Line 3: | Line 3: | ||
During a calibration, | During a calibration, {{RadiMation}} stores all equipment under test used during the calibration. If the test equipment used can return it’s serial number through the IEEE bus, this serial number will be recorded. | ||
Furthermore, all settings (like start frequency, stop frequency, probe response time, frequency step size, calibration type (FWD, transmitted or signal generator level, etc.) are recorded. | Furthermore, all settings (like start frequency, stop frequency, probe response time, frequency step size, calibration type (FWD, transmitted or signal generator level, etc.) are recorded. | ||
Furthermore, the following graphs are stored: | Furthermore, the following graphs are stored: | ||
*The achieved field strength (or current) during calibration. | * The achieved field strength (or current) during calibration. | ||
*The signal generator level, necessary to achieve this field strength (or current level) | * The signal generator level, necessary to achieve this field strength (or current level) | ||
*The Forward power, necessary to achieve this field strength (or current level) | * The Forward power, necessary to achieve this field strength (or current level) | ||
*The transmitted power, necessary to achieve this field strength (or current level) | * The transmitted power, necessary to achieve this field strength (or current level) | ||
{{Note|When a signal level calibration is made and no power meters are used, the forward and reflected power graphs are not stored. | {{Note|When a signal level calibration is made and no power meters are used, the forward and reflected power graphs are not stored. | ||
Line 22: | Line 22: | ||
===Frequency interpolation of Calibration files=== | ===Frequency interpolation of Calibration files=== | ||
{{RadiMation}} allows you to use a frequency step size during a test, which is smaller than the frequency step size during the calibration. For this purpose {{RadiMation}} uses a linear interpolation between the frequency point which is just below the desired test point and the frequency point which is directly above the test point. Of coarse frequency interpolation will reduce the accuracy of the test level, and should therefore only be used during pre-compliance tests. | |||
===Amplitude extrapolation of calibration files=== | ===Amplitude extrapolation of calibration files=== | ||
{{RadiMation}} allows you to run a test at a different test level as used during the calibration. For this purpose {{RadiMation}} uses a linear extrapolation from the calibration level to the desired level. | |||
If, for example at a certain frequency point 100 Watt (+50 dBm) is needed to generate a field of 10 V/m (which is determined during calibration) and one wants to use this calibration file to run a test at 3 V/m, RadiMation® will use a power level of 50 DBm - 20 log (10/3) = 40 dBm (10 Watt). Of coarse, amplitude interpolation will reduce the accuracy of the test level and should therefore only be used during pre-compliance tests. | If, for example at a certain frequency point 100 Watt (+50 dBm) is needed to generate a field of 10 V/m (which is determined during calibration) and one wants to use this calibration file to run a test at 3 V/m, RadiMation® will use a power level of 50 DBm - 20 log (10/3) = 40 dBm (10 Watt). Of coarse, amplitude interpolation will reduce the accuracy of the test level and should therefore only be used during pre-compliance tests. | ||
{{warning|The amplitude extrapolation will only be correct if the power amplifier works in its linear region. | {{warning|The amplitude extrapolation will only be correct if the power amplifier works in its linear region. | ||
{{RadiMation}} supports three different ways of power calibration. During the configuration of the calibration, the desired method should be selected. During a test, when a calibration file is selected, {{RadiMation}} automatically selects the method used during calibration.}} | |||
==Calibration methods== | ==Calibration methods== | ||
Line 53: | Line 51: | ||
===Radiated immunity calibration=== | ===Radiated immunity calibration=== | ||
Calibration of the radiated immunity test set-up is necessary when tests in accordance to the substitution method are performed. | Calibration of the [[radiated immunity]] test set-up is necessary when tests in accordance to the substitution method are performed. | ||
During a substitution test, first the empty anechoic room is calibrated (i.e.: the power to the transmitting antenna is recorded which is necessary to achieve the desired field strength in the room). During the test the EUT is placed inside the anechoic chamber, and the same power is applied to the transmitting antenna. | During a substitution test, first the empty anechoic room is calibrated (i.e.: the power to the transmitting antenna is recorded which is necessary to achieve the desired field strength in the room). During the test the [[EUT]] is placed inside the anechoic chamber, and the same power is applied to the transmitting antenna. | ||
