m (→Zoom function: Fix grammar) |
| (31 intermediate revisions by 5 users not shown) | |||
| Line 13: | Line 13: | ||
We recommend to use a transient limiter or to disconnect the receiving device when the LISN is switched on or off. The used LISN, spectrum analyser and/or measurement receiver can be selected under the configuration, tests site pull down menu. | We recommend to use a transient limiter or to disconnect the receiving device when the LISN is switched on or off. The used LISN, spectrum analyser and/or measurement receiver can be selected under the configuration, tests site pull down menu. | ||
The measuring device can be configured for the desired test through the Attenuator, RBW, VBW and sweep time settings. | The measuring device can be configured for the desired test through the Attenuator, RBW, VBW and sweep time settings. | ||
Measurements can be performed in peak, average and quasi peak mode. For Peak mode tests, the number of sweeps to be made, can be selected. | Measurements can be performed in peak, average and quasi peak mode. For Peak mode tests, the number of sweeps to be made, can be selected. | ||
| Line 47: | Line 48: | ||
In this window the operator can select which lines have to be measured. The operator can either select to measure just one LINE at a time (and have the possiblity to put them in a sequence) or select two or more lines in the same test. | In this window the operator can select which lines have to be measured. The operator can either select to measure just one LINE at a time (and have the possiblity to put them in a sequence) or select two or more lines in the same test. | ||
When two or more LINES are selected in the same test, all data will be displayed in one graph thus | When two or more LINES are selected in the same test, all data will be displayed in one graph thus losing information. However the peak list table will show on which LINE a peak was measured. | ||
If the emission graphs of all separate LINES are required, separate tests must be configured for each LINE. These tests can be put in a sequence, eliminating the need for a user intervention. | If the emission graphs of all separate LINES are required, separate tests must be configured for each LINE. These tests can be put in a sequence, eliminating the need for a user intervention. | ||
| Line 61: | Line 62: | ||
{{ScreenElementDescription|Measure time|Sets the measure time of the receiver.}} | {{ScreenElementDescription|Measure time|Sets the measure time of the receiver.}} | ||
When a parameter is set to auto, {{RadiMation}} will determine the most | When a parameter is set to auto, {{RadiMation}} will determine the most optimal settings for this parameter. If the parameter is set to coupled, the parameter will be set to automatic internally in the receiver. | ||
===Max. hold measurements=== | ===Max. hold measurements=== | ||
In this window the number | In this window the number of peak hold sweeps for each measurement can be selected. If this value is set to, for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweepspeed. For spectrum analysers and CW signals you can use a relatively low amount of peaksweeps, for example 10. But when you have discontinuous signals like a spark bridge, you may need 100 peaksweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope. | ||
===Traces window=== | ===Traces window=== | ||
| Line 72: | Line 73: | ||
===Detectors window=== | ===Detectors window=== | ||
In the detectors window the test engineer can configure the detectors used during the final emission measurements of the detected peaks. The first value is the time constant of the detector while the second value is used for the measure time at each frequency point for this detector. The time constant is send to the receiver, to set the sample time of the detector. To select a detector, check the desired box. | |||
In the detectors window the test engineer can configure the detectors used during the emission measurements. The first value is the time constant of the detector while the second value is used for the measure time at each frequency point for this detector. | |||
===Peak detection window=== | ===Peak detection window=== | ||
| Line 85: | Line 85: | ||
{{ScreenElementDescription|Test equipment|In the test site window the test engineer can select which equipment list will be used during this test.}} | {{ScreenElementDescription|Test equipment|In the test site window the test engineer can select which equipment list will be used during this test.}} | ||
{{ScreenElementDescription|Test engineer|In the test engineer window, the test engineer can select | {{ScreenElementDescription|Test engineer|In the test engineer window, the test engineer can select their own name. The test engineer name will be stored by the test results.}} | ||
