Difference between revisions of "Chapter 7"
(→Frequency window) |
(→Multiband) |
Line 88: | Line 88: | ||
== Multiband == | == Multiband == | ||
+ | The radiated emission Multiband test, is a very flexible and powerful test to perform radiated emission tests. | ||
+ | When the multiband test is performed, it will first perform a prescan measurement on all the specified locations. After this prescan is finished the temporary results are shown, and some peaks are automatically detected and shown. At this moment those peaks are NOT measured yet with the final measurement detectors. The end-user now has the possibility to add additional peaks, or do some other peak investigation/changes. | ||
+ | When the end-user is satisfied with the list of detected peaks, he can continue with the final measurement, by pressing the {{ScreenElement|Automatic}} button again. | ||
+ | Then the peaks will be measured with the detector that are selected in the {{ScreenElement|Final peak measurement}} box. Only after those measurements are finalized, those detector values will become available in the list of detected peaks. | ||
[[Image:Radiated Emission Multi Band Configuration Window.png]] | [[Image:Radiated Emission Multi Band Configuration Window.png]] | ||
− | The test engineer can configure a complete radiated emission test in the Radiated Emission configuration menu. The test engineer can change the frequency range from one frequency band or the RBW of several bands. It all depends on the selection of bands. When several bands are selected and some options are blank, this means that these settings in bands are different. For example if the test engineer selects both frequency band in | + | The test engineer can configure a complete radiated emission test in the Radiated Emission configuration menu. The test engineer can change the frequency range from one frequency band or the RBW of several bands. It all depends on the selection of bands. When several bands are selected and some options are blank, this means that these settings in bands are different. For example if the test engineer selects both frequency band in the window above, the frequency range will be blank. If the engineer sets the start and stop frequency, all the selected bands will have the same frequency setting. One of the advantages of this feature is the quick settings of several settings like the RBW or the peak detection. |
===Frequency window=== | ===Frequency window=== |
Revision as of 10:17, 27 October 2016
Radiated emission
Introduction
- Multiband
- Manual Mode
- Burst
- Anechoic chamber
- Fixed height
- GTEM
- Mil Std Magnetic
- Manual mode (Singleband)
- Merge
- X + k * S
Starting a radiated emission test
Loading a TSF file
Test results screen
You can access the test results of the performed test by selecting the test in the EUT window and then press the Info button. For the Multiband test, the window that is appearing looks a lot like the manual mode screen, the reason that it looks the same is because it is. You can continue testing after you finished a test, just select the testsite you want to use and you can continue testing.
Testsite information
An radiated emission test, requires that a testsite is used that can have the following components.
Device name | Tab in RadiMation® configuration window |
Spectrum analyser | Devices 2 |
Pre amplifier | Devices 2 |
Antenna | Devices 1 |
Turn table | Devices 1 |
Antennatower | Devices 1 |
1 | Cable preamp -> analyser | Cables |
2 | Cable antenna -> preamp | Cables |
Multiband
The radiated emission Multiband test, is a very flexible and powerful test to perform radiated emission tests. When the multiband test is performed, it will first perform a prescan measurement on all the specified locations. After this prescan is finished the temporary results are shown, and some peaks are automatically detected and shown. At this moment those peaks are NOT measured yet with the final measurement detectors. The end-user now has the possibility to add additional peaks, or do some other peak investigation/changes. When the end-user is satisfied with the list of detected peaks, he can continue with the final measurement, by pressing the Automatic button again. Then the peaks will be measured with the detector that are selected in the Final peak measurement box. Only after those measurements are finalized, those detector values will become available in the list of detected peaks.
Frequency window
Turntable window
Start Angle | Sets the start position of the turntable. |
End Angle | Sets the stop position of the turntable. |
Steps | Sets the number of steps for one turntable rotation. |
Turn during measurements | This will force the turntable to continuously rotate during the test and will disable steps setting. |
EUT angle offset | The offset of the EUT compared to the zero angle of the table. |
Optimize Angle | Optimized settings for searching the correct angle. |
Turntable window optimize angle settings
Area | Range over which the peak search will be performed (0 to 360 degrees max.). |
Step size | The angle step size. |
Antenna tower window
Max. height | Sets maximum height of the antenna tower. |
Min. height | Sets the minimum height of the antenna tower. |
Steps | Sets the number of steps to go from min. height to max. height. |
Move during measurements | This will force the Antenna Tower to continuously move during the test and will disable steps setting. |
Optimize height | The software will automatically detect on which antenna height maximum emission will occur. |
Optimize height settings
Area | Range over which the peak search will be performed (0 to Max height). |
Step size | The height step size. |
Antenna window
Polarisation
- Horizontal
- Vertical
- Both
Distance
Receiver settings
Traces window
Detectors | The detector settings (Peak, Average, Quasi-peak, RMS) used during sweeptime. This setting is generic. So it is possible to select multiple detectors, but only one detector can be selected due to the device limitations. In that case for each enabled detector the amount of sweeps will be performed by RadiMation®. |
Number of sweeps | The number op peak hold sweeps for each measurement can be selected. If this value is set to, for example 10, the software will measure the frequency band 10 times for each turntable and antenna tower position. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweep speed. For spectrum analysers and CW signals you can use a relatively low amount of peak sweeps, for example 10. But when you have discontinues signals like a spark bridge, you may need 100 peak sweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope. |
Peak detection window
No peaks | When no final peak measurements are required select the option no peaks. |
Maximum Peaks | With the max. peaks option, the user can limit the number of peak signals the software will detect and re-measure. |
Test site window
Limit lines window
All peaks above “xx” below the limit lines | In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 5 dB, the software will detect all peaks, which are higher than 5 dB below the limit line. |
Change Order window
So for Site 1 you want:
- Band A RBW 9 kHz
- Band B RBW 9 kHz
- Band A RBW 120 kHz
- Band B RBW 120 kHz
And for test site 2:
- Band A RBW 9 kHz
- Band B RBW 9 kHz
- Band A RBW 9 kHz
- Band B RBW 9 kHz
- RBW
- Testsite
- Frequency band
- Band A RBW 9 kHz
- Band B RBW 9 kHz
Manual Mode
Receiver settings
Turn table
Antenna tower
Test site
Final measurement
Peaks window
Take sweeps | Triggers the receiver of analyzer to take the configured amount of sweeps. |
Detect peaks | Detects the configured amount of sweeps in the measured spectrum. |
Measure Peaks | Measures all selected peaks with the configured settings in the final measurement window. |
Delete Peaks | Delete the selected peaks from the list. |
Peak actions | Allows several actions to be performed on the detected peaks, like select all and deselect all. |
Anechoic chamber
Frequency window
Stop | Sets stop frequency |
Turntable window
Start position | Sets the start angle of the turntable. |
Stop position | Sets the stop angle of the turntable. |
# | Sets the number of steps for one turntable rotation. |
Peak search | The software will automatically detect on which angle maximum emission will occur. |
- If a turntable device is specified in the selected test-site, RadiMation will control the turntable to turn to the specified angle.
