RadiMation Application Note 156: Difference between revisions
(50 intermediate revisions by 2 users not shown)
__NOTOC__
__NOTOC__
== How to perform a radiated emission substitution test ==
This application note explains how to perform a radiated emission substitution test with {{RadiMation}}, as for example mentioned in ETSI EN 300-320 C5.3 and C5.4
This application note explains how to perform a radiated emission substitution test with {{RadiMation}}, as for example mentioned in [[ETSI EN 300 328]] C5.3 and C5.4
The exact requirements and test methods for the ETSI EN 300-320 are specified in the [[ETSI EN 300-320]].
The exact requirements and test methods are specified in the [[ETSI EN 300 328]].
== Necessary equipment ==
The following devices are necessary to perform a radiated emission substitution test:
* '''Signal generator'''
* Signal generator
* '''Transmitting Antenna'''
* Transmitting Antenna
* '''Receiving Antenna'''
* Receiving Antenna
* '''FWD power meter'''
* Forward power meter
* '''Coupler''' (0 dB coupling factor)
* Coupler (0 dB coupling factor)
== General remarks ==
For higher quality and accuracy, it is still advised to use normal measurements practices like calibrating and zeroing the used powermeter.
* For higher quality and accuracy, it is still advised to use normal practices like powermeter zeroing and calibration
Cabling and system setup of all components from the receive antenna to the receiver that is used during Phase B, should be exactly the same during the normal emission measurements. Changes in cabling or antenna should result in a redetermination of the antenna correction.
* Cabling and system setup of all components from the receive antenna to the reciever that is used during Phase B, should be exactly the same during the normal emission measurements. Changes in cabling or antenna should result in a redetermination of the antenna correction.
== Performing the measurements ==
To perform the Radiated emission substitution measurements, the antenna correction factor of the test setup must first be determined.
There are three phases (A, B & C) to determine the correct antenna factor
This is done in three phases called (A, B & C).
=== Phase A: Determine the power being applied to the antenna===
==== Purpose: ====
==== Purpose ====
Purpose of this phase is to determine the power that is going into the transmit antenna during Phase B.
Very often an attenuator is used between the signal generator and the antenna, to reduce mismatches.
This phase is used to correct for cable, attenuator and signal generator losses.
==== Test set up ====
[[File:PHASE-A_setup.png]]
* '''A1'''. Build a setup with signal generator outside the chamber and the transmit antenna in the chamber.
* '''A2'''. Do not connect the transmit antenna yet, but connect the receiver on the location of the transmit antenna.
==== Set up: ====
* '''A3'''. Sweep with a RI Fixed power test 0 dBm signal power over the frequency range. The antenna selected in the testsite of this test is not relevant. A coupler with a 0 dB coupling factor should also be used.
* A1. Build a setup with signal generator outside the chamber and the transmit antenna in the chamber
* '''A4'''. Also measure the forward power using the receiver. The power values measured as the forward power meter are the power values that are applied to the transmit antenna during phase C.
* A2. Do not connect the transmit antenna yet, but connect the receiver on the location of the transmit antenna
==== Test configuration ====
* A3. Sweep with a RI Fixed power test 0 dBm signal power over the frequency range. The antenna selected in the testsite of this test is not relevant. A coupler with a 0 dB coupling factor should also be used
To perform the '''phase A''' test, create a radiated immunity multiband test.
The radiated immunity multiband test, can be started by selecting from the menu:
{{Menu|Tests|Radiated Immunity|Multiband}}
[[File:RI-FIXED-POWER-0dBm-SignalPower.png]]
* A4. Also measure the forward power using the receiver. The power values measured as the forward power meter are the power values that are applied to the transmit antenna during phase C
{{ScreenElementDescription|Start|The start frequency of the test. For example 1 GHz.}}
{{ScreenElementDescription|End|The stop frequency of the test. For example 6 GHz.}}
{{ScreenElementDescription|Frequency Step|The frequency step.}}
{{ScreenElementDescription|Dwell Time|The dwell time.}}
{{ScreenElementDescription|Frequency Change Mode|Constant.}}
{{ScreenElementDescription|Modulation|The modulation}}
{{ScreenElementDescription|Test equipment|The equipment needed for the test.}}
{{ScreenElementDescription|Location Type|Antenna height, distance, polarization, angle.}}
{{ScreenElementDescription|Test Level|Specify the power closed loop method test level.}}
{{ScreenElementDescription|Inputs|Specify the forward power measurement.}}
=== Phase B: Determine the overall transfer of the measurement system ===
==== Purpose: ====
==== Purpose ====
Purpose of this phase is to measure the overall transfer from the transmit antenna to the receiver.