{{ScreenElementDescription|Equipment icon|By clicking on the equipment icon, the equipment list can be viewed and edited.}} | {{ScreenElementDescription|Equipment icon|By clicking on the equipment icon, the equipment list can be viewed and edited.}} | ||
| Line 95: | Line 95: | ||
===Test options window=== | ===Test options window=== | ||
In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full | In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full traceability, all test data should be stored. | ||
===Test site information=== | ===Test site information=== | ||
[[Image:Conducted Emission Test Setup LISN.png]] | [[Image:Conducted Emission Test Setup LISN.png]] | ||
{| | {|class="wikitable" | ||
!Device name || Tab in {{RadiMation}} configuration window | ! # ||Device name || Tab in {{RadiMation}} configuration window | ||
|Spectrum analyser || | | ||Spectrum analyser || devices 2 | ||
|Pre amplifier || | | ||Pre amplifier || devices 2 | ||
| | | || LISN || devices 2 | ||
! | !colspan="3" |'''Cables''' | ||
|1 || | |1 ||cable Preamp -> analyser || Cables | ||
|2 || | |2 || cable antenna -> Preamp || Cables | ||
{{Note|Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during emission | {{Note|Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during emission measurements. When in doubt contact your reseller.}} | ||
measurements. | |||
When in doubt contact your reseller. | |||
==Absorbing clamp measurements== | ==Absorbing clamp measurements== | ||
Absorbing clamp measurements can be performed | Absorbing clamp measurements can be performed fully automatically when a clamp positioner is used. The test engineer can configure a complete absorbing clamp test in the Conducted Emission-Absorbing clamp configuration menu. To enter this menu, please select | ||
{{menu|Tests|Conducted emission|Absorbing clamp}} | {{menu|Tests|Conducted emission|Absorbing clamp}} | ||
| Line 151: | Line 143: | ||
{{ScreenElementDescription|Max. distance|Sets the maximum clamp position.}} | {{ScreenElementDescription|Max. distance|Sets the maximum clamp position.}} | ||
{{ScreenElementDescription|Min distance|Sets the minimum clamp position.}} | {{ScreenElementDescription|Min. distance|Sets the minimum clamp position.}} | ||
{{ScreenElementDescription|#|Number | {{ScreenElementDescription|#|Number of locations from {{ScreenElement|Max. distance}} to {{ScreenElement|Min. distance}} that should be measured.}} | ||
{{ScreenElementDescription|Peak search|The software will automatically detect on which | {{ScreenElementDescription|Peak search|The software will automatically detect on which clamp distance the maximum measurement value of each peak is present.}} | ||
When | When {{ScreenElement|Peak search}} is selected, the user can configure the method that is used during the peak search. In the peak search configuration window, the following configurations can be made: | ||
In the peak search configuration window, the following configurations can be made: | |||
{{ScreenElementDescription|Area|Range over which the peak search will be performed | {{ScreenElementDescription|Area|Range over which the peak search will be performed. The initially detected clamp distance of the peak will be the middle of this {{ScreenElement|Area}}, and the total search area for this peak will be half of this area up and half of this area down.}} | ||
{{ScreenElementDescription|Step size|The step size for each clamp movement.}} | {{ScreenElementDescription|Step size|The step size for each clamp movement.}} | ||
{{ScreenElementDescription|Threshold|This value | {{ScreenElementDescription|Threshold|This value is only used during the {{ScreenElement|Intelligent}} mode. It influences how long the movement is continued. The movement will stop when the measurement value is lower than the specified {{ScreenElement|Threshold}} below the highest known measurement value of the searched peak.}} | ||
{{ScreenElementDescription|Mode|The mode setting determines the way a peak is detected. | {{ScreenElementDescription|Mode|The mode setting determines the way a peak is detected. The following modes are available: | ||
The following modes are available: | * {{ScreenElement|Intelligent}} | ||
*Intelligent | * {{ScreenElement|Complete area}} | ||
*Complete area | |||
The Intelligent mode has the advantage that it stops with the scan as soon as a maximum measurement value has been determined. This could improve overall testtime as less movement of the clamp is possible. The Complete area mode has the advantage that the complete specified {{ScreenElement|Area}} is measured. | |||
====Intelligent==== | |||
