| Line 86: |
Line 86: |
| === Calibration result === | | === Calibration result === |
| When the calibration is performed, the engineer is asked, to save the results of that calibration to a calibration (.CAL) file. It is possible to give the .CAL file a clear and distinct name. | | When the calibration is performed, the engineer is asked, to save the results of that calibration to a calibration (.CAL) file. It is possible to give the .CAL file a clear and distinct name. |
| == Verification ==
| CS114 requires to perform a verification of the setup with the current probe in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit.
| === Verification equipment ===
| [[File:CS114_Verification.png]]
| The configuration of the verification test site should contain the following devices:
| {|class="wikitable"
| !# || Device name || Tab in testsite configuration window || note
| | || Signal Generator || Devices 1 || The signal generator to use
| | || Amplifier || Devices 1 || The amplifier to use
| | || Coupler || Devices 1 || The coupler to use
| | || Forward power meter || Devices 1 || The forward power meter to use
| | || Sensor powermeter || Devices 2 || The power meter or analyser to use for measuring the current
| | || Current Sensor || Devices 2 || The current sensor to use with transfer factor attached to the driver
| | || Injection device || Devices 2 || The injection clamp to use
| !colspan="4"|Cables
| |① || Cable current -> power meter || Cables || Cable with a correction file specified for the cable loss
| |② || Cable current -> power meter || Cables || Cable with the specified loss of the used attenuator
| |③ || Cable current -> power meter || Cables || Cable with a correction file specified for the cable loss
| === Configure the verification ===
| The verification can be performed in {{RadiMation}} by doing an actual conducted immunity test. This test can be started by selecting from the menu:
| {{Menu|Tests|Conducted Immunity|Multiband}}
| Create a new multiband conducted immunity test and define three bands with the following settings:
| {|class="wikitable"
| ! Band || Frequency range || Step size
| | Band 1 || 10 kHz - 1 MHz || 10 %
| | Band 2 || 1 MHz - 30 MHz || 2 %
| | Band 2 || 30 MHz - 200 MHz || 1 %
| [[File:CS114 Verification Setup.png]]
| {{ScreenElementDescriptionStart}}
| {{ScreenElementDescription|Start|The start frequency of the verification. For example 10 kHz.}}
| {{ScreenElementDescription|End|The stop frequency of the verification. For example 1 MHz.}}
| {{ScreenElementDescription|Frequency Step|The frequency step, see 5.12.3.4.c.3.}}
| {{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II of the [[MIL-STD-461]].}}
| {{ScreenElementDescription|Frequency Change Mode|The method that is used to change from one to the next frequency: {{ScreenElement|Constant}}.}}
| {{ScreenElementDescription|Modulation|Modulation is optional during the verification.}}
| {{ScreenElementDescription|Test equipment|The equipment needed for the verification.}}
| {{ScreenElementDescription|Location Type|Injection clamp.}}
| {{ScreenElementDescription|Test Level|Specify the test level {{ScreenElement|Current substitution method}}.}}
| {{ScreenElementDescription|Inputs|Add the {{ScreenElement|Current sensor}} as an input.}}
| {{ScreenElementDescriptionEnd}}
| Each band needs a test level defined with the test level {{ScreenElement|Current substitution method}} and the appropriate calibration file attached.
| [[File:CS114 Current Testlevel Configuration.png]]
| {{ScreenElementDescription|Description|The description of the test level.}}
| {{ScreenElementDescription|Calibration file|Select the calibration file created during the calibration procedure.}}
| {{ScreenElementDescription|Test level|{{ScreenElement|Variable current}} and specify the correction file for the applicable test level created during the calibration procedure.}}
| {{ScreenElementDescription|Calibration method|Select Forward Power.}}
| {{ScreenElementDescription|Tolerance|Specify the tolerance to use.}}
| Add an input {{ScreenElement|Current sensor}}, this will ensure that the current is measured during the test.
| === Run the verification test ===
| When all settings are configured press {{ScreenElement|Start Test}} to run the verification test.
| [[File:CS114 Verification Result.png]]
| === Verification result ===
| With the result of the verification it should be possible to determine if the measured current is within the 3 dB tolerance of the current test limit as specified in clause 5.12.3.4.c of the MIL-STD-461.
| == EUT Testing == | | == EUT Testing == |
How to perform a MIL-STD-461 CS101, Conducted susceptibility, power leads[edit]
This application note explains how the MIL-STD-461 CS101, Conducted susceptibility, power leads test can be performed with RadiMation®
MIL-STD-461 CS101 is a test method used in military and aerospace environments to assess the conducted susceptibility of equipment to low-frequency interference, 30 Hz - 150 kHz, ensuring it operates reliably in the presence of conducted disturbances.
