Current sensor transfer impedance determination method: Difference between revisions
= Theory = | |||
[[Image:Current sensor Transfer impedance Prove.PNG]] | [[Image:Current sensor Transfer impedance Prove.PNG]] | ||
The left impedance is the signal generator which is generating enough power for 1 ampere. | |||
This 1Amp. generates <math>P = I^2*R=50 \ Watt</math> in the right impedance. | |||
The current sensor has 1 ohm transfer impedance, this means 1 ampere generates 1 Volt on the measuring part below. | |||
The power in the lower 50 ohm impedance is <math>P = \frac{U^2}{R} = 20 \ mWatt</math> | |||
= Reference measurement = | |||
[[Image:Current sensor Transfer impedance Ref measurement.PNG]] | [[Image:Current sensor Transfer impedance Ref measurement.PNG]] | ||
= Probe measurement = | |||
[[Image:Current sensor Transfer impedance final measurement.PNG]] | [[Image:Current sensor Transfer impedance final measurement.PNG]] | ||
= Calculation = | |||
<Math>Correction \ factor (dB)= P_{Measured} - P_{Reference} + 33.98</Math> | |||
<Math>P_{Measured}</Math> and <Math>P_{Reference}</Math> in dBm. | |||
{{note|This method is not a replacement for a real calibration.}} | {{note|This method is not a replacement for a real calibration.}} |
Revision as of 15:19, 23 January 2009
Theory[edit]
This 1Amp. generates in the right impedance.
Reference measurement[edit]
Probe measurement[edit]
Calculation[edit]
and in dBm.
Note: | This method is not a replacement for a real calibration. |