Release Notes/Solved items Version 2013.1: Difference between revisions
{{:Release Notes/Solved items Version 2013.1.23}} | |||
{{:Release Notes/Solved items Version 2013.1.22}} | |||
{{:Release Notes/Solved items Version 2013.1.21}} | {{:Release Notes/Solved items Version 2013.1.21}} | ||
{{:Release Notes/Solved items Version 2013.1.20}} | {{:Release Notes/Solved items Version 2013.1.20}} |
Revision as of 13:14, 11 November 2013
RadiMation version 2013.1.23
- #3834: If overload check for receiver, not scanning receiver
- #4212: Adjustments (advanced) settings have no effect with calibration power sensor (4418)
- #4372: Test information time is incorrectly set
- #4739: Report Generator: time, date and special character is wrong
- #4820: Test data can't be opened, error 'No data document could be created for the test identifier'
- #4897: RadiMation is not responding when the Database connection location is opened
- #4935: Driver R&S unit SCIU
- #4960: Frequency settings are hidden and not usable anymore in manual mode
- #5050: Manual mode frequency adjustment (arrow button) is pressed multiple times
- #5056: During test it is not possible to add or remove graph lines
- #5057: Settings to change graph line colors are not retained
- #5070: R&S SMA 100 device driver check crashes RadiMation
- #5073: R&S HCC should read a limit during a check to ensure that the correct device is connected
- #5074: Word report generator can cause an exception caused by delayed unloading
- #5076: Graphs button is disabled during a multiband immunity test
- #5106: Driver: Schaffner/Chase CBL6111D antenna
- #5108: Driver: HP 59307A switch matrix
- #5109: Driver: AR 1000LM24 Amplifier
- #5110: Driver: AR 500W1000M3 Amplifier
- #5120: Error from RadiPower is not correctly reported
- #5122: Driver: R&S unit SCIU switch matrixes
- #5124: RPR2018C timeout errors
- #5130: RadiPower can become suddenly unresponsive
- #5131: Incorrect resizing is causing unreadable items RPR3006W
- #5132: Wrong times are shown in the eventlog of a calibration
- #5134: RadiTower stays in Remote after test is closed
- #5135: Control of the RadiTower enters an endless loop
- #5137: The field sensor device driver configuration should be extended with 'Monitoring measurement settings'
- #5138: Add possibility in the Maturo device driver to select the device unit
- #5139: Driver: R&S FSEA as powermeter
- #5145: Wrong ranges of Holaday HI 4433-MSE field probe
- #5147: R&S unit SCIU Amplifier isn't going to operate
RadiMation version 2013.1.22
- #4059: Axis selection for the FI7000
- #4584: Signal generator HP 83752A is not set for modulation
- #5119: Report generator code: FORWARD POWER LIST is not exporting any data
RadiMation version 2013.1.21
- #3220: Scaling of the Tsf selection window
- #4593: Driver: EMR-300 does not check Battery status
- #4744: Measurement is done in dBm with Antenna Terminal selected , when MIL STD 461E specifies the limit in dBuV
- #4909: After measurement no result(s) are shown
- #4994: R&S SMA 100A Invalid buffer mask specified
- #5040: Wrong graph scaling cuts off measurement results after test and RG
- #5066: RS ZNB-40 command and remote Queue overflow errors
- #5100: Communication problem with R&S SMC100A (via USB)
- #5101: ODBC database is not remembered and set to ACT again
- #5102: RMS value in Powermeter burst measurement is not corrected
- #5113: Test Equipment is set to 'none', but was correctly selected before test
RadiMation version 2013.1.20
- #2848: Regulated testlevel should be above requested testlevel if testlevel could not be regulated within the tolerance
- #4910: Closed loop electrical field regulation is always assuming a linear system
- #5033: Driver: Tilt Antenna Mast TAM 4.0-P and a NCD controller
- #5036: Not all limit lines are shown in an emission test
- #5063: Power delivery test is extremely slow
RadiMation version 2013.1.19