====One-point calibration==== | ====One-point calibration==== | ||
During a 1-point calibration, | During a 1-point calibration, {{RadiMation}} for each frequency point increases the power to the antenna, until the desired field strength is reached. Then, depending on the type of calibration RadiMation® records the Signal generator level, FWD power and/or transmitted power in the calibration file. | ||
To start a radiated immunity one-point calibration please select: | To start a [[radiated immunity]] one-point calibration please select: | ||
{{menu|Calibration|Radiated Immunity|1 Point Calibration}} | {{menu|Calibration|Radiated Immunity|1 Point Calibration}} | ||
Line 113: | Line 111: | ||
===16 point field homogeneity verification=== | ===16 point field homogeneity verification=== | ||
The EN 61000-4-3 states that the anechoic room must have a | The EN 61000-4-3 (and other standards) states that the anechoic room must have a homogeneous field distribution in a 1.5 m. x 1.5 m. square area, 80 cm. above the ground. | ||
The field is defined as | The field is defined as homogeneous if, at 12 of 16 points in the 1.5 m. x 1.5 m. square the field strength lies between the desired value and the desired value plus 6 dB (see the figure below). | ||
If, for example, the field homogeneity check is made at a field strength of 10 V/m, 12 of the 16 points must lie between 10 V/m and 20 V/m. | If, for example, the field homogeneity check is made at a field strength of 10 V/m, 12 of the 16 points must lie between 10 V/m and 20 V/m. | ||
For this purpose, sixteen | For this purpose, sixteen '1 point calibrations' must be performed. After all 16 calibrations are carried out, the 16 point field homogeneity verification software can be used to verify the field homogeneity in the room. | ||
The sixteen-point verification is started by selecting: | The sixteen-point verification is started by selecting: | ||
Line 136: | Line 133: | ||
[[Image:Radiated Immunity Uniform Field Area Calculation Settings.png]] | [[Image:Radiated Immunity Uniform Field Area Calculation Settings.png]] | ||
The | The {{RadiMation}} package supports different ways to calculate the uniformity of a field, to determine the 4 points (out of the 16 points) which must be ignored for each test frequency. | ||
* EN 61000-4-3 1995 | |||
* EN 61000-4-3 2002 | |||
* EN 61000-4-3 2002 Amendment 1 | |||
* EN 61000-4-3 2006 | |||
#Use one time averaging | #Use one time averaging | ||
#Use four time averaging | #Use four time averaging | ||
One time averaging | === One time averaging === | ||
With this method, the software, for each frequency point, will take the average of all 16 test points, and will determine the four points which differ the most from the average. For this frequency, these four test points will be removed from the calibration file. | With this method, the software, for each frequency point, will take the average of all 16 test points, and will determine the four points which differ the most from the average. For this frequency, these four test points will be removed from the calibration file. | ||
Four time averaging | === Four time averaging === | ||
With this method, the software, for each frequency point, will use the following procedure: | With this method, the software, for each frequency point, will use the following procedure: | ||
Line 154: | Line 158: | ||
#The software takes the average of the remaining 13 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file. | #The software takes the average of the remaining 13 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file. | ||
After | After {{RadiMation}} has calculated the complete calibration file, the following information will be displayed: | ||
Graphical representation of the calibration results: | Graphical representation of the calibration results: | ||
{{RadiMation}} will display 16 coloured dots on the screen, each dot representing a test point in the anechoic room. At the side of these dots, the test frequency can be incremented, which will directly show the new information for the 16 points as represented in the picture below. | |||
[[Image:Radiated Immunity Uniform Field Area Calculation Result Window.png]] | [[Image:Radiated Immunity Uniform Field Area Calculation Result Window.png]] | ||
Line 187: | Line 191: | ||
===Calibration information that is stored after verification=== | ===Calibration information that is stored after verification=== | ||
Now that Radimation has determined the homogeneity, what is the information that is stored in the calibration file? And why? | Now that {{Radimation}} has determined the homogeneity, what is the information that is stored in the calibration file? And why? | ||
The method below describes for the signal generator level but also applies for the forward power and the net power calibration. | The method below describes for the signal generator level but also applies for the forward power and the net power calibration. | ||
