Difference between revisions of "Chapter 12"
(→System Calibration) |
Line 506: | Line 506: | ||
{{Note|When inserting the calibration in the EUT compression configuration the same test levels are used. It is not possible to change the value, so select the settings with care.}} | {{Note|When inserting the calibration in the EUT compression configuration the same test levels are used. It is not possible to change the value, so select the settings with care.}} | ||
− | ==EUT | + | ==EUT Calibration== |
− | + | EUT calibrations are performed on devices that are considered to be an EUT. For example, during an attenuation calibration, the attenuation of an attenuator, cable or coupler can be measured. | |
− | EUT calibrations are performed on devices that are considered to be an EUT. For example, during an attenuation calibration, the attenuation of an attenuator, cable or coupler can be measured | ||
===EUT Attenuation Calibration=== | ===EUT Attenuation Calibration=== | ||
+ | During an attenuation calibration, the attenuation of an attenuator, cable or coupler can be measured. It is also possible to use the calibration to determine the gain of an amplifier. The attenuation will in this case have a negative value. | ||
+ | ====EUT Attenuation Calibration==== | ||
Before an EUT calibration can be performed, the calibration system first needs to be calibrated using a system calibration. After calibrating the EUT there is a possibility to save the measured attenuation into a correction file. | Before an EUT calibration can be performed, the calibration system first needs to be calibrated using a system calibration. After calibrating the EUT there is a possibility to save the measured attenuation into a correction file. | ||
Line 517: | Line 518: | ||
[[Image:Flowchart EUT Attenuation Calibration.png]] | [[Image:Flowchart EUT Attenuation Calibration.png]] | ||
− | ===EUT calibration set up=== | + | ====EUT calibration set up==== |
− | + | To open the configuration of a calibration, as shown below, make sure the correct EUT window is open. Then select from the menu: | |
− | To open the configuration of a calibration, | ||
{{menu|Calibration|E.U.T. Calibration|Attenuation / Gain}} | {{menu|Calibration|E.U.T. Calibration|Attenuation / Gain}} | ||
Select the correct calibration file and all the other setting will automatically be set. | Select the correct calibration file and all the other setting will automatically be set. | ||
− | The EUT calibration is performed with the same configuration settings as the system | + | The EUT calibration is performed with the same configuration settings as the configuration of the system calibration that was selected. This will ensure the lowest error, and the highest accuracy during the calibration: |
#There is no interpolation error possible between frequency points. | #There is no interpolation error possible between frequency points. | ||
#The calibration is performed at the same signal generator level preventing extra amplitude errors. | #The calibration is performed at the same signal generator level preventing extra amplitude errors. | ||
Line 548: | Line 548: | ||
In the information screen you can create a correction file for the attenuation measured. After pressing the “Create correction file” button a save dialog will appear, allowing you to select the desired path and filename for the correction file. This correction file can be attached to the calibrated device. For example, when the frequency response of a cable is measured, the correction file can be attached to the cable device driver, in the configuration menu of the cable. | In the information screen you can create a correction file for the attenuation measured. After pressing the “Create correction file” button a save dialog will appear, allowing you to select the desired path and filename for the correction file. This correction file can be attached to the calibrated device. For example, when the frequency response of a cable is measured, the correction file can be attached to the cable device driver, in the configuration menu of the cable. | ||
− | ==EUT Compression Calibration== | + | ===EUT Compression Calibration=== |
− | EUT compression calibrations are performed on amplifiers that are considered to be | + | EUT compression calibrations are performed on amplifiers that are considered to be an [[EUT]]. The goal of this calibration is to determine the 1 dB or 2 dB compression point. The calibration also allows to calibrate any other compression point, because the required compression levels can be configured in the calibration. It is thus also possible to measure for example the 1 dB and the 3 dB compression points during a single calibration. |
− | Before a EUT calibration can be performed, it is best to first perform the system calibration. This will | + | Before a EUT calibration can be performed, it is best to first perform the system calibration. This will decrease the overall error, which is depending on the devices you are using. |
The flowchart below describes the testing procedure used for EUT calibration. | The flowchart below describes the testing procedure used for EUT calibration. | ||
[[Image:Flowchart EUT Calibration Routine.png]] | [[Image:Flowchart EUT Calibration Routine.png]] | ||
− | To open the configuration of a calibration, open as showed below, make sure the correct EUT window is open | + | To open the configuration of a calibration, open as showed below, make sure the correct EUT window is open, and then select from the menu: |