- #4843: Envelope trace mode for a RadiPower is not working anymore after a Burst mode measurement
- #4929: Driver: R&S HCC turntable + absorbing clamp controller
- #4996: Last bursts in a powermeter burst graph is showing a higher value
- #4999: Long burst measurements with the RadiPower will result in not enough memory errors
- #5061: Invalid buffer mask Agilent N9010A Scanning Receiver over USB (VISA Alias)
RadiMation version 2013.1.18
- #1368: Use blanking input for SMY02 PM modulation
- #4062: R&S SMY-01 input blanking for pulse modulation
- #4255: Driver: strange VBW value is set
- #4256: Calibration asks for resistor and shows voltage level
- #4260: Measuring deviations despite correct .CAL
- #4267: Starting Level Signal Power can trigger RF power on without interlock
- #4273: Power meter is read continuously with modulation applied in Manual Mode
- #4547: Setting the reference level has no effect on the ESMI
- #4548: Can select only one peak at the time
- #4592: Powermeter virtual instrument is not always unloaded from memory
- #4616: MXE alignment issue and is not released from remote
- #4842: Graph lines from a previous VI Powermeter window are still shown under the 'Graphs' button
- #4932: Driver: Injection device FCC F51
- #4992: Configure a maximum of 360 degrees for a Manual Controlled Turn Table
- #4995: Customization of the Graphline color in the Powermeter window is not working
- #5008: Dropdown of the threshold level for a burst test is using an unlogical order
- #5022: Innco CO3000 is started at 100cm instead of 90cm
- #5037: RG error when requesting the Equipment Table information
- #5047: Pulsed Immunity explanation
- #5054: Mil Std CAL text is not completely visible
- #5059: Innco CO3000 control multiple devices with one GPIB address without addressing
- #5064: 18-channel AD-convertor device driver for the RadiSense to be used during justation
- #5067: Agilent 8753ES can't change GPIB address
RadiMation version 2013.1.17
- #4944: Extend R&S ZNB driver to be used during antenna calibrations
- #5028: R&S SME: Use fixed attenuator mode to reduce attenuator relais switching
RadiMation version 2013.1.16
- #2541: RF setting of the marconi 2023 does not check for errors
- #4352: Driver: device driver check Marconi 2022 (ID1343) doesn't check anything
- #4594: Limited Signal power is never adjusted anymore in Multiband
- #4905: Test data of some previous versions are corrupted and can't be opened anymore
- #4930: Driver: Bonn amplifier BLWA 0810-400
- #4936: Changing AT polarization does not seems to work
- #4954: Where to define Cables for the EFT testing
- #4967: How to configure switch matrix changes
- #4968: Driver: HI4433LFH Holaday field probe
- #4969: CANoe interface object not initialized correctly
- #4970: Marconi 2023/24 error: Measurement using PM is not possible
- #4972: AR FP3000A is not responding
- #4973: RadiSense CTR1001A IV error unknown command
- #4976: Stop button doesn't work in Network Analyser system calibration
- #4977: Report generator raises index error for S-parameter test
- #4978: Export of graph data to excel is not using the correct comma
- #4979: Phase information is not retrieved from the R&S ZNB
- #4982: Driver: Amplifier Research AT4510
- #4985: Driver: AR 60S4G8
- #4986: Driver: AR 40S1G4
- #4987: Driver: 350S1G4
- #4993: Calculate the RMS value of a Burst measurement
- #5020: ESI26B Spectrum Analyser Query unterminated error
- #5021: R&S ESIB 40 receiver Query unterminated error
- #5023: SMY-01 PM with ext blanking
- #5024: Option window to Enable software configuration EMCO2090 is not shown anymore
- #5030: R&S SME: Device driver should use 0.01 dB Output resolution
RadiMation version 2013.1.15
- #1680: The resolution for the reflection coefficient should be increased from 0.001 to 0.0001
- #2232: Device driver to control the R&S ESI(B) series analysers/receivers over LAN
- #2758: ZVK measuring times are not inserted in side the machine, it is always 1 sec.