In this mode the software increases the distance of the clamp from the initial point with a step equal to {{ScreenElement|Step size}} as long as the measurement value is not more than {{ScreenElement|Threshold}} below the already known maximum measurement value. As soon as a measurement value is detected that is {{ScreenElement|Threshold}} below the already known maximum value, the clamp movement is reversed until the initial point and from this initial point the distance of the clamp is decreased with a step equal to {{ScreenElement|Step size}} until a measurement value is measured that is {{ScreenElement|Threshold}} below the already known maximum measurement value. The step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak. | |||
{{Warning|This might cause a measurement error if the initial number of steps ({{ScreenElement|#}}) is set to a low value.}} | |||
====Complete area==== | |||
In this mode the complete area around the initial point and defined by {{ScreenElement|Area}} is scanned for a peak. The clamp is moved to one end of the to be measured area, and is then moved with the specified {{ScreenElement|Step size}} to the other end of the to be measured area. On each step the measurement value is determined, and the step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak. | |||
{{Warning|This might cause a measurement error if the {{ScreenElement|Area}} is set to a smaller value than the original stepsize (given by the stop position minus start position divided by the number of steps).}} | |||
| Line 179: | Line 182: | ||
| Line 190: | Line 193: | ||
| Line 204: | Line 205: | ||
| Line 214: | Line 215: | ||
| Line 220: | Line 221: | ||
[[Image:Conducted Emission Test Setup Absorbing Clamp.png]] | [[Image:Conducted Emission Test Setup Absorbing Clamp.png]] | ||
!Device name || Tab in {{RadiMation}} configuration window | !# ||Device name || Tab in {{RadiMation}} configuration window | ||
|Pre amplifier || | | || Pre amplifier || devices 2 | ||
|Current sensor || | | || Current sensor || devices 2 | ||
|1 || | |1 || cable preamp -> analyser || Cables | ||
==Current Probe Measurement== | |||
The test engineer can perform a conducted emission test in the conducted emission configuration Menu. To enter this menu, please select Tests > Conducted Emission -> Current probe | |||
[[File:CurrentProbeMeasurement.png]] | |||
===Frequency Window=== | |||
In the frequency window, the operator can enter a frequency range to be measured or select a frequency list. | |||
When the Frequency range is selected, the following items can be entered: | |||
{{ScreenElementDescription|Start|Sets start frequency}} | |||
{{ScreenElementDescription|Stop|Sets stop frequency}} | |||
When the frequency list is selected the operator can select a frequency list file by browsing to the file location. With the start-stop frequency the software searches for the peaks, while with a frequency list the frequencies on the list are measured. | |||
===Max. Hold Measurement/Peak Sweeps=== | |||
The number of peak hold sweeps for each measurement can be selected. If this value is set to, for example 10, the software will measure the frequency band 10 times. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweep speed. For spectrum analysers and CW signals you can use a relatively low amount of peak sweeps, for example 10. But when you have discontinuous signals like a spark bridge, you may need 100 peak sweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope. | |||
===Peak detection window=== | |||
{{ScreenElementDescription|All peaks above “xx” below the limit lines|In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 6 dB, the software will detect all peaks, which are higher than 6 dB below the limit line.}} | |||
{{ScreenElementDescription|Max. Peaks|With the max. peaks option, the user can limit the number of peak signals the software will detect and re-measure.}} | |||
===Receiver Settings=== | |||
{{ScreenElementDescription|Reference level|Sets the reference level of the receiver}} | |||
{{ScreenElementDescription|Attenuator|Sets the receiver attenuator level}} | |||
{{ScreenElementDescription|RBW|Sets the Resolution BandWidth filter}} | |||
{{ScreenElementDescription|VBW|Sets the Video BandWidth filter}} | |||
{{ScreenElementDescription|Sweep|Sets the sweep time of the receiver}} | |||
{{ScreenElementDescription|Pre-amp|Sets the pre-amp gain}} | |||
{{ScreenElementDescription|Step frequency|Sets the frequency step size of the receiver}} | |||
{{ScreenElementDescription|Measure time|Sets the measure time of the receiver}} | |||
When a parameter is set to auto, {{RadiMation}} will determine the most optimal settings for this parameter. If the parameter is set to coupled, the parameter will be set to automatic internally in the receiver. | |||
However, using the 'Auto' setting for the attenuator can let the analyser decides to use a different attenuator setting during (re-)measuring of a peak, which as a result can lead to measuring a higher Quasi-Peak value than the Peak value. This is related to the used analyser when auto attenuator is used. If it is not desired that the analyser can decide to switch to a different attenuator setting, then this should be disabled by the test engineer in {{RadiMation}} by using a fixed attenuator value. | |||
