Line 54: Line 54:
During the one-point calibrations, {{RadiMation}} will always use the 'constant field strength calibration method' that is described by the standards. This method has several advantages:
* Some fieldsensors have problems with their accuracy at the overlapping of their measurements ranges. To avoid these problems it is better to always level to the same field strength level, which will result in the usage of the same measurement range of the field sensor.
* With the constant field strength calibration method, it is more obvious if the required field strength level can be achieved. If the 'constant power calibration method' is used, there may be a big dip or peak in the calibration. This can result in interpolation errors. For example, a constant power calibratoin at 50 dBm can generate a field of 100 V/m at 100 MHz, and in the same calibration it is possible that a field of 80 V/m is generated at 800 MHz, due to the amplifier/powermeters/cables and couplers characteristics. If with these calibration results, a substitution is performed for a fixed field of 110 V/m, the error that is made as a result of the extrapolation from 100 V/m will be much smaller then from 80 V/m.
* With the constant field strength calibration method, it is more obvious if the required field strength level can be achieved. If the 'constant power calibration method' is used, there may be a big dip or peak in the calibration. This can result in interpolation errors. For example, a constant power calibration at 50 dBm can generate a field of 100 V/m at 100 MHz, and in the same calibration it is possible that a field of 80 V/m is generated at 800 MHz, due to the amplifier/powermeters/cables and couplers characteristics. If with these calibration results, a substitution is performed for a fixed field of 110 V/m, the error that is made as a result of the extrapolation from 100 V/m will be much smaller then from 80 V/m.
* The linearity error from the field sensor is much larger than the linearity error of the powermeter. When a 'constant field calibration method' is used, the error caused by the linearity of the fieldsensor is reduced to a minimum because the field sensor is always levelled to the same measurement value. The powermeter will perform measurements over a larger amplitude range, but the dynamic range of all powermeters is large enough to have a minimal linearity error. So when you the calibration is leveling on the field sensor, the linearity error of the calibration is minimal.
* The linearity error from the field sensor is much larger than the linearity error of the powermeter. When a 'constant field calibration method' is used, the error caused by the linearity of the fieldsensor is reduced to a minimum because the field sensor is always levelled to the same measurement value. The powermeter will perform measurements over a larger amplitude range, but the dynamic range of all powermeters is large enough to have a minimal linearity error. So when you the calibration is levelling on the field sensor, the linearity error of the calibration is minimal.
===One-point calibration===
Revision as of 10:06, 13 January 2009
  • The achieved field strength (or current or voltage) during calibration.
  • The signal generator level, necessary to achieve this field strength (or current or voltage level)
  • The forward power, necessary to achieve this field strength (or current or voltage level)
  • The transmitted power, necessary to achieve this field strength (or current or voltage level)
Note: When a signal level calibration is made and no power meters are used, the forward and reflected power graphs are not stored.
When a forward power calibration is made and no reflected power meter is used, the reflected power graph is not stored.
   Menu.svg File
      Menu.svg Open
         Menu.svg Calibration
Frequency interpolation[edit]
RadiMation® allows you to use a frequency step size during a test, that is different than the frequency step size during the calibration. To have optimal results, RadiMation® will use a linear interpolation between the frequency point which is just below the desired test point and the frequency point which is directly above the test point, to correctly calculate the needed power values at the desired frequency. For the most accurate results it is however advised to perform the calibration in the same frequency range and frequency steps, as will be used during the test. RadiMation® will not (and cannot) allow that a test is performed on frequencies that is outside the calibrated frequency range.
Amplitude extrapolation[edit]
RadiMation® allows you to run a test at a different test level as used during the calibration. To facilitate this, RadiMation® will use a linear extrapolation from the calibration level to the desired level.
If, for example, at a certain frequency point 100 Watt (+50 dBm) is needed to generate a field of 10 V/m (which is determined during calibration), it is possible to use this calibration file to run a test at 3 V/m. RadiMation® will recalculate the needed power level to: 50 dBm - 20 log (10/3) = 40 dBm (10 Watt). Of coarse, amplitude extrapolation will reduce the accuracy of the test level and should therefore only be used during pre-compliance tests. For the most accurate results it is therefore advised to perform the calibration on the same test level, as will be used during the test.
Warning: The amplitude extrapolation will only be correct if the power amplifier works in its linear region.
Calibration methods[edit]
When a signal generator level calibration is used, only the signal generator level is recorded during the calibration. During a test, the signal power that was recorded during a calibration will be re-established. This method does not take amplifier instability or fluctuation into account. During this calibration method no forward and reflected powermeters are required. This method is less accurate than the Forward power and Net power calibrations and therefore is not recommended for full compliance tests. Because no powermeters are used, this is method is the fastest method to perform a test.
