  • #5730: PMM 9010F Incorrect start and/or stop frequency
  • #8391: Removing one LLF while there is another one with the same description in the list is removing both items
  • #8859: Driver: Keysight N9322C spectrum analyser
  • #9057: Report generator does not show units for environmental data of PI tests
  • #9070: Pre amplifier stays off R&S ESR26
  • #9092: Driver: BONN BSA 0101-50D
  • #9096: RPA0901A-050: Amplifier status timeout. Device refuses to change to the new requested status
  • #9099: Driver: Prana MT400 DC internal powermeter driver
  • #9107: specifying a pulse width time as '60 u' is not selecting us
  • #9120: Prana SX 120/100: Device Driver Function Not Implemented.
  • #9124: ESR 7 "Execution error*DISP:TRAC:X:SPAC LOG
  • #9135: Exporting EUT calibrationtest crashes radimation
  • #9146: AMP3(Low) 0.8G to 3.2GHz (SX120/100): Invalid Device Driver Handle
  • #9147: AMP2 upto1GHz 400Watt RS (MT400): Invalid Device Driver Handle
  • #9152: How to export a table with final measurements
  • #9165: Amplifier (ID1647) device not connected
  • #9166: Signal power went too high (-6 dBm) during power delivery test
  • #9175: RF Amplifier (ID1647): Device Not Connected
  • #9176: Signal power went too high (-6 dBm) during power delivery test
  • #9180: External Pulse modulation is not turned off
  • #9190: DARE!! Instruments RSS1006A: Zero operation took longer than 60 seconds.
  • #9193: DARE!! Instruments RadiField RFS2006B: Device Not Connected.
  • #9262: Bonn BLWA 0830-160/100/40D doesn't switch to other band
RadiMation version 2017.2.10
  • #5637: Excel Export error Object reference not set to an instance of an object
  • #5832: Updated Radiwiki Agilent U2000 series page
  • #5973: Absorbing clamp test raises System.OutOfMemoryException and crashes
  • #5995: Shield to Ref. is not listed as a EFT cable test
  • #6135: Driver: Teseq ITS6006 for IEC61000-4-3 & CISPR11 as Switch Matrix
  • #6696: Driver: Schloeder SFT1400/SFT1420
  • #6697: Driver: Schloeder CWG500
  • #6979: Create a correct application manifest file for the executables
  • #7035: Driver: Teseq ITS6006 functionalities for EUT monitoring interface
  • #7706: Start and Stop frequency Report generator code behavior.
  • #7949: Unable to select peaks from the graph
  • #7997: Driver: R&S URV35 Level Meter
  • #8005: Replace icons to text for ReportGenerator Test pass, fail or not-tested.
  • #8035: Ability to use the R&S URV35 Powermeter by RS232
  • #8039: Remove and replacing LLF from test data file 1 is also removing the LLF from another test data file in the EUT
  • #8087: Driver: R&S ESW26 internal high-pass filters 2.4 and 5.6 GHz
  • #8105: Powermeter virtual instruments keep disabling device specific settings after first measurement
  • #8131: RadiMation memory leak when triggering powermeter
  • #8183: Rohde & Schwarz, ESIB 26: Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO)
  • #8267: Burst Measurement does not show the combined graph data in the graph
  • #8524: Incorrect mouse curser value shown in graph
  • #8649: Missing error description Radisense10 Err 723
  • #8659: Driver: Keysight N9914A Receiver spectrum analyser & powermeter
  • #8660: Object reference error in the VI Powermeter window while measuring pulsed power using the RPR2006P
  • #8707: Driver: BONN BSA 0140-200
  • #8708: Driver: SCHAFFNER PCJ 9201B
  • #8714: Driver: MOCS-MFD + OTM301
  • #8716: RadiCentre justation, powermeter freezes
  • #8747: R&S ESRP cannot be used in combination with 100 Hz RBW
  • #8777: Driver: Switchmatrix: TOYO Corporation NS4900-series
  • #8806: Amplifier ETS-Lindgren 8100-060
  • #8808: Problems controlling the INNCO CO2000 controller when no Tilt mast is connected
  • #8810: Frequency Range limited AC DC coupling
  • #8826: Missing time-out / retry for Prana amplifier device driver
  • #8827: Prana RFOff check
  • #8838: Marconi 2024 Retry: Query Unterminated
  • #8840: Amplifier stays in operate
  • #8848: Wrong states PRANA MT400
  • #8877: Count column is always shown in the event-log
  • #8878: "Report Error" Error: Please fill in all the user information fields
  • #8879: Driver: RadiAmp amplifiers
  • #8883: Changing Engineer Value was either too large or too small for an Int32.
