Current sensor transfer impedance determination method: Difference between revisions
Line 35: | Line 35: | ||
== Verification == | == Verification == | ||
[[Image:Current sensor Transfer impedance Prove.PNG]] | |||
The left impedance is the signal generator which is generating enough power for 1 ampere. | The left impedance is the signal generator which is generating enough power for 1 ampere. |
Revision as of 15:59, 23 January 2009
Theory[edit]
This 1Amp. generates in the right impedance.
The current sensor has 1 ohm transfer impedance, this means 1 ampere generates 1 Volt on the measuring part below.
The power in the lower 50 ohm impedance is
So 1 ohm:
Reference measurement[edit]
Probe measurement[edit]
Calculation[edit]
and in dBm.
Example[edit]
On 10 MHz we have the following information:
- Calibration: 0 dBm.
- Measurement: -27,96 dBm.
So:
Verification[edit]
This 1 Ampere generates in the right impedance.
The current sensor has 2 ohm transfer impedance, this means 1 ampere generates 2 Volt on the measuring part below.
So 2 ohm:
The difference is
Correction to a 1 ohm impedance is
Note: | This method is not a replacement for a real calibration as it may be performed by a none traceable device |