When a Radiated Immunity [[Chapter_12#One-point_calibration|1 point calibration]] or a Radiated Immunity Fixed Field test is done, which used the measurement value of the field probe to regulate the test level, the regulation is done on the average of the field probes that are measured. It is thus required that at least one electrical field strength probe is selected to be included in the calculation of the average field. | |||
This error happens when you are doing a Radiated Immunity calibration and do not have at least one averaging probe selected | This error happens when you are doing a Radiated Immunity calibration or a Radiated Immunity Fixed Field and do not have at least one averaging probe selected. | ||
For an explanation on how to select the average probes see: [[Chapter_12#One-point_calibration| 1-point calibration]]. | |||
== Solution == | == Solution == | ||
Re-open your Radiated Immunity calibration | * Re-open your Radiated Immunity 1 point calibration and go to 'Field Probes' and make sure that at least one of the probes has the averaging checkbox checked. | ||
* Re-open your Radiated Immunity Fixed Field configuration and go to 'Inputs' and to 'Field Probes' and make sure that at least one of the probes has the averaging checkbox checked. | |||
[[Category:RadiMation Error]] | [[Category:RadiMation Error]] | ||
[[Category:RadiMation]] | [[Category:RadiMation]] | ||
Latest revision as of 09:32, 19 January 2022
There is no average probe selected[edit]
When a Radiated Immunity 1 point calibration or a Radiated Immunity Fixed Field test is done, which used the measurement value of the field probe to regulate the test level, the regulation is done on the average of the field probes that are measured. It is thus required that at least one electrical field strength probe is selected to be included in the calculation of the average field.
For an explanation on how to select the average probes see: 1-point calibration.
Solution[edit]
- Re-open your Radiated Immunity 1 point calibration and go to 'Field Probes' and make sure that at least one of the probes has the averaging checkbox checked.