{{Warning| All these settings are generic. Therefore, always verify that the configured settings are also supported by the used hardware to ensure it is correctly used and no measurement faults are introduced. Use a fixed value to disabled interference of equipment intelligence.}} | |||
===Traces=== | |||
When | In the traces window, the test engineer can select a maximum of four traces, which can be measured and displayed at the same time | ||
===Test Options=== | |||
{{ScreenElementDescription|Record Data|Select to record all raw data or only the results. When choosing only the results you will only have the data from the found peaks and the data of the final measurement when this is performed}} | |||
{{ScreenElementDescription|Narrow/Broad Band Discrimination|Selects the method used to determine if a signal is broad band or narrow band. There are three options: Generic (The standard of your analyser)/CISPR 25 (Broad/Narrow band discrimination according to the CISPR 25 standard)/MilStandard(Broad/Narrow band discrimination according to the MilStandard}} | |||
===Limit lines window=== | |||
In the limit line window the operator can select one or more limit lines to be used during the test. | |||
===Test site window=== | |||
{{ScreenElementDescription|Test equipment|In the test site window the test engineer can select which equipment list will be used during this test.}} | |||
{{ScreenElementDescription|Test engineer|In the test engineer window, the test engineer can select their own name. The test engineer name will be stored by the test results.}} | |||
==Starting a test== | ==Starting a test== | ||
| Line 273: | Line 327: | ||
==Manual mode== | ==Manual mode== | ||
For product improvements, it | For product improvements, it quite often is required to operate the test setup in manual mode. This can be done either by operating the measurement receiver (or spectrum analyser) manually or by the {{RadiMation}} software. In our opinion it is no use to take over the buttons of the receiver without adding any extra functionality. For this reason a number of handy features are included in manual mode. | ||
| Line 288: | Line 342: | ||
{{ScreenElementDescription|Pre-amp|Sets the pre-amp gain.}} | {{ScreenElementDescription|Pre-amp|Sets the pre-amp gain.}} | ||
The left column in the receiver settings menu shows the parameters as configured by the test engineer while the right column shows the actual receiver configuration. | The left column in the receiver settings menu shows the parameters as configured by the test engineer while the right column shows the actual receiver configuration. | ||
| Line 318: | Line 372: | ||
====Traces==== | ====Traces==== | ||
In the “traces” menu, a maximum | In the “traces” menu, a maximum of four traces can be switched on. Each trace can be appointed to be a Max hold trace, Average trace, Write trace or be switched either to View mode or just left Blank. | ||
The figure below shows the traces menu: | The figure below shows the traces menu: | ||
[[Image:Conducted Emission Manual Mode Traces Tab.png]] | [[Image:Conducted Emission Manual Mode Traces Tab.png]] | ||
====Detectors==== | ====Detectors==== | ||
| Line 339: | Line 393: | ||
====Zoom function==== | ====Zoom function==== | ||
When the operator wants to “zoom in” on a certain signal just use the mouse to drag a box around the signal of interest. The {{RadiMation}} software will fully | When the operator wants to “zoom in” on a certain signal just use the mouse to drag a box around the signal of interest. The {{RadiMation}} software will fully automatically zoom in to this signal. The screen, as well as the measurement receiver will be zoomed. | ||
This reduces the number of actions for the operator significantly. | This reduces the number of actions for the operator significantly. | ||
| Line 366: | Line 420: | ||
===Short cut keys=== | ===Short cut keys=== | ||
The following short cut keys can be used during manual mode. | The following short cut keys can be used during manual mode. | ||
{| | {| class="wikitable" | ||
!Key || Function | !Key || Function | ||
Latest revision as of 16:10, 6 February 2025
Conducted Emission[edit]
| Warning: | Extreme care should be taken when the LISN measuring output is switched from one phase to the other. The measurement output of the LISN is capacitive coupled to the mains line and therefore switching between Phase and Neutral will cause a voltage transient of 310 V at the input of the receiver. When no extra protection is used, this transient can easily destroy the mixer diode(s) of the measuring device.