The software has determined the remaining 12 points for each frequency according to the chapter | The software has determined the remaining 12 points for each frequency according to the chapter '16 point field homogeneity verification'. | ||
With only 1 signal generator at hand its impossible to set all the 12 levels. So you would say take the average of all the levels and you are done. This would be right for about 50% of the verifications. But with a simple example it can be proven that this is not the right method. | With only 1 signal generator at hand its impossible to set all the 12 levels. So you would say take the average of all the levels and you are done. This would be right for about 50% of the verifications. But with a simple example it can be proven that this is not the right method. | ||
Line 230: | Line 234: | ||
A better way is to use the highest value. When you perform a substitution with the highest level you can be 100% that the field is at least 10 V/m at 12 of the 16 points. | A better way is to use the highest value. When you perform a substitution with the highest level you can be 100% that the field is at least 10 V/m at 12 of the 16 points. | ||
That is why Radimation saves the highest level in the calibration file. | That is why {{Radimation}} saves the highest level in the calibration file. | ||
==Triplate calibration== | ==Triplate calibration== | ||
The triplate calibration needs to be performed in order to perform the triplate test. With the triplate calibration you determine the impedance of the triplate, which is needed to calculate the field. | The [[triplate]] calibration needs to be performed in order to perform the triplate test. With the triplate calibration you determine the impedance of the triplate, which is needed to calculate the field. | ||
During the calibration the field is set on a constant level, there are several reasons for doing so. | During the calibration the field is set on a constant level, there are several reasons for doing so. | ||
#Some fieldsensors have problems with their ranges, to avoid these problems it is better to field level. | # Some fieldsensors have problems with their ranges, to avoid these problems it is better to field level. | ||
#With fixed field you are aways sure that the required level is there, if you use fixed power there may be a big dip or peak in the calibration. This can give interpolation errors. For example a fixed power test at 50 dBm generates a field of 100 V/m at 100 MHz but generates a field of 80 V/m at 800 MHz due to the amplifier/powermeters/cables and couplers. Suppose Radimation uses these values to generate a fixed field of 110 V/m. The error made when extrapolation from 100V/m will be much smaller then from 80 V/m. | # With fixed field you are aways sure that the required level is there, if you use fixed power there may be a big dip or peak in the calibration. This can give interpolation errors. For example a fixed power test at 50 dBm generates a field of 100 V/m at 100 MHz but generates a field of 80 V/m at 800 MHz due to the amplifier/powermeters/cables and couplers. Suppose {{Radimation}} uses these values to generate a fixed field of 110 V/m. The error made when extrapolation from 100V/m will be much smaller then from 80 V/m. | ||
#Error field sensor | # Error field sensor vs. error powermeter, when performing a fixed power you can have problems with the linearity of your system. In general you can say that your powersensors are better in the linear response then your fieldsensor, so when you are leveling on the field sensor you are sure that the linearity error is reduced to the minimum. | ||
===One point calibration=== | ===One point calibration=== | ||
Line 249: | Line 253: | ||
To define the fieldprobe position in height use the antenna distance. Polarization does not influence this calibration. | To define the fieldprobe position in height use the antenna distance. Polarization does not influence this calibration. | ||
A description of the rest of the settings can be found in the manual chapter | A description of the rest of the settings can be found in the manual in the chapter 'Radiated immunity calibration'. | ||
[[Image:Radiated Immunity Test Setup Calibration Triplate.png]] | [[Image:Radiated Immunity Test Setup Calibration Triplate.png]] | ||
!Device name || Tab in Radimation configuration window | !Device name || Tab in {{Radimation}} configuration window | ||
|Signal Generator || Standard 1 | |Signal Generator || Standard 1 | ||
Line 274: | Line 278: | ||
|+'''Cables''' | |+'''Cables''' | ||
!# || Device name || Tab in Radimation configuration window | !# || Device name || Tab in {{Radimation}} configuration window | ||
|1 || Sign-> Amp || Cables | |1 || Sign-> Amp || Cables | ||
Line 287: | Line 291: | ||
{{Note|Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement | {{Note|Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during immunity measurements. | ||
