both
Distance[edit]
Receiver settings[edit]
Sets the reference level of the receiver. |
Sets the receiver attenuator level. |
Sets the Resolution Band Width filter. |
Sets the Video Band Width filter. |
Sets the sweep time of the receiver. |
Sets the frequency step size of the receiver. |
Sets the measure time of the receiver. |
Sets the pre-amp gain. |
Warning: | When older analyzers and receivers are used that switch between measuring in a linear or logarithmic scale during the pre-scan and final-measurement phases, it is advised to not use a fixed reference level |
Traces window[edit]
The detector settings (Peak, Average, Quasi-peak, RMS) used during sweeptime. This setting is generic. So it is possible to select multiple detectors, but only one detector can be selected due to the device limitations. In that case for each enabled detector the amount of sweeps will be performed by RadiMation®. |
The number of peak hold sweeps for each measurement can be selected. If this value is set to for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps depends on the type of signal and the setting of the sweep speed. For spectrum analysers and CW signals you can use a relatively low amount of peak sweeps, for example 10. But when you have discontinuous signals like a spark bridge, you may need 100 peak sweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope. |
Final peak measurement[edit]
Peak detection window[edit]
When no final peak measurements are required select the option no peaks. |
With the max. peaks option, the user can limit the number of peak signals the software will detect and re-measure. |
Test site window[edit]
By clicking on the equipment icon, the equipment list can be viewed and edited. |
Limit lines window[edit]
In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 5 dB, the software will detect all peaks which are higher than 5 dB below the limit line. |
Change Order window[edit]
- Band A RBW 9 kHz
- Band B RBW 9 kHz
- Band A RBW 120 kHz
- Band B RBW 120 kHz
And for test site 2:
- Band A RBW 9 kHz
- Band B RBW 9 kHz
- Band A RBW 9 kHz
- Band B RBW 9 kHz
- RBW
- Testsite
- Frequency band
- Band A RBW 9 kHz
- Band B RBW 9 kHz
Note: | If you have any questions regarding the setting of the change order please contact your reseller. |
Manual Mode[edit]
Turn table[edit]
Antenna tower[edit]
Test site[edit]
Final measurement[edit]
Peaks window[edit]
Maximum allowed amount of peaks to be automatically determined. |
Detects the configured amount of sweeps in the measured spectrum. |
Measures all selected peaks with the configured settings in the final measurement window. |
Delete the selected peaks from the list. |
Amount of sweeps to be measured. |
Triggers the receiver of analyser to take the configured amount of sweeps. |
Start performing of taking measurements. This is a heavy process and it is normal when more CPU is used by the computer. |
Each new measurement will overwrite the previous measurement data. |
Only higher new measurement values will overwrite the previous measurement data. |
Removes all previous measurement data. |
Anechoic chamber[edit]
Sets start frequency |
Sets stop frequency |
Sets the start angle of the turntable. |
Sets the stop angle of the turntable. |
Number of locations from Start postion to Stop position that should be measured. |
The software will automatically detect on which turntable angle the maximum measurement value of each peak is present. |
Range over which the peak search will be performed. The initially detected turntable angle of the peak will be the middle of this Area, and the total search area for this peak will be half of this area clockwise and half of this area counter-clockwise. |
The step size for each turntable movement. |
This value is only used during the Intelligent mode. It influences how long the movement is continued. The movement will stop when the measurement value is lower than the specified Threshold below the highest known measurement value of the searched peak. |
Intelligent[edit]
Warning: | This might cause a measurement error if the initial number of steps (#) is set to a low value. |
Complete area[edit]
Warning: | This might cause a measurement error if the Area is set to a smaller value than the original stepsize (given by the stop position minus start position divided by the number of steps). |
Number of locations from Max. height to Min. height that should be measured. |
The software will automatically detect on which antenna height the maximum measurement value of each peak is present. |
Range over which the peak search will be performed. The initially detected antenna height of the peak will be the middle of this Area, and the total search area for this peak will be half of this area up and half of this area down. |
The step size for each antenna height movement. |
Intelligent[edit]
Complete area[edit]
Polarisation[edit]
- Horizontal
- Vertical
- Both
Distance[edit]
Sets the reference level of the receiver |
Sets the receiver attenuator level |
Sets the Resolution BandWidth filter |
Sets the Video BandWidth filter |
Sets the sweep time of the receiver |
Sets the pre-amp gain |
Sets the frequency step size of the receiver |
Sets the measure time of the receiver |
When a parameter is set to auto, RadiMation® will determine the most optimal settings for this parameter. If the parameter is set to coupled, the parameter will be set to automatic internally in the receiver. However, using the 'Auto' setting for the attenuator can let the analyser decide to use a different attenuator setting during (re-)measuring of a peak, which as a result can lead to measuring a higher Quasi-Peak value than the Peak value. This is related to the used analyser when auto attenuator is used. If it is not desired that the analyser can decide to switch to a different attenuator setting, then this should be disabled by the test engineer in RadiMation® by using a fixed attenuator value.
