RadiMation Application Note 120: Difference between revisions
Line 8: Line 8:
This measurement can be done with two different methods:
# using a Vector Network Analyzer
# using a Vector Network Analyser
# using a signal generator and spectrum analyzer (optionally using an amplifier)
# using a signal generator and spectrum analyser (optionally using an amplifier)
In this application note it is explained how to perform the measurement using the signal generator method, because that type of equipment is more often available.
Line 16: Line 16:
== Reference measurement (system calibration) ==
To improve the accuracy of the final attenuation measurement, first a reference measurement is needed to determine the intrinsic attenuation of the system setup, without the EUT present. This thus accurately determines the attenuation of the used cables, any used artificial networks, other used equipment, and the amplitude difference between the signal generator and the spectrum analyzer. This reference measurement is therefor also often called the system calibration.
To improve the accuracy of the final attenuation measurement, first a reference measurement is needed to determine the intrinsic attenuation of the system setup, without the EUT present. This thus accurately determines the attenuation of the used cables, any used artificial networks, other used equipment, and the amplitude difference between the signal generator and the spectrum analyser. This reference measurement is therefor also often called the system calibration.
This reference measurement can be performed in {{RadiMation}} by using the Attenuation/Gain system calibration.
In a setup where a spectrum analyzer or receiver is used to perform the measurement of the generated signal, the setup is:
In a setup where a spectrum analyser or receiver is used to perform the measurement of the generated signal, the setup is:
[[File:AttenuationGainSystemCalibration.png]]
Line 39: Line 39:
[[File:HV-LVCouplingAttenuationSystemCalibrationTSF.png]]
When this system calibration is started, the signal generator will generate every frequency, and the generated power will be measured by the spectrum analyzer.
When this system calibration is started, the signal generator will generate every frequency, and the generated power will be measured by the spectrum analyser.
At the end of the system calibration a 'Save Calibration As' dialog is shown, which allows to save the result of the calibration to a .CAL calibration file.
It is advised to specify a useful location and filename for this system calibration file, so it clearly identifies the purpose and reason of the system calibration.
Line 63: Line 63:
[[File:HV-LVCouplingAttenuationEUTCalibrationTSF.png]]
When this EUT calibration is started, the signal generator will be set to every frequency again, and the generated power will be also be measured by the spectrum analyzer.
When this EUT calibration is started, the signal generator will be set to every frequency again, and the generated power will be also be measured by the spectrum analyser.
Once the EUT calibration is finished, the results of this EUT Calibration is stored in the EUT file, and available as one of the performed {{ScreenElement|Tests}} in the EUT file. Selecting the corresponding test result and pressing on {{ScreenElement|Info}} will show the EUT Calibration test result.
Line 92: Line 92:
The {{RadiMation}} Attenuation/Gain calibration can be used to very accurately determine the Coupling attenuation between the HV and LV ports of components that are used in motor vehicles.
Some standards describe that the HV-LV coupling attenuation should be determined between different combinations of lines. That can be achieved by performing the above procedure multiple times where the line on which the signal generator or spectrum analyzer is connected is different. These EUT Calibrations can all be performed in the same EUT file. The {{ScreenElement|Note}} of the EUT calibration can then for example be used to specify which combination of lines is measured.
Some standards describe that the HV-LV coupling attenuation should be determined between different combinations of lines. That can be achieved by performing the above procedure multiple times where the line on which the signal generator or spectrum analyser is connected is different. These EUT Calibrations can all be performed in the same EUT file. The {{ScreenElement|Note}} of the EUT calibration can then for example be used to specify which combination of lines is measured.
According the CISPR25:2016-10 (as specified in Annex I), conducted emission measurements should be performed, where the determined HV-LV coupling attenuation has influence on the limits that should be used during the conducted emission measurements on the HV system.
Based on the determined HV-LV coupling attenuation the correct emission limit can be determined, which then can be used during a normal conducted emission measurement performed with {{RadiMation}}.
Revision as of 09:30, 3 January 2020
How to measure the HV-LV coupling attenuation[edit]
This RadiMation® application note explains how the high-voltage (HV) to low-voltage (LV) Coupling attenuation can be measured with RadiMation®. This HV-LV coupling attenuation is needed for several automotive standards, for example the BMW GS 95002-3:2015-12 or the CISPR 25:2016-10
The HV-LV coupling attenuation measurement as it is described in these standards, is designated for components that are connected to the high voltage (above 60 V) system and the low voltage system of a motor vehicle, and has the purpose to determine the coupling between the two voltage systems through the component.
The principle of this measurement is that a component (the EUT) is connected on one side to the LV network and on the other side to the HV network. Then an attenuation measurement should be performed to the determine the coupling between the HV and LV networks that is caused by EUT.
  1. using a Vector Network Analyser
  2. using a signal generator and spectrum analyser (optionally using an amplifier)
A simple Attenuation/Gain calibration, which is also used to determine the attenuation of cables and other components, can be used for this measurement.
