===Detectors window===
In the detectors window the test engineer can configure the detectors used during the emission measurements. The first value is the time constant of the detector while the second value is used for the measure time at each frequency point for this detector.
===Peak detection window===
Line 102: Line 101:
{|class="wikitable"
{|class="wikitable"
!Device name || Tab in {{RadiMation}} configuration window
! # ||Device name || Tab in {{RadiMation}} configuration window
|Spectrum analyser || devices 2
|  ||Spectrum analyser || devices 2
|Pre amplifier || devices 2
|  ||Pre amplifier || devices 2
|LISN || devices 2
|  || LISN || devices 2
{|class="wikitable"
|+'''Cables'''
!# || Device name || Tab in {{RadiMation}} configuration window
!colspan="3" |'''Cables'''
|1 ||cable Preamp -> analyser || Cables
Line 198: Line 193:
Line 229: Line 222:
{|class="wikitable"
{|class="wikitable"
!#  ||Device name || Tab in {{RadiMation}} configuration window
|  || Pre amplifier || devices 2
|Current sensor || devices 2
|  || Current sensor || devices 2
{|class="wikitable"
|+'''Cables'''
|1 || cable preamp -> analyser || Cables
Line 288: Line 277:
When a parameter is set to auto, {{RadiMation}} will determine the most optimum settings for this parameter. If the parameter is set to coupled, the parameter will be set to automatic internally in the receiver.
However, using the 'Auto' setting for the attenuator can let the analyzer decides to use a different attenuator setting during (re-)measuring of a peak, which as a result can lead to measuring a higher Quasi-Peak value than the Peak value. This is related to the used analyser when auto attenuator is used. If it is not desired that the analyzer can decide to switch to a different attenuator setting, then this should be disabled by the test engineer in {{RadiMation}} by using a fixed attenuator value.
However, using the 'Auto' setting for the attenuator can let the analyser decides to use a different attenuator setting during (re-)measuring of a peak, which as a result can lead to measuring a higher Quasi-Peak value than the Peak value. This is related to the used analyser when auto attenuator is used. If it is not desired that the analyser can decide to switch to a different attenuator setting, then this should be disabled by the test engineer in {{RadiMation}} by using a fixed attenuator value.
{{Warning| All these settings are generic. Therefore, always verify that the configured settings are also supported by the used hardware to ensure it is correctly used and no measurement faults are introduced. Use a fixed value to disabled interference of equipment intelligence.}}
Line 431: Line 420:
===Short cut keys===
{| class="wikitable"
!Key || Function
!Key || Function
Latest revision as of 09:04, 3 January 2020
Conducted Emission
Introduction
Warning: Extreme care should be taken when the LISN measuring output is switched from one phase to the other. The measurement output of the LISN is capacitive coupled to the mains line and therefore switching between Phase and Neutral will cause a voltage transient of 310 V at the input of the receiver. When no extra protection is used, this transient can easily destroy the mixer diode(s) of the measuring device.
We recommend to use a transient limiter or to disconnect the receiving device when the LISN is switched on or off. The used LISN, spectrum analyser and/or measurement receiver can be selected under the configuration, tests site pull down menu.
LISN measurements
Frequency window
ScreenElementDescription.png Start Sets start frequency
ScreenElementDescription.png Stop Sets stop frequency
LISN line window
Receiver settings
ScreenElementDescription.png Reference level Sets the reference level of the receiver.
ScreenElementDescription.png Attenuator Sets the receiver attenuator level.
ScreenElementDescription.png RBW Sets the Resolution BandWidth filter.
ScreenElementDescription.png VBW Sets the Video BandWidth filter.
ScreenElementDescription.png Sweep Sets the sweep time of the receiver.
ScreenElementDescription.png Pre-amp Sets the pre-amp gain.
ScreenElementDescription.png Step frequency Sets the frequency step size of the receiver.
ScreenElementDescription.png Measure time Sets the measure time of the receiver.
Max. hold measurements
Traces window
Detectors window
ScreenElementDescription.png All peaks above “xx” below the limit lines In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 6 dB, the software will detect all peaks, which are higher than 6 dB below the limit line.
ScreenElementDescription.png Max. Peaks Under max. peaks, the user can limit the number of peak signals the software will detect and re-measure.
Test site window
ScreenElementDescription.png Test equipment In the test site window the test engineer can select which equipment list will be used during this test.
ScreenElementDescription.png Test engineer In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results.
ScreenElementDescription.png Equipment icon By clicking on the equipment icon, the equipment list can be viewed and edited.
Limit lines window
Test options window
Spectrum analyser devices 2
Pre amplifier devices 2
LISN devices 2
Cables
1 cable Preamp -> analyser Cables
2 cable antenna -> Preamp Cables
Note: Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during emission measurements. When in doubt contact your reseller.