- #4041: Antenna mast Frankonia FC 02 not working
- #4324: Error Advantest TR4172 not all bytes are written to the GPIB
- #4464: Driver: tracking generator R&S ESPI3 receiver with option FSP-B9
- #4554: R&S ESPI7 ethernet communication
- #4658: Network analyser measurement: Swap noise and full through measurement
- #4659: Network analyser measurement: Support Full 1 port measurment, which only measures a single port
- #4687: Strange graphs are included in the report if the frequency has been changed to an already tested frequency in RI Single band manual mode
- #4689: Emco 2090 is controlled with an offset
- #4804: Simultaneously remote control LISN ESH3-Z and use Spectrum Analyser of R&S ESAI/ESMI doesn't work
- #4904: No graphical display of results and no results are stored
- #4911: Help Technical Support items are not working
- #4912: Advantest TR4172 is not using Video Average detector
- #4913: Wrong dates are shown in the EUT window
- #4915: Antenna tower limits are not retained for the EMCO2090
- #4917: Driver: R&S ZNB 40 Network analyser
- #4918: RadiSense 4.51 probe is not correctly detected
- #4920: Configuration of magnetic field substitution test level is problematic
- #4921: Unhandled exception after finishing RI substitution tests
- #4922: SMA100A not working correctly
- #4925: Powermeter measurement with the R&S FSP below 300 Hz measures the zero feed through signal
- #4927: Driver: pre-amplifier Teseq LNA 6901
- #4928: Driver: pre-amplifier R&S AFS
- #4931: Driver: Injection device Mess Electr NNBM8125
- #4937: StandardDetails window is not automatically resizing
- #4939: Default Address Information screen text is cut off
- #4946: Text in Calibration Import window not completely visible
- #4949: Does RadiMation support time domain scan
- #4951: Is it possible to use a calibration file at 20vm to generate a field of 3vm
- #4956: How can we run 5.8.38 and 2013.1.15 on the same PC
- #4957: Files generated in 2013.1.15 are not readabel in 5.8.38
- #4961: How to add attenuators for calibration
- #4971: How and with which RM version can we control the power meter RPR3006P?
- #4980: The 'number of sweeps' setting in a network analyser test is not working
RadiMation version 2013.1.14
- #693: There is no logical order in the selection of the controls, if the selection is moved with the 'TAB' key
- #3235: Cancel button in report generation is not cancel but skip
- #4047: First point from final measurement is not measured correctly
- #4112: Final measurement at wrong frequency
- #4202: HI field probe readout raises errors
- #4216: Incorrect QP measurement on first peak measurement
- #4287: Mast sometimes doesn't move with HD100 antenna controller
- #4414: Maturo NCD no VISA available
- #4669: Disable the spell checker during the generation of the report
- #4710: Only the first line from NOTE is showed in test list
- #4736: RG: Object reference not set to an instance of an object
- #4746: First Measured Peak is not correct with R&S ESVS 10 / ESHS 10
- #4793: Driver: Apsin 6010 Signal Generator
- #4818: Add direct link in Menu Help for 'Latest device drivers...'
- #4876: RunMacro(AutoFitThisTable) is auto fitting the wrong table
- #4877: Agilent 4440A is not working correctly in an attenuation/gain calibration test
- #4878: HP8542E is resetted twice during the initialization
- #4880: RadiGen doesn't always activate the AM modulation
- #4881: Driver: GPIB and VISA support for the AnaPico APSIN signal generators
- #4882: Detector measurement time is not correctly set in the PMM9010
- #4886: HI-4450 probe error without description
- #4887: Accents and special characters are replaced with question marks icons
- #4889: Endless loops Maturo NCD AT after communication is out of sync
- #4890: Position Maturo NCD AT is not alway set exact by driver
- #4891: Maturo NCD TT minimum step size of 3 degrees
- #4893: VISA (GPIB) controlled RGN error Invalid buffer mask specified
- #4894: RadiMation Free can't connect with RPR Power meter
RadiMation version 2013.1.13
- #1085: Field measurement during CI testing
- #2657: A graph of an AD-Input should auto scale the y-axis to also show the limits that are relevant for the AD-Input
- #2853: Multiband Report generator code for an ad-channel should also be available as a pre-determined code
- #2904: EUT ad channel monitoring scale is wrong, no information is displayed.