When in doubt contact your reseller. | When in doubt contact your reseller. | ||
Line 305: | Line 309: | ||
<math>Z(f)=h^2*\frac{{E_{avg}}^2(f)}{P_{mid}(f)} [\Omega]</math> | <math>Z(f)=h^2*\frac{{E_{avg}}^2(f)}{P_{mid}(f)} [\Omega]</math> | ||
'''Formula 12 1: Impedance''' | '''Formula 12-1: Impedance''' | ||
<math>P_{mid}(f)=\frac{P_{net}(f)[W] + P_{out}(f)[W]}{2}</math> | <math>P_{mid}(f)=\frac{P_{net}(f)[W] + P_{out}(f)[W]}{2}</math> | ||
'''Formula 12 2: Middle power calculation''' | '''Formula 12-2: Middle power calculation''' | ||
==Conducted immunity calibration== | ==Conducted immunity calibration== | ||
Line 335: | Line 339: | ||
<math>U_{powerhead}=U_{rms} - 6[dB]-20*^{10}Log\frac{50}{150}= U_{rms}-15,6[dB]</math> | <math>U_{powerhead}=U_{rms} - 6[dB]-20*^{10}Log\frac{50}{150}= U_{rms}-15,6[dB]</math> | ||
Formula 12 3: Expected powermeter reading | '''Formula 12-3: Expected powermeter reading''' | ||
'''Example:''' | '''Example:''' | ||
Line 347: | Line 351: | ||
Line 369: | Line 373: | ||
Line 384: | Line 388: | ||
Revision as of 15:26, 12 January 2009
During a calibration, RadiMation® stores all equipment under test used during the calibration. If the test equipment used can return it’s serial number through the IEEE bus, this serial number will be recorded.
- The achieved field strength (or current) during calibration.
- The signal generator level, necessary to achieve this field strength (or current level)
- The Forward power, necessary to achieve this field strength (or current level)
- The transmitted power, necessary to achieve this field strength (or current level)
Note: | When a signal level calibration is made and no power meters are used, the forward and reflected power graphs are not stored.
When a forward power calibration is made and no reflected power meter is used, the reflected power graph is not stored. |
All can be viewed and printed by the operator.
To view calibration data, open the calibration file through the File, Open, calibration pull down menu. When a calibration window is opened, the calibration configuration data, as well as the Equipment used, is displayed.
Frequency interpolation of Calibration files[edit]
RadiMation® allows you to use a frequency step size during a test, which is smaller than the frequency step size during the calibration. For this purpose RadiMation® uses a linear interpolation between the frequency point which is just below the desired test point and the frequency point which is directly above the test point. Of coarse frequency interpolation will reduce the accuracy of the test level, and should therefore only be used during pre-compliance tests.
Amplitude extrapolation of calibration files[edit]
RadiMation® allows you to run a test at a different test level as used during the calibration. For this purpose RadiMation® uses a linear extrapolation from the calibration level to the desired level.
Warning: | The amplitude extrapolation will only be correct if the power amplifier works in its linear region.
RadiMation® supports three different ways of power calibration. During the configuration of the calibration, the desired method should be selected. During a test, when a calibration file is selected, RadiMation® automatically selects the method used during calibration. |
Calibration methods[edit]
During signal generator level calibration only the signal generator level is recorded during the calibration. This method does not take amplifier instability into account. During this calibration method no power meters are required. This method is less accurate than the FWD power and transmitted power calibrations and therefore not recommended for full compliance tests.
During a Forward power calibration, the forward power to the transmitting antenna is recorded. During a test, the FWD power to the antenna is kept the same as during the calibration. By this method, amplifier drift will have no influence on the test.
Net power calibration[edit]
During a Net power calibration, the Forward and Reflected power to the antenna are recorded. The difference between the FWD power and Reflected power is the Transmitted power. During a test, the Transmitted power is kept the same as during calibration. When a large metallic, EUT is placed in front of the antenna, some RF energy is reflected back into the antenna, resulting in a worse VSWR. When only FWD power is calibrated, this effect is not compensated for. When transmitted power is calibrated, the by the antenna transmitted power is kept the same as during calibration.
Radiated immunity calibration[edit]
Calibration of the radiated immunity test set-up is necessary when tests in accordance to the substitution method are performed. During a substitution test, first the empty anechoic room is calibrated (i.e.: the power to the transmitting antenna is recorded which is necessary to achieve the desired field strength in the room). During the test the EUT is placed inside the anechoic chamber, and the same power is applied to the transmitting antenna.
