How to perform a MIL-STD-461 RS103, radiated susceptibility, electric field[edit]
This application note explains how the MIL-STD-461 RS103, radiated susceptibility, electric field test can be performed with RadiMation®
The exact requirements and test methods for the RS103 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to perform the RS103 test:
- Signal generator
- Amplifier
- Antenna
- Field probe
EUT Testing[edit]
EUT Testing equipment[edit]
Device name |
Tab in testsite configuration window |
note
Signal Generator |
Devices 1 |
The signal generator to use
Amplifier |
Devices 1 |
The amplifier to use
Antenna |
Devices 1 |
The antenna to use
Field Sensor 1 |
Field probes |
The field sensor to use
Configure the EUT test[edit]
To perform the test on the EUT, create a radiated immunity multiband test.
Create a test setup configuration file for the following settings:
Frequency range |
Frequency step |
Polarization
2 MHz to 30 MHz |
1% |
Vertical
30 MHz - 18 GHz |
0.5% |
Both (Horizontal and vertical)
18 GHz - 40 GHz |
0.25% |
The radiated immunity multiband test, can be configured by selecting from the menu:
Tests
Radiated Immunity
Multiband
New
Start
The start frequency of the test. For example 2 MHz.
End
The stop frequency of the test. For example 30 MHz.
Frequency Step
The frequency step, see Table III.
Dwell Time
The dwell time.
Frequency Change Mode
Constant.
Modulation
The modulation to use see below.
Test equipment
The equipment needed for the radiated susceptibility test.
Location Type
Antenna height, distance, polarization and angle.
Test Level
Specify the Electrical field closed loop method test level. Also configured a "Variable field"
Inputs
No inputs are needed.
At TestLevel click Add to add a new Test level and select TestLevel - Electrical field closed loop method
Description
The description of the test level.
Test level
Variable field, and specify the correction file for the applicable test level. When the level is one value this value can be specified with Constant field
Tolerance
Specify the tolerance to use.
Probes
Check Probe 1 to measure with the first probe.
To configure the pulse modulation settings click Config next to the modulation settings, click the PM tab, and enter the pulse modulation settings as defined in 5.21.3.4.c.1.a
Activate modulation
Check this checkbox to enable the pulse modulation.
Frequency
The PM frequency 1 kHz.
Duty Cycle
A duty cycle of 50%.
Also apply modulation out of the dwell-time period
Check this checkbox to enable modulation outside the dwell-time.
When all settings are configured press Start Test to run the test.
EUT Test Result[edit]
The RadiMation® Radiated Immunity multiband test can be used to perform the RS103 EUT test.