- If no turntable device is specified in the selected test-site, RadiMation will assume that the EUT is already positioned on that specified angle. No actions on the turntable will be taken.
As long as a single angle is measured, and no turntable device is specified in the test-site, it is possible to use the Multiband Radiated Emission test without a turntable. So, if you specify the same 'Start Angle' and 'End Angle', and only specify a single 'Step' ('1'), then the test will assume that only that single specified angle should be measured.
Threshold | This value determines the acceptance criteria for a peak. |
Mode | The mode setting determines the way a peak is detected. |
- Intelligent
- Complete area
The Intelligent mode has the advantage that it stops with the scan as soon as a maximum has been determined. This could improve overall testtime as less movement of the clamp is possible. The Complete area mode has the advantage that the specified "Area" is measured, and that it ensures that the real maximum of the measurement value is measured.
Intelligent
Complete area
# | Sets the number of steps to go from min. height to max. height. |
Peak search | The software will automatically detect on which antenna height maximum emission will occur. |
Area | Range over which the peak search will be performed (0 to n meter). |
Step size | The distance for each step. |
- Intelligent
- Complete area
Intelligent
Complete area
Antenna window
Polarisation
- Horizontal
- Vertical
- Both
Distance
Receiver settings
Max. hold measurements
Traces window
Detectors window
Test site window
Limit lines window
Test options window
Record data | In the test options window the operator can select whether all RAW data must be stored or only the measurement results are stored. It should be noted that for full tracebility, all test data should be stored. |
Narrow band / broad band discrimination | RadiMation® supports two types of narrow band / broad band determination. |
Limits | Distortion type |
Starting a Anechoic Chamber test
Loading a TSF file
Test Procedure
Info screen
Settings | In the settings window all the used settings are displayed. |
Graph control | In the graph control window the scale of the graph can be changed. You can show or hide the table by selecting or deselecting the show table checkbox. |
Trace | In the trace window you can control the traces that are shown in the graph by selecting or deselecting the checkbox. If during a measurement a trace is not used then the check box is grayed out. |
Detector | The detector window gives you the possibility to show and hide the values measured on the peaks. If during a measurement a type of detector is not used for determining the value on the peak the checkbox is grayed out. |
Table | In the table all the measured peaks are showed. The table also shows the limit values for that frequency point when a limit was given at the beginning of the test. |
Test site information
Pre amplifier | Standard 2 |
Antenna | Standard 1 |
2 | Ant->Pre amp | Cables |
Measurement of peak amplitude and peak frequency
Fixed height
Introduction
Frequency window
Turntable window
- Intelligent
- Complete area
Intelligent
Complete area
Min. Height | Sets the minimum height of the antenna tower. |
- Intelligent
- Complete area
Intelligent
Complete area
Antenna window
Polarisation
- Horizontal
- Vertical
- Both
Distance
Receiver settings
Traces window
Detectors window
Max. Peaks | With the ‘max. peaks” option, the user can limit the number of peak signals the software will detect and re-measure. |
Test site window
Limit lines window
Test options window
Starting a fixed height test
Loading a TSF file
Test Procedure
GTEM
Correlation method
Measurement procedure
Test configuration
Frequency range selection
Frequency list selection
OATS settings
G-TEM settings
Receiver settings
Traces window
Detectors window
Max. Peaks
Test site window
Limit lines window
Test options window
Starting a G-TEM test
Loading a TSF file
Receiver settings
Reference | Sets the reference level of the receiver |
RBW | Sets the Resoution Band Width filter |
Data correction
Traces Menu
Traces
Detectors
- Peak detector
- Average detector
- Quasi peak detector
- RMS detector
Turntable and antenna tower position
Zoom function
X-axis | When this option is switched on, the frequency settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph. |
Y-axis | When this option is switched on, the amplitude settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph. |
Limit lines
Merge
X + k * S