Line 87: | Line 87: | ||
The configuration of the eut test site should contain the following devices: | The configuration of the eut test site should contain the following devices: | ||
! Device name || Tab in testsite configuration window || note | !# || Device name || Tab in testsite configuration window || note | ||
| Signal | | || Signal Generator || Devices 1 || The signal generator to use | ||
| Amplifier || Devices 1 || The amplifier to use | | || Amplifier || Devices 1 || The amplifier to use | ||
| Coupler || Devices 1 || The coupler to use | | || Coupler || Devices 1 || The coupler to use | ||
| Forward power meter || Devices 1 || The forward power meter to use | | || Forward power meter || Devices 1 || The forward power meter to use | ||
| Sensor | | || Sensor powermeter || Devices 2 || The power meter or analyser to use for measuring the current | ||
| Current | | || Current Sensor || Devices 2 || The current sensor to use with transfer factor attached to the driver | ||
| Injection device || Devices 2 || The injection clamp to use | | || Injection device || Devices 2 || The injection clamp to use | ||
!colspan="4"|Cables | |||
| Cable current -> power meter || Cables || Cable ( | |① || Cable current -> power meter || Cables || Cable (1) with a correction file specified for the cable loss | ||
| Cable current -> power meter || Cables || Cable (3) with a correction file specified for the cable loss | |② || Cable current -> power meter || Cables || Cable (2) with the specified loss of the used attenuator | ||
|③ || Cable current -> power meter || Cables || Cable (3) with a correction file specified for the cable loss | |||
Revision as of 15:59, 2 November 2023
How to perform a MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test[edit]
Necessary equipment[edit]
The following devices are necessary to run this test:
- Signal generator
- Amplifier
- Coupler
- Forward power meter
- Current sensor
- 50 Ohm load
- 50 Ohm attenuator
- Current injection device
Coupler | Devices 1 | The coupler to use | |
Forward power meter | Devices 1 | The forward power meter to use | |
Sensor powermeter | Devices 2 | The power meter or analyser to use for measuring the current | |
Current Sensor | Devices 2 | The current sensor to use with transfer factor attached to the driver | |
Injection device | Devices 2 | The injection clamp to use | |
Jig | Devices 2 | The jig to use | |
Cables | |||
① | Cable current -> power meter | Cables | Cable (1) with a correction file specified for the cable loss |
② | Cable current -> power meter | Cables | Cable (2) with the specified loss of the used attenuator |
③ | Cable current -> power meter | Cables | Cable (3) with a correction file specified for the cable loss |
Verification[edit]
Cables | |||
EUT Testing[edit]
Cables | |||
Calibration procedure[edit]
Create a correction file for the test level[edit]
- File
- New
- Correction
- New
- File
First create a correction file in RadiMation® and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.'). In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured.
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBuA.
Create calibration test setup files[edit]
Band | Frequency range | Step size |
Band 1 | 10 kHz - 1 MHz | 5 % |
Band 2 | 1 MHz - 30 MHz | 1 % |
Band 2 | 30 MHz - 200 MHz | 0.5 % |
- Calibration
- System calibration
- Conducted immunity
- Calibration
This example shows the settings for band 1.
Start | The start frequency of the test. For example 10 kHz. |
End | The stop frequency of the test. For example 1 MHz. |
Step | The frequency step, in this case 5%. |
Test level | The test level, in this case current. |
Variable | The variable test level, select the correction file created earlier. |
Tolerance | The regulation tolerance. |
Calibration method | Forward power. |
Test equipment | Calibration test equipment. |
Test engineer | The engineer that performed the calibration. |
Note: | If the current unit is shown as uA, it can be changed to dBuA by clicking Units, select Current and click Edit to change this. |
Calibration result[edit]
Verification[edit]
Create the Test Setup File (TSF)[edit]
CS114 requires to perform a verification of the setup with the probes in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit. The verification can be performed in RadiMation® by doing an actual conducted immunity test. This test can be started by selecting from the menu:
- Tests
- Conducted Immunity
- Multiband
- Tests
Band 1 | 10 kHz - 1 MHz | 10 % |
Band 2 | 1 MHz - 30 MHz | 2 % |
Band 2 | 30 MHz - 200 MHz | 1 % |
Calibration file | Select the calibration file to use. |
Test level | Variable current, and specify the correction file for the applicable test level. |
Calibration method | Select Forward Power. |
Frequency Step | The frequency step, see 5.12.3.4.c.3. |
Dwell Time | The dwell time specified see TABLE II. |
Modulation | Modulation is optional during the verification. |
Test equipment | The equipment needed for the verification. |
Location Type | Injection clamp. |
Test Level | Specify the current substitution method. |
Inputs | Add the current sensor as an input. |
Verification result[edit]
EUT Testing[edit]
- Tests
- Multiband
- Tests
Frequency Step | The frequency step, see Table III. |
Modulation | Configure Pulse Modulation with 1 kHz and 50% duty cycle. |
Test equipment | The equipment needed for the conducted immunity test. |
Test Level | Specify the current substitution method. Also add a "Measured Current Limit" |
Activate modulation | Check this checkbox to enable the pulse modulation. |
Frequency | The PM frequency 1 kHz. |
Duty Cycle | A duty cycle of 50%. |
Calibration file | Select the calibration file to use for this band. |
Description | The description of the limit |
Constant current | The current to limit on. |
EUT Test Result[edit]