Net power calibration[edit]
During a Net power calibration, the Forward and Reflected power to the antenna are recorded. The difference between the Forward power and Reflected power is the Transmitted power. When net power is calibrated, then during the test, the transmitted power by the antenna is kept the same as during calibration. This method has the advantage, that the transmitted power is adjusted to be the same as during the calibration, also when a large metallic, EUT is placed in front of the antenna. In that situation some RF energy can be reflected back into the antenna, resulting in a worse VSWR. If in this situation a forward power calibration or a signal generator level calibration was used, this effect is not compensated for.
  • Some fieldsensors have problems with their accuracy at the overlapping of their measurements ranges. To avoid these problems it is better to always level to the same field strength level, which will result in the usage of the same measurement range of the field sensor.
  • With the constant field strength calibration method, it is more obvious if the required field strength level can be achieved. If the 'constant power calibration method' is used, there may be a big dip or peak in the calibration. This can result in interpolation errors. For example, a constant power calibration at 50 dBm can generate a field of 100 V/m at 100 MHz, and in the same calibration it is possible that a field of 80 V/m is generated at 800 MHz, due to the amplifier/powermeters/cables and couplers characteristics. If with these calibration results, a substitution is performed for a fixed field of 110 V/m, the error that is made as a result of the extrapolation from 100 V/m will be much smaller then from 80 V/m.
  • The linearity error from the field sensor is much larger than the linearity error of the powermeter. When a 'constant field calibration method' is used, the error caused by the linearity of the fieldsensor is reduced to a minimum because the field sensor is always levelled to the same measurement value. The powermeter will perform measurements over a larger amplitude range, but the dynamic range of all powermeters is large enough to have a minimal linearity error. So when you the calibration is levelling on the field sensor, the linearity error of the calibration is minimal.
One-point calibration[edit]
      Menu.svg Radiated Immunity
         Menu.svg 1 Point Calibration
The one-point calibration is also able to perform a calibration of a triplate setup. The triplate calibration is required, to perform the triplate test. With the triplate calibration it is also possible to determine the impedance of the triplate, which is needed to calculate the field. The difference between a anechoic chamber calibration and a triplate calibration is the out power. To perform a triplate calibration the Net-power calibration method as well as the out-power method needs to be selected.
Device name Tab in RadiMation®configuration window
Signal Generator Standard 1
Amplifier Standard 1
Coupler Standard 1
Power meter Pfwd Standard 1
Power meter Refl Standard 1
Antenna Standard 1
Field sensor Field probes
Cables
1 Sign-> Amp Cables
2 Amp-> Coupler Cables
3 Coup->Ant Cables
4 Coup->Pfwd Cables
5 Coup->Pref Cables
The EN 61000-4-3 (and other standards) state that the anechoic chamber must have a homogeneous field distribution in a 1.5 m x 1.5 m square area, 80 cm. above the ground.
The field is defined as homogeneous if, at 12 of 16 points in the 1.5 m x 1.5 m square the field strength lies between the desired value and the desired value plus 6 dB (see the figure below).
The 'Uniform Field Area calculation' is started by selecting from the menu:
      Menu.svg System Calibration
         Menu.svg Radiated immunity
            Menu.svg Uniform Field Area calculation
In the Uniform Field Area calculation, After a calibration TSF file is selected a typical 16 point calibration TSF will appear like the one shown in the picture below.
The RadiMation® package supports different ways to calculate the uniformity of a field, to determine the 4 points (out of the 16 points) which must be ignored for each test frequency.
  • EN 61000-4-3 1995 (Average Power)
  • ISO 11451-2
  • J1113-25
  • EN 61000-4-3 2002 (Average Field)
  • EN 61000-4-3 2002 Amendment 1 (Maximum Power)
  • EN 61000-4-3 2006
Uniform Field Area calculations[edit]
  1. Use one time averaging
  2. Use four time averaging
One time averaging[edit]
Four time averaging[edit]
  1. The software takes the average of all 16 test-points, and will determine the test point, which differs the most from the average. For this frequency, this point will be removed from the calibration file.
  2. The software takes the average of the remaining 15 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file.
  3. The software takes the average of the remaining 14 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file.
  4. The software takes the average of the remaining 13 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file.
RadiMation® will display 16 coloured dots on the screen, each dot representing a test point in the anechoic chamber. At the side of these dots, the test frequency can be incremented, which will directly show the new information for the 16 points as represented in the picture below.
  1. Green dot, This point is one of the 12 points, which are taken into account for the calibration file.
  2. Yellow dot, This point is one of the 4 points, which differ the most from the average of the 16 points.
  3. Red dot, This point is one of the 12 remaining points, however, it has an error of more than 6 dB (i.e. the anechoic chamber does not meet the field homogeneity requirements for this point!).
  1. Overall error graph (to verify if 12 of the 16 points lie within the 6 dB error)
  2. Signal generator level graph.
  3. Forward power level graph (only if the forward or net power calibration method is used).
  4. Net power level graph (only if the net power calibration method is used).
  1. Calculated field graphs: When the final calibration file is applied, these graphs display the calculated field for the selected test point location.
