RadiMation Application Note 165: Difference between revisions
(20 intermediate revisions by 2 users not shown)
Line 19: Line 19:
[[File:CI_Calibration_setup.png]]
{{Warning|Make sure that the sensor power meter is connected to the EUT-port of the injection device.}}
Line 45: Line 47:
=== Optional Attenuator ===
Before deciding whether an attenuator is required in front of the sensor power meter, the power level at the output of the calibration jig must first be calculated.
Before deciding whether an attenuator is required in front of the sensor power meter, the power level at the output of the calibration jig must be calculated first.
==== Why a factor of 6 applies ====
Line 60: Line 62:
:<math>L_{\text{port, dBµV}} = L_{\text{test, dBµV}} - 15.56\,\text{dB}</math>
==== Conversion to power in dBm ===
==== Conversion to power in dBm ====
:<math>P_{\text{dBm}} = L_{\text{port, dBµV}} - 107\,\text{dB}</math>
Line 73: Line 75:
! Test Level (V, rms) !! Test Level (dBµV) !! Jig Output (dBµV) !! Power (dBm)
| 1.0 || 120.0 || 104.44 || −2.56
| 1 || 120.0 || 104.44 || −2.56
| 3.0 || 129.54 || 113.98 || +6.98
| 3 || 129.54 || 113.98 || +6.98
| 10.0 || 140.0 || 124.44 || +17.44
| 10 || 140.0 || 124.44 || +17.44
=== Configure the calibration ===
==== Conclusion ====
Using the calculated power levels, it is straightforward to determine the amount of attenuation required for a given sensor power meter.
For example, the '''Raditeq RPR2006C''' has a specified measurement range up to +10&nbsp;dBm.
* At 1&nbsp;V and 3&nbsp;V test levels (−2.56&nbsp;dBm and +6.98&nbsp;dBm), no attenuation is required.
* At 10&nbsp;V test level (+17.44&nbsp;dBm), the input exceeds the safe range by approximately 7.5&nbsp;dB.
In this case, adding at least a 10&nbsp;dB attenuator ensures the sensor operates within its specified limits. Make sure to correct for this 10&nbsp;dB by creating a cable device driver with a correction file attached to it, then configure this cable in the test site: {{ScreenElement|Cables}}, {{ScreenElement|Cable current -> power meter}}.
=== Configure the injection device and jig device driver ===
In order for RadiMation to calibrate on the correct test level, the injection device and jig device driver must be configured correctly.
==== Jig device driver ====
For a 150&nbsp;Ω system the resistance is 150&nbsp;Ω or uses a jig factor of 9.5&nbsp;dB.
[[Image:Calibration_jig_factor_9.5_dB_for_150_Ohm_system.png]]
[[Image:Calibration_jig_factor_150_Ohm_for_150_Ohm_system.png]]
==== Injection device driver ====
For the injection device driver the impedance should also be set to 150&nbsp;Ω if that is configurable.
[[Image:CI_injection_device.png]]
=== Configure the Conducted immunity calibration ===
The calibration can be configured and started from the menu by selecting:
{{Menu|Calibration|System calibration|Conducted immunity}}
Line 98: Line 123:
When the calibration has finished, {{RadiMation}} will ask to store the calibration file.
== Troubleshooting ==
{{Warning|When a message is shown: {{ScreenElement|Signal power went too high (-4 dBm) during power delivery test}}, a system verification must be carried out to confirm that the hardware is functioning correctly and is correctly connected.}}
== EUT Test ==
== EUT Test ==
Line 167: Line 195:
=== Conclusion ===
=== Conclusion ===
The IEC 61000-4-6 calibration can be performed by using the Conducted immunity calibration. The RadiMation® Conducted immunity multiband test can then be used to perform the IEC 61000-4-6 EUT test with a calibration file in the EUT test.
Latest revision as of 11:29, 25 November 2025
How to perform an IEC 61000-4-6, Conducted immunity test[edit]
This application note explains how the IEC 61000-4-6 conducted immunity test can be performed with RadiMation®. The IEC 61000-4-6 conducted immunity test method is used for electrical and electronic equipment that may be subjected to radio-frequency disturbances conducted along cables coming from radio transmitters in the frequency range from 150 kHz up to 80 MHz. The exact requirements, calibration and EUT test methods are described in the IEC 61000-4-6 standard.
Necessary equipment[edit]
  • Signal generator
  • Amplifier
  • Coupler
  • Forward power meter
  • Injection device
  • Calibration jig with 100 Ohm resistors
  • 50 Ohm load
  • Attenuator in front of sensor power meter (optional)
  • Sensor power meter
Calibration procedure[edit]
The calibration is performed to determine the power required to match the corresponding voltage test level seen at the output of the injection device.
Warning: Make sure that the sensor power meter is connected to the EUT-port of the injection device.