  • #8886: Amplifier: Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO).
  • #8890: Unable to create new TSF for Radiated Immunity Substitution test
  • #8891: Missing harmonics graphs
  • #8897: An unknown error is raised, when the RadiCentre is not yet ready with initialization
  • #8901: ESCI Scanning Receiver Option not available;DET1 CAV
  • #8902: Dragging the marker of an emission graph: The method or operation is not implemented
  • #8910: RPR3006W wrong command Mode 4
  • #8912: Object reference clsIDocumentBase.UpdateAllViews update graph
  • #8913: RepGen VDI test is not included in ||EUT|ALL IMMUNITY TESTS||
  • #8914: CE MB infinit zoom peak loop
  • #8919: Get amplifier information Object reference error
  • #8923: ErrorReport dialog should show a red cross when not all the fields are filled in
  • #8937: Manual controlled antenna tower doesn't show a message for the horizontal polarization at the begin of the test
  • #8939: Driver: Configurable Turntable
  • #8947: How to control a camera PTZ positioner using the configurable turntable device driver
  • #8958: Add a required checkbox to the ErrorReport that the end-user agrees that his data is used
  • #8964: Anritsu Command or parameter is not (correctly) confirmed and is probably raising an error. EMF
  • #8982: Destination array was not long enough. Check destIndex and length, and the array's lower bounds
  • #8985: HI6153 checkdevice passed while not connected
  • #8990: Repeat count is not remembered in MB EFT TSF
  • #8997: Y-axis of the multiband emission height graph is shown logarithmically to 100 m
  • #9030: Dongle failure after installation of Windows 10 April 2018 Update
  • #9034: Missing shortcut keycodes
  • #9035: ||PEAK TABLE|| report generator code does not include peak number and frequency
  • #9036: Before and After actions in testsite are showing strange DARE.RadiMat.clsCommand.... names
  • #9038: Sunol SC99V antenna tower should by default use a small delay
  • #9048: Driver: Amplifier NoiseKen NA80MF3G201003
  • #9064: Show a clear error that the R&S rsnrpz_32.dll cannot be loaded
  • #9073: Driver: R&S FSV3 tracking generator
  • #9074: Unable to add TSF sub directory in Voltage Dips and Interrupts
  • #9077: Surge Random angle on Thermo Scientific EMCPro is always 2 degrees
  • #9080: RadiMation 2017.2.10 raises Error nr.: 536871935 on Windows 7 SP1
  • #9091: Driver: Schwarzbeck 420NJ
  • #9128: RadiField RFS2006B Error 574. Maximum number of retries reached on communication over the link.
  • #9138: Driver: CHASE CBL8111
  • #9143: Driver: Aaronia Hyperlog 7060
  • #9149: Updated Radiwiki page for the Agilent U2000 series
  • #9155: EMI radiated sweep very slow with test receiver R&S ESPI3
  • #9162: Radimation - CAN-Bus link
  • #9164: CISPR14 voltage probe transducer not added?
  • #9174: Object reference not set to an instance of an object.
