Line 158: Line 158:
| 33.1 dBm @ 500 MHz || 37.4 dBm @ 1000 MHz || 37.9 dBm @ 1500 MHz  || 38.4 dBm @ 2000 MHz || --- @ 2500 MHz
It is possible that one or more of the higher harmonics have not been measured due to limitations of the allowed frequency range of the used equipment (which can be coupler, spectrum analyzer, antenna and cables). For a CW frequency of 500 MHz, the 5th harmonic is generated at 2500 MHz. If however the used spectrum analyzer is only able to measure up to 2000 MHz, it is not possible to measure the 5th harmonic of 500 MHz, and thus the value in the table will be empty. Also the graphline of the 5th harmonic will then only be drawn up to 400 MHz, which corresponds to the 5th harmonic at 2000 MHz, which is the highest measurable frequency of that spectrum analyzer.
Revision as of 10:45, 27 June 2022
As the harmonics are thus present on higher frequencies than the CW signal, all the equipment that is connected to the output of the amplifier up to and including the used spectrum analyser should be able to be used on those higher frequencies. For an amplifier of 80 MHz - 1000 MHz, measuring up to the 6th harmonic, it is thus necessary that the measurement equipment is usable up to 6 GHz.
Overview[edit]
There are three possible setups of the used equipment to measure amplifier harmonics. Each with their own advantages, and small differences in the used equipment.
  1. Measure the power of all harmonics using a spectrum analyser.
  2. Measure the broadband power with a powermeter and the harmonics using a spectrum analyser.
  3. Measure the actual field harmonics using a receive antenna.
Device name Tab in testsite configuration window suitable for harmonic frequencies note
Signal Generator Devices 1 no
Amplifier Devices 1 no This is the amplifier of which the harmonics are being measured.
Coupler Devices 1 yes
Forward power meter Devices 1 yes The powermeter device driver for the analyser that is being used to measure the harmonics.
Coupler Devices 1 no
Forward power meter Devices 1 no The powermeter device driver for the broadband power measurement of the forward power.
Sensor power meter Devices 2 yes The powermeter device driver for the analyser that is being used to measure the harmonics.
Calibration antenna Devices 1 yes The antenna that is used for the receiving of the generated electrical fields on the harmonics.
Spectrum analyser Devices 2 yes The device driver for the analyser that is being used to measure the harmonics.
   Menu.svg Calibration
      Menu.svg EUT Calibration
         Menu.svg Harmonics
ScreenElementDescription.svg Test Equipment In the test site window the test engineer can select which equipment list will be used during this test.
Calibration procedure[edit]
  1. Set the signal generator to the first frequency and a low power.
  2. Measure the power using the forward powermeter. If the power is within the desired tolerance, continue at step 4.
  3. Adjust the signal generator carrier level. If the forward power was too high, reduce the carrier level, otherwise increase the carrier level.
  4. Use the spectrum analyser to measure the first harmonic
  5. Use the spectrum analyser to measure the other harmonics that should be measured
  6. Increase the signal generator carrier frequency to the next frequency. continue at step 2.
Carrier frequency 100 MHz 200 MHz 300 MHz 400 MHz 500 Mhz
Forward power 56.0 dBm 56.1 dBm 56.2 dBm 56.1 dBm 56.0 dBm
1st harmonic 55.9 dBm @ 100 MHz 55.8 dBm @ 200 MHz 55.9 dBm @ 300 MHz 55.8 dBm @ 400 MHz 55.9 dBm @ 500 MHz
3rd harmonic 37.8 dBm @ 300 MHz 42.2 dBm @ 600 MHz 52.7 dBm @ 900 MHz 43.1 dBm @ 1200 MHz 39.2 dBm @ 1500 MHz
5th harmonic 33.1 dBm @ 500 MHz 37.4 dBm @ 1000 MHz 37.9 dBm @ 1500 MHz 38.4 dBm @ 2000 MHz --- @ 2500 MHz
Test results[edit]
