Difference between revisions of "RadiMation Application Note 160"
Jump to: navigation, search
(Define the test level)
(Create a correction file for the test level by selecting the menu:)
Line 31: Line 31:
{{Menu|File|New|Correction}}
First, create a correction file in {{RadiMation}} and specify the applicable test level for the BCI test based on the test requirements. The levels are typically specified in terms of current in mA for different frequency bands (see 'ISO 11452-4 BCI test severity levels ').
In this example the levels for test severity level 1 are configured.
Ensure that the correction file has a {{ScreenElement|Frequency}} column and a {{ScreenElement|Current}} column with the unit set to mA.
Revision as of 12:14, 17 September 2024
How to perform an ISO 11452-4 BCI (Bulk Current Injection) test
This application note explains how the ISO 11452-4 BCI (Bulk Current Injection) test can be performed with RadiMation®.
The ISO 11452-4 standard is used to test the immunity of automotive electronic components to conducted RF (radio frequency) disturbances injected into the wiring harness. The test frequency typically ranges from 1 MHz to 400 MHz, using both continuous and amplitude-modulated signals. The test ensures that the electronic devices can withstand the RF disturbances without performance degradation or failure.
The exact requirements and test methods for the BCI test are specified in the ISO 11452-4.
Necessary equipment
The following devices are required to run the calibration, verification, and EUT (Equipment Under Test) test:
  • Signal generator
  • Amplifier
  • Coupler
  • Forward power meter
  • Current sensor
  • 50 Ohm load
  • 50 Ohm attenuator
  • Current injection device
  • Current injection calibration jig
Tips
Note: If the current unit is shown as μA, it can be changed to dBμA by clicking Units, select Current and click Edit to change this.
The ISO 11452-4 BCI test requires test levels to be defined for different frequency ranges. The test level may vary depending on the application class or specific test plan, and these levels are set up as a correction file in RadiMation®.
   Menu.png File
      Menu.png New
         Menu.png Correction
First, create a correction file in RadiMation® and specify the applicable test level for the BCI test based on the test requirements. The levels are typically specified in terms of current in mA for different frequency bands (see 'ISO 11452-4 BCI test severity levels '). In this example the levels for test severity level 1 are configured.
Ensure that the correction file has a Frequency column and a Current column with the unit set to mA.
Store the correction file for the test level by selecting the menu:
   Menu.png File
      Menu.png Save correction Save the test level as a correction file on disk.
This correction file will be used during the calibration and EUT testing.
Calibration procedure
The calibration procedure is necessary to characterize losses in the setup and determine the correct power level for injection.
Calibration equipment
The configuration of the calibration test setup should include the following equipment:
# Device name Tab in test site configuration window note
Coupler Devices 1 The directional coupler to use
Forward power meter Devices 1 The forward power sensor to use
Power meter Devices 2 The power meter or analyzer for measuring current
Current probe Devices 2 The current monitoring probe used with a transfer factor applied
Injection clamp Devices 2 The bulk current injection (BCI) clamp
Calibration jig Devices 2 The jig used for the current injection calibration
Cables
Cable from current probe to power meter Cables Cable with a correction file specified for cable loss
Cable from signal generator to injection clamp Cables Cable with specified loss of the used attenuator
Cable from injection clamp to current probe Cables Cable with a correction file specified for cable loss
Configure the calibration
The BCI test covers three primary frequency bands. The following settings can be configured according to ISO 11452-4 requirements:
Band 1 1 MHz - 30 MHz 1 %
Band 2 30 MHz - 200 MHz 1 %
Band 3 200 MHz - 400 MHz 0.5 %
These three bands should be calibrated individually. Calibration can be started by selecting:
   Menu.png Calibration
      Menu.png System calibration
         Menu.png Conducted immunity
This example shows the configuration for Band 1; similar configurations can be created for the other bands.
ScreenElementDescription.png Start The start frequency of the calibration (e.g., 1 MHz).
ScreenElementDescription.png End The stop frequency of the calibration (e.g., 30 MHz).
ScreenElementDescription.png Step The frequency step, in this case, 1%.
ScreenElementDescription.png Test level The test level, in this case current.
ScreenElementDescription.png Variable The variable test level, select the correction file created earlier.
ScreenElementDescription.png Tolerance The regulation tolerance.
ScreenElementDescription.png Calibration method The power measurement that should result from the calibration: Forward power.
ScreenElementDescription.png Test equipment The calibration test equipment.
ScreenElementDescription.png Test engineer The engineer performing the calibration.
Calibration result
Once calibration is complete, the engineer will save the results in a calibration (.CAL) file. A separate calibration file will be generated for each frequency band.
EUT Testing
EUT Testing equipment
The configuration for EUT testing should contain the following devices:
Power meter Devices 2 The analyzer for measuring current
Current probe Devices 2 The current probe with transfer factor applied
Injection clamp Devices 2 The bulk current injection clamp
Configure the EUT test
To perform the EUT test, configure the conducted immunity multiband test by selecting from the menu:
   Menu.png Tests
      Menu.png Conducted Immunity
         Menu.png Multiband
The following bands should be configured:
ScreenElementDescription.png Start The start frequency of the test. For example 10 kHz.
ScreenElementDescription.png End The stop frequency of the test. For example 1 MHz.
ScreenElementDescription.png Frequency Step The frequency step, see Table III.
ScreenElementDescription.png Dwell Time The dwell time.
ScreenElementDescription.png Frequency Change Mode Constant.
ScreenElementDescription.png Modulation Configure Pulse Modulation with 1 kHz and 50% duty cycle.
ScreenElementDescription.png Test equipment The equipment needed for the conducted immunity test.
ScreenElementDescription.png Location Type Injection clamp.
ScreenElementDescription.png Test Level Specify the current substitution method. Also add a "Measured Current Limit"
ScreenElementDescription.png Inputs No inputs are needed.
To configure the pulse modulation settings click Config next to the modulation settings, click the PM tab, and enter the pulse modulation settings as specified in ISO 11452-4.
ScreenElementDescription.png Activate modulation Check this checkbox to enable the pulse modulation.
ScreenElementDescription.png Frequency The PM frequency 1 kHz.
ScreenElementDescription.png Duty Cycle A duty cycle of 50%.
ScreenElementDescription.png Also apply modulation out of the dwell-time period Check this checkbox to enable modulation outside the dwell-time.
ScreenElementDescription.png Description The description of the test level.
ScreenElementDescription.png Calibration file Select the calibration file to use for this band.
ScreenElementDescription.png Test level Variable current, and specify the correction file for the applicable test level.
ScreenElementDescription.png Calibration method Select Forward Power.
ScreenElementDescription.png Tolerance Specify the tolerance to use.
EUT Test Result
Once the test is complete, the results are stored in a .LOG file. The test results can be reviewed by selecting the corresponding test from the list and clicking Info.
Conclusion
Using RadiMation®, you can effectively perform the ISO 11452-4 BCI test, ensuring compliance with automotive RF immunity standards and ensuring product robustness in the presence of RF interference.
[[Category Application Note]]