[[File:Phase_B_setup.png]]
* '''B1'''. Build a setup with a signal generator outside the chamber and the transmit antenna in the chamber
==== Set up: ====
* '''B2'''. Place the receiver outside the chamber and connect it to the receive antenna in the chamber
* B1. Build a setup with a signal generator outside the chamber and the transmit antenna in the chamber
* '''B3'''. Sweep with a RI Fixed power test 0 dBm signal power over the frequency range. This can be the same test configuration as is used during step A3. The antenna selected in the test site of this test is not relevant. A coupler with a 0 dB coupling factor should also be used
* B2. Place the receiver outside the chamber and connect it to the receive antenna in the chamber
* '''B4'''. Measure the forward power using the receiver. The power values measured as the forward power meter will be used to determine the correct correction factor
* B3. Sweep with a RI Fixed power test 0 dBm signal power over the frequency range. This can be the same test configuration as is used during step A3. The antenna selected in the test site of this test is not relevant. A coupler with a 0 dB coupling factor should also be used
* B4. Measure the forward power using the receiver. The power values measured as the forward power meter will be used to determine the correct correction factor
To perform the '''phase B''' test, create a radiated immunity multiband test.
=== Phase C: Calculate the correct antenna correction factor ===
==== Purpose: ====
==== Purpose ====
Purpose of this phase is to use the measurement results obtained during phase A and B to determine the correction factor for the total system that can be used during an emission substitution measurement
Purpose of this phase is to use the measurement results obtained during '''phase A''' and '''B''' to determine the correction factor for the total system that can be used during an emission substitution measurement.
This measurement measures the difference between the transmitted power and the received power at the receiving antenna.
The chamber characteristics, which is the overall transfer factor of the chamber using the applicable measurement equipment, is represented by this difference.
==== Calculation: ====
==== Calculations ====
For each frequency use the following calculation:
  C = P_transmit + G_dipole - P_transfer
<math>C = P_{transmit} + G_{dipole} - P_{transfer}</math>
* '''C''': The antenna correction factor
* <math>C</math>: The antenna correction factor
* '''P_transmit''': The forward power during Phase A
* <math>P_{transmit}</math>: The forward power during '''Phase A'''
* '''G_dipole''': the Transmit Antenna Factor of the transmit antenna compared to a dipole
* <math>G_{dipole}</math>: the Transmit Antenna Factor of the transmit antenna compared to a dipole
* '''P_transfer''': The forward power during Phase B
* <math>P_{transfer}</math>: The forward power during '''Phase B'''
The '''G_dipole''' can be calculate from the calibrated Antenna Factor of the transmit antenna with:
The <math>G_{dipole}</math> can be calculate from the calibrated Antenna Factor of the transmit antenna with:
  G_dipole = 20 * log10(f) - AF - 29.79 - 2.15
<math>G_{dipole} = 20 * log^{10}(f) - AF - 29.79 - 2.15</math>
* <math>f</math>: The frequency, expressed in MHz
* <math>AF</math> The antenna factor of the antenna, expressed in dB/m
* '''f''': The frequency, expressed in MHz
* '''AF''': The antenna factor of the antenna, expressed in dB/m
(The factor 2.15 is to correct from an isotropic antenna to a dipole antenna)
==== Application: ====
==== Application ====
The calculated '''C''' can be used as the correction of the antenna during the real measurement, and can be placed in a correction file that is attached to the receiving antenna.
=== The measurement with the EUT ===
=== The EUT measurements ===
==== Purpose ====
Measuring the emission of the EUT using the antenna correction factors as determined in the previous phases.
==== Test Set up ====
[[File:EUT-Measurement.png]]
* On the location of the transmit antenna the EUT is positioned.
* The determined antenna correction factor (C) should be applied to the antenna that is used in the testsite during the measurement. No other corrections should be included in the testsite. Correction of the cables, preamplifiers and attenuators should not be done, because all these components were also present during phase B, and are thus already corrected for in the determined antenna correction factor (C).
[[File:ReceiveAntennaCorrection.png]]
Perform a multi-band radiated emission measurement, with the 'Antenna Spurious and harmonics' location type. The values that are measured are the correct Effective radiated power measurements.
* Perform a multi-band radiated emission measurement, with the 'Antenna Spurious and harmonics' location type. The values that are measured are the correct Effective radiated power measurements.
The radiated emission multi-band test, can be started by selecting from the menu:
{{Menu|Tests|Radiated emission|Multiband}}
[[File:RE-ANTENNA-SPURIOUS-AND-HARMONIC-OUTPUT.png]]
For more information regarding the radiated emission test configuration, please see: [https://wiki.radimation.com/wiki/index.php/Chapter_7#Multiband Radiated emission multiband test configuration]
== Conclusion ==
== Conclusion ==
Line 103: Line 157:
The provided formulae and guidelines offer a methodology for calibrating the system, ensuring accurate measurements in real-world scenarios.
It is crucial to adhere to the general remarks, maintaining consistency in setup and utilizing calibrated antenna factors.
It is strongly adviced to adhere to the general remarks, maintaining consistency in setup and utilizing calibrated antenna factors.
This application note serves as a comprehensive guide, enabling users to perform these emission substitution measurements with confidence and accuracy.