Coupler Devices 1 Coupler
Forward power meter Devices 1 Forward power meter
Injection device Devices 2 The injection clamp or CDN
Calibration Jig Devices 2 The calibration jig to use
Sensor power meter Devices 2 Power meter
Cables
Cable current -> power meter Cables Cable with the specified loss of the used attenuator
Optional Attenuator[edit]
Why a factor of 6 applies[edit]
Two effects reduce the test voltage before it appears at the 50 Ω measurement port:
  • Impedance conversion (150 Ω → 50 Ω): factor of 3
  • Open-circuit to loaded conversion: factor of 2
Combined, this gives a total factor of 6:
Conversion from test level (dBµV) to jig output (dBµV)[edit]
Conversion to power in dBm[edit]
Short form[edit]
By combining the steps:
Examples[edit]
The following table shows the results for common test levels:
Test Level (V, rms) Test Level (dBµV) Jig Output (dBµV) Power (dBm)
1 120.0 104.44 −2.56
3 129.54 113.98 +6.98
10 140.0 124.44 +17.44
For example, the Raditeq RPR2006C has a specified measurement range up to +10 dBm.
  • At 1 V and 3 V test levels (−2.56 dBm and +6.98 dBm), no attenuation is required.
  • At 10 V test level (+17.44 dBm), the input exceeds the safe range by approximately 7.5 dB.
In this case, adding at least a 10 dB attenuator ensures the sensor operates within its specified limits. Make sure to correct for this 10 dB by creating a cable device driver with a correction file attached to it, then configure this cable in the test site: Cables, Cable current -> power meter.
Configure the injection device and jig device driver[edit]
Jig device driver[edit]
For a 150 Ω system the resistance is 150 Ω or uses a jig factor of 9.5 dB.
Injection device driver[edit]
For the injection device driver the impedance should also be set to 150 Ω if that is configurable.
Configure the Conducted immunity calibration[edit]
   Menu.svg Calibration
      Menu.svg System calibration
         Menu.svg Conducted immunity
ScreenElementDescription.svg Start The start frequency of the calibration. For example 150 kHz.
ScreenElementDescription.svg End The stop frequency of the calibration. For example 80 MHz.
ScreenElementDescription.svg Step The frequency step, in this case 1%.
ScreenElementDescription.svg Test level The test level, in this case Voltage.
ScreenElementDescription.svg Voltage The voltage test level, 1 V, 3 V or 10 V.
ScreenElementDescription.svg Calibration method The power measurement that should be the result of the calibration: Forward power
ScreenElementDescription.svg Test equipment Calibration test equipment.
ScreenElementDescription.svg Test engineer The engineer that performed the calibration.
When the calibration has finished, RadiMation® will ask to store the calibration file.
Troubleshooting[edit]
Warning: When a message is shown: Signal power went too high (-4 dBm) during power delivery test, a system verification must be carried out to confirm that the hardware is functioning correctly and is correctly connected.
EUT Test[edit]
EUT testing equipment[edit]
The configuration of the RadiMation® EUT test site should contain the following devices:
Forward power meter Devices 1 The forward power meter to use
Injection device Devices 2 The injection clamp or CDN to use
Configure the EUT test[edit]
To perform the actual test on the EUT, create a conducted immunity multiband test. The conducted immunity multiband test configuration, can be opened by selecting from the menu:
   Menu.svg Tests
      Menu.svg Conducted Immunity
         Menu.svg Multiband
ScreenElementDescription.svg Start The start frequency of the test. For example 150 kHz.
ScreenElementDescription.svg End The stop frequency of the test. For example 80 MHz.
ScreenElementDescription.svg Frequency Step The frequency step, 1%.
ScreenElementDescription.svg Dwell Time The dwell time.
ScreenElementDescription.svg Frequency Change Mode The method that is used to change from one to the next frequency: Constant.
ScreenElementDescription.svg Modulation Configure Amplitude Modulation with 1 kHz and 80 % duty cycle.
ScreenElementDescription.svg Test Level Specify the Voltage substition method
ScreenElementDescription.svg Inputs No inputs are needed, this is optional.
To configure the modulation settings click Config next to the modulation settings, click the AM tab, and enter the amplitude modulation settings.
ScreenElementDescription.svg Activate modulation Check this checkbox to enable the amplitude modulation.
ScreenElementDescription.svg Frequency The AM frequency 1 kHz.
ScreenElementDescription.svg Duty Cycle A depth of 80 %.
ScreenElementDescription.svg Also apply modulation out of the dwell-time period No need to apply modulation out of the dwell-time period.
At TestLevel click Add to add a new Test level and select TestLevel - Voltage substitution method
ScreenElementDescription.svg Calibration file Select the calibration file created during the calibration procedure.
ScreenElementDescription.svg Test level Constant voltage, 1 V, 3 V or 10 V.
When all settings are configured press Start Test to run the EUT test.
EUT test result[edit]