How to perform a MIL-STD-461 RS103, radiated susceptibility, electric field[edit]
This application note explains how the MIL-STD-461 RS103, radiated susceptibility, electric field test can be performed with RadiMation®
RS103 is a test method used in military and aerospace environments to evaluate the susceptibility of electronic equipment to radiated electromagnetic fields across a frequency range of 2 MHz to 40 GHz. It primarily uses pulse modulation to simulate high-intensity electromagnetic interference, ensuring that devices can withstand these conditions without malfunctioning.
The exact requirements and test methods for the RS103 are specified in the MIL-STD-461. This application note demonstrates how to setup a closed loop test. When testing closed loop, no calibration is required.
Necessary equipment[edit]
The following devices are necessary to perform the RS103 test:
- Signal generator
- Amplifier
- Antenna
- Field probe
Define the test level[edit]
If the limit in "Table XI RS103" is the same value for the whole frequency range the creation and usage of the correction file can be skipped. Usage of the correction file is only applicable to variable limits.
Create a correction file for the test level[edit]
- File
- New
- Correction
- New
- File
First create a correction file in RadiMation® and specify the test level (See 'Table XI RS103 limits.').
- File
- Save correction
- File
Store the correction file with the RS103 limits as file on disk.
This correction file with the test level will be used during EUT testing.
EUT Testing[edit]
EUT Testing equipment[edit]
The configuration of the EUT test site should contain the following devices:
Antenna | Devices 1 | The antenna to use | |
Field Sensor 1 | Field probes | The field sensor to use |
Configure the EUT test[edit]
To perform the test on the EUT, create a radiated immunity multiband test.
Create a test setup configuration file for the following settings:
Frequency range | Frequency step | Polarization |
2 MHz to 30 MHz | 1% | Vertical |
30 MHz - 18 GHz | 0.5% | Both (Horizontal and vertical) |
18 GHz - 40 GHz | 0.25% | Both (Horizontal and vertical) |
- Tests
- Radiated Immunity
- Multiband
- New
- Multiband
- Radiated Immunity
- Tests
Start | The start frequency of the test. For example 2 MHz. |
Stop | The stop frequency of the test. For example 30 MHz. |
Frequency Step | The frequency step, for the applicable steps see "Table III susceptibility scanning" in MIL-STD-461G. |
Modulation | The modulation to use see below. |
Test equipment | The equipment needed for the radiated susceptibility test. |
Location Type | Antenna height, distance, polarization and angle. |
Test Level | Specify the Electrical field closed loop method test level and configure a Constant field or Variable field |
At TestLevel click Add to add a new Test level and select TestLevel - Electrical field closed loop method
Test level | Check Variable field and specify the correction file for the applicable test level. When the level is one fixed value check Constant field and specify the test level |
Probes | Check Probe 1 to measure with the first probe. |
To configure the pulse modulation settings click Config next to the modulation settings, click the PM tab, and enter the pulse modulation settings as defined in 5.21.3.4.c.1.a
Activate modulation | Check this checkbox to enable the pulse modulation. |
Frequency | The PM frequency 1 kHz. |
Duty Cycle | A duty cycle of 50%. |
Also apply modulation out of the dwell-time period | Check this checkbox to enable modulation outside the dwell-time. |
When all settings are configured press Start Test to run the test.
EUT test result[edit]
The RadiMation® Radiated Immunity multiband test can be used to perform the RS103 EUT test.