This application note serves as a comprehensive guide, empowering users to perform these emission substitution measurements with confidence and accuracy.
[[Category:RadiMation]]
Latest revision as of 10:25, 9 September 2024
How to perform a radiated emission substitution test[edit]
This application note explains how to perform a radiated emission substitution test with RadiMation®, as for example mentioned in ETSI EN 300 328 C5.3 and C5.4
The exact requirements and test methods are specified in the ETSI EN 300 328.
Necessary equipment[edit]
  • Signal generator
  • Transmitting Antenna
  • Receiving Antenna
  • Forward power meter
  • Coupler (0 dB coupling factor)
General remarks[edit]
Performing the measurements[edit]
Phase A: Determine the power being applied to the antenna[edit]
Purpose[edit]
Purpose of this phase is to determine the power that is going into the transmit antenna during Phase B. Very often an attenuator is used between the signal generator and the antenna, to reduce mismatches. This phase is used to correct for cable, attenuator and signal generator losses.
Test set up[edit]
  • A1. Build a setup with signal generator outside the chamber and the transmit antenna in the chamber.
  • A2. Do not connect the transmit antenna yet, but connect the receiver on the location of the transmit antenna.
  • A3. Sweep with a RI Fixed power test 0 dBm signal power over the frequency range. The antenna selected in the testsite of this test is not relevant. A coupler with a 0 dB coupling factor should also be used.
  • A4. Also measure the forward power using the receiver. The power values measured as the forward power meter are the power values that are applied to the transmit antenna during phase C.
Test configuration[edit]
To perform the phase A test, create a radiated immunity multiband test. The radiated immunity multiband test, can be started by selecting from the menu:
   Menu.svg Tests
      Menu.svg Radiated Immunity
         Menu.svg Multiband
ScreenElementDescription.svg Start The start frequency of the test. For example 1 GHz.
ScreenElementDescription.svg End The stop frequency of the test. For example 6 GHz.
ScreenElementDescription.svg Frequency Step The frequency step.
ScreenElementDescription.svg Dwell Time The dwell time.
ScreenElementDescription.svg Frequency Change Mode Constant.
ScreenElementDescription.svg Modulation The modulation
ScreenElementDescription.svg Test equipment The equipment needed for the test.
ScreenElementDescription.svg Location Type Antenna height, distance, polarization, angle.
ScreenElementDescription.svg Test Level Specify the power closed loop method test level.
ScreenElementDescription.svg Inputs Specify the forward power measurement.
Phase B: Determine the overall transfer of the measurement system[edit]
Purpose[edit]
Test set up[edit]
  • B1. Build a setup with a signal generator outside the chamber and the transmit antenna in the chamber
  • B2. Place the receiver outside the chamber and connect it to the receive antenna in the chamber
  • B3. Sweep with a RI Fixed power test 0 dBm signal power over the frequency range. This can be the same test configuration as is used during step A3. The antenna selected in the test site of this test is not relevant. A coupler with a 0 dB coupling factor should also be used
  • B4. Measure the forward power using the receiver. The power values measured as the forward power meter will be used to determine the correct correction factor
To perform the phase B test, create a radiated immunity multiband test. The radiated immunity multiband test, can be started by selecting from the menu:
   Menu.svg Tests
         Menu.svg Multiband
Phase C: Calculate the correct antenna correction factor[edit]
Purpose[edit]
Purpose of this phase is to use the measurement results obtained during phase A and B to determine the correction factor for the total system that can be used during an emission substitution measurement. This measurement measures the difference between the transmitted power and the received power at the receiving antenna. The chamber characteristics, which is the overall transfer factor of the chamber using the applicable measurement equipment, is represented by this difference.
Calculations[edit]
  • : The antenna correction factor
  • : The forward power during Phase A
  • : the Transmit Antenna Factor of the transmit antenna compared to a dipole
  • : The forward power during Phase B
The can be calculate from the calibrated Antenna Factor of the transmit antenna with:
  • : The frequency, expressed in MHz
  • The antenna factor of the antenna, expressed in dB/m
Application[edit]
The calculated C can be used as the correction of the antenna during the real measurement, and can be placed in a correction file that is attached to the receiving antenna.
The EUT measurements[edit]
Purpose[edit]
Test Set up[edit]
  • On the location of the transmit antenna the EUT is positioned.
  • The determined antenna correction factor (C) should be applied to the antenna that is used in the testsite during the measurement. No other corrections should be included in the testsite. Correction of the cables, preamplifiers and attenuators should not be done, because all these components were also present during phase B, and are thus already corrected for in the determined antenna correction factor (C).
   Menu.svg Tests
      Menu.svg Radiated emission
         Menu.svg Multiband
For more information regarding the radiated emission test configuration, please see: Radiated emission multiband test configuration
This application note outlines a three-phase process for determining the antenna correction factor in emission substitution measurements using RadiMation.
By systematically addressing power application to the antenna, measuring the overall transfer of the measurement system, and calculating the correct correction factor, users can achieve precise and reliable results.