Warning: | All these settings are generic. Therefore, always verify that the configured settings are also supported by the used hardware to ensure it is correctly used and no measurement faults are introduced. Use a fixed value to disable interference of equipment intelligence. |
Max. hold measurements[edit]
Traces window[edit]
Detectors window[edit]
In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 6 dB, the software will detect all peaks, which are higher than 6 dB below the limit line. |
In the test engineer window, the test engineer can select their own name. The test engineer name will be stored by the test results. |
Test options window[edit]
RadiMation® supports two types of narrow band / broad band determination. |
Limits | Distortion type |
Loading a TSF file
Test Procedure[edit]
Info screen[edit]
Note: | The pass/fail criteria, shown in the tables do not take the measure uncertainty of the test setup into account. |
In the settings window all the used settings are displayed. |
In the graph control window the scale of the graph can be changed. You can show or hide the table by selecting or deselecting the show table checkbox. |
In the trace window you can control the traces that are shown in the graph by selecting or deselecting the checkbox. If during a measurement a trace is not used then the checkbox is grayed out. |
The detector window gives you the possibility to show and hide the values measured on the peaks. If during a measurement a type of detector is not used for determining the value on the peak the checkbox is grayed out. |
In the table all the measured peaks are shown. The table also shows the limit values for that frequency point when a limit was given at the beginning of the test. |
Test site information[edit]
Cables | ||
1 | cable Preamp -> analyser | Cables |
2 | cable Antenna -> Preamp | Cables |
Note: | Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during emission measurements.
Fixed height[edit]
Intelligent[edit]
Complete area[edit]
Intelligent[edit]
Complete area[edit]
Polarisation[edit]
- Horizontal
- Vertical
- Both
Distance[edit]
Traces window[edit]
With the ‘max. peaks” option, the user can limit the number of peak signals the software will detect and re-measure. |
Loading a TSF file
RadiMation® now performs a maxhold on all the sweeps and determines the peaks. Then it determines the position a peak was found. The turntable position is changed to that position to re-measure the peak. The antenna tower is changed in height according to the settings.
Pre amplifier | Devices2 | |
Cables | ||
1 | cable amplifier ->Spectrum | Cables |
2 | cable antenna ->Preamp | Cables |
GTEM[edit]
Correlation method[edit]
Measurement procedure[edit]
Tests
Radiated emission
G-TEM
Test configuration[edit]
OATS settings[edit]
G-TEM settings[edit]
Sets the Resolution BandWidth filter. |
Sets the Video BandWidth filter. |
Traces window[edit]
Max. Peaks
Starting a G-TEM test[edit]
Cables | ||
1 | cable Preamp->analyser | Cables |
2 | cable Antenna->Preamp | Cables |
For product improvements, it quit often is required to operate the test set-up in manual mode. This can be done either by operating the measurement receiver (or spectrum analyser) manually or by the RadiMation® software. In our opinion it is no use to take over the buttons of the receiver without adding any extra functionality. For this reason a number of handy features are included in manual mode.
Sets the Resolution Band Width filter |
Data correction[edit]
The semi- manual mode of RadiMation® uses the PC screen to display the emission data real-time on the computer screen. However the measured data is corrected for antenna factors and cable losses.
[edit]
RadiMation® can detect peaks automatically. From the “peaks” TAB, choose “Search peak” to detect the maximum signal. The software will measure the detected peak. |
The measured values can be added to the peak list by pressing the “Add Peak” button. |
By pressing the “Search Angle” button, the software starts a turntable scan to determine at which turn table angle maximum emission takes place. |
By pressing the “Search Height” button, the software starts an antenna height scan to determine at which antenna height maximum emission takes place. |
Traces Menu[edit]
Traces[edit]
Detectors[edit]
- Peak detector
- Average detector
- Quasi peak detector
- RMS detector
Zoom function[edit]
RadiMation® features a live zoom function. When the operator wants to “zoom in” on a certain signal just use the mouse to drag a box around the signal of interest. The RadiMation® software will fully automatically zoom in to this signal. The screen, as well as the measurement receiver will be zoomed. This reduces the number of actions of the operator significantly.
When this option is switched on, the frequency settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph. |
When this option is switched on, the amplitude settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph. |
Limit lines[edit]
[edit]
Merge[edit]
X + k * S[edit]