Absorbing clamp measurements
   Menu.png Tests
      Menu.png Conducted emission
         Menu.png Absorbing clamp
Frequency window
ScreenElementDescription.png Start Sets start frequency.
ScreenElementDescription.png Stop Sets stop frequency.
Clamp movement window
ScreenElementDescription.png Max. distance Sets the maximum clamp position.
ScreenElementDescription.png Min. distance Sets the minimum clamp position.
ScreenElementDescription.png # Number op locations from Max. distance to Min. distance that should be measured.
ScreenElementDescription.png Peak search The software will automatically detect on which clamp distance the maximum measurement value of each peak is present.
ScreenElementDescription.png Area Range over which the peak search will be performed. The initially detected clamp distance of the peak will be the middle of this Area, and the total search area for this peak will be half of this area up and half of this area down.
ScreenElementDescription.png Step size The step size for each clamp movement.
ScreenElementDescription.png Threshold This value is only used during the Intelligent mode. It influences how long the movement is continued. The movement will stop when the measurement value is lower than the specified Threshold below the highest known measurement value of the searched peak.
ScreenElementDescription.png Mode The mode setting determines the way a peak is detected. The following modes are available:
  • Intelligent
  • Complete area
The Intelligent mode has the advantage that it stops with the scan as soon as a maximum measurement value has been determined. This could improve overall testtime as less movement of the clamp is possible. The Complete area mode has the advantage that the complete specified Area is measured.
Intelligent
Warning: This might cause a measurement error if the intial number of steps (#) is set to a low value.
Complete area
Warning: This might cause a measurement error if the Area is set to a smaller value than the original stepsize (given by the stop position minus start position divided by the number of steps).
Receiver settings
Traces window
Detectors window
Test site window
Limit lines window
Test options window
Current sensor devices 2
Cables
1 cable preamp -> analyser Cables
Current Probe Measurement
Frequency Window
Max. Hold Measurement/Peak Sweeps
The number op peak hold sweeps for each measurement can be selected. If this value is set to, for example 10, the software will measure the frequency band 10 times. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweep speed. For spectrum analysers and CW signals you can use a relatively low amount of peak sweeps, for example 10. But when you have discontinues signals like a spark bridge, you may need 100 peak sweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope.
ScreenElementDescription.png Max. Peaks With the max. peaks option, the user can limit the number of peak signals the software will detect and re-measure.
Receiver Settings
Warning: All these settings are generic. Therefore, always verify that the configured settings are also supported by the used hardware to ensure it is correctly used and no measurement faults are introduced. Use a fixed value to disabled interference of equipment intelligence.
Traces
Test Options
ScreenElementDescription.png Record Data Select to record all raw data or only the results. When choosing only the results you will only have the data from the found peaks and the data of the final measurement when this is performed
ScreenElementDescription.png Narrow/Broad Band Discrimination Selects the method used to determine if a signal is broad band or narrow band. There are three options: Generic (The standard of your analyser)/CISPR 25 (Broad/Narrow band discrimination according to the CISPR 25 standard)/MilStandard(Broad/Narrow band discrimination according to the MilStandard
Limit lines window
Test site window
Starting a test
Loading a TSF file
   Menu.png Tests
         Menu.png CISPR LISN
  1. The presence of a valid software key is checked.
  2. The software settings are checked (i.e. is all test equipment used within its parameters?)
  3. The presence of all test equipment is checked.
  4. All relevant equipment will be initialized.
Manual mode
Receiver settings
ScreenElementDescription.png Reference Sets the reference level of the receiver.
ScreenElementDescription.png RBW Sets the Resolution Band Width filter.
ScreenElementDescription.png VBW Sets the Video Band Width filter.
Data correction
Peak menu
ScreenElementDescription.png Search peak RadiMation® can detect peaks automatically. From the “peaks” TAB, choose “Search peak” to detect the maximum signal. The software will measure the detected peak.
ScreenElementDescription.png Add peak The measured values can be added to the peak list by pressing the “Add Peak” button.
ScreenElementDescription.png Search angle By pressing the “Search Angle” button, the software starts a turntable scan to determine at which turn table angle maximum emission takes place.
ScreenElementDescription.png Search Height By pressing the “Search Height” button, the software starts an antenna height scan to determine at which antenna height maximum emission takes place.
Traces menu
Traces
Detectors
  • Peak detector
  • Average detector
  • Quasi peak detector
  • RMS detector
LISN line selection
Zoom function
Limit lines
Short cut keys
Key Function
F3 Turn engine of the clampmover left, full speed.
F4 Turn engine of the clampmover left, quarter speed.
F5 Turn engine of the clampmover right, quarter speed.
F6 Turn engine of the clampmover right, full speed.
F7 Frequency offset down.
F8 Frequency offset up.
F9 Center frequency down.
F10 Center frequency up.
Export menu
Merge function
X+k*S calculation