One-point calibration[edit]
During a 1-point calibration, RadiMation® for each frequency point increases the power to the antenna, until the desired field strength is reached. Then, depending on the type of calibration RadiMation® records the Signal generator level, FWD power and/or transmitted power in the calibration file. To start a radiated immunity one-point calibration please select:
- Calibration
- Radiated Immunity
- 1 Point Calibration
- Radiated Immunity
- Calibration
File:Select TSF File 1 Point Calibration Window.png
Device name | Tab in Radimation configuration window |
Signal Generator | Standard 1 |
Amplifier | Standard 1 |
Coupler | Standard 1 |
Power meter Pfwd | Standard 1 |
Power meter Refl | Standard 1 |
Antenna | Standard 1 |
Field sensor | Field probes |
# | Device name | Tab in Radimation configuration window |
1 | Sign-> Amp | Cables |
2 | Amp-> Coupler | Cables |
3 | Coup->Ant | Cables |
4 | Coup->Pfwd | Cables |
5 | Coup->Pref | Cables |
Note: | Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuraccy during immunity measurements.
16 point field homogeneity verification[edit]
- Calibration
- Radiated immunity
- 16 Point Calibration
- Radiated immunity
- Calibration
The user is requested to select a calibration TSF file from the list of available TSF files as shown below.
File:Select TSF File 16 Point Calibration Window.png
After a calibration TSF file is selected a typical 16 point calibration TSF will appear like the one shown in the picture below.
The RadiMation® package supports different ways to calculate the uniformity of a field, to determine the 4 points (out of the 16 points) which must be ignored for each test frequency.
- EN 61000-4-3 1995
- EN 61000-4-3 2002
- EN 61000-4-3 2002 Amendment 1
- EN 61000-4-3 2006
- Use one time averaging
- Use four time averaging
One time averaging[edit]
Four time averaging[edit]
- The software takes the average of all 16 test-points, and will determine the test point, which differs the most from the average. For this frequency, this point will be removed from the calibration file.
- The software takes the average of the remaining 15 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file.
- The software takes the average of the remaining 14 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file.
- The software takes the average of the remaining 13 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file.
RadiMation® will display 16 coloured dots on the screen, each dot representing a test point in the anechoic room. At the side of these dots, the test frequency can be incremented, which will directly show the new information for the 16 points as represented in the picture below.
- Green dot, This point is one of the 12 points, which are taken into account for the calibration file.
- Yellow dot, This point is one of the 4 points, which differ the most from the average of the 16 points.
- Red dot, This point is one of the 12 remaining points, however, it has an error of more than 6 dB (i.e. the room does not meet the field homogeneity requirements for this point!).
- Overall error graph (to verify if 12 of the 16 points lie within the 6 dB error)
- Signal generator level graph.
- Forward power level graph (only if the forward or net power calibration method is used).
- Net power level graph (only if the net power calibration method is used).
- Calculated field graphs: When the final calibration file is applied, these graphs display the calculated field for the selected test point location.
- Forward power graphs: Forward power used to achieve the desired field strength. This graph is corrected for the calibration (field) uncertainty. If the calibration is made at a test level of 10 V/m with a uncertainty of 1 V/m, the software will calculate the real forward power to achieve exactly 10 V/m (even if the calibration is made at for example 9.4 V/m).
- Error graphs: These graphs display the errors for the selected test point location. On frequencies where the test point was not taken into account for the calibration file, the error is set to zero. (i.e. this frequency point is one of the 4 points which where ignored).
Calibration information that is stored after verification[edit]
Now that RadiMation® has determined the homogeneity, what is the information that is stored in the calibration file? And why? The method below describes for the signal generator level but also applies for the forward power and the net power calibration.
Point | Signal level (dBm) |
1 | -10 |
2 | -10 |
3 | -8 |
4 | -12 |
5 | -8 |
6 | -12 |
7 | -8 |
8 | -12 |
9 | -11 |
10 | -9 |
11 | -13 |
12 | -7 |
The average will give –10 dBm.
This means that with points 3,5,7,10 and 12 you are not generating enough power, because for these points you need a higher signal level. Worst case (number 12), you will generate only around 7 V/m instead of 10 V/m.
That is why RadiMation® saves the highest level in the calibration file.
Triplate calibration[edit]
- Some fieldsensors have problems with their ranges, to avoid these problems it is better to field level.