{{menu|Calibration|E.U.T. Calibration|Gain Compression}} | {{menu|Calibration|E.U.T. Calibration|Gain Compression}} | ||
Select the correct calibration file, and all the other setting will automatically be set. | Select the correct calibration file, and all the other setting will automatically be set. | ||
The EUT calibration is performed with the same configuration settings as the system settings. This is done to insure the lowest error. | The EUT calibration is performed with the same configuration settings as the system settings. This is done to insure the lowest error. | ||
Line 572: | Line 569: | ||
#Error budget calculation is possible | #Error budget calculation is possible | ||
When no calibration file has been selected the settings can be set as desired. | When no calibration file has been selected the settings can be set as desired. | ||
− | + | Calibration of an amplifier can result in damaging the amplifier itself! We therefore suggest the following security precautions, during these type of calibrations: | |
− | * | + | * {{ScreenElement|Signal power amplifier protection}}} should be set high enough for testing. The maximum input level is default 0 dBm when no correction file is attached to the amplifier. With a correction file the maximum input can be set higher then 0 dBm. |
− | *Forward power amplifier protection should be used when the amplifier has an output power protection. The fail safe option is setting the limitation to the maximum and decrees this with one step size of the test. For example the limitation is set on 56,5 dBm (450 Watt) and the step size is 0.3 dB, than the limitation should be set on 56,2 dBm (420 Watt). This way the changes of the amplifier turning off during testing will become significantly smaller. | + | * Forward power amplifier protection should be used when the amplifier has an output power protection. The fail safe option is setting the limitation to the maximum and decrees this with one step size of the test. For example the limitation is set on 56,5 dBm (450 Watt) and the step size is 0.3 dB, than the limitation should be set on 56,2 dBm (420 Watt). This way the changes of the amplifier turning off during testing will become significantly smaller. |
− | * | + | * {{ScreenElement|Start level}} of the signal generator should be set on 20 dB lower then the lowest point specified by the manufacturer. |
− | * | + | * {{ScreenElement|Stop level}} of the signal generator should be the maximum input of the amplifier, most of the times this level will not be reached. |
− | {{note|Please note that these settings are guidelines, en should be | + | {{note|Please note that these settings are guidelines, en should be treated as such.}} |
− | === Reviewing measurement data === | + | ==== Reviewing measurement data ==== |
In the EUT calibration the standard deviation of the following items are calculated: | In the EUT calibration the standard deviation of the following items are calculated: |
Revision as of 20:21, 12 January 2009
Contents
- 1 Calibration
- 1.1 Frequency interpolation of Calibration files
- 1.2 Amplitude extrapolation of calibration files
- 1.3 Calibration methods
- 1.4 Radiated Immunity calibration
- 1.5 Conducted immunity calibration
- 1.6 System Calibration
- 1.7 EUT Calibration
Calibration
RadiMation® allows you to use a frequency step size during a test, that is different than the frequency step size during the calibration. To have optimal results, RadiMation® will use a linear interpolation between the frequency point which is just below the desired test point and the frequency point which is directly above the test point, to correctly calculate the needed power values at the desired frequency. For the most accurate results it is however advised to perform the calibration in the same frequency range and frequency steps, as will be used during the test. RadiMation® will not (and cannot) allow that a test is performed on frequencies that is outside the calibrated frequency range.
RadiMation® allows you to run a test at a different test level as used during the calibration. To facilitate this, RadiMation® will use a linear extrapolation from the calibration level to the desired level.
If, for example, at a certain frequency point 100 Watt (+50 dBm) is needed to generate a field of 10 V/m (which is determined during calibration), it is possible to use this calibration file to run a test at 3 V/m. RadiMation® will recalculate the needed power level to: 50 dBm - 20 log (10/3) = 40 dBm (10 Watt). Of coarse, amplitude extrapolation will reduce the accuracy of the test level and should therefore only be used during pre-compliance tests. For the most accurate results it is therefore advised to perform the calibration on the same test level, as will be used during the test.
Calibration methods
Signal generator level calibration
When a signal generator level calibration is used, only the signal generator level is recorded during the calibration. During a test, the signal power that was recorded during a calibration will be re-established. This method does not take amplifier instability or fluctuation into account. During this calibration method no forward and reflected powermeters are required. This method is less accurate than the Forward power and Net power calibrations and therefore is not recommended for full compliance tests. Because no powermeters are used, this is method is the fastest method to perform a test.