LISN measurements[edit]
Frequency window[edit]
| Sets start frequency |
| Sets stop frequency |
LISN line window[edit]
Receiver settings[edit]
| Sets the pre-amp gain. |
Max. hold measurements[edit]
Traces window[edit]
Detectors window[edit]
Peak detection window[edit]
| Under max. peaks, the user can limit the number of peak signals the software will detect and re-measure. |
Test site window[edit]
Limit lines window[edit]
Test options window[edit]
Test site information[edit]
| Spectrum analyser | devices 2 | |
| Pre amplifier | devices 2 | |
| LISN | devices 2 | |
| Cables | ||
| 2 | cable antenna -> Preamp | Cables |
| Note: | Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during emission measurements. When in doubt contact your reseller. |
Tests
Conducted emission
Absorbing clamp
| Sets start frequency. |
| Sets stop frequency. |
Clamp movement window[edit]
| Sets the maximum clamp position. |
| Sets the minimum clamp position. |
| Number of locations from Max. distance to Min. distance that should be measured. |
| The software will automatically detect on which clamp distance the maximum measurement value of each peak is present. |
| Range over which the peak search will be performed. The initially detected clamp distance of the peak will be the middle of this Area, and the total search area for this peak will be half of this area up and half of this area down. |
| The step size for each clamp movement. |
Intelligent[edit]
In this mode the software increases the distance of the clamp from the initial point with a step equal to Step size as long as the measurement value is not more than Threshold below the already known maximum measurement value. As soon as a measurement value is detected that is Threshold below the already known maximum value, the clamp movement is reversed until the initial point and from this initial point the distance of the clamp is decreased with a step equal to Step size until a measurement value is measured that is Threshold below the already known maximum measurement value. The step on which the highest measurement value is measured, will be the optimum peak search location that is found for that peak.
Complete area[edit]
Traces window[edit]
| Current sensor | devices 2 | |
| Cables | ||
| 1 | cable preamp -> analyser | Cables |
Frequency Window[edit]
Receiver Settings[edit]
| Sets the pre-amp gain |
When a parameter is set to auto, RadiMation® will determine the most optimal settings for this parameter. If the parameter is set to coupled, the parameter will be set to automatic internally in the receiver. However, using the 'Auto' setting for the attenuator can let the analyser decides to use a different attenuator setting during (re-)measuring of a peak, which as a result can lead to measuring a higher Quasi-Peak value than the Peak value. This is related to the used analyser when auto attenuator is used. If it is not desired that the analyser can decide to switch to a different attenuator setting, then this should be disabled by the test engineer in RadiMation® by using a fixed attenuator value.
| Warning: | All these settings are generic. Therefore, always verify that the configured settings are also supported by the used hardware to ensure it is correctly used and no measurement faults are introduced. Use a fixed value to disabled interference of equipment intelligence. |
Traces[edit]
Test Options[edit]
| Select to record all raw data or only the results. When choosing only the results you will only have the data from the found peaks and the data of the final measurement when this is performed |
| Selects the method used to determine if a signal is broad band or narrow band. There are three options: Generic (The standard of your analyser)/CISPR 25 (Broad/Narrow band discrimination according to the CISPR 25 standard)/MilStandard(Broad/Narrow band discrimination according to the MilStandard |
Starting a test[edit]
Loading a TSF file[edit]
Tests
Conducted emission
CISPR LISN
Manual mode[edit]
For product improvements, it quite often is required to operate the test setup in manual mode. This can be done either by operating the measurement receiver (or spectrum analyser) manually or by the RadiMation® software. In our opinion it is no use to take over the buttons of the receiver without adding any extra functionality. For this reason a number of handy features are included in manual mode.
Data correction[edit]
[edit]
[edit]
Traces[edit]
Detectors[edit]
- Peak detector
- Average detector
- Quasi peak detector
- RMS detector
LISN line selection[edit]
Zoom function[edit]
When the operator wants to “zoom in” on a certain signal just use the mouse to drag a box around the signal of interest. The RadiMation® software will fully automatically zoom in to this signal. The screen, as well as the measurement receiver will be zoomed. This reduces the number of actions for the operator significantly.
Limit lines[edit]
Short cut keys[edit]
| Key | Function |
| F3 | Turn engine of the clampmover left, full speed. |
| F4 | Turn engine of the clampmover left, quarter speed. |
| F5 | Turn engine of the clampmover right, quarter speed. |
| F6 | Turn engine of the clampmover right, full speed. |
| F7 | Frequency offset down. |
| F8 | Frequency offset up. |
| F9 | Center frequency down. |
| F10 | Center frequency up. |
[edit]
Merge function[edit]
X+k*S calculation[edit]
