RadiMation Application Note 155: Difference between revisions
Line 85: | Line 85: | ||
{{ScreenElementDescription|Tolerance|Specify the tolerance to use.}} | {{ScreenElementDescription|Tolerance|Specify the tolerance to use.}} | ||
When all settings are configured press {{ScreenElement|Start Test}} to run the test. | |||
=== EUT Test Result === | |||
Once the test is finished, the results of this test are stored in the EUT file and available as one of the performed {{ScreenElement|Tests}} in the EUT file. Selecting the corresponding test result and pressing on {{ScreenElement|Info}} will show the test results again. | |||
[[File:CS109 Test Result.png]] | |||
== Conclusion == | |||
The {{RadiMation}} Conducted Immunity multiband test can be used to perform the CS109 test with a variable current closed loop method test level. | |||
[[Category:RadiMation Application Note]] |
Revision as of 16:36, 29 November 2023
How to perform a MIL-STD-461 CS109, Conducted susceptibility, structure current test[edit]
This application note explains how the MIL-STD-461 CS109, Conducted susceptibility, structure current test can be performed with RadiMation®
The exact requirements and test methods for the CS109 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to perform the CS109 test:
- Signal generator
- Amplifier
- Coupling Transformer
- Current sensor
- 0.5 Ohm resistor
- Isolation Transformer
Tips[edit]
Note: | If the current unit is shown as uA, it can be changed to dBuA by clicking Units, select Current and click Edit to change this. |
Define the test level[edit]
Create a correction file for the test level[edit]
- File
- New
- Correction
- New
- File
First create a correction file in RadiMation® and specify the test level (See 'FIGURE CS109-1. CS109 limit.').
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBuA.
- File
- Save correction
- File
This correction file with the test level will be used during EUT testing.
EUT Testing[edit]
EUT Testing equipment[edit]
Sensor powermeter | Devices 2 | The power meter or analyser to use for measuring the current | |
Current Sensor | Devices 2 | The current sensor to use with transfer factor attached to the driver | |
Injection device | Devices 2 | The injection device to use | |
Cables | |||
① | Cable current -> power meter | Cables | Cable (1) with a correction file specified for the cable loss |
Configure the EUT test[edit]
To perform the actual test on the EUT, create a conducted immunity multiband test with 1 band. The conducted immunity multiband test, can be started by selecting from the menu:
- Tests
- Conducted Immunity
- Multiband
- Tests
Start | The start frequency of the test. For example 30 Hz. |
End | The stop frequency of the test. For example 100 kHz. |
Frequency Step | The frequency step, see Table III. |
Modulation | CS109 does not specify a modulation to use. |
Test equipment | The equipment needed for the conducted immunity test. |
Location Type | Injection clamp. |
Test Level | Specify the current substitution method. Also add a "Measured Current Limit" |
At TestLevel click Add to add a new Test level and select TestLevel - Current closed loop method
Test level | Variable current, and specify the correction file for the applicable test level. |
When all settings are configured press Start Test to run the test.
EUT Test Result[edit]
The RadiMation® Conducted Immunity multiband test can be used to perform the CS109 test with a variable current closed loop method test level.