Forward power calibration
During a Net power calibration, the Forward and Reflected power to the antenna are recorded. The difference between the Forward power and Reflected power is the Transmitted power. When net power is calibrated, then during the test, the transmitted power by the antenna is kept the same as during calibration. This method has the advantage, that the transmitted power is adjusted to be the same as during the calibration, also when a large metallic, EUT is placed in front of the antenna. In that situation some RF energy can be reflected back into the antenna, resulting in a worse VSWR. If in this situation a forward power calibration or a signal generator level calibration was used, this effect is not compensated for.
Calibration of the radiated immunity test set-up is necessary when tests in accordance to the substitution method are performed. During a substitution test, first the empty anechoic room is calibrated (i.e.: the power to the transmitting antenna is recorded which is necessary to achieve the desired field strength in the room). During the test the EUT is placed inside the anechoic chamber, and the same power is applied to the transmitting antenna.
One-point calibration
During a 1-point calibration, RadiMation® for each frequency point increases the power to the antenna, until the desired field strength is reached. Then, depending on the type of calibration RadiMation® records the Signal generator level, FWD power and/or transmitted power in the calibration file. To start a radiated immunity one-point calibration please select:
File:Select TSF File 1 Point Calibration Window.png
Device name | Tab in RadiMation®configuration window |
Signal Generator | Standard 1 |
Amplifier | Standard 1 |
Coupler | Standard 1 |
Power meter Pfwd | Standard 1 |
Power meter Refl | Standard 1 |
Antenna | Standard 1 |
Field sensor | Field probes |
1 | Sign-> Amp | Cables |
2 | Amp-> Coupler | Cables |
3 | Coup->Ant | Cables |
4 | Coup->Pfwd | Cables |
5 | Coup->Pref | Cables |
Triplate calibration
The triplate calibration needs to be performed in order to perform the triplate test. With the triplate calibration you determine the impedance of the triplate, which is needed to calculate the field.
- Some fieldsensors have problems with their ranges, to avoid these problems it is better to field level.
- With fixed field you are aways sure that the required level is there, if you use fixed power there may be a big dip or peak in the calibration. This can give interpolation errors. For example a fixed power test at 50 dBm generates a field of 100 V/m at 100 MHz but generates a field of 80 V/m at 800 MHz due to the amplifier/powermeters/cables and couplers. Suppose RadiMation® uses these values to generate a fixed field of 110 V/m. The error made when extrapolation from 100V/m will be much smaller then from 80 V/m.
- Error field sensor vs. error powermeter, when performing a fixed power you can have problems with the linearity of your system. In general you can say that your powersensors are better in the linear response then your fieldsensor, so when you are leveling on the field sensor you are sure that the linearity error is reduced to the minimum.
One point calibration
To start a one-point calibration, close all EUT windows. And select the one point calibration.
A new window will appear, allowing you to set up the calibration. The difference between a chamber calibration and a triplate calibration is the out power. To perform a triplate calibration Net-power needs to be selected as well as outpower.
To define the fieldprobe position in height use the antenna distance. Polarization does not influence this calibration. A description of the rest of the settings can be found in the manual in the chapter 'Radiated immunity calibration'.
Merge five one point calibration into one final calibration
Formula 12-1: Impedance
Formula 12-2: Middle power calculation
16 point field homogeneity verification / Uniform Field Area calculation
The EN 61000-4-3 (and other standards) states that the anechoic room must have a homogeneous field distribution in a 1.5 m. x 1.5 m. square area, 80 cm. above the ground.
The field is defined as homogeneous if, at 12 of 16 points in the 1.5 m. x 1.5 m. square the field strength lies between the desired value and the desired value plus 6 dB (see the figure below).
For this purpose, sixteen '1 point calibrations' must be performed. After all 16 calibrations are carried out, the 16 point field homogeneity verification software can be used to verify the field homogeneity in the room.
The sixteen-point verification is started by selecting:
The user is requested to select a calibration TSF file from the list of available TSF files as shown below.
File:Select TSF File 16 Point Calibration Window.png
After a calibration TSF file is selected a typical 16 point calibration TSF will appear like the one shown in the picture below.
The RadiMation® package supports different ways to calculate the uniformity of a field, to determine the 4 points (out of the 16 points) which must be ignored for each test frequency.
