| Line 155: | Line 155: | ||
{{ScreenElementDescription|Record data|In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full tracebility, all test data should be stored.}} | {{ScreenElementDescription|Record data|In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full tracebility, all test data should be stored.}} | ||
{{ScreenElementDescription|Narrow band / broad band discrimination| | {{ScreenElementDescription|Narrow band / broad band discrimination|{{RadiMation}} supports two types of narrow band / broad band determination.}} | ||
The first one is called 'Generic'. During the determination of the peak, the so called ‘zooming’. | The first one is called 'Generic'. During the determination of the peak, the so called ‘zooming’. {{RadiMation}} verifies if the peak is bigger or equal to twice the RBW filter. If so, the peak is determined broad band. Otherwise it is called narrow band. | ||
The second method is called 'Mil Standard'. This method is also based on the RBW filter. But now the filter is changed to determine the type of peak. First the peak is measured with the current RBW, then it is made 10 times bigger. If multiplying the current RBW gives a RBW that is bigger then 1 MHz the current RBW is divided by 10. A delta RBW is calculated according to the formula below. | The second method is called 'Mil Standard'. This method is also based on the RBW filter. But now the filter is changed to determine the type of peak. First the peak is measured with the current RBW, then it is made 10 times bigger. If multiplying the current RBW gives a RBW that is bigger then 1 MHz the current RBW is divided by 10. A delta RBW is calculated according to the formula below. | ||
Revision as of 10:47, 17 February 2011
Radiated emission[edit]
The RadiMation® package support three main radiated emission methods:
- Anechoic chamber method
- Fixed height method
- G-TEM method
Anechoic chamber[edit]
Frequency window[edit]
| Sets start frequency |
| Sets stop frequency |
Turntable window[edit]
| The software will automatically detect on which angle maximum emission will occur. |
| The angle step size. |
| This value determines the acceptance criteria for a peak. |
| The mode setting determines the way a peak is detected. |
- Intelligent
- Complete area
Intelligent[edit]
| Warning: | This might cause a measurement error if the number of steps is set to a low value. |
Complete area[edit]
| Warning: | This might cause a measurement error if the “Area” is set to a low value relative to the area given by the stop position minus start position dived by number of steps. |
Antenna tower window[edit]
| Range over which the peak search will be performed (0 to n meter). |
| The distance for each step. |
- Intelligent
- Complete area
Intelligent[edit]
Complete area[edit]
| Warning: | This might cause a measurement error if the “Area” is set to a low value relative to the area given by the max. height minus min. height dived by number of steps. |
Antenna window[edit]
Polarisation[edit]
- Horizontal
- Vertical
- Both
Distance[edit]
Receiver settings[edit]
| Sets the pre-amp gain |
Max. hold measurements[edit]
Traces window[edit]
Detectors window[edit]
Peak detection window[edit]
Test site window[edit]
| In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results. |
Limit lines window[edit]
Test options window[edit]
| In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full tracebility, all test data should be stored. |
| Limits | Distortion type |
Loading a TSF file
Test Procedure[edit]
Radimation now performs a maxhold on all the sweeps and determines the peaks. Then it determines the position a peak was found. And changes the turntable and antenna to that current position to re-measure the peak. When multiple peaks have been detected, the most compact way is determined to measure these peaks.
Info screen[edit]
| In the trace window you can control the traces that are shown in the graph by selecting or deselecting the checkbox. If during a measurement a trace is not used then the check box is grayed out. |
| In the table all the measured peaks are showed. The table also shows the limit values for that frequency point when a limit was given at the beginning of the test. |
Test site information[edit]
| Device name | Tab in Radimation configuration window |
| Spectrum analyser | Standard 2 |
| Pre amplifier | Standard 2 |
| Antenna | Standard 1 |
| # | Device name | Tab in Radimation configuration window |
| 1 | Pre amp->Spectrum | Cables |
| 2 | Ant->Pre amp | Cables |
When using a spectrum analyser for emission measurements, Radimation first performs a full span max. hold measurement with the configured trace detector. The number of max. hold peak sweeps can be defined by the user. From the obtained max hold graph, Radimation determines a number of peaks, which will be further investigated. The number of peak signals to be measured can be defined by the user. For each peak from the graph, Radimation will zoom in, by reducing the span in several steps. This must be done while span inaccuracy of the analyser in large frequency spans will give slightly incorrect frequency information. After the exact peak frequency is determined, Radimation will perform the configured Peak,Quasi peak, Average and / or RMS measurements. Due to the above, the finally measured frequency and amplitude of a peak can slightly differ from the max hold graph. This is not an error, but merely an indication that the final measurements are more accurate. The peak table always shows the final measured values (most accurate).
Fixed height test method[edit]
In the turn table window, the following parameters can be configured:
- Intelligent
- Complete area
Intelligent[edit]
Complete area[edit]
- Intelligent
- Complete area
Intelligent[edit]
Complete area[edit]
Polarisation[edit]
- Horizontal
- Vertical
- Both
Distance[edit]
In this box, the operator can enter the distance between antenna and EUT. RadiMation uses the distance information for reference only.
Traces window[edit]
Loading a TSF file
Radimation now performs a maxhold on all the sweeps and determines the peaks. Then it determines the position a peak was found. The turntable position is changes to that position to re-measure the peak. The antenna tower is changed in height according to the settings.
measurements. When in doubt contact your reseller. |
G-TEM Method[edit]
Correlation method[edit]
Measurement procedure[edit]
Tests
Radiated emission
G-TEM
Test configuration[edit]
OATS settings[edit]
G-TEM settings[edit]
| Sets the pre-amp gain. |
Traces window[edit]
Max. Peaks
Starting a G-TEM test[edit]
Loading a TSF file[edit]
Manual mode[edit]
| Sets the Resoution Band Width filter |
Data correction[edit]
[edit]
| RadiMation can detect peaks automatically. From the “peaks” TAB, choose “Search peak” to detect the maximum signal. The software will measure the detected peak. |
Traces Menu[edit]
Traces[edit]
Detectors[edit]
- Peak detector
- Average detector
- Quasi peak detector
- RMS detector
Zoom function[edit]
Limit lines[edit]
[edit]
Merge function[edit]
X+k*S Calculations[edit]
Anechoic chamber (Multi band)[edit]
The area is defined 50% left of the starting position and 50% right. So if the area is 30 degrees, RadiMation will scan between starting position plus and minus 15 degrees.
| The height step size. |
The area is defined 50% above of the starting position and 50% below. So if the area is 1 meter, RadiMation will scan between starting position plus and minus 50 cm.
Polarisation[edit]
- Horizontal
- Vertical
- Both
Distance[edit]
Traces window[edit]
| The number op peak hold sweeps for each measurement can be selected. If this value is set to, for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweep speed. For spectrum analysers and CW signals you can use a relatively low amount of peak sweeps, for example 10. But when you have discontinues signals like a spark bridge, you may need 100 peak sweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope. |
Final peak measurement[edit]
Change Order window[edit]
So for Site 1 you want:
- Band A RBW 9 kHz
- Band B RBW 9 kHz
- Band A RBW 120 kHz
- Band B RBW 120 kHz
And for test site 2:
- Band A RBW 9 kHz
- Band B RBW 9 kHz
- Band A RBW 9 kHz
- Band B RBW 9 kHz
- RBW
- Testsite
- Frequency band
- Band A RBW 9 kHz
- Band B RBW 9 kHz
Info screen[edit]
| Note: | Continuing the test is only possible with the selected test site. RadiMation will not select the correct test site for you. All measurements are done with the selected test site. If you want automatic changes of the test site, restart the configured TSF file. |
Manual Mode (Multi band)[edit]
Turn table[edit]
Antenna tower[edit]
Test site[edit]
Before manual mode can be used the test site needs to be selected so that Radimation® knows which devices to use.
Final measurement[edit]
Peaks window[edit]
| Triggers the receiver of analyzer to take the configured amount of sweeps. |
| Allows several actions to be performed on the detected peaks, like select all and deselect all. |
