How to perform a MIL-STD-461 RS103, radiated susceptibility, electric field[edit]
This application note explains how the MIL-STD-461 RS103, radiated susceptibility, electric field test can be performed with RadiMation®
The exact requirements and test methods for the RS103 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to perform the CS109 test:
- Signal generator
- Amplifier
- Antenna
- Field probe
Define the test level[edit]
Create a correction file for the test level[edit]
File
New
Correction
First create a correction file in RadiMation® and specify the test level (See 'Table XI RS103 limits.').
File
Save correction
This correction file with the test level will be used during EUT testing.
EUT Testing[edit]
EUT Testing equipment[edit]
| # | Device name | Tab in testsite configuration window | note |
| Signal Generator | Devices 1 | The signal generator to use | |
| Amplifier | Devices 1 | The amplifier to use | |
| Field Sensor 1 | Field probes | The field sensor to use |
Configure the EUT test[edit]
To perform the test on the EUT, create a radiated immunity multiband test.
Create a test setup configuration file for the following frequency ranges:
| Frequency range | Polarization |
| 2 MHz to 30 MHz | Vertical |
| 30 MHz - 18 GHz | Both (Horizontal and vertical) |
| 18 GHz - 40 GHz | Both (Horizontal and vertical) |
Tests
Radiated Immunity
Multiband
| The start frequency of the test. For example 2 MHz. |
| The stop frequency of the test. For example 30 MHz. |
| The frequency step, see Table III. |
| The dwell time. |
| Constant. |
| The modulation to use see below. |
| The equipment needed for the radiated susceptibility test. |
| Injection clamp. |
| Specify the Electrical field closed loop method test level. Also configured a "Variable field" |
| No inputs are needed. |
At TestLevel click Add to add a new Test level and select TestLevel - Electrical field closed loop method
| The description of the test level. |
| Variable field, and specify the correction file for the applicable test level. When the level is one value this value can be specified with Constant field |
| Specify the tolerance to use. |
| Check Probe 1 to measure with the first probe. |
When all settings are configured press Start Test to run the test.