- EN 61000-4-3 1995 (Average Power)
- ISO 11451-2
- J1113-25
- EN 61000-4-3 2002 (Average Field)
- EN 61000-4-3 2002 Amendment 1 (Maximum Power)
- EN 61000-4-3 2006
Uniform Field Area calculations
- Use one time averaging
- Use four time averaging
One time averaging
Four time averaging
- The software takes the average of all 16 test-points, and will determine the test point, which differs the most from the average. For this frequency, this point will be removed from the calibration file.
- The software takes the average of the remaining 15 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file.
- The software takes the average of the remaining 14 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file.
- The software takes the average of the remaining 13 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file.
RadiMation® will display 16 coloured dots on the screen, each dot representing a test point in the anechoic room. At the side of these dots, the test frequency can be incremented, which will directly show the new information for the 16 points as represented in the picture below.
- Green dot, This point is one of the 12 points, which are taken into account for the calibration file.
- Yellow dot, This point is one of the 4 points, which differ the most from the average of the 16 points.
- Red dot, This point is one of the 12 remaining points, however, it has an error of more than 6 dB (i.e. the room does not meet the field homogeneity requirements for this point!).
- Overall error graph (to verify if 12 of the 16 points lie within the 6 dB error)
- Signal generator level graph.
- Forward power level graph (only if the forward or net power calibration method is used).
- Net power level graph (only if the net power calibration method is used).
- Calculated field graphs: When the final calibration file is applied, these graphs display the calculated field for the selected test point location.
- Forward power graphs: Forward power used to achieve the desired field strength. This graph is corrected for the calibration (field) uncertainty. If the calibration is made at a test level of 10 V/m with a uncertainty of 1 V/m, the software will calculate the real forward power to achieve exactly 10 V/m (even if the calibration is made at for example 9.4 V/m).
- Error graphs: These graphs display the errors for the selected test point location. On frequencies where the test point was not taken into account for the calibration file, the error is set to zero. (i.e. this frequency point is one of the 4 points which where ignored).
Calibration information that is stored after verification
Now that RadiMation® has determined the homogeneity, what is the information that is stored in the calibration file? And why? The method below describes for the signal generator level but also applies for the forward power and the net power calibration.
The software has determined the remaining 12 points for each frequency according to the chapter '16 point field homogeneity verification'.
Point | Signal level (dBm) |
1 | -10 |
2 | -10 |
3 | -8 |
4 | -12 |
5 | -8 |
6 | -12 |
7 | -8 |
8 | -12 |
9 | -11 |
10 | -9 |
11 | -13 |
12 | -7 |
The average will give –10 dBm.
This means that with points 3,5,7,10 and 12 you are not generating enough power, because for these points you need a higher signal level. Worst case (number 12), you will generate only around 7 V/m instead of 10 V/m.
That is why RadiMation® saves the highest level in the calibration file.
Calibration methods
Voltage calibration
Calibration procedure:
Formula 12-3: Expected powermeter reading
Example:
Test level according to IEC 1000-4-6: 10 Vrms (140 dBuV) Measured value (U powerhead): 140 dBuV - 15,6 dB = 124 dBuV or = 17 dBm (in a 50 Ω system)
Injection device | Standard 2 |
Calibration Jig | Standard 2 |
Sensor power meter | Standard 2 |
6 | Curr->Sens | Cables |
Current calibration
Verification procedure:
System Calibration
- Open/Short/Through calibration of network analyzer before S-parameter measurement can be performed.
Attenuation system calibration
Formula 12-4: Dynamic range
System calibration set up
In the configuration window you can define the amount of sample that should be taken. The fastest way is one sample but this will disable the calculation of the standard deviation.
Power meter setting
Formula 12-5: Average power calculation
Signal power setting
- Dynamic range
- Reference attenuation
- Reference noise floor
Formula 12-6: Standard deviation
Formula 12-7: Standard deviation of the dynamic range
System Compression Calibration
Power meter setting
Formula 12-8: Average power calculation
EUT Calibration
- Dynamic range
- EUT attenuation
Formula 12-10: Standard deviation of the attenuation
- Signal power amplifier protection} should be set high enough for testing. The maximum input level is default 0 dBm when no correction file is attached to the amplifier. With a correction file the maximum input can be set higher then 0 dBm.
- Small signal gain power
- Compression points power