  • #9183: Subject: LeCroy WaveSurfer Channel B: Binary block buffer couldn't be found
  • #9191: ErrorReport: "LUMILOOP LS Probe 1.2 Y-axis (mode 0): Setting the 'Computer Interface Serial Number' raised an error. Is the identifier correct and present in the system? "
  • #9198: Report: Template uses macro(AutoFitThisTable)
  • #9223: Kalmus 757LC (Hi Band): Impossible to get description string for this Error! Probably an unknown Windows Error. Code: 236446416 (0x0E17E2D0)
  • #9225: Software Protection Key could not be found on Windows 10 update
  • #9235: Driver: Schloeder EMCL-20 & Schloeder EMCL-35
  • #9255: Improve driver of R&S SME 06 to use High frequency accuracy
RadiMation version 2017.2.9
  • #4069: Driver: Generic Switch Matrix / EUT Controller
  • #5541: Explanation on how to determine the antenna diagram of an antenna
  • #7652: Configurable Injection Device custom frequency range is not retained
  • #8163: Driver: AR 200S1G6
  • #8236: Check device fail Narda 3006
  • #8257: Crash with Fluke 8845A Multimeter
  • #8262: Burst Measurement not working anymore: Unable to cast object of type 'DARE.RadiMat.clsSettings' to type 'DARE.RadiMat.IDataStorageGraphLineChecker'
  • #8453: Possibility to select AC or DC coupling in the advanced configuration window of the Tektronix TDS-series oscilloscopes
  • #8601: Driver: Audio analyser U8903B
  • #8641: Driver: tektronix MSO3014 as Oscilloscope
  • #8724: Driver: AFJ LS16C and LT32C LISN's
  • #8740: Driver: haefly PSurge 4.1
  • #8769: Driver: Anritsu MS2830A
  • #8776: Driver: passive devices
  • #8808: Problems controlling the INNCO 2000 controller
  • #8829: Prana MT 400 over ethernet issues
  • #8836: Enable the sequence list control during testing for PI tests.
  • #8856: Use ESW analyser in Receiver input mode when using it as powermeter.
  • #8869: selecting an item in a grid-list is not selecting another item
  • #8874: Driver: FP-Surge 32.1 IP4A
  • #8885: QP columns in REMB are not shown in Peak table by default
  • #8895: Monitor and control current during CI MILSTD tests
  • #8915: AM modulation using square wave is not applied to Agilent 332x0A
  • #8926: Driver: Lecroy HDO09204
  • #9015: Driver: Schwarzbeck automotive antennas
  • #9022: Driver: Passive antenna driver: R&S HL050
  • #9049: KC Standard
  • #9058: Driver: Amplifier BONN BLWA 0830-160/100/40D
  • #9061: Driver: Antenna HL046E
RadiMation version 2017.2.8
  • #8044: Expand measuring points for Network Analyser (S-Parameter) test
  • #8286: Extend R&S ESR3 driver to allow selection of CISPR-Average or (old-)Average as Average detector
  • #8287: Extend R&S ESCI driver to allow selection of CISPR-Average or (old-)Average as Average detector
  • #8513: ESPI DET2 CAV cispr option not available
  • #8544: Rohde & Schwarz SME 06: Specific System Error: 105, "Frequency underrange"
  • #8598: Compression gain calibration slow
  • #8692: Multiband RE MB 32 different positions is too slow.
  • #8701: Report generator Example of example EUT raises an error
  • #8761: Driver: Schaffner / Teseq SMR4503
  • #8764: Frequency and time column empty CI manual adding points
  • #8788: Turntable angles are not saved to/loaded from the Multiband Immunity TSF file
  • #8790: Unable to close test, after selecting Remote Desktop
  • #8813: Preselector of Keysight N9038A MXE cannot be disabled
  • #8816: Turn table control no longer working
  • #8842: Driver: Add missing TESEQ CDN
RadiMation version 2017.2.7
  • #6095: Driver: Bonn Elektronik BTA 0122-3000
  • #6203: Signal power went too high (-6dBm) during power delivery test with Amplifier Bonn BTA 0122-3000
  • #8001: Exporting codes with band numbers should be sorted on the start frequency.
  • #8206: Integrate the french translation
  • #8215: R&S ESPI 3: Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO)
  • #8694: Missing coupling for NSG3040
  • #8706: Keysight N5171B Narrow pulse modulation
  • #8739: Report Generator code infinit loop
  • #8758: R&S SML Set Carrier level is not checked
  • #8759: Missing equipment in testsite FieldSensor
  • #8765: Missing first band in CE MB test
  • #8771: R&S ESW Input switching
  • #8846: Driver: RadiLink 8 channel as cable
  • #8857: Field strength is not reached. Calculation uniform field area.