  1. Forward power graphs: Forward power used to achieve the desired field strength. This graph is corrected for the calibration (field) uncertainty. If the calibration is made at a test level of 10 V/m with a uncertainty of 1 V/m, the software will calculate the real forward power to achieve exactly 10 V/m (even if the calibration is made at for example 9.4 V/m).
  1. Error graphs: These graphs display the errors for the selected test point location. On frequencies where the test point was not taken into account for the calibration file, the error is set to zero. (i.e. this frequency point is one of the 4 points which where ignored).
Merge five one point calibration into one final calibration[edit]
         Menu.svg 16 Point Calibration
Formula 12-1: Impedance
Formula 12-2: Middle power calculation
Calibration information that is stored after verification[edit]
Now that RadiMation® has determined the homogeneity, what is the information that is stored in the calibration file? And why? The method below describes for the signal generator level but also applies for the forward power and the net power calibration.
The software has determined the remaining 12 points for each frequency according to the chapter '16 point field homogeneity verification'.
Point Signal level (dBm)
1 -10
2 -10
3 -8
4 -12
5 -8
6 -12
7 -8
8 -12
9 -11
10 -9
11 -13
12 -7
The average will give –10 dBm.
This means that with points 3,5,7,10 and 12 you are not generating enough power, because for these points you need a higher signal level. Worst case (number 12), you will generate only around 7 V/m instead of 10 V/m.
That is why RadiMation® saves the highest level in the calibration file.
Voltage calibration[edit]
Calibration procedure:
Formula 12-3: Expected powermeter reading
Example:
Test level according to IEC 1000-4-6: 10 Vrms (140 dBuV) Measured value (U powerhead): 140 dBuV - 15,6 dB = 124 dBuV or = 17 dBm (in a 50 Ω system)
Coupler Standard 1
Injection device Standard 2
Calibration Jig Standard 2
Sensor power meter Standard 2
Cables
1 Sign-> Amp Cables
3 Coup->Ant Cables
4 Coup->Pfwd Cables
5 Coup->Pref Cables
6 Curr->Sens Cables
Current calibration[edit]
Verification procedure:
System Calibration[edit]
  1. Attenuation measurements of cables, couplers or other attenuators.
  2. 1-16 point calibration of anechoic chamber before a radiated immunity test can be performed.
  3. Calibration of conducted immunity setup before a conducted immunity test can be performed.
  4. Open/Short/Through calibration of network analyzer before S-parameter measurement can be performed.
Formula 12-4: Dynamic range
System calibration set up
         Menu.svg Attenuation / Gain
In the configuration window you can define the amount of sample that should be taken. The fastest way is one sample but this will disable the calculation of the standard deviation.
Power meter setting[edit]
Formula 12-5: Average power calculation
Note: The power measurement settings done in the device driver configuration screen are the settings for one sample. This is done because each power meter has a different leveling and measuring time. For the most optimal settings look at the device driver chapter for explanation and tips.
Signal power setting[edit]
  • Dynamic range
  • Reference attenuation
  • Reference noise floor
Formula 12-6: Standard deviation
Formula 12-7: Standard deviation of the dynamic range
Note: Setting the power meter to be infinitely accurate is only valid in a relative measurement. The error caused by the power meter and the signal generator are corrected in the signal generator.
         Menu.svg Gain Compression
Formula 12-8: Average power calculation
Note: When inserting the calibration in the EUT compression configuration the same test levels are used. It is not possible to change the value, so select the settings with care.
EUT Calibration[edit]
EUT calibration set up[edit]
      Menu.svg E.U.T. Calibration
  1. There is no interpolation error possible between frequency points.
  2. The calibration is performed at the same signal generator level preventing extra amplitude errors.
  3. Error budget calculation is possible
  • Dynamic range
  • EUT attenuation
Formula 12-9: Standard deviation of the dynamic range
Formula 12-10: Standard deviation of the attenuation
To open the configuration of a calibration, open as showed below, make sure the correct EUT window is open, and then select from the menu:
Calibration of an amplifier can result in damaging the amplifier itself! We therefore suggest the following security precautions, during these type of calibrations:
  • Signal power amplifier protection} should be set high enough for testing. The maximum input level is default 0 dBm when no correction file is attached to the amplifier. With a correction file the maximum input can be set higher then 0 dBm.
  • Forward power amplifier protection should be used when the amplifier has an output power protection. The fail safe option is setting the limitation to the maximum and decrees this with one step size of the test. For example the limitation is set on 56,5 dBm (450 Watt) and the step size is 0.3 dB, than the limitation should be set on 56,2 dBm (420 Watt). This way the changes of the amplifier turning off during testing will become significantly smaller.
  • Start level of the signal generator should be set on 20 dB lower then the lowest point specified by the manufacturer.
  • Stop level of the signal generator should be the maximum input of the amplifier, most of the times this level will not be reached.
Note: Please note that these settings are guidelines, en should be treated as such.
  • Small signal gain power
  • Compression points power