RadiMation Application Note 155: Difference between revisions
(4 intermediate revisions by the same user not shown) | |||
Line 17: | Line 17: | ||
== Tips == | == Tips == | ||
{{Note|If the current unit is shown as | {{Note|If the current unit is shown as μA, it can be changed to dBμA by clicking {{ScreenElement|Units}}, select {{ScreenElement|Current}} and click {{ScreenElement|Edit}} to change this.}} | ||
== Define the test level == | == Define the test level == | ||
Line 24: | Line 24: | ||
First create a correction file in {{RadiMation}} and specify the test level (See 'FIGURE CS109-1. CS109 limit.'). | First create a correction file in {{RadiMation}} and specify the test level (See 'FIGURE CS109-1. CS109 limit.'). | ||
Ensure that the correction file has a {{ScreenElement|Frequency}} column and a {{ScreenElement|Current}} column with the unit set to | Ensure that the correction file has a {{ScreenElement|Frequency}} column and a {{ScreenElement|Current}} column with the unit set to dBμA. | ||
[[File:CS109_Test_level.png]] | [[File:CS109_Test_level.png]] | ||
Line 54: | Line 54: | ||
!colspan="4"|Cables | !colspan="4"|Cables | ||
|① || Cable current -> power meter || Cables || Cable | |① || Cable current -> power meter || Cables || Cable with a correction file specified for the cable loss | ||
=== Configure the EUT test === | === Configure the EUT test === | ||
To perform the | To perform the test on the EUT, create a conducted immunity multiband test. | ||
The conducted immunity multiband test, can be started by selecting from the menu: | The conducted immunity multiband test, can be started by selecting from the menu: | ||
{{Menu|Tests|Conducted Immunity|Multiband}} | {{Menu|Tests|Conducted Immunity|Multiband}} | ||
Line 94: | Line 94: | ||
The {{RadiMation}} Conducted Immunity multiband test can be used to perform the CS109 test with a | The {{RadiMation}} Conducted Immunity multiband test can be used to perform the CS109 test with a current closed loop method test level and a variable test level specified in a correction file. | ||
[[Category:RadiMation Application Note]] | [[Category:RadiMation Application Note]] |
Latest revision as of 08:48, 5 September 2024
How to perform a MIL-STD-461 CS109, Conducted susceptibility, structure current test[edit]
This application note explains how the MIL-STD-461 CS109, Conducted susceptibility, structure current test can be performed with RadiMation®
The exact requirements and test methods for the CS109 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to perform the CS109 test:
- Signal generator
- Amplifier
- Coupling Transformer
- Current sensor
- 0.5 Ohm resistor
- Isolation Transformer
Tips[edit]
Note: | If the current unit is shown as μA, it can be changed to dBμA by clicking Units, select Current and click Edit to change this. |
Define the test level[edit]
Create a correction file for the test level[edit]
- File
- New
- Correction
- New
- File
First create a correction file in RadiMation® and specify the test level (See 'FIGURE CS109-1. CS109 limit.').
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBμA.
- File
- Save correction
- File
This correction file with the test level will be used during EUT testing.
EUT Testing[edit]
EUT Testing equipment[edit]
Sensor powermeter | Devices 2 | The power meter or analyser to use for measuring the current | |
Current Sensor | Devices 2 | The current sensor to use with transfer factor attached to the driver | |
Injection device | Devices 2 | The injection device to use | |
Cables | |||
① | Cable current -> power meter | Cables | Cable with a correction file specified for the cable loss |
Configure the EUT test[edit]
To perform the test on the EUT, create a conducted immunity multiband test. The conducted immunity multiband test, can be started by selecting from the menu:
- Tests
- Conducted Immunity
- Multiband
- Tests
Start | The start frequency of the test. For example 30 Hz. |
End | The stop frequency of the test. For example 100 kHz. |
Frequency Step | The frequency step, see Table III. |
Modulation | CS109 does not specify a modulation to use. |
Test equipment | The equipment needed for the conducted immunity test. |
Location Type | Injection clamp. |
Test Level | Specify the current substitution method. Also add a "Measured Current Limit" |
At TestLevel click Add to add a new Test level and select TestLevel - Current closed loop method
Test level | Variable current, and specify the correction file for the applicable test level. |
When all settings are configured press Start Test to run the test.
EUT Test Result[edit]
The RadiMation® Conducted Immunity multiband test can be used to perform the CS109 test with a current closed loop method test level and a variable test level specified in a correction file.