Reference measurement (system calibration)[edit]
This reference measurement can be performed in RadiMation® by using the Attenuation/Gain system calibration. In a setup where a spectrum analyser or receiver is used to perform the measurement of the generated signal, the setup is:
In RadiMation® then the system calibration can be started by selecting from the menu:
   Menu.svg Calibration
      Menu.svg System Calibration
         Menu.svg Attenuation / Gain
In the Attenuation System Calibration dialog, the required settings can be configured. The used standard often specifies most of these settings. This can for example be:
  • Start: 0.1 MHz
  • End: 1000 MHz
  • Previous Frequency: 2 %, which will then use a logarithmic step with a frequency increment that is 2 % of the previous frequency.
  • Signal power: 0 dBm, for a generated signal carrier level of 0 dBm.
  • Amount of samples: 4, which will perform 4 power measurements on every frequency, and will use the average of those 4 samples, as the result on that frequency.
  • Enable determination of dynamic range: Off, to disable that the maximum possible measurable attenuation is determined.
  • Test equipment: The name of the testsite that should be used. This testsite should at least have a selected signal generator and a selected forward powermeter. The device that is selected as the forward powermeter in the testsite will be used as the measurement device.
This then results in the following configuration of the Attenuation System Calibration:
When this system calibration is started, the signal generator will generate every frequency, and the generated power will be measured by the spectrum analyser. At the end of the system calibration a 'Save Calibration As' dialog is shown, which allows to save the result of the calibration to a .CAL calibration file. It is advised to specify a useful location and filename for this system calibration file, so it clearly identifies the purpose and reason of the system calibration.
EUT measurement[edit]
Once the Attenuation/Gain system calibration has been performed, and thus the reference measurement is known, the EUT can be installed and connected in the same measurement setup.
The converter between the coax cable and the HV port or LV port of the EUT, are depending on the connectors used by the EUT, and should probably be custom made, but should be designed in such way that they have a low as possible attenuation.
Then the coupling of the EUT between the HV and LV networks should be determined. This measurement can be performed in RadiMation® by using the Attenuation/Gain EUT Calibration.
During the configuration Attenuation/Gain EUT calibration, the previously performed system calibration should be selected. And because the system calibration is selected, the results of the EUT Calibration are automatically corrected with the measurement results from the system calibration. The reported EUT attenuation thus already is the desired attenuation coupling.
In RadiMation® when an EUT calibration itself is performed, first an EUT file needs to be opened. Once an EUT file is opened, the Attenuation/Gain measurement for the EUT can be started by selecting from the menu:
      Menu.svg EUT Calibration
In the Attenuation EUT Calibration dialog, only the previously created system calibration file should be selected in the System Calibration field. All the other settings will be retrieved from the selected Attenuation/Gain System calibration file.
This then for example results in the following configuration of the Attenuation EUT Calibration:
When this EUT calibration is started, the signal generator will be set to every frequency again, and the generated power will be also be measured by the spectrum analyser. Once the EUT calibration is finished, the results of this EUT Calibration is stored in the EUT file, and available as one of the performed Tests in the EUT file. Selecting the corresponding test result and pressing on Info will show the EUT Calibration test result.
The measured attenuation during the EUT calibration (which is the injected signalpower minus the measured forwardpower) is compared to the attenuation that was measured during the system calibration, and the real attenuation of the EUT is then calculated by subtracting those values.
Reporting[edit]
For the reviewing of the measurement result or the generation of the report, the measurement result can be accessed and used. Open the test result of the EUT Calibration, which shows the Attenuation/Gain result window:
The following graphs are available from this window:
  • Attenuation: The real attenuation of the EUT, already corrected with the reference attenuation that was determined during the system calibration.
  • Dynamic Range: The possible maximum attenuation that can be measured with the used test and measurement equipment.
  • System Calibration: Signal power: The signal power level that was used during the system calibration.
  • System Calibration: Received power (Ref. Att.): The reading of the forward powermeter during the system calibration, when the carrier level of the signal generator was active.
  • System Calibration: Received power (Ref. noise floor): The reading of the forward powermeter during the system calibration, when the carrier level of the signal generator was turned off. The difference between the two System Calibration: Received power graphs is used to calculate the System Calibration: Dynamic Range graph.
  • System Calibration: Reference attenuation: The attenuation that was calculated during the system calibration.
  • System Calibration: Dynamic range: The possible maximum attenuation that can be measured with the used test and measurement equipment.
The Attenuation graph is the most important graph because it specifies the determined real attenuation of the EUT, which is the HV-LV Attenuation coupling that is requested.
The RadiMation® Attenuation/Gain calibration can be used to very accurately determine the Coupling attenuation between the HV and LV ports of components that are used in motor vehicles.
Some standards describe that the HV-LV coupling attenuation should be determined between different combinations of lines. That can be achieved by performing the above procedure multiple times where the line on which the signal generator or spectrum analyser is connected is different. These EUT Calibrations can all be performed in the same EUT file. The Note of the EUT calibration can then for example be used to specify which combination of lines is measured.
According the CISPR25:2016-10 (as specified in Annex I), conducted emission measurements should be performed, where the determined HV-LV coupling attenuation has influence on the limits that should be used during the conducted emission measurements on the HV system. Based on the determined HV-LV coupling attenuation the correct emission limit can be determined, which then can be used during a normal conducted emission measurement performed with RadiMation®.