- #3031: AD Graph 1 does not work in multi band
- #3236: Max limit is not viewable when all the measurements are above the limit
- #3564: R&S ESVS device driver should be updated to use double parameters
- #3867: Net power and calculated current on the CAN-bus
- #4167: Power meter driver for fsiq / FSEK and fsp
- #4171: Report generator: keyword for step frequency is not working
- #4460: RadiCentre (RadiTower) errors
- #4582: Wrong peak is measured with final peak measurement selection
- #4615: Error 'Settings conflict: step size clipped to nearest value allowed'
- #4622: Error printout ||MAX HOLD PEAKS GRAPH||
- #4642: Modulation settings are not correctly saved
- #4675: Graph keyword gives 'Object variable or With block variable not set' error
- #4682: Selected peaks do not match with selected peaks in the peak table
- #4711: Peak zoom-in is not working correctly if a limitline file is specified
- #4728: All peaks above the limit dB value is not retained
- #4745: RadiMation shortcut should have version number
- #4753: HP8542E Receiver is measuring only one sweep and only one band
- #4764: RadiCentre problems
- #4824: Setting conflict error with a Agilent Scanning Receiver
- #4859: RadiPower misses the last part of a pulsed signal
- #4864: Driver: EMGen 7003-001
- #4865: Controlling a RadiGen is raising a dialog with the message: Operation is NOT completed
- #4868: Remaining test time is shown in 12:04:30 AM notation
- #4870: Installation on a PC cannot be completed due to network location configuration
- #4871: Limit values of ad-channels are not used in the scaling of the ad-channel graph
- #4872: Installation of drivers.exe generates an error regarding registration of RADCANoeWrapper
- #4879: How should the magnetic field antenna factor be specified?
- #4883: Wrong readings during conducted immunity fixed current test
- #4888: RadiCentre (RadiTower) timeout expired before OPC
- #4895: Driver: RSS1006A Device Not Connected
- #4896: RSW1021N and RSW1024S errors 'Not all bytes are written to the GPIB board within the timeout period'
RadiMation version 2013.1.12
- #4108: Report generator keywords for final measurement
- #4743: Driver: no communication and motor not moving with Emco 2090
- #4813: RM crashes during search of DARE products
- #4822: Unit for currentprobe is not set to dBuA, but remains dBuV
- #4837: Initializing a powermeter while on the envelope or burst mode tab, is not working correctly
- #4844: Changing the sample rate for the RadiPower in envelope trace mode is not working, if 1E+3 notation is used
- #4857: RadiPower RPR3006P measurement should be corrected for the Modulation bandwidth that is used by the EUT
- #4866: Antenna tower goes down to hard limit and gets stuck
RadiMation version 2013.1.11
- #4406: No coupling available with Synergy database
- #4817: Menu Help > Suggestion URL is opening DARE homepage
- #4819: Continuity between LLF and COR
- #4848: Error message when printout a Multiband test
- #4856: IF overload detected error with R&S ESU 8 EMI
RadiMation version 2013.1.10
- #4242: Issue measuring time setting in MB
- #4269: Settings (RBW) differ in comparison with measurement results
- #4541: Driver: Bonn amplifier BSA 0125-125
- #4596: Problem with PMM in latest version
- #4716: Reference Level is not correctly set during final measurement
- #4718: COR file graph is not correctly drawn
- #4750: RadiMate AD converter is not working
- #4752: PMM 9010 timeout errors
- #4754: EMCO 4825/2 LISN error Start frequency was too low
- #4758: Detector Keywords for report generator template
- #4759: PMM is not detecting that it isn't getting any real data
- #4763: PMM 9010 Error detected during start sweep measurement
- #4765: RI Uniform Field Area calculation form
- #4766: Frequency Change Mode configuration button is not present
- #4781: Driver: Agilent N5171B Signal generator
- #4782: Driver: AR 1000W1000D amplifier
- #4783: Driver: AR 2500A225 amplifier
- #4785: Configuration to remote control RSW1002E on 12V
- #4786: Not possible to printout a MB test
- #4791: RI Fixed power test in a sequence doesn't remember that a frequency list should be used
- #4798: R&S ESAI EMC receiver
- #4803: Move and Turn during measurement settings are disappeared
- #4807: RDM1004A with RadiCentre 1 is not connecting
- #4816: Driver: Wandel & Goltermann EMR-10 + EMR-11
RadiMation version 2013.1.9
- #4285: Can't change order of the frequency band (MB) for sequence testing
- #4606: Merging graphs is loosing correction
- #4695: Removing an information item, selects an unlogical next information item
- #4704: Test window opens to large for the screen; buttons are not visible