- With fixed field you are aways sure that the required level is there, if you use fixed power there may be a big dip or peak in the calibration. This can give interpolation errors. For example a fixed power test at 50 dBm generates a field of 100 V/m at 100 MHz but generates a field of 80 V/m at 800 MHz due to the amplifier/powermeters/cables and couplers. Suppose RadiMation® uses these values to generate a fixed field of 110 V/m. The error made when extrapolation from 100V/m will be much smaller then from 80 V/m.
- Error field sensor vs. error powermeter, when performing a fixed power you can have problems with the linearity of your system. In general you can say that your powersensors are better in the linear response then your fieldsensor, so when you are leveling on the field sensor you are sure that the linearity error is reduced to the minimum.
One point calibration[edit]
To start a one-point calibration, close all EUT windows. And select the one point calibration.
- Calibration
- Radiated Immunity
- Radiated Immunity
- Calibration
A new window will appear, allowing you to set up the calibration. The difference between a chamber calibration and a triplate calibration is the out power. To perform a triplate calibration Net-power needs to be selected as well as outpower.
To define the fieldprobe position in height use the antenna distance. Polarization does not influence this calibration. A description of the rest of the settings can be found in the manual in the chapter 'Radiated immunity calibration'.
Device name | Tab in RadiMation® configuration window |
Note: | Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during immunity measurements.
Merge five one point calibration into one final calibration[edit]
- Calibration
- Radiated Immunity
- Radiated Immunity
- Calibration
Voltage calibration[edit]
Calibration procedure:
Example:
Test level according to IEC 1000-4-6 : 10 Vrms (140 dBuV) Measured value (U powerhead) : 140 dBuV - 15,6 dB = 124 dBuV or = 17 dBm (in a 50 Ω system)
Injection device | Standard 2 |
Calibration Jig | Standard 2 |
Sensor power meter | Standard 2 |
6 | Curr->Sens | Cables |
Current calibration[edit]
Verification procedure:
When a current probe with an transfer impedance of 1 Ω is used, a voltage of 33 mV should be measured (or ~90 dBuV).
System Calibration[edit]
System calibrations are performed on a test setup, before the test setup can be used to perform the actual measurement. For example:
- Calibration of conduct immunity setup before a conducted immunity text can be performed.
- Open/Short/Through calibration of network analyzer before S-parameter measurement can be performed.
Formula 12 4: Dynamic range
System calibration set up
- Calibration
- System Calibration
- Attenuation / Gain
- Calibration
Power meter setting[edit]
Formula 12.5: Average power calculation
Note: | The power measurement settings done in the device driver configuration screen are the settings for one sample. This is done because each power meter has a different leveling and measuring time. For the most optimal settings look at the device driver chapter for explanation and tips. |
Signal power setting[edit]
- Dynamic range
- Reference attenuation
- Reference noise floor
Formula 12.6: Standard deviation
Formula 12.7: Standard deviation of the dynamic range
Note: | Setting the power meter to be infinitely accurate is only valid in a relative measurement. The error caused by the power meter and the signal generator are corrected in the signal generator. |
- Calibration
- Gain Compression
- Calibration
Formula 12-8: Average power calculation
Note: | When inserting the calibration in the EUT compression configuration the same test levels are used. It is not possible to change the value, so select the settings with care. |
EUT calibration set up[edit]
- Calibration
- E.U.T. Calibration
- Calibration
- Dynamic range
- EUT attenuation
Formula 12 9: Standard deviation of the dynamic range
Formula 12 10: Standard deviation of the attenuation
EUT compression calibrations are performed on amplifiers that are considered to be a EUT. The test goal is to determine the x dB compression point. Yes, you read correctly you can define yourself which compression point you want to measure. It is also possible to measure for example the 1 dB and the 3 dB compression point in 1 test. 12.11.1 EUT Compression Calibration
Before a EUT calibration can be performed, it is best to first perform the system calibration. This will decrees the overall error, which is depending on the devices you are using. The flowchart below describes the testing procedure used for EUT calibration.
- Calibration
- Gain Compression
- Calibration
General setting guide lines:
- Signal power amplifier protection should be set high enough for testing. The maximum input level is default 0 dBm when no correction file is attached to the amplifier. With a correction file the maximum input can be set higher then 0 dBm.
Note: | Please note that these settings are guidelines, en should be treaded as such. |
- Small signal gain power
- Compression points power