Keysight Technologies N9038A General: Difference between revisions
(3 intermediate revisions by the same user not shown) | |||
Line 11: | Line 11: | ||
The N9000-series analysers have an option to perform the alignment automatically when it detects that an alignment is needed. It is however undesired to perform the alignment during the actual test, as this cause a period of time that no measurements can be performed, and it causes differences in the amplitude measurement within the same test, before and after the alignment. | The N9000-series analysers have an option to perform the alignment automatically when it detects that an alignment is needed. It is however undesired to perform the alignment during the actual test, as this cause a period of time that no measurements can be performed, and it causes differences in the amplitude measurement within the same test, before and after the alignment. | ||
To prevent that the alignment happens automatically during the actual emission test, the {{RadiMation}} device driver for the N9000-series analyser uses the following steps: | To prevent that the alignment happens automatically during the actual emission test, the {{RadiMation}} device driver for the N9000-series analyser uses the following steps: | ||
# A check is done to determine if an alignment is needed on the instrument. If this is the case, the error 'Device is not aligned. Please align device and restart the measurement' is raised. | # A check is done to determine if an alignment is needed on the instrument. If this is the case, the error [[RadiMation_Error:_805315268|'Device is not aligned. Please align device and restart the measurement']] is raised. | ||
# The auto alignment is disabled | # The auto alignment is disabled | ||
# The actual emission measurement is performed | # The actual emission measurement is performed | ||
# At the end of the measurement when the used equipment is being deinitialised, the automatic alignment is activated again. | # At the end of the measurement when the used equipment is being deinitialised, the automatic alignment is activated again. | ||
== External mixer == | |||
If option EXM is present in the {{PAGENAME}}, it is possible to connect and use an external mixer to the {{PAGENAME}}. The mixer has to be connected to the 'IF Input/1s LO Output' port of the {{PAGENAME}}, and desired configuration and use of the mixer can be activated and configured on the display of the {{PAGENAME}} itself. Also see chapter 6.2.13 of the [https://www.keysight.com/nl/en/assets/9018-04666/user-manuals/9018-04666.pdf Keysight X-series Signal Analyzers manual] for the details on how this should be configured on the {{PAGENAME}}. | |||
Depending on the desired use case, it could help to also activate the 'Signal ID' function and use the 'Image Suppress' setting as the 'Signal ID Mode', to suppress signal images. | |||
When RadiMation is controlling the {{PAGENAME}}, the analyser is preset, however the configuration of the external mixer input will remain selected. Thus manually configuring the external mixer on the {{PAGENAME}} before RadiMation is started, will also allow RadiMation to use the mixer. The only change in the configuration of RadiMation that is needed, is to modify the frequency range of the device driver to match the actually measured frequency range. This can be done on {{ScreenElement|Frequency range}} tab of the advanced configuration dialog. | |||
If for example the '[https://www.keysight.com/nl/en/product/M1971V/waveguide-harmonic-mixer-smart-mixer-50-75-ghz.html Keysight USB-M1971E-001 V-Band]' mixer is connected, the {{ScreenElement|Start frequency}} and {{ScreenElement|Stop frequency}} of the device driver should be modified to respectively 50 GHz and 75 GHz. Of course also the used antenna should be matching with the to be measured frequency range. In the TSF, the actually measured frequency range (50 GHz - 75 GHz) can be configured, and this will control the {{PAGENAME}} to perform the correct measurement. | |||
== 'Advanced' configuration == | == 'Advanced' configuration == | ||
Line 38: | Line 46: | ||
{{ScreenElementDescription|Has UKB Option|Specifies if the UKB option is present in the device, which will allow to perform measurements on a lower frequency range.}} | {{ScreenElementDescription|Has UKB Option|Specifies if the UKB option is present in the device, which will allow to perform measurements on a lower frequency range.}} | ||
{{:DeviceDriverResolutionTab}} | |||
{{:DeviceDriverCommunicationTab}} |
Latest revision as of 08:14, 29 May 2024
Warning: | At least firmware version A.12.09 should be installed on the Keysight Technologies N9038A General, otherwise the preamplifier cannot be controlled correctly in EMI receiver mode. |
Alignment[edit]
- A check is done to determine if an alignment is needed on the instrument. If this is the case, the error 'Device is not aligned. Please align device and restart the measurement' is raised.
- The auto alignment is disabled
- The actual emission measurement is performed
- At the end of the measurement when the used equipment is being deinitialised, the automatic alignment is activated again.
External mixer[edit]
If option EXM is present in the Keysight Technologies N9038A General, it is possible to connect and use an external mixer to the Keysight Technologies N9038A General. The mixer has to be connected to the 'IF Input/1s LO Output' port of the Keysight Technologies N9038A General, and desired configuration and use of the mixer can be activated and configured on the display of the Keysight Technologies N9038A General itself. Also see chapter 6.2.13 of the Keysight X-series Signal Analyzers manual for the details on how this should be configured on the Keysight Technologies N9038A General. Depending on the desired use case, it could help to also activate the 'Signal ID' function and use the 'Image Suppress' setting as the 'Signal ID Mode', to suppress signal images.
When RadiMation is controlling the Keysight Technologies N9038A General, the analyser is preset, however the configuration of the external mixer input will remain selected. Thus manually configuring the external mixer on the Keysight Technologies N9038A General before RadiMation is started, will also allow RadiMation to use the mixer. The only change in the configuration of RadiMation that is needed, is to modify the frequency range of the device driver to match the actually measured frequency range. This can be done on Frequency range tab of the advanced configuration dialog.
If for example the 'Keysight USB-M1971E-001 V-Band' mixer is connected, the Start frequency and Stop frequency of the device driver should be modified to respectively 50 GHz and 75 GHz. Of course also the used antenna should be matching with the to be measured frequency range. In the TSF, the actually measured frequency range (50 GHz - 75 GHz) can be configured, and this will control the Keysight Technologies N9038A General to perform the correct measurement.
'Advanced' configuration[edit]
The 'Advanced' configuration dialog of the Keysight Technologies N9038A General device driver has several tabs to configure the device driver.
Settings[edit]
Select peak detector | Specifies the standard that should be used for the peak detector. Possible standards are: 'EMC Standard' or 'MIL Standard'. |
Select input | Specifies the input that should be used. Depending the exact model used, it is possible that the selection of the input is disabled. |
Use preselector | If the preselector should be enabled or not. |
Use EMI detector when possible (more accurate, but slower) | Specifies if the EMI average or non-EMI average detector should be used. The EMI average detector is more accurate, but also slower than the non-EMI average detector. Depending on the available options in the measurement device it is possible to select the EMI average detector. When this option is activated and the EMI average detector is available, the EMI average detector will be selected. Otherwise the non-EMI average detector will be selected. |
Input[edit]
AC Coupling | Specifies that AC coupling of the input should be selected. |
DC Coupling | Specifies that DC coupling of the input should be selected. |
Has UKB Option | Specifies if the UKB option is present in the device, which will allow to perform measurements on a lower frequency range. |
The scan resolution of the Keysight Technologies N9038A General can be selected. The scan resolution is the number of points that will be present in a single sweep of the analyzer.
Scan resolution | Specify the number of points that should be used by the Keysight Technologies N9038A General for a sweep. It may be possible to select a known value from the dropdown list. |