- #4721: Exit from a multiband test requires two clicks on Close button
- #4722: Advantest R3361B Analyser event timeout
- #4723: Graph lines are not correctly enabled and disabled
- #4735: 4446A preselector is searched when deselected in driver settings
RadiMation version 2013.1.8
- #1534: Large System fonts will make menu labels unreadable
- #4250: Text is not readable in Final peak measurements frame
- #4374: .NET: crash during continuous powermeter readout
- #4415: PMM OR03 communication issue
- #4451: New default installation on Windows 7 is not able to write to the CONFDVDR directory
- #4453: No quasi-peak detector with Analyser N9000A
- #4468: Window Client Logo: install to the correct folders by default
- #4478: Driver: Signal generator Wiltron 68369B (10M - 40GHz)
- #4480: Driver: Field probe Chauvin-Arnoux EF2A
- #4490: Driver: Agilent N9010A Signal Analyser Scanning Receiver
- #4527: Driver: EMI average detector setting
- #4543: Strange Quasi-Peak measurements with Agilent N9000A
- #4573: Printing the graph in the Powermeter Virtual instrument raises an error
- #4608: Correction of the powermeter in Burst mode
- #4609: Customization of the graph is not remembered for the burst test
- #4618: ESU with two bands and one sweep is not performing measurement
- #4628: First point ADC measurement is sometimes wrong
- #4632: Number of sweeps to be taken is not visible on small screen
- #4640: An object reference error is raised when open ESD test
- #4645: CE Multiband hangs when first band is using different trace detectors compared to second band
- #4646: CE Multiband keeps repeating the last band
- #4650: Graph control is always showing logaritmic
- #4653: Not possible to use CTR+A inside Note option in TSF
- #4678: setup program is aborted when .NET framework is going to be downloaded on a german pc
- #4691: Surge status window, doesn't show Pass/Fail option
- #4694: Possibility to remove multiple 'Information items'
RadiMation version 2013.1.7
- #4095: Before action is deleted after the test has finished
- #4425: Long delays between multiple tests in a very large sequence
- #4515: Error during zeroing of RadiSense
- #4532: Limit Lines displayed outside graph in report
- #4610: Driver: D.A.R.E!! Instruments RPR3006P
- #4614: Limit lines overlap the legend
- #4619: Driver: ETS HI-2200 Field Probe
- #4623: Cannot use R&S NRPZ51 above 6 GHz
- #4624: No coupler device driver is present for the internal coupler in the AR20ST1G18A
- #4627: Driver: ESPI3 receiver is not set to reference level from TSF
- #4629: In some specific situations the wrong test files are deleted
- #4647: Stopping an emission measurement with a R&S ESI, very often generates an EABO GPIB Error
- #4648: Driver: Anapico APSIN20G
- #4671: Driver: Agilent 4418B
- #4672: Insert specific attachements for report generator
- #4673: Driver: Agilent N1914A
- #4683: Driver: Prana AP32 MT225
RadiMation version 2013.1.6
- #4604: R&S ESI as powermeter takes a very long time to measure 20 Hz
- #4603: TSF is marked as changed, even if nothing is changed
- #4602: Customization of the graphs restored to the default if 'Use this as default' is presssed
- #4600: Adjust the amplifier protection
- #4583: System calibration Attenuation / Gain Unhandled exeption error
- #4581: Frequency unit in the limitline configuration seems to be specified in Hz, but is interpreted as MHz
- #4580: Changes in the unit of a limitline are not directly updated in the grid with limitline values
- #4579: Limitlines with a dBpT unit are still shown as dBuV/m limitlines in a multiband emission test
- #4578: Limitlines configuration window doesn't show very small/big frequencies
- #4577: Multiband test is not able to change to the next testsite due to frequency limitations in the testsite
- #4501: H-Field with no measurement data
- #4500: Receiver is set to wrong reference level
- #4378: Error Software Protection Key when upgrading software
- #4371: Higher value then 1000 is not displayed in COR file
- #4278: Crash during initialising R&S FSH6 Scanning Receiver
- #4169: AD Channels are swapped
- #4146: Changed sorting of AD channel configuration is confusing AD Convertor readout
- #4057: Copy to clipboard is forgetting the last item
- #4031: AD-Channels are changing after saving EUT
- #3455: Error message is displayed when changing from lower band to higher band with different equipment
- #3124: Double clicking the main window for reviewing the test sometimes opens a different test
- #2285: Port RadiMation to modern development tools (.NET)
- #708: new file is saved while No is selected.