RadiMation version 2017.2.6
  • #7042: Implement disabling of SRQ's for Bonn amplifiers
  • #8489: RadiLog crashes on Windows 7 SP1 32-bit
  • #8529: Com already opened with 2 AD Channels
  • #8551: ETS amplifier from 8000 series cannot be controlled over GPIB
  • #8650: Warning message on Amplifier screen: RadiMation hangs
  • #8589: ETS-Lindgren 8100-015 amplifier has additional states
  • #8617: Possibility to clear multiple graph customizations at once
  • #8639: Cannot select 0.1 mA current testlevel in MIL-STD Conducted Susceptibility calibration
  • #8677: More than 1000 MB of RAM usage in a MB emission test since version 2017.2.5
  • #8682: Flashing band selection
  • #8683: Supporting the Preamplifier of the new TDEMI X depending on version.
  • #8684: Object reference RE CHamber peak measurement
  • #8685: AD Channel colors are the same
  • #8686: ETS-Lindgren 8000-004 always remains in power-on, and cannot be powered off during closing of RadiMation
  • #8687: Excel export table shifted.
  • #8688: Internal/external refference toggle Marconi/IFR 2023
  • #8689: Note field cannot be larger then 1000 characters.
  • #8699: ESPI 3 Single trace returns 0
  • #8710: C# Events Object reference
  • #8717: HAMEG HMS1010 Undefined header
  • #8729: Lumiloop Range selection
  • #8862: Cannot find sensor powermeter in the currently active testsite
  • #8921: Spectrum Analyser missing in test site
RadiMation version 2017.2.5
  • #8653: Singleband emission data cannot be loaded anymore
  • #8654: Correction is applied twice during final peak measurement in MB GTEM
  • #8656: MB Emission graphs show a lot of crosses when an ambient line is also used
RadiMation version 2017.2.4
  • #5003: Exporting a graphline as limitline and also add it to test itself is not working
  • #7706: Start and Stop frequency Report generator code behaviour.
  • #8229: Too much duplicated data is represented in the table
  • #8440: Source array cannot be assigned to destination array type
  • #8471: Performing final measurement on peaks during MB GTEM test is changing too often to the X EUT side
  • #8516: Ambient difference line is not detector specific
  • #8530: AD Channel 2 times in result table
  • #8540: Save TSF As Readonly is not working
  • #8570: After Action from testsite is shown twice
  • #8571: Also show the 'System Calibration' calibration graphs during the Attenuation/Gain EUT calibration
  • #8591: Correction files stores values as strings since 2017.1
  • #8592: Logarithmic interpolation on the value in a correction file should not be allowed for <= 0 values
  • #8629: Apply column custumization to the report generator
  • #8651: When a correction file is opened, often the logarithmic value interpolation is automatically selected
RadiMation version 2017.2.3
  • #3142: 10 Hz frequency steps not possible with Marconi 2024 signal generator
  • #6216: Error incorrect step frequency raised while setting should be disabled and not be checked at all
  • #6217: Stepsize setting is no longer disabled when an Analyser is used
  • #7798: Performance of the System calibration Attenuation / Gain is still not satisfying
  • #8062: Outside Points
  • #8157: Measurement of power with 100 Hz RBW on 2 GHz is not working
  • #8402: Cell is still kept active when COR is being modified, causing it that the current value is not being saved when using main menu items
  • #8439: RE Chamber, object reference at closing test finished
  • #8494: RadiLog error reports are not being send
  • #8502: Showing the graphs from the Attenuation/Gain system calibration in the EUT Calibration results
  • #8507: High accuracy cable/attenuation gain measurements
  • #8509: Error: 'Index was outside the bounds of the array', in the Powermeter Virtual Instrument
  • #8528: Multiple fixes frmVIBase, frmVIPowerMeter
RadiMation version 2017.2.2
  • #875: GTEM support accoording EN61000-4-20
  • #7804: Possibility to ignore unwanted ambients signals in emission measurements
  • #8498: Driver: AR Amp 60S1G6
  • #8499: Driver: AR Coupler DC7200A (1-6GHz)
  • #8500: Driver: FCC-BCICF-150 - CAL Jig
  • #8501: Driver: F-2000-32mm Current Clamp
  • #7695: R&S ESPC returns error The spectrum analyser is incapable of using the requested span. Correct the frequency settings(s) to within the hardware limits
  • #8065: Folder Selection improvements
  • #8159: Add Virtual Instrument window for a cable
  • #8232: Update the CD Start menu for 2017.2
  • #8245: VI SpecificSettings control resizes incorrectly
  • #8260: Device specific paremeters period update
  • #8261: LED Control in device specific settings