The exact requirements and test methods for the CS101 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to run the calibration and EUT test:
- Signal generator
- Amplifier
- Coupler
- Forward power meter (optional)
- Oscilloscope or measurement receiver
- 0.5 Ω resistor
Define the calibration test level[edit]
The CS101 EUT test is limited by the power levels specified in CS101-2, applied through a 0.5 Ω resistor.
During the calibration RadiMation® is levelling on a voltage and the corresponding signal source is recorded which is used during EUT testing to limit on.
To calculate the corresponding voltages based on the power levels the following formula can be used:
This gives the following voltages:
| Frequency |
Power |
Voltage
| 30 Hz |
80 Watts |
136 dBµV
| 5 kHz |
80 Watts |
136 dBµV
| 150 kHz |
0.09 Watts |
106.5 dBµV
File
New
Correction
Note:
Make sure to add the Vrms column with Columns/Units and remove the Correction column. In addition change the Unit to dBµV.
File
Save correction
Calibration procedure[edit]
The calibration is performed to determine the signal source required for the corresponding power level through the 0.5 Ω resistor, which will be used during EUT testing as limit.
Calibration equipment[edit]
Device name |
note
Signal Generator |
Devices 1 |
Amplifier |
Devices 1 |
Oscilloscope |
Devices 1 |
Oscilloscope
Sensor power meter |
Devices 2 |
measurement receiver as power meter (in case no oscilloscope is used)
The configuration of the calibration can be created by selecting the menu:
Calibration
System calibration
Start
The start frequency of the calibration. 30 Hz.
End
The stop frequency of the calibration. 150 kHz.
Step
The frequency step, 5%.
Test level
The test level, in this case Voltage.
Variable
The variable test level, select the correction file created before 'CS101-2 power limit.cor'.
Tolerance
The regulation tolerance.
Calibration method
The power measurement that should be the result of the calibration: Signal generator level
Test equipment
Calibration test equipment.
Test engineer
The engineer that performed the calibration.
Calibration result[edit]
EUT Testing[edit]
EUT Testing equipment[edit]
Device name |
note
Signal Generator |
Devices 1 |
Amplifier |
Devices 1 |
Coupler |
Devices 1 |
The coupler to use
Devices 1 |
Sensor powermeter |
Devices 2 |
Current Sensor |
Devices 2 |
Injection device |
Devices 2 |
| Cables
| ① |
Cables |
| ② |
Cables |
| ③ |
Cables |
Configure the EUT test[edit]
To perform the actual test on the EUT, create a conducted immunity multiband test with 3 bands.
The conducted immunity multiband test, can be started by selecting from the menu:
Tests
Conducted Immunity
Multiband
| Band |
Frequency range |
Step size
| Band 1 |
10 kHz - 1 MHz |
5 %
| Band 2 |
1 MHz - 30 MHz |
1 %
| Band 2 |
30 MHz - 200 MHz |
0.5 %
Start
The start frequency of the test. For example 10 kHz.
End
The stop frequency of the test. For example 1 MHz.
Frequency Step
The frequency step, see Table III.
Dwell Time
The dwell time specified see TABLE II of the MIL-STD-461.
The method that is used to change from one to the next frequency: Constant.
Modulation
Configure Pulse Modulation with 1 kHz and 50% duty cycle.
Test equipment
The equipment needed for the conducted immunity test.
Location Type
Injection clamp.
Test Level
Specify the Current substition method and add a limit Measured Current Limit
Inputs
No inputs are needed.
Activate modulation
Check this checkbox to enable the pulse modulation.
Frequency
The PM frequency 1 kHz.
Duty Cycle
A duty cycle of 50%.
Also apply modulation out of the dwell-time period
Check this checkbox to enable modulation outside the dwell-time.
Description
Calibration file
Select the calibration file to use for this band.
Test level
Variable current, and specify the correction file for the applicable test level.
Calibration method
Tolerance
Description
The description of the limit
Constant current
The current to limit on.
EUT Test Result[edit]
The MIL-STD-461G CS114 calibration can be performed by using the Conducted immunity calibration.
The RadiMation® Conducted immunity multiband test can then be used to perform the MIL-STD-461G CS114 verification and the EUT test with a calibration file and the current sensor used as a limit in the EUT test.