RadiMation version 2013.1.5
- #4569: Reading out a powermeter in Free mode, is using maximum CPU load
RadiMation version 2013.1.4
- #4585: Possibility to dump the waveform of a LeCroy oscilloscope to a CSV file
- #4574: ESIB 26 preamp settings are not used
- #4536: Unhandled exception for object name icon
- #4535: Setting the preamplifier to 0 dB doesn't seem to be remembered
- #4533: Resizing main EUT test data columns is not correctly done
- #4504: Looping device check
- #4428: Error with ESI40 as powermeter
- #4396: Administrator password is unexpectedly changed/corrupted
- #4337: TDEMI soft-reset
- #4336: Abort scan on TDEMI
- #3573: Amplifier control panel can not be resized so you can see the complete status message
- #3304: Driver: AR 350AH1 (GPIB + USB)
RadiMation version 2013.1.3
- #4330 RI-SB manual mode, some texts are partly hidden
- #4529 Buttons are aligned on the right side, not very close to the other controls on the window
- #4444 .NET: cosmetic changes to setting windows
- #4368 not completely readable receiver settings should be extended
- #4249 .NET: window dimensions of settings can be completely resized
- #4513: The cancel button is not responding on the environmental window that is shown during the initialization
- #4512: Graph and table controls are not correctly initialized
- #4511: A 'Save As' dialog is displayed when the Powermeter window is closed in RadiMation Free
RadiMation version 2012.1.23
- #4491: Wrong LLF is removed
- #4485: Serial communication using Agilent IO libraries is not working
- #4459: .NET: visualisation frequency bands is not working correctly
- #4450: After installation it is still required to download and install .Net Framework 4
- #4448: Move message of restart of PC to the end of the setup program
- #4446: .NET: X+k*s module; issue with selection, calculation and crash
- #4445: Driver: LAN control of Maturo NCD Antenna Tower
- #4443 .NET: default window is showing events while unchecked
- #4441: .NET: crash after closing a new LLF without saving
- #4439 unlogically tab sequence for Default Address Information
- #4437: .NET: no information and wrong deleting of Cables
- #4436: .NET: no confirmation and wrong deleting of AD-channel(s)
- #4434: .NET: no menu and message descriptions dutch language
- #4433: .NET: adding and deleting two folders at once
- #4427: Driver: Agilent 54832B as an oscilloscope
- #4426: Driver: Agilent 54641A as an oscilloscope
- #4419: .NET: crash customized query customer database configuration
- #4407: .NET: crash while looking up EUT data
- #4402: .NET: typo tab Reports information line and Eut information
- #4393: .NET: Not all TSF files are deleted leaving a messed up EUT file and test summary
- #4383: .NET: confusing highlight selection
- #4376: .NET: LLF is not correctly showed
- #4373: .NET: when closing 'Select TSF'-window the 'New TSF'-window is opened
- #4367: .NET: crash after closing an adjusted test
- #4366: .NET: not possible to Save TSF when adjusting a test
- #4365: .NET: no 'save changes?' is asked when closing an adjusted test
- #4362: .NET: no units for modulation settings
- #4360: .NET: adjustment of constant power settings is not correctly done
- #4348: .NET: report generator arrow up/down doesn't work
- #4347: .NET: report generator can't use whole selection
- #4344: .NET: object reference error when using peak info table
- #4343: .NET: row is shifting up completely
- #4339: .NET: deleting a TSF file from list causes crash
- #4335: The up and down buttons in the Report Generator Configuration window are not working
- #4330: .NET: crash when test is directly closed after message
- #4318: .NET: some columns are reduced to an unreadable size
- #4315: .NET: some buttons, labels and text are misplaced
- #4314: .NET: when re-opening test re-selecting test equipment is needed
- #4313: .NET: folder name is changed to no description available
- #4312: .NET: screen unexpectedly disappear to background
- #4311: column width, location and sorting are unexpectedly changed
- #4310: sorting peak table information is illogical
- #4292: Can't include tests for X+k*s calculation
- #4289: .NET: Show Pulse Shape has wrong default window size/can be full resized in ESD, EFT and Surge
- #4288: Absence tab labels in ESD setting window
- #4266: Error and 'Loading Data' with ESD testing
- #4228: Issues frequency range, axis is not aligned ESMI receiver
- #4225: Issue Walter and Golterman field sensor EMR300
- #4203: Ignored settings powermeter gives incorrect measurement
- #4197: Send Crash dump dialog is shown when RadiMation is closed with visible sequence file
- #4191: HP33250A signal generator switching between -22 and -56dBm
- #4185: Readout of WG EMR 30 is not working
- #4138: Driver: Frankonia FC 05
- #4036: How should the CISPR25 measurement (with dBuV unit) be performed?
- #3813: using kV instead of v unit in ESD test
- #3712: Peak table sorting by frequency does not work correctly
- #3710: Sorting of peak table not saved
- #3638: Driver: Rohde & Schwarz HFH2-Z2
- #3600: sorting of peak table
- #3483: Sorting peaks according to frequency does not work correctly
- #2556: Changing different templates in the report window
- #1833: The peak-list at the end of the measurement should be sorted automatically (by Frequency)
RadiMation version 2012.1.22
- #4408: .NET: wrong frequencies measured during final peak measurement
- #4405: .NET: changes EUT info not recognized: no save confirmation
- #4404: .NET: specify error to 'can't write to file' or change in warning
- #4403: .NET: adjust the default row width
- #4395: .NET: Coax set Probably an unknown RadiMation Error
- #4382: Auto launch RadiMation doesn't close setup
- #4375: .NET: report generator creates strange signs
- #4369: Driver: PMM EP-600 with X, Y and Z-axis support
- #4350: .NET: report generator can't open when clicking on the Template file
- #4323: Information check failed error should be warning
- #4279: Wrong numeric data is exported
- #3931: Easily create a reference graph in emission measurements
- #3688: showing old measurement results during new measurement
- #3246: How to get the displayed data
RadiMation version 2012.1.21
- #4416: Change build-up of image points graph to logarithmic
RadiMation version 2012.1.20
- #4381: Driver: PMM EP-603
- #4357: .NET: modulation can't be turned off
- #4353: final peak measurement is done on wrong frequency with ESI40
- #2635: Starting level is not used when disabling the power delivery test
- #2630: Powerdelivery test cannot be disabled for Amplifier harmonic radical test
- #2113: Driver: RadiGen
RadiMation version 2012.1.19
- #4355: RadiPower is not detected correctly if NI-VISA is not installed
- #4351: .NET: report generator error object reference
- #4349: RadiPower device driver raises error: VISA received more event information of the specified type than the configured queue size could hold
- #4342: .NET: incompatibility of test results
- #4341: Driver: HI-6100 Field Monitor
- #4334: Driver: Amplifier Research FM 5004 communication issue
- #4331: When CTRL+A is pressed some unknown windows are shown
- #4327: Activate the calculated gain regulation as the default amplifier regulation
- #4326: .NET: improve view table information
- #4325: .NET: enlarge name information
- #4322: .NET: crash and wrong window name customer database configuration
- #4309: .NET: user color and thickness settings are not working
- #4308: .NET: equipment list is not updated
- #4295: Advanced options hows items as bold when it shouldn't
- #4294: Can't remove last EUT Information
- #4293: Remove new EUT Information does delete all
- #4291: Can't add Basic or Product standards
- #4286: Crash while checking communication R&S FSH3
- #4234: RadiCentre: unknown device type
- #4233: Readout of the RadiMate 4 hangs for a minute
- #4190: Zero frequency issue in the graph, testing from 40Hz
RadiMation version 2012.1.18
- #4363: Starting Level Signal generator can't be configurated
- #4297: Report Generator settings unexpectedly pop-up the Select template window
- #4296: Scrolling is disabled in EUT Information window
- #4249: Window dimensions of settings can be completely resized
- #4247: Display graph isn't working properly
- #4206: RadiMation is deleting the wrong test files
- #4078: Driver: HI6153
RadiMation version 2012.1.17
- #4298: Frequency range error NBM-520 Broadband Field Meter
- #4263: Claim that AR 50S1G4 doesn't exist
- #4262: Driver: Amplifier AR 40S1G4
- #4240: Error when calibration file is selected that is using another amplifier
- #4239: Not practical to have Multiband tests in a sequence
- #4229: Standard button selection after opening test data from EUT
- #4227: Driver: EMCO 5402 GTEM Cell
- #4224: Error report generation Data can not be pasted in Excel
- #4199: Showing the height graph in the RE Multiband test crashes the software
- #4192: Verify if CS106 test is not supported by RadiMation
- #3738: GUI pauses RI test
RadiMation version 2012.1.15
- #4222: Several errors with .NET version on Windows XP
RadiMation version 2012.1.11
- #4186: Command error (4198) during generation of report
RadiMation version 2012.1.9
- #4142: Testsite in band in multiband emission is reactivated even when no peaks should be detected
- #2952: Checkbox is not grayed out when viewing test results of Pulsed Immunity EFT
RadiMation version 2012.1.6
- #4080: Vertical grid lines of the graph are drawn on top of the measurements, which could hide measurement results
- #3457: Multiband menu name change
- #3396: Status screen to small for messages
- #2674: Popup menu in multiband test with list of test-sites is not sorted
- #611: Immunity Manual mode directly accessible from main menu
RadiMation version 2012.1.4
- #4438: .NET: two menu options will open same window
- #4178: TDEMI IF-Overload check
RadiMation version 2011.1.5
- #3453: The "Status" column, in the "Test Initialisation" dialog, is not sized correctly by default
RadiMation version 5.9.0
- #3829: Driver for Amplifier Bonn BSA1050-250/100
- #3828: Unit is single band graph do not change after changing the unit in the unit configuration of the result file
- #3824: Configurable alarm detection is always reporting alarm, even when the correct response is returned.
- #3819: Network Analyser EUT Calibration, using a R&S ZVK, crashes with an overflow
- #3782: Driver: Antenna Schwarzbeck STLP 9128D and Elektromectrics EM 6954
- #3780: Driver: Pre amplifier Schwarzbeck BBV 9718
- #3755: RadiMation is setting up the ESPI with incorrect RBW during CI-testing
- #3746: Peak zoom-in doesn't seem to work in multiband emission
- #3745: Turntable doesn't turn a complete turn when it starts on a negative angle
- #3736: I can not select apply modulation outside dwell time when selecting pulse
- #3735: Driver: SCHWARZBECK TEMZ 5231
- #3734: Driver: schwarzbeck antenna's FESP 5132 + FESP 5133-7/41
- #3732: Driver: Milmega Rf1000
- #3724: Radisense 4 over RS232 does not set the filter setting
- #3720: During emission test: crosses instead of line
- #3701: Multiband Radiated Emission: Default value in dropdown list for Turntable Angle specifies 235°
- #3652: Driver: Kalmus LA200U
- #3651: Driver: Kalmus 122FC
- #3637: Driver: Logimetrics AN 8100Q
- #3636: Driver: Logimetrics AN8200G
- #3615: Multiband-Conducted Immunity: data is not displayed
- #3544: Device driver for Narda NBM550 uses the wrong baudrate
- #3447: Conducted Immunity Tests are slow in 5.5.7 in comparison to 5.3.18
- #3354: CARAD interface: two commands get combined to one
- #2597: In multiband immunity the cancel button only reacts on the end of the dwelltime
- #507: Overload indication of an analyser/receiver is not detected
RadiMation version 5.8.39
- #4411: